I have a new Ds4014 that I have updated to the latest firmware and made a few other tweaks.
The firmware update procedure states to run a self-cal after the firmware update.

Is this a good idea? The scope is brand new. Is the self-cal going to overwrite the factory calibration or does it do something different?

many thanks

Lee
Title: Re: ds4014 self-cal
Post by: DaJMasta on July 27, 2018, 03:31:53 pm
Self cal routines don't write over factory calibration, which is generally protected behind a password and is in a service menu.  In fact, a run of the self-cal after a warm up period is the first step in the full calibration for the DS4000 series.
Title: Re: ds4014 self-cal
Post by: Fungus on July 27, 2018, 03:44:19 pm
Is this a good idea?

Yes.

The scope is brand new. Is the self-cal going to overwrite the factory calibration or does it do something different?

The "factory calibration" of machines with self-cal is probably just somebody at the factory pressing the "self-cal" button.

Your manual says:
(https://www.eevblog.com/forum/testgear/ds4014-self-cal/?action=dlattach;attach=485033;image)

Key words: "At any time" - you should really do it before any important measurement.

Or ... just relax. These things are only about 3% accurate (see manual).

Use it to look at wiggly lines on screen, the numbers are really just ballpark figures.
Title: Re: ds4014 self-cal
Post by: lmclaren on July 28, 2018, 12:05:23 am
thank you,

Off to buy some bits to make the 5-1 cable.