My HP8566B has developed a problem where the trace when looking at a signal (either externally applied or from the 100 MHz CAL port) is noticeably wobbly at spans below 100 MHz.  At a span of 5 MHz or below the trace is a broad mess.  It looks like one of the LOs has come unlocked.  Any ideas on which one?  How to check?  I have used 8566s at work and home for many years but never had to repair one, so any thoughts would be appreciated.
73 Dave W2KV
Online cncjerry
Re: HP8566B wobbly trace at narrow spans
« Reply #1 on: November 21, 2017, 06:10:41 pm »
Have you checked the RF deck power supply voltages with a scope?
Offline AF6LJ
« Reply #2 on: November 22, 2017, 02:47:50 am »
That's a good start.
Check all the power supply voltages first.
Also you might use AC line synch for the sweep and see if the trace is stable with line frequency synch. that will tell you if it is power supply related or not.
Sue AF6LJ