Author Topic: EEVBlog 121GW Discussion thread  (Read 778579 times)
Offline npelov
  • Country: bg
    • Microlab.info
Re: EEVBlog 121GW Discussion thread
« Reply #1200 on: November 30, 2018, 04:35:40 pm »
First a suggestion - it would be nice to have a change log in txt format in the firmware zip. The last firmware version is not present in the change log in the manual, because (probably) the manual is still being updated. Having a copy of the change log with the binary file let's you know what's fixed and what to test after updating.
And another thing - the 50MOhm range is currently slow. If it needs to be slow for stability I'm ok with that, but it should be "manual only" range, because it's slowing down all resistance measurements. You don't measure resistance 5-50 MOhms that often - it's ok if it's not included in auto ranging. Or at least have a setup option if it should be included in auto-ranging.
Also the 50 ohm range is redundant. It does not give you any more information than 500 Ohm. When shorting the probes  (even the gold coated ones) the last digit cannot show the same result twice (not even close) for measurements few minutes apart, without temperature changes. The short term measurements show difference more than 10mOhms, which makes the last digit useless (and distracting). I think 500 ohm is pretty good compared to other DMMs, although as many other ranges you have to wait few seconds for the last digit to settle. My request is if possible to turn 50 Ohm range off in SETUP.
I'm happy with the DMM - it has few interesting features that are not present on most. Of course you pay the price of low burden voltage - zero offset is drifting all over the pace on most of the ranges. It must be calibrated pretty often. It could be the case with full scale error, but I haven't tested it yet over time. The other annoying thing is that it takes quite a few seconds for the last digit to settle. I haven't tested this since fw 1.26 - ohms still settles slow. On most of the other DMMs I have the digits stop moving withing 1 second, even the cheap Chinese ones ($40). I think there is something wrong with the filtering on some ranges.
Offline CDaniel
« Reply #1201 on: November 30, 2018, 09:34:52 pm »
This issues are known for a long time if you read the posts .
Autorange is slow .
50ohm range may be redundant because they did it wrong . The protection thermistor PTC3 little drift with temperature is responsible for getting different results all the time . Maybe there is some self heating from the measurement current itself , not just the internal temperature variation .  2 wire low ohm measurement is not that simple to do it right .
Later edit
I played a little with the board out and ... it's not the thermistor heating up ... The problem is that the measurement current is low ( maybe too low )  and it's using the lowest range possible with 10 times internal amplification ( in HY3131 configuration is like that ) . Every solder joint is a thermocouple at this  low level . That's why it drifts all the time and can't be stable .
With  the thermistors shunted , any solder joint I heat up ( in input zone ) results in massive 1000 counts drift . Thats 1 ohm . In other ranges the drift is lower , as expected . What's impossible to say at this point is if they use HY3131 correctly in 50ohm range as in configuration pdf document .
« Last Edit: December 02, 2018, 01:17:26 pm by CDaniel »
The following users thanked this post: TikhonovC
Offline EEVblog
« Reply #1202 on: December 03, 2018, 01:11:41 am »
The question is are the new batch of meters any better than the kickstarter meter?
Is the new meter still using the spacer?
Has the new meter any new components and specification?
Does the new meter still have the existing firmware bugs?
Dave is there any difference with the new over the old?
The latest meters shipping have two hardware changes. A copper shield, and a two transistor daughter board to replace the TVS.
They still use the spacer shims until existing blank PCB stock is used up.
Firmware is the current release 1.57
Specs have not changed.
The following users thanked this post: 1anX, Marco1971
Online IanB
« Reply #1203 on: December 03, 2018, 01:18:17 am »
Can we see pictures, especially of the shield?
These seem like things that might be relevant retro-fits.
The following users thanked this post: gnavigator1007
Offline EEVblog
« Reply #1204 on: December 03, 2018, 01:33:46 am »
Attached.
The shield is a multi-layer plastic sandwich construction. You could DIY, but watch for shorts and clearances etc. Connected to the negative battery clip on the other side.
Again, this will only be for existing stock of blank PCBs.
I don't believe the DIY transistor mod is recommended because it likely requires recalibration.
The following users thanked this post: ChunkyPastaSauce, gnavigator1007, 1anX, Marco1971
Online joeqsmith
« Reply #1205 on: December 03, 2018, 02:05:59 am »
Sure, there's some new shield and clamping...   but what's this I see!!!????
Is that a TI CD4053B with an absolute maximum Vsupply of 20V??  The same part I have said several times that I went to in an attempt to get the meter to survive some basic transients???!!!!   Looks like someone is trying to spoil my fun.    If they get the clamping right, this meter should do just fine.
Looking forward to seeing the meter finished up.   :-+
Offline EEVblog
« Reply #1206 on: December 03, 2018, 02:55:33 am »
Yes, sorry, forgot about that change.
Offline 1anX
« Reply #1207 on: December 03, 2018, 03:30:37 am »
Joe, does changing to this spec chip alter calibration?
What "basic transients" or voltage levels did you apply during your testing which destroyed the 4053?
Offline EEVblog
« Reply #1208 on: December 03, 2018, 03:57:05 am »
No, it doesn't. It's a DIY mod if you are really that keen, I personally would not bother.
I can't speak for Joe's tests, but FYI the previous chip passed all UL CAT III standard tests just fine.
Offline 1anX
« Reply #1209 on: December 03, 2018, 04:23:13 am »
Thanks for the info Dave!
I have bought the BM235 and the 121GW from you and have given the BM235 to my son for his automotive work. I'm looking to buy another 121GW type spec meter and have been considering the Brymen 869. How long before the next gen 121GW becomes available with the new PCB and updates?
I'm quite happy to hardware mod my current 121GW and would love to see a video on EEVBlog on the modding and re-calibrating this meter. I'm sure plenty of others with the kickstarter meter would also like to mod and update this meter with a instructional video from you!
Still need to buy another meter, but making the current kickstarter meter a project would be fun and educational.
« Last Edit: December 03, 2018, 04:30:26 am by 1anX »
Online joeqsmith
« Reply #1210 on: December 03, 2018, 04:37:39 am »
Changing to the CD4053B in of itself would do nothing to improve the meters robustness.   We are talking about a couple of volts vs thousands of volts we are applying.    The problem is coming up with a clamp that is not set so low that it interferes with the high voltage supply used for the diode check.  Yet it can't be so high that it allows us to exceed the absolute Vs rating of the mux.   You have a 15-16V supply and an IC rated for 18V (I think) absolute max.   Increasing this margin a couple of volts eased the effort of sorting out the clamp.
I am hoping that what led them to use the TI part was to try to improve the meter's robustness.  I won't know how it behaves until I run one.
The video for the prototype is still on-line.  It was first damaged with a 2KV peak, 100us FWHH with a 2ohm source.   There were a lot of low cost meters that were damaged at these low levels.    For those wanting to point out the 87V as a gold standard, make sure you check all the facts first.   Again, the data is here:
https://docs.google.com/spreadsheets/d/1cXzYpIoyVm9QJUju4KXqM22CEQZP3_xwWvDyeVwxTy4/edit#gid=400910915
Video of the 121 proto can be found here:
https://youtu.be/X28bwdTBW8g?list=PLZSS2ajxhiQDBDdtQNjVnGxShaVQ3nUMY&t=1444
Offline ChunkyPastaSauce
« Reply #1211 on: December 03, 2018, 04:47:44 am »
Attached.
The shield - a fix for the jumping values when interacting with the membrane buttons?
Offline EEVblog
« Reply #1212 on: December 03, 2018, 05:39:02 am »
Well past this batch of 700, they have a lot of blank boards yet. Can't give a timeline.
Most people do not have the capability to calibrate the meter, it requires precise generated values from a proper meter calibrator.
Offline EEVblog
« Reply #1213 on: December 03, 2018, 05:40:12 am »
Correct.
Offline GeoffreyF
« Reply #1214 on: December 03, 2018, 12:58:39 pm »
Is there a suggested process to make (or purchase from EEVBlog) a shield in earlier models?
US Amateur Extra W1GCF.
Offline EEVblog
« Reply #1215 on: December 03, 2018, 02:24:29 pm »
I could ask Kane to send me a whole bunch I guess for those that want them.
Offline 1anX
« Reply #1216 on: December 03, 2018, 09:29:10 pm »
Dave, start a list of those wanting the button shield and add me to it!
I would appreciate the added stability it would bring to the meter.
Offline darik
« Reply #1217 on: December 03, 2018, 09:37:14 pm »
I'm interested.
Offline TikhonovC
  • Country: ru
« Reply #1218 on: December 04, 2018, 04:23:53 pm »
And a two transistor daughter board to replace the TVS.
Offline lowimpedance
  • Watts in an ohm?
« Reply #1219 on: December 04, 2018, 10:47:52 pm »
As Dave stated above this would mean a re-calibration which uses equipment that most users would not have ready access to, so is best to leave as is. Of course having the PCB available to those that do have the required gear would be nice too  ;).
The following users thanked this post: Marco1971
Offline TikhonovC
  • Country: ru
« Reply #1220 on: December 05, 2018, 07:25:50 am »
I think that the cost of this board is 2-5 USD.
The cost of the shield is 5-10 USD.
Shipping costs will be at least 10-15 USD.
This PCB can be added to the package. For future calibration. Anyway it can be useful.
« Last Edit: December 05, 2018, 07:35:34 am by TikhonovC »
Offline CDaniel
« Reply #1221 on: December 05, 2018, 07:32:31 am »
That's the output protection for resistence / capacitance ranges , so it's not that difficult to recalibrate . Only high resistance ranges and maybe low capacitance range will be affected anyway .
« Last Edit: December 05, 2018, 07:39:49 am by CDaniel »
Offline Brumby
« Reply #1222 on: December 05, 2018, 07:35:33 am »
The comment was not on how easy/difficult recalibration might be - but about the equipment required to do so properly.
Offline CDaniel
« Reply #1223 on: December 05, 2018, 07:40:38 am »
equipment = some high precision resistors and maybe capacitors
« Last Edit: December 05, 2018, 07:45:54 am by CDaniel »
Offline Brumby
« Reply #1224 on: December 05, 2018, 07:44:12 am »
What?  No high precision voltage standards?
Therein lies the question: what equipment and standards are required?  Do you know?