It sounds like all the power supply voltages are good. Kind of need another clue.
What you need is a faded dimmed trace, in X-Y mode measure some voltages to compare to when the trace has good brightness.
If it's temperature related, hot air or freeze-spray (not on the HV section OK, frost conducts!) might be worth trying.
For the horizontal deflection transistor running hot, with the (H) POS ADJ control centered, I would make sure VR807 is reasonable and VR806 if you like to adjust for x10MAG centering. In X-Y mode you can measure the X and Y plates voltages. Each +,- amp is symmetrical. 150BTB31A X-axis 19-21.6 VDC/cm, Y-axis 10.8-13.2 VDC/cm and a 14cm screen.
I think if a V or H deflection is not balanced but the front panel control is cranked, it might cause the (defl amp) bias current to be upset. Or if trying to cover for a lazy deflection transistor or DC offset in the sweep oscillator with the front panel control.
I thought if the deflection plates (both in the pair) had low voltage, the acceleration would be a bit less, but the astigmatism would go off too.
Maybe look at C2025 (CRT pin 12) I think that's the 1st acceleration anode. Cap should be over 100V rating
Otherwise, it might be the beam blanking is drifting and upsetting the intensity (grid bias), and not easy to troubleshoot. Or try inject a signal on the Z-axis and see how it does.
The CRT bias tells a lot, but measuring it is too difficult as -2kV common-mode exists and a DMM would be dangerous.
If instead you measure the (HV) cathode (CRT pin2) and then control grid (CRT pin3), the difference is the CRT bias, beam cutoff is typically -25 to -65V for the 150BT131.
It might tell, with a maxxed out intensity control if the CRT is biased weak.