Hello EEVblog
This is my first post :D
I have just got a brand new DP832 for my lab, and I began testing it. I can see that channel 1 and 3 is spot on within spec, however channel two is off by approximating 7-10mV. My big consern is that when all outputs is turned off, I can see that on channel two there is 7mV on the out put and running 143uA. I do not know if this i normal or I got a faulty unit?
Looking forward to hear from you guys/girls, and please say if you need more information :)
Kind Regards
Emil
-
Mine had similar offsets. Technically within published specs but a little bit annoying.
I adjusted them out using my new 9999 count $22 AN8009.
Rigol calibration document attached below.