Hi Gandalf_Sr,
I think the left connector is the forward port and the right is the reverse
The forward parameters S11 and S21 are measured with the source signal coming out of port 1 and when measuring
reverse parameters S22 and S12 the analyzer switches the source generator to port 2.
It looks like I might also need a calibration disk for the 8753E which I don't have - do I need one?
If you had exactly one of the calibration kits which are listed on your analyzer in the CalKit selection menu then you don't
need a calibration disk because the coefficients of these are already stored in the analyzer. But with a non-HP calibration
kit to be exact, one should enter user-defined coefficients for this cal kit. But doing this seems to be very complicated.
First, because it is not clear how to determine the correct coefficients and secondly because it is complicated to enter these
coefficients into the machine. This is the reason why I suggested to select the HP 85035C CalKit on the 8753 together with
your SMA calibration kit as a starting point because they are very similar. At least up to 1 or 2 GHz this should not give too much error.
I also need a simple step-by-step calibration guide that includes pictures of what to connect where
I also had about the same difficulties when I started using a VNA 3years ago. I suggest to start with a simple introductory exercise.
Measuring the inductivity of a small air core inductor with some turns of enameled copper wire, something like this:
http://www.cettechnology.com/products/air-coil-inductors/This can be done by measuring the S11 Parameter. We use only Port 1. First you have to decide which frequency range:
You select Start and Stop frequency. I suggest Start: 30 MHz Stop: 300 MHz. Later you should also eventually modify the IF-BW,
source power, averaging and the like. It is important to set up these parameters before calibrating. This because a modification
of some of these parameters makes the calibration obsolete.
Select the Cal Kit:
[Cal ] --> <Cal Kit> --> <Select Cal Kit> --> <3.5mm D>
Then you can calibrate:
[Cal] --> <Calibrate Menu> --> <S11 1-port>
Connect your OPEN reference. Press <OPEN>. Wait for beep.
Connect your SHORT reference. Press <SHORT>. Wait for beep.
Connect your LOAD reference. Press <LOAD>. Wait for beep.
Press <DONE>
The machine is calibrated now. To measure our coil we solder it onto one of these SMA connectors
http://uk.rs-online.com/web/p/sma-connectors/6163422/But before soldering the coil we connect this connector on the VNA without the coil and we enter the following:
[Meas] --> <s11>
[Format] --> <Smith Chart>
Now we should see a small vertical trace at the right side of the chart. This is essentially caused by the extension
of our measuring cables by the electrical length of this SMA connector. We try now to compensate this by pressing
[Cal] --> <More> --> <Port Extensions> --> <Extension ON> --> <Extension PORT1>
Now we turn the knob clockwise until we see a single point, the ideal open.
Now we solder our coil onto the terminals of this SMA connector. (Please remove the connector to do this to avoid
static discharge of the soldering iron)
Reconnect the connector with the coil and press
[Marker]
Turn the knob to a frequency of interest and read the value of the inductivity on the top of the screen.
Shouldn't he have selected a Cal kit before he did the calibration?
Yes, absolutely. This is not clear. At some point he is pressing the Preset Button. On my machine this selects the 7mm Cal Kit as default.
There are 2 other things in the video which I find not very good:
1. He should avoid the mechanical stress on the small connector on port 2 with the filter pcb suspended on it by using a second cable.
2. When he is removing the calibration standards from the measuring cable he is turning the standard ccw against the cable connector.
This also should be avoided. Only the nut of the cable connector should be rotated.
Once the full two-port calibration was done, wouldn't it make sense to save it somewhere?
Yes, this could be done but personally I don't have much confidence in the long time stability of a calibration.