Author Topic: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)  (Read 151815 times)

0 Members and 2 Guests are viewing this topic.

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #125 on: January 29, 2014, 07:33:34 am »
Is there any significant change on the battery's life after the replacement ?

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #126 on: January 29, 2014, 04:44:33 pm »
Besides my personal 189, we have one here at work that is used in the RF Lab daily. I inquired as to it's history of battery life, the main user says he thinks it typically lasts over one year. We have not however done an inspection of this one yet . I'm a bit confused on the exact purpose of the SuperCap. Does it's usefullness only come into play during battery swaps, when there is no other voltage present to maintain the memories? Mine is without one now, but if I just power off, and back on, it still seem to retain memory.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #127 on: January 29, 2014, 04:51:50 pm »
That cap's purpose is just to retain the dmm clock during batteries replacement, according to Fluke.

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #128 on: January 30, 2014, 05:07:00 pm »
Thanks...   that's good to confirm,  I was coming to that conclusion. I think I'll just leave it out. Obviously it caused more problems for Fluke, than it solved.
 

Offline tony3d

  • Frequent Contributor
  • **
  • Posts: 393
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #129 on: January 30, 2014, 06:31:15 pm »
According to CDI The life expectancy seems to be 2.8 years. That is dependent on several factors, one humidity must remain below 50%. Kinda strange for a meter meant to last 10 years!
 

Offline stuartk

  • Regular Contributor
  • *
  • Posts: 102
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #130 on: February 13, 2014, 04:11:03 pm »
I cracked the case on my 289 purchased in 2007 and found the leaking capacitor. It was giving me intermittent loss of timekeeping.

I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

They have replaced the cap, changed my scuffed up battery cover, changed the batteries and re-calibrated it, with a certificate and documentation.

All for free.  :)

Kudos to Fluke Canada and thank you.

Stuart
 

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #131 on: February 13, 2014, 04:20:54 pm »
Has anyone found the Fluke part number for this capacitor?   When one has a meter repaired under warranty, is this information given?  I'd expect not, but since these aren't available in the US in quantities less than 2000, I was considering ordering from Fluke.  However, Fluke will do no parts research - one has to know the part number.

 

Offline stuartk

  • Regular Contributor
  • *
  • Posts: 102
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #132 on: February 13, 2014, 08:39:54 pm »
No they didn't. They only said replaced capacitor C145...
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #133 on: February 13, 2014, 08:46:50 pm »
Has anyone found the Fluke part number for this capacitor?

Search back a few pages, I replaced mine with a $2 part from Digikey.

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #134 on: February 13, 2014, 08:56:47 pm »
Has anyone found the Fluke part number for this capacitor?

Search back a few pages, I replaced mine with a $2 part from Digikey.

I saw that.  It looks like that's the only alternative for a non-warranty unit.

My 189 has the larger capacitor pictured earlier in this thread.  It seems to be OK, but I didn't unsolder it to look underneath.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #135 on: February 14, 2014, 02:59:06 am »
I cracked the case on my 289 purchased in 2007 and found the leaking capacitor. It was giving me intermittent loss of timekeeping.

I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

They have replaced the cap, changed my scuffed up battery cover, changed the batteries and re-calibrated it, with a certificate and documentation.

All for free.  :)

Kudos to Fluke Canada and thank you.

Stuart

Stuart, closeup photo of the newly replaced cap please.

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #136 on: February 14, 2014, 04:34:29 am »
I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

All for free.  :)
Hi Stuart,

Are you the original owner of the meter and did you purchase it from an authorized dealer?

edit: The other advantage of sending it back to Fluke for repair, especially if you owned the meter for a few years, is that you might get calibration thrown in for free.
« Last Edit: February 14, 2014, 04:43:35 am by retiredcaps »
 

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #137 on: February 16, 2014, 06:13:30 pm »
I cracked the case on my 289 purchased in 2007 and found the leaking capacitor. It was giving me intermittent loss of timekeeping.

I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

They have replaced the cap, changed my scuffed up battery cover, changed the batteries and re-calibrated it, with a certificate and documentation.

All for free.  :)

Kudos to Fluke Canada and thank you.

Stuart

Stuart, closeup photo of the newly replaced cap please.


Probably would void the calibration :)
 

Offline stuartk

  • Regular Contributor
  • *
  • Posts: 102
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #138 on: February 17, 2014, 12:38:48 am »
Yep,

I'm the original owner. I was going to crack the case, but they put the calibration sticker and a tamper sticker over the case.

I think they were trying to send a message to me after I sent them a picture of my leaking cap! ;D

I suspect I might mess up the calibration if I crack it in anyhow....

Stuart
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #139 on: February 17, 2014, 05:20:33 am »
I'm the original owner. I was going to crack the case, but they put the calibration sticker and a tamper sticker over the case.
Well, thanks for reporting back.  I wouldn't break the seals either if it were my meter.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #140 on: February 17, 2014, 08:17:14 am »
+1 , thanks for the update.

I guess we'll have to wait at least until next year to be able to peek the replaced cap in Stuart's 289.  :P

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #141 on: February 18, 2014, 01:28:56 am »
One, or two things seem pretty certain. One funky Super Cap does not ruin the Fluke reputation. Fluke equipment holds it's value, particularly the handhelds. They get re-sold for quite respectable sums on a regular basis. This, in contrast to the junk meters that you toss in the trash.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #142 on: February 18, 2014, 02:00:14 am »
Agree, apart from the ultra-cap problem, have to admit this 28x series build quality is excellent.

I was very impressed when seeing Martin tested his 287 in freezing cold in the chill box, the measurement result was still spot on with it's spec at operating temp -20 °C to 55 °C.  :-+
« Last Edit: February 18, 2014, 02:05:38 am by BravoV »
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #143 on: February 18, 2014, 12:41:27 pm »
hi guys..
i have a 1 year old fluke287 and it consumes battery much faster..i opened the case and i saw the same cap corroding.. i wonder if a 0.06F supercap i have could be a good replacement?
Always learn how to break and fix things electronics!
 

Offline kaz911

  • Super Contributor
  • ***
  • Posts: 1052
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #144 on: February 27, 2014, 09:17:16 am »
hi guys..
i have a 1 year old fluke287 and it consumes battery much faster..i opened the case and i saw the same cap corroding.. i wonder if a 0.06F supercap i have could be a good replacement?

Don't fix them. I just handed a 289 and a 287 into Fluke and they will be repaired on warranty despite being quite old :) Fluke has "Limited World Wide" warranty on them - and they have a list of serial numbers for which repair is free of charge. As I only paid $200 for my 289 since it was not perfect - I'm now very happy that has been fixed on warranty as well.

And to quote the Calibration girl - 287/289's are really well made apart from the Cap issue - so they are usually in spec/calibration even after 6+ years - they have so far not seen a working 287/289 that was NOT within spec if it had not been "mishandled"
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #145 on: February 27, 2014, 09:45:12 am »

Don't fix them. I just handed a 289 and a 287 into Fluke and they will be repaired on warranty despite being quite old :) Fluke has "Limited World Wide" warranty on them - and they have a list of serial numbers for which repair is free of charge. As I only paid $200 for my 289 since it was not perfect - I'm now very happy that has been fixed on warranty as well.

And to quote the Calibration girl - 287/289's are really well made apart from the Cap issue - so they are usually in spec/calibration even after 6+ years - they have so far not seen a working 287/289 that was NOT within spec if it had not been "mishandled"

well the problem is i cant find any fluke service center here in Philippines :--..so im thinking the best thing to do is to replace the cap myself just like others did..i love my fluke but i hate the way it consumes the battery :palm:
« Last Edit: February 27, 2014, 09:47:44 am by elex_enthusiast »
Always learn how to break and fix things electronics!
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #146 on: February 27, 2014, 10:36:28 am »
well the problem is i cant find any fluke service center here in Philippines :--..so im thinking the best thing to do is to replace the cap myself just like others did..i love my fluke but i hate the way it consumes the battery :palm:

Are you sure the cap is bad, the 28x is a battery hog even with a good cap? 

Mine was drawing something like 10mA when powered off, if you have the means you might test yours.  If you look back in this thread there is a spec, in the uA is what it should draw when working correctly.

Offline kaz911

  • Super Contributor
  • ***
  • Posts: 1052
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #147 on: February 27, 2014, 10:53:05 am »

Don't fix them. I just handed a 289 and a 287 into Fluke and they will be repaired on warranty despite being quite old :) Fluke has "Limited World Wide" warranty on them - and they have a list of serial numbers for which repair is free of charge. As I only paid $200 for my 289 since it was not perfect - I'm now very happy that has been fixed on warranty as well.

And to quote the Calibration girl - 287/289's are really well made apart from the Cap issue - so they are usually in spec/calibration even after 6+ years - they have so far not seen a working 287/289 that was NOT within spec if it had not been "mishandled"


well the problem is i cant find any fluke service center here in Philippines :--..so im thinking the best thing to do is to replace the cap myself just like others did..i love my fluke but i hate the way it consumes the battery :palm:

I handed mine into a distributor - not Fluke direct. So any distributor can verify warranty recalls

Mine has been off to Holland for repair as the local guys does not do any repairs. All at no cost to me.
 

Offline kaz911

  • Super Contributor
  • ***
  • Posts: 1052
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #148 on: February 27, 2014, 10:56:22 am »
and the quick way to check CAP problem - put in new batteries - check time and date is fine. Leave on for a couple of minutes. Shut it down. Remove the batteries - count to 10 - plug batteries back in. If it asks for time/date - you have a super-cap issue.
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #149 on: February 27, 2014, 12:20:05 pm »
yes it does ask to reset the time/date..
Mine was drawing something like 10mA when powered off, if you have the means you might test yours.

unfortunately i have no other reliable meter i can use to measure its current draw but im pretty sure the cap is bad..
« Last Edit: February 27, 2014, 01:48:10 pm by elex_enthusiast »
Always learn how to break and fix things electronics!
 