Author Topic: DS1074Z - Decode RS232 ASCII problem  (Read 2530 times)
Offline matkarTopic starter
  • Country: si
  • Sixty percent of the time it works EVERY time.
« on: October 10, 2015, 09:46:06 am »
I've found a problem with the decoder. When decoding character "8" it actually displays "9" when in ASC mode. Other modes (HEX, BIN, DEC) return the correct value.
I've tested on 38400, 8N1.
I have the original firmware it came with the scope: 00.02.01.SP1
Is this a known issue? Has it been resolved in newer firmware updates?
Offline Karel
Re: DS1074Z - Decode RS232 ASCII problem
« Reply #1 on: October 10, 2015, 11:34:34 am »
No problem here, running latest fw 00.04.03.SP1.
  • Country: si
« Reply #2 on: October 10, 2015, 09:12:13 pm »
Thank you Karel.
Obviously it's time for an update...
Offline H.O
« Reply #3 on: October 11, 2015, 08:13:00 am »
Have you tried all other characters and is it only '8' that's wrong?
I'm thinking perhaps your actual baudrate doesn't match what the scope is set to so it misinterprets.
I don't have a DS1054Z myself, just thinking out loud....
Offline smithnerd
« Reply #4 on: October 11, 2015, 12:00:30 pm »
This bug was fixed in version: 00.04.02.03.00.
"4. RS232 decoder shows ‘9’ for ‘8’ input in ASCII format."
(According to the firmware release notes).
