This is set, IIRC, to 40 mHz. Pressing "MEASURE" and then the FREQ: button, produces the low signal or low rep rate response.
I have the service manual, so I suppose what i need to do is figure out where to look for problem.
best,
John
This your image is not after you have run measurement.
What it show with this setting tell me that this is not result after auto measurement. (I mean V/div and t/div what can see in image)
Btw, how much you have there DC offset in signal.
But first take signals off. Under vertical V/div knobs there is buttons. Push both upper buttons together for start offs/balance null process and wait it ends.
Also please turn CH1, CH2, Timebase "Variable" to detend position, now they are not as can see in image.. (keep these normally always in detend position until nessessary to change and after then return always back to detend for avoid mistakes, this is normal lab procedure)
After then select CH1 GND
Adjust trace position to vertical center.
Select CH1 DC, 1V/div
Connect your signal generator. Adjust signal generator level to something like 6Vpp (6 div)
Adjust signal generator DC offset for well vertical centered signal (DC offset)
Adjust freg to around example 40kHz
Change scope input DC / DC50ohm and watch signal level change around 1:2
If not, then 50ohm input is defect (this is not very rare) If it is defect then continue with normal DC input.
Adjust signal generator level is around 6 divisions (1V/div)
Adjust enough low trace and text brightness for good picture with camera.
Take image from screen and after then do not change anything.
Press measure and after then select frequency
take image from screen.
Press measure and select voltage
take image from screen.
(my preliminary suspect is that CAL is not ok)