Hi!

After updating my meter from 1.21 to 1.51 my calibration-data is partially killed.  |O
I restored my backup, but that didn't fix the problem.

Is there something else, I can do, except sending it in for a recalibration?

Can someone post his cal-file, so that I can test it to exclude a hardware-problem?

Here is my cal-file, decoded with the script I found in the forum:
Misc Factors:
  Internal thermistor: 393
  HY3131 Frequency: 0 Hz

DCV
  5.0000V: Offset: 102, Gain: 1/61.18204
  50.000V: Offset: 328, Gain: 1/62.82518
  500.00V: Offset: 257, Gain: 1/62.9736
  1000.0V (and LowZ): Offset: 96, Gain: 1/63.086832

ACV
  5.0000V: Offset: 21961, Gain: 1/62.2256
           Frequency nonlinearity coefficients:
             100kHz: 500mV: 943, 5V: -123
             20kHz:  500mV: -55, 5V: -952
  50.000V: Offset: 22433, Gain: 1/62.19126
           Frequency nonlinearity coefficients:
             100kHz: 5V: 871, 50V: -609
             20kHz:  5V: -121, 50V: -1461
  500.00V: Offset: 21403, Gain: 1/62.3602
           Frequency nonlinearity coefficients:
             10kHz: 50V: 382, 500V: 3201
  1000.0V (and LowZ): Offset: 22470, Gain: 1/622.383488
           Frequency nonlinearity coefficients:
             10kHz: 60V: 45, 600V: 400

1ms Peak: Offset: -147, Gain: 1/1.082722

DCmV
  50.000mV: Offset: -714, Gain: 1/6.21588
  500.00mV: Offset: 702, Gain: 1/62.76796

ACmV
  50.000mV: Offset: 2500, Gain: 1/61.99774
            Frequency nonlinearity coefficients:
              100kHz: 5mV: 847, 50mV: -1239
  500.00mV: Offset: 2070, Gain: 1/62.01416
            Frequency nonlinearity coefficients:
              100kHz: 50mV: 669, 500mV: -2327

Temperature: Offset: 454, Gain: 1/6.22358

Resistance
  50.000Ω: Offset: 577, Gain: 1/7.52888
  500.00Ω (and Continuity): Offset: 395, Gain: 1/75.30344
  5.0000kΩ: Offset: 172, Gain: 1/75.376824
  50.000kΩ: Offset: 108, Gain: 1/75.51892
  500.00kΩ: Offset: 47, Gain: 1/77.42072
  5.0000MΩ: Offset: 78, Gain: 1/124.598984
  50.000MΩ: Offset: 90, Gain: 1/122.64012
            Nonlinearity coefficient: 87

Diode: Offset: 0, Gain: 1/31.1203

Capacitance
  10.00nF: Offset: 8389, Gain: 1/14.7948
  100.0nF: Offset: 8349, Gain: 1/147.3008
  1.000µF: Offset: 0, Gain: 1/1478.56512
  10.00µF: Offset: 0, Gain: 1/1694.696
  100.0µF: Offset: 0, Gain: 1/16842.54464
  10.00mF: Offset: 0, Gain: 1/1245.14904

ACµVA
  5V*50µA:   Volts Offset: 21163, Gain: 1/62.24492
              Amps Offset: 5024, Gain: 1/124.720136
  50V*50µA:  Volts Offset: 20983, Gain: 1/62.59712
              Amps Offset: 5127, Gain: 1/124.889504
  5V*500µA:  Volts Offset: 21158, Gain: 1/62.2558
              Amps Offset: 2, Gain: 1/125.11836
  50V*500µA: Volts Offset: 20946, Gain: 1/62.60438
              Amps Offset: 6, Gain: 1/125.25884

ACmVA
  5V*5mA:   Volts Offset: 21176, Gain: 1/62.22568
             Amps Offset: 5082, Gain: 1/126.22944
  50V*5mA:  Volts Offset: 20970, Gain: 1/62.20692
             Amps Offset: 5123, Gain: 1/126.30036
  5V*50mA:  Volts Offset: 18032, Gain: 1/62.29684
             Amps Offset: 1507, Gain: 1/126.392024
  50V*50mA: Volts Offset: 20956, Gain: 1/62.20952
             Amps Offset: 5, Gain: 1/126.49368

ACVA
  5V*500mA:   Volts Offset: 21272, Gain: 1/62.20432
               Amps Offset: 5274, Gain: 1/133.666424
  50V*500mA:  Volts Offset: 21018, Gain: 1/62.23128
               Amps Offset: 5314, Gain: 1/133.766344
  5V*10A:     Volts Offset: 21711, Gain: 1/60.7784
               Amps Offset: 11, Gain: 1/134.668496
  50V*10A:    Volts Offset: 21210, Gain: 1/61.67086
               Amps Offset: 3, Gain: 1/134.740608

ACµA
  50.000µA: Offset: 2717, Gain: 1/125.04032
  500.00µA: Offset: 2953, Gain: 1/125.11756

DCµA
  50.000µA: Offset: 1160, Gain: 1/75.37924
  500.00µA: Offset: -357, Gain: 1/12.66064

ACmA
  5.0000mA: Offset: 2977, Gain: 1/126.307704
  50.000mA: Offset: 2891, Gain: 1/126.409416

DCmA
  5.0000mA: Offset: -2873, Gain: 1/76.152296
  50.000mA: Offset: -256, Gain: 1/12.79048

ACA
  500.00mA: Offset: 3133, Gain: 1/133.77768
  5.0000A:  Offset: 3169, Gain: 1/133.84856
  10.000A:  Offset: 35, Gain: 1/134.454192

DCA
  500.00mA: Offset: -2877, Gain: 1/80.66664
  5.0000A:  Offset: -258, Gain: 1/13.54746
  10.000A:  Offset: -256, Gain: 1/135.479504

DCµVA
  5V*50µA:   Volts Offset: 51, Gain: 1/61.18784
              Amps Offset: -2059, Gain: 1/75.448664
  50V*50µA:  Volts Offset: 197, Gain: 1/62.87052
              Amps Offset: -2321, Gain: 1/75.4398
  5V*500µA:  Volts Offset: 50, Gain: 1/61.18944
              Amps Offset: -418, Gain: 1/12.66652
  50V*500µA: Volts Offset: 174, Gain: 1/62.87176
              Amps Offset: -424, Gain: 1/12.6673

DCmVA
  5V*5mA:   Volts Offset: 5, Gain: 1/61.19048
             Amps Offset: -2816, Gain: 1/76.14716
  50V*5mA:  Volts Offset: 143, Gain: 1/62.83512
             Amps Offset: -2920, Gain: 1/76.179936
  5V*50mA:  Volts Offset: -13, Gain: 1/61.20266
             Amps Offset: -229, Gain: 1/12.7908
  50V*50mA: Volts Offset: 141, Gain: 1/62.83952
             Amps Offset: -313, Gain: 1/12.79828

DCVA
  5V*500mA:   Volts Offset: -60, Gain: 1/61.1095
               Amps Offset: -2779, Gain: 1/80.678224
  50V*500mA:  Volts Offset: 135, Gain: 1/62.83042
               Amps Offset: -2732, Gain: 1/80.70308
  5V*10A:     Volts Offset: 692, Gain: 1/59.77082
               Amps Offset: -253, Gain: 1/135.526096
  50V*10A:    Volts Offset: 141, Gain: 1/62.6857
               Amps Offset: -357, Gain: 1/135.5664


Thanks for your support.
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5780
  • Country: ca
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #1 on: October 18, 2018, 08:27:34 pm »
You should put this thread in the right gw121 issues : https://www.eevblog.com/forum/testgear/eevblog-121gw-multimeter-issues/
 

Offline bicycleguy

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #2 on: October 19, 2018, 12:55:22 am »
Here's a text file like yours and the BIN.  Looks fairly similar, why do you think it's killed?
 

Offline KomposterTopic starter

  • Newbie
  • Posts: 3
  • Country: de
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #3 on: October 22, 2018, 03:07:04 am »
I will do some tests and post test results tomorrow, when beeing back in the workshop.
 

Offline KomposterTopic starter

  • Newbie
  • Posts: 3
  • Country: de
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #4 on: October 24, 2018, 12:46:14 am »
Having 2V DC on the input, gives me
in R0 2V DC
in R1 2VDC
in R2 2V DC
in R3 287V DC
in R4 OFL

Having 5V AC (50Hz) in the input, gives me
in R0 5V AC
in R1 a small voltage, counting down to 0V
in R2 same
in R3 same
in R4 same

If I raise DC voltage in autorange-mode
54V are OK, 100V on input results in 348V on the meter, 200V on input gives 408V on the meter.

I also tried the cal-data from bicycleguy - if my meter accepted the data - the result was the same.

I'm also confused, whats described in the manual, p52 ... both descriptions end up with
"The calibration data will now save to the SD card". My meter everytime shows "save"
at the end and I don't know, what happens ...

I think, something is seriously wrong here. Anything else, I can do without running a complete calibration-procedure?
I'm thinking about sending the meter in for a check.
Here in Germany the meters were shipped from a company called Welectron - the also offer calibration-services for test equipment.

*confused*
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 11601
  • Country: us
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #5 on: October 25, 2018, 02:29:57 am »
Hi!

After updating my meter from 1.21 to 1.51 my calibration-data is partially killed.  |O
I restored my backup, but that didn't fix the problem.

...

I am not sure what is meant by "I restored my backup".  It sounds like you are messing about with both the calibration file as well as the firmware version.  Did you reinstall both the 1.21 firmware along with your original calibration file?  And when you say install, did you actually install them into the meter or just copy them to the card?  I sounds like you may be confused how to do this so I wanted to make sure.

Did you make these same measurements prior to making the changes?  I would assume you have this data or would not be blaming the firmware upgrade. 

I have not really done anything with the newer versions of firmware beyond 1.26.  If there was a major problem like this with the firmware, I would have thought a lot of people would be posting about it but maybe not.   



Offline bicycleguy

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #6 on: October 25, 2018, 02:52:14 am »
I'm confused too.  There's no way putting my cal data in your meter could do anything good :scared:

I thought you were just going to look at it to see if yours looked hosed.
 

Offline Brumby

  • Supporter
  • ****
  • Posts: 12281
  • Country: au
Re: 121GW - calibration killed after Firmware-Update?!
« Reply #7 on: October 25, 2018, 03:08:03 am »
There's no way putting my cal data in your meter could do anything good :scared:

That would worry me too.  The whole purpose of having individual cal data is to address the variations in components from unit to unit.

The best case scenario of doing that would be to bring such a meter into readings that would look sane - but there is no way in the wide, wide world of sports that I would consider that meter accurate.  No way.  Not ever.  I just would not even try loading up someone else's cal data.

Quote
I thought you were just going to look at it to see if yours looked hosed.
This is a reasonable thing to do.  Just a side by side desk check.
 


Share me