Hi all,
I just acquired a new to me HP 6285 0-20V / 5A psu. I plan on using it to compliment my 6237B. I have ran into some situations where I needed more current output for DC motors and such and the 6237B wasn't up to the task. I also wanted to make a project out of it, sort of a restore / learning project if you will. I might have bitten off more then I can chew though hehe.
Symptoms:
When I first got it, the volt / amp meter was broken and just kind of flap-en in the breeze. I got that fixed and zeroed. It seems to be working well.
The volt / amp meter selector switch works, but it is really worn out. I will eventually need to replace this, and will at a later date.
Onto the real issues:
When the PSU is first turned on, it seems to work just peachy, but after about 10mins or so the regulation really starts to go bonkers. At first the output voltage starts to vary in the uV range and steadily gets worse over time. After about 15 to 20 mins, all the variations will stop and the output voltage will settle down to about -300mV, where it pretty much stays till after it cools down. This is with the constant current adjustment wide open, and with the voltage adjustment wide open.
My questions:
This PSU really needs to be recapped, because I'm pretty sure that it has all of the original capacitors installed. I don't see a single new capacitor in it. As far as the film capacitors are concerned, can I just replace them with new modern day ceramics, or should I use something else?
With the loss of regulation, I started checking the reference voltages, and noticed that Q600 is getting quite warm compared to all of the others in the reference circuit. I can touch it and hold onto it for a bit, but it gets quite uncomfortable after about 10 secs or so. However, when the PSU becomes unstable, I don't see a noticable change in the reference voltages. So it leads me to believe that Q600 is okay. Q600 is an PNP, and I checked it for shorting, and it seems to check okay.
The reference voltage specs are as follows a long with what I have measured:
Test point (TP 29) = 9.4V +- 0.4V. I measured 9.07 Under normal operation and under the fault
TP 30: 15.4V + 0.5V = 15.1, which I also measured under both fault and normal operation
TP 29: -6.2V +0.3 = 6.07, which I also measured under both fault and normal operation
Now I just assume that these voltages are within spec, but the service manual of says +0.5 and +0.3 for TP 30 and TP 29? So are these really out of spec, or can it also be +-? I am a little confused in regards to this.
I also noticed that the zener diode VR600, or altleast I think that is what it is has been replaced. Its a more modern day component, then the rest of the diodes that are on board. There is a third component next to VR600 that I not sure what it is. Any suggestions on the part? its not on the lay out, but it also looks like a capacitor. (see included pictures)
Any suggestions on how you think I should proceed would be greatly appreciated.
rcowboy