Armadillo, thank you, when I opened the oscilloscope, I've found that VR540 seems to be broken by previous owner: its cap was just separated from its body. I'll buy a new one tomorrow and try to adjust.
>Have you recently done anything to the oscilloscope?
No, just opened its housing.
you may want to check the DELAYED TIMEBASE setting. did you make sure that it is switched to something like "MAIN ONLY" or "OFF"?
Hmm, how can I do this? If you mean the MAIN/MIX/DELAY switch, I've used it as described in operator's manual.
UPD:
When step 25 says set TIME/DIV to 10 us/DIV. it must mean the DELAYED TB speed, it doesn't make much sense the have the MAIN on 10us/Div and the DELAYED on the 100us/Div.
Only the 1st photo looks wrong to me, sometimes there's a switch setting to make the MAIN or DELAYED part of the trace the brightest.
Thank you, I understand now. I've tried to switch DELAY to 10 us, now it shows something that supposed to be the edge, but it is not as showed in manual, see the attachment. Even at 2 us Delay I'm unable to make it longer.