Author Topic: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)  (Read 151622 times)

0 Members and 1 Guest are viewing this topic.

Offline JohnDeeTopic starter

  • Newbie
  • Posts: 5
I have a Fluke 189 multimeter.  It suddenly died (nothing shows on the LCD display).  I changed the 4 batteries (the old batteries were only slightly below normal).  The two fuses checked OK.  Inspection of the circuit board shows some leaking on a small (6.5mm diameter) surface mount battery.  The marking of this two-tab coin cell has the logo of Matsushita (now Panasonic) and the numbers 3709.  I could not identify this cell on the web in order to find a suitable replacement.  I am hoping somebody can help...
« Last Edit: November 28, 2013, 04:28:11 pm by JohnDee »
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount coin cell
« Reply #1 on: November 04, 2013, 11:52:54 pm »
Inspection of the circuit board shows some leaking on a small (6.5mm diameter) surface mount battery.  The marking of this two-tab coin cell has the logo of Matsushita (now Panasonic) and the numbers 3709.  I could not identify this cell on the web in order to find a suitable replacement.
I suggest you clean up the leakage with a q-tip/toothbrush with a generous amount of IPA.

I remember seeing an ebay auction for what you are talking about ...

http://www.ebay.com/itm/281103728372

Right now, I can't find any more information regarding the specifications.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount coin cell
« Reply #2 on: November 05, 2013, 12:31:30 am »
 

Offline JohnDeeTopic starter

  • Newbie
  • Posts: 5
Re: Fluke 189 with leaking surface mount coin cell
« Reply #3 on: November 05, 2013, 12:42:33 pm »
You are right !!!   It is indeed a Panasonic EN gold capacitor.  I see that it is available at Mouser ($2.84 unit price).  I will order it and hopefully it will fix the problem.  I also read the Panasonic Technical Guide you sent me and reading at sections 4.1 and 4.2 on expected lifetime it seems to me that selecting that capacitor was a poor design choice with a surface mounted component having to be changed after barely a few years of usage...
Thanks for the great help !
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #4 on: November 05, 2013, 01:38:12 pm »
Damn, is this leaking supercap at the board is a serious problem plaguing the fluke dmm 180 & 280 series ?  :-\

As a new owner of 287,  this makes me worry as I just knew it.

JohnDee, please post the photo of that leaked cap and the damages at the pcb if any.
« Last Edit: November 05, 2013, 01:40:59 pm by BravoV »
 

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
Re: Fluke 189 with leaking surface mount coin cell
« Reply #5 on: November 05, 2013, 07:23:51 pm »
Damn, is this leaking supercap at the board is a serious problem plaguing the fluke dmm 180 & 280 series ?  :-\

As a new owner of 287,  this makes me worry as I just knew it.

JohnDee, please post the photo of that leaked cap and the damages at the pcb if any.
there is some internal pictures of the 287 here in the forum its hard to find btw.
 

Offline uoficowboy

  • Regular Contributor
  • *
  • Posts: 186
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #6 on: November 05, 2013, 08:06:47 pm »
Interesting - I had noticed that my 189 did not have the battery life that it used to have. Still lasting months... but it used to last years. So the consensus is that replacing this component will fix that?
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #7 on: November 05, 2013, 08:21:55 pm »
I have a Fluke 189 multimeter.  It suddenly died (nothing shows on the LCD display).  I changed the 4 batteries (the old batteries were only slightly below normal).  The two fuses checked OK.  Inspection of the circuit board shows some leaking on a small (6.5mm diameter) surface mount battery.  The marking of this two-tab coin cell has the logo of Matsushita (now Panasonic) and the numbers 3709.  I could not identify this cell on the web in order to find a suitable replacement.  I am hoping somebody can help...
From what I've found, that's a supercapacitor rated at 3.3V and 0.2F. Looks like they're discontinued, but perhaps one of these may do the job: Digikey
 

Online mikeselectricstuff

  • Super Contributor
  • ***
  • Posts: 13741
  • Country: gb
    • Mike's Electric Stuff
Re: Fluke 189 with leaking surface mount coin cell
« Reply #8 on: November 05, 2013, 08:32:21 pm »
Just had a look in my 287 - some signs of corrosion on the cap
Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
Re: Fluke 189 with leaking surface mount coin cell
« Reply #9 on: November 05, 2013, 08:41:10 pm »
Just had a look in my 287 - some signs of corrosion on the cap
how old is your unit?
 

Online mikeselectricstuff

  • Super Contributor
  • ***
  • Posts: 13741
  • Country: gb
    • Mike's Electric Stuff
Re: Fluke 189 with leaking surface mount coin cell
« Reply #10 on: November 05, 2013, 08:47:19 pm »
About 6 years old I think
Youtube channel:Taking wierd stuff apart. Very apart.
Mike's Electric Stuff: High voltage, vintage electronics etc.
Day Job: Mostly LEDs
 

Offline JohnDeeTopic starter

  • Newbie
  • Posts: 5
Re: Fluke 189 with leaking surface mount coin cell
« Reply #11 on: November 05, 2013, 10:38:26 pm »
To answer BravoD's question below, the supercap on my Fluke189 looks very much like the one in mikeselectricstuff's 287 picture below. 
The original Panasonic supercap (P/N EECEN0F204AK) is still available in the USA and Canada from Digikey, Allied and others but with a min buy qty of 2,000pcs.  May as well buy a new DMM...  However I did a datasheet comparison and ELNA P/N DSK-3R3H204T614-H2L (corresponding to Digikey 604-1147-1-ND at $2.81ea, no min qty) as suggested by tsmith35 below looks like a good candidate for replacement.  I will probably go with that one.
 

Offline ElektroQuark

  • Supporter
  • ****
  • Posts: 1244
  • Country: es
    • ElektroQuark
Re: Fluke 189 with leaking surface mount coin cell
« Reply #12 on: November 05, 2013, 11:01:38 pm »
What does FLUKE say about this?
Warranty?
;)

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #13 on: November 06, 2013, 01:37:08 am »
there is some internal pictures of the 287 here in the forum its hard to find btw.
A huge high res photo of 287's gut -> Fluke 287 DMM Teardown and photos , although its not very clear whether its using Panasonic cap, can't identify the (M) Matsushita logo like Mike's.


Just had a look in my 287 - some signs of corrosion on the cap
About 6 years old I think
Mike, looks like you need to de-solder that cap off, worry it will going to burst out the goo once the container structure collapsed eaten by rust.  |O


My two 287s are manufactured in Jan and April 2013, will open it up to see which brand, its very unlikely they're using Panasonic Goldcap like Mike's since Panasonic already discontinued that for a while.

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Fluke 287 with corroded ultra cap
« Reply #14 on: November 06, 2013, 05:09:59 am »
My 287 is manufactured in April 2013 (Box label) with serial no 23780XXX.

The ultra cap's date code 1204 is also showing corroded sign, not as bad as Mike's though. Definitely not Panasonic (Matsushita brand) with that (M) logo like Mike's.   ???

Does this thing leak ?  Anyone knows what brand is that ?

Dave, this issue should be a good material for a short video blog.

Close up shots attached below.
« Last Edit: November 06, 2013, 01:48:56 pm by BravoV »
 

Offline Tasman

  • Contributor
  • Posts: 40
  • Country: au
Re: Fluke 189 with leaking surface mount coin cell
« Reply #15 on: November 06, 2013, 10:00:18 am »
Just had a look inside my 187 - it does not have one of these capacitors. 
Seems like they are for the logging memory backup.  Glad I didn't cough up the extra for the 189. :)
 

Offline Svuppe

  • Regular Contributor
  • *
  • Posts: 92
  • Country: dk
Re: Fluke 189 with leaking surface mount coin cell
« Reply #16 on: November 06, 2013, 10:16:58 am »
Ahh, good news. I was going to open my 187 tonight, but thanks to you I don't have to.

-Mikael
 

Online amyk

  • Super Contributor
  • ***
  • Posts: 8264
Re: Fluke 189 with leaking surface mount coin cell
« Reply #17 on: November 06, 2013, 11:11:11 am »
The original Xbox had the same problem with its supercapacitor... you should definitely replace it as it can cause rather serious damage to the PCB if it bursts: http://dannygalaga.com/xbox.html

There's also a rather strange paragraph in the datasheet given earlier:
Quote
7. Brown extraction
Brown extraction may occur around the sealing part depending on use conditions. This extraction is electrically nonconductive.
Therefore, it does not infl uence in the electric characteristics.
« Last Edit: November 06, 2013, 11:17:42 am by amyk »
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #18 on: November 06, 2013, 11:16:13 am »
However I did a datasheet comparison and ELNA P/N DSK-3R3H204T614-H2L (corresponding to Digikey 604-1147-1-ND at $2.81ea, no min qty) as suggested by tsmith35 below looks like a good candidate for replacement.  I will probably go with that one.

Actually, with Digikey's shipping, you might as well buy a dozen and sell the remaining caps off on eBay for $6.88 each. ;D
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #19 on: November 06, 2013, 12:42:26 pm »
There's also a rather strange paragraph in the datasheet given earlier:
Quote
7. Brown extraction
Brown extraction may occur around the sealing part depending on use conditions. This extraction is electrically nonconductive.
Therefore, it does not influence in the electric characteristics.

Interesting, I missed that part.

If the brown goo even grew into like Mike's is harmless, then its fine, but at JohnDee's its leaking and ruined the board and surrounding components as he claimed, wish he could share some close up photo here.

Offline Kjelt

  • Super Contributor
  • ***
  • Posts: 6460
  • Country: nl
Re: Fluke 189 with leaking surface mount coin cell
« Reply #20 on: November 06, 2013, 02:35:37 pm »
How about the calibration settings? Are those also going to be affected?
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #21 on: November 06, 2013, 04:21:40 pm »
Fluke 289 cap, board rev 14, cap has been dead for years.

Offline somlioy

  • Regular Contributor
  • *
  • Posts: 128
  • Country: no
Re: Fluke 189 with leaking surface mount coin cell
« Reply #22 on: November 06, 2013, 05:00:07 pm »
My fluke 289
« Last Edit: March 22, 2014, 01:46:07 pm by somlioy »
 

Offline JohnDeeTopic starter

  • Newbie
  • Posts: 5
Re: Fluke 189 with leaking surface mount coin cell
« Reply #23 on: November 06, 2013, 05:30:50 pm »
Just to clarify things up, the supercap on my 189 is just as bad as the what has been shown in the previous pictures.  It has NOT leaked onto the board or on other components.
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #24 on: November 06, 2013, 05:34:20 pm »
How about the calibration settings? Are those also going to be affected?

Nope, cal will be fine.

Offline ModemHead

  • Frequent Contributor
  • **
  • Posts: 731
  • Country: us
  • No user-serviceable parts inside.
    • Mr. ModemHead
Re: Fluke 189 with leaking surface mount coin cell
« Reply #25 on: November 06, 2013, 06:30:55 pm »
Here's what my 189 looks like.  It's a dry brown fuzz on this one.
 

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
Re: Fluke 189 with leaking surface mount coin cell
« Reply #26 on: November 06, 2013, 07:01:26 pm »
fffs i will take mine apart in this weekend. if its leaking can i contact the distributor to get them replace the thing within warranty? as itsd will seriously decrase the life of the instrument
 

Offline uoficowboy

  • Regular Contributor
  • *
  • Posts: 186
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #27 on: November 06, 2013, 07:53:05 pm »
Here's what my 189 looks like.  It's a dry brown fuzz on this one.
That's pretty much identical to the two 189s I looked at today.

Is there any consensus on if these parts need to be replaced or not? Any means of testing them? I would be very sad if my 189 were to die!
 

alm

  • Guest
Re: Fluke 189 with leaking surface mount coin cell
« Reply #28 on: November 06, 2013, 07:58:37 pm »
The note from the datasheet suggests that this behavior is normal for this particular cap and might not indicate its imminent failure.

I just checked a fairly early 189 and it had a different surface mount supercap (with plastic sleeve) that did not show any signs of leakage.
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #29 on: November 06, 2013, 08:19:55 pm »
fffs i will take mine apart in this weekend. if its leaking can i contact the distributor to get them replace the thing within warranty? as itsd will seriously decrase the life of the instrument
From the Fluke warranty page:

Industrial Products Limited Lifetime Warranty
Lifetime is defined as seven years after Fluke discontinues manufacturing the product, but the warranty period shall be at least ten years from date of purchase. *(Lifetime Warranty applies to products manufactured after October 1996). The warranty does not cover manuals, fuses, disposable batteries, damage from neglect, misuse, contamination, alteration, accident or abnormal conditions of operation or handling, including failures caused by use outside of the product's specifications, or normal wear and tear of mechanical components. This warranty covers the original purchaser only and is not transferable. This warranty covers the LCD for 10 years only (state-of-the-art for LCDs). To establish original ownership proof of purchase is required (20, 70, 80, 170, 180 and 280 models).
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #30 on: November 06, 2013, 08:29:31 pm »
My 289's cap was completely shorted.  I'd kinda been wondering why new batteries only lasted 2-3 days before going ka-put   :-//, however not enough to actually spend the time tearing it down!  I've modified mine with a 2.1mm DC jack, it's been powered by a power supply for years sitting on the bench.

Just ordered a new cap from Digikey for $2.50, for 5 minutes of down time I'll be back in business!

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
Re: Fluke 189 with leaking surface mount coin cell
« Reply #31 on: November 06, 2013, 08:31:22 pm »
i know what the warranty states but what is that damn component falls under?
 

alm

  • Guest
Re: Fluke 189 with leaking surface mount coin cell
« Reply #32 on: November 06, 2013, 09:19:14 pm »
I've modified mine with a 2.1mm DC jack, it's been powered by a power supply for years sitting on the bench.

Just ordered a new cap from Digikey for $2.50, for 5 minutes of down time I'll be back in business!
Just note that this defeats most of the meter's protection circuit unless the barrel plug, lead and power supply offer the same isolation as the DMM. If you connect the common lead to mains, the barrel connector and power supply will probably also be at mains potential, something common DC barrel connectors and wiring are not designed to protect from. Not a big deal for separated extra-low voltage (eg. Arduino) circuits, as long as your aware and clearly indicate on the meter that the CAT IV 600 V (or even a CAT I 300 V) rating no longer applies.
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #33 on: November 06, 2013, 09:34:14 pm »
Wait a minute, there's life above 3.3v?     :scared:

All kidding aside, absolutely, this is on my personal bench, otherwise I wouldn't allow it.

alm

  • Guest
Re: Fluke 189 with leaking surface mount coin cell
« Reply #34 on: November 06, 2013, 09:44:43 pm »
Some voltage references go all the way up to 7V! But fair enough, you have to get really creative to get shocked by low voltage DC. I would probably affix a label to the front notifying anyone picking it up of the mod, but I can certainly see the convenience of a wall-powered 289.
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #35 on: November 06, 2013, 10:36:43 pm »
i know what the warranty states but what is that damn component falls under?
??? It's not a battery, so it seems pretty straightforward.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #36 on: November 07, 2013, 01:16:02 am »
My 289's cap was completely shorted.  I'd kinda been wondering why new batteries only lasted 2-3 days before going ka-put   :-//, however not enough to actually spend the time tearing it down!  I've modified mine with a 2.1mm DC jack, it's been powered by a power supply for years sitting on the bench.

Great to know this, thanks for sharing. For sure I will be watching the batteries consumption rate, if its getting significantly faster than it used to.

Other thought, with that shorted cap, just curious if that heated up cap will increase the ambient temp inside the DMM significantly ? Cause the DMM case is quite thick and made from 2 layers of different plastics which is quite effective to seal out the heat from inside.  :-//

Assuming we're not going to chase for warranty from Fluke, on the cap replacement, is there any better alternative brand/type of supercap that won't leak and has much more longer life than just <5  years ? Size wise for the replacement cap is not that critical since there should be enough space.
« Last Edit: November 07, 2013, 01:19:57 am by BravoV »
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #37 on: November 07, 2013, 01:30:10 am »
Great to know this, thanks for sharing. For sure I will be watching the batteries consumption rate, if its getting significantly faster than it used to.

Other thought, with that shorted cap, just curious if that heated up cap will increase the ambient temp inside the DMM significantly ? Cause the DMM case is quite thick and made from 2 layers of different plastics which is quite effective to seal out the heat from inside.  :-//

Assuming we're not going to chase for warranty from Fluke, on the cap replacement, is there any better alternative brand/type of supercap that won't leak and has much more longer life than just <5  years ? Size wise for the replacement cap is not that critical since there should be enough space.

Mine had a phantom draw, I knew it was there just never bothered to find it cause it was bench powered, easy enough to test by measuring standby current.

Since it's only the settings and date/time that is stored by the supercap, reliability really isn't that much of a concern, just replace the cap if and when it goes bad, especially since you can buy like and kind for about $2.50.

Your correct though, there is plenty of room if you'd like to retrofit a higher voltage piece, hopefully with longer life.

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #38 on: November 07, 2013, 02:28:32 am »
Since it's only the settings and date/time that is stored by the supercap, reliability really isn't that much of a concern
Agree, its just I don't like the idea of knowing that the cap is keep growing nasty brown goo, that may leak or burst it's liquid once the container is weakened enough by corrosion caused by that brown stuff.  |O

Possible candidates are CDE Cornell Dubilier type ELDR ultracap (Datasheet) which highlights the long-life feature than ordinary ultracap.



Same capacitance as original cap at 0.2F 3.6V, and cost $2.79 a pop at Digikey -> EDLRD224H3R6C, has two terminals styles like above pics, just abit bigger than original cap and not much, and I think its still manageable like by bending it's legs to fit in the solder pads.

« Last Edit: November 07, 2013, 02:31:59 am by BravoV »
 

Offline true

  • Frequent Contributor
  • **
  • Posts: 329
  • Country: us
  • INTERNET
Re: Fluke 189 with leaking surface mount coin cell
« Reply #39 on: November 07, 2013, 03:46:56 am »
I've repaired a couple of these (that otherwise wouldn't turn on) with 22uF MLCCs. As long as batteries are in the unit it works fine...I think they may lose settings during battery change but it's been a while since I've had to change them, maybe I should check?
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #40 on: November 07, 2013, 08:32:32 am »


Same capacitance as original cap at 0.2F 3.6V, and cost $2.79 a pop at Digikey -> EDLRD224H3R6C, has two terminals styles like above pics, just abit bigger than original cap and not much, and I think its still manageable like by bending it's legs to fit in the solder pads.

Yup, this one should fit no problem, and probably last longer like you said.    :-+

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #41 on: November 08, 2013, 04:50:57 pm »
For anyone wondering, Digikey part#P11068CT-ND is a perfect fit for the original on the 289.

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #42 on: November 09, 2013, 01:43:31 pm »
For anyone wondering, Digikey part#P11068CT-ND is a perfect fit for the original on the 289.

Did you purchase that ? Digikey currently don't carry it anymore for cut tape, and also I can't find the datasheet at Panasonic for this product type, another discontinued series ?  :-//
 
The following users thanked this post: CharlieD

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #43 on: November 09, 2013, 01:48:29 pm »
For anyone wondering, Digikey part#P11068CT-ND is a perfect fit for the original on the 289.

Did you purchase that ? Digikey currently don't carry it anymore for cut tape, and also I can't find the datasheet at Panasonic for this product type, another discontinued series ?  :-//

Sure did:


Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #44 on: November 10, 2013, 04:21:52 pm »
Damn, I guess I was blind when I browse the Digikey.  :palm:

Btw, that Panasonic cap doesn't have any datasheet, and at the DK page its rated with lifetime @ Temp. 500 Hrs @ 60°C.

Just fyi, at the CDE type ELDR ultracap I pointed previously, after the exposure of 2000 h at +85 ºC ,still capable of maintaining ±30% of the initial measured value.

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #45 on: November 10, 2013, 04:25:17 pm »
Just fyi, at the CDE type ELDR ultracap I pointed previously, after the exposure of 2000 h at +85 ºC ,still capable of maintaining ±30% of the initial measured value.

You are correct Sir, I should have put forth a bit more research into retrofitting a better cap, time got the best of me.

I'll be doing this exercise again in the future.    |O

Offline smoothtalker

  • Regular Contributor
  • *
  • Posts: 76
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #46 on: November 12, 2013, 10:13:11 am »
is this leaking cap issue affecting all fluke 289? i'm disappointed with fluke how can this happen to their top of the line modal.

I'm thinking of buying 289, but now i guess i've to stick with 87v. fluke 289 is disappointing.. leaking cap, the ghosting dot matrix lcd, the poor battery life. expensive usb cable, and lack of thermal probe compared to 87v etc.

other than graphing are there advantages for 289 over 87v? ok i can see loz and crest factor calculation.
« Last Edit: November 12, 2013, 10:15:43 am by smoothtalker »
 
The following users thanked this post: FrankE

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #47 on: November 12, 2013, 10:41:44 am »
Really depends on your requirements.  I am not happy with my 289, it's huge and bulky, battery life is not good.  If they made the 289 in a compact bench unit I'd be a happy camper.

Note, there is an available TC probe for the 289, which I do use all the time.

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #48 on: November 13, 2013, 05:15:51 am »
I suspect the decaying cap has something to do with poor battery life ? :-//

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #49 on: November 13, 2013, 11:19:06 am »
I suspect the decaying cap has something to do with poor battery life ? :-//

Very true, but even with a properly working cap the battery life is not too good IMO.  If memory serves the 289 is rated for just under 200 hours, with no back-light use.  Probably fine for light duty use, but in heavy industrial applications your having to change the batteries every 2 week.

Here in the dark depths of a power plant, the auto power off and auto back-light off are always disabled. 

The 287/289 should have come with a rechargeable pack, charging nightly or every few nights would have sealed the deal.
 
The following users thanked this post: FrankE

Online amyk

  • Super Contributor
  • ***
  • Posts: 8264
Re: Fluke 189 with leaking surface mount coin cell
« Reply #50 on: November 13, 2013, 11:43:38 am »
I think officially, they're supposed to leak slightly over their rated lifetime, according to the product datasheet. :-\
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount coin cell
« Reply #51 on: November 13, 2013, 12:01:03 pm »
I think officially, they're supposed to leak slightly over their rated lifetime, according to the product datasheet. :-\

That's disappointing.  As Bravo stated, there is plenty of physical room, heck that meter is HUGE, they should have used a more suitable cap, with hopefully no leakage.

Offline turbo!

  • Regular Contributor
  • *
  • Posts: 89
Re: Fluke 189 with leaking surface mount coin cell
« Reply #52 on: November 13, 2013, 09:02:23 pm »
 :wtf: :wtf:
I opened my up and guess what, its doing it too :(
 

Offline turbo!

  • Regular Contributor
  • *
  • Posts: 89
Re: Fluke 189 with leaking surface mount coin cell
« Reply #53 on: November 13, 2013, 09:05:22 pm »
is this leaking cap issue affecting all fluke 289? i'm disappointed with fluke how can this happen to their top of the line modal.

I'm thinking of buying 289, but now i guess i've to stick with 87v. fluke 289 is disappointing.. leaking cap, the ghosting dot matrix lcd, the poor battery life. expensive usb cable, and lack of thermal probe compared to 87v etc.

other than graphing are there advantages for 289 over 87v? ok i can see loz and crest factor calculation.

The user interface to use the peak min/max in 189 is crap and capacitor test algorithm (for which one of the 87V's patent is for) on 87V is better, but my 189 is plagued with this leaking cap and the 87V is plagued with GSM interference susceptibility.

Great eh?
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #54 on: November 14, 2013, 04:17:31 am »
:wtf: :wtf:
I opened my up and guess what, its doing it too :(


Looks like there are many variations of these Matshusita caps installed at Fluke DMM aren't they ?

Yours doesn't seem to have that brown goo sticking on it. Maybe that black shrinked plastic wrapper is protecting the outer shell ?

Is it dead ?
« Last Edit: November 14, 2013, 04:58:35 am by BravoV »
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16276
  • Country: za
Re: Fluke 189 with leaking surface mount coin cell
« Reply #55 on: November 14, 2013, 04:29:46 am »
I had Panasonic cordless phones with a similar supercap that was used to store data like number lists and such in them Those caps never did that, the only failures were the leads breaking off and the cap rattling around loose in the case. 0.022F 5V cap with through hole leads. I retired the last one ( mine) last year as I had run out of spare parts for it, it did 15 years of work. I had a pile that I used for RF modules, antenna's, microphones, speakers, cases and keypad rubbers. The handset flap latches were weak, so I modified it to use a small magnet ( CDROM lens motor) and a reed switch to replace the switch as well. These were expensive units, costing around $150 each when new.
 

Offline turbo!

  • Regular Contributor
  • *
  • Posts: 89
 

Offline Napalm2002

  • Regular Contributor
  • *
  • Posts: 152
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #57 on: November 27, 2013, 02:04:24 am »
Has anyone come up with an ideal replacement?
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #58 on: November 27, 2013, 02:11:13 am »
Check post #38, although it is not a fully ideal replacement regarding it's size, but check both datasheets, the CDE Cornell Dubilier type ELDR ultracap can be easily soldered with a bit of tweaking, and since the DMM still has plenty of spare space in there, it should not be a problem.

When it comes to long term reliability, it beats that shitty Matshusita cap out of the water, simply no contest here.


Btw, that Panasonic cap doesn't have any datasheet, and at the DK page its rated with lifetime @ Temp. 500 Hrs @ 60°C.

Just fyi, at the CDE type ELDR ultracap I pointed previously, after the exposure of 2000 h at +85 ºC ,still capable of maintaining ±30% of the initial measured value.
« Last Edit: November 27, 2013, 02:21:11 am by BravoV »
 

Offline JohnDeeTopic starter

  • Newbie
  • Posts: 5
Re: Fluke 189 with leaking surface mount coin cell
« Reply #59 on: November 28, 2013, 12:36:46 am »
By the way, I just replaced the supercap on my Fluke 189 and it is now back to life !!  I used the Elna supercap suggested by tsmith35 (reply #7; Digikey PN 604-1147-1-ND) but also bought as a backup the CDE Cornell supercap suggested by BarvoV (reply #38) just in case.  The Elna has the exact same footprint which made the re-soldering a simple task.  As the originator of the initial query, I wish to thank all contributors to this blog for having helped me solve my problem - Cheers !
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount coin cell
« Reply #60 on: November 28, 2013, 12:47:32 am »
By the way, I just replaced the supercap on my Fluke 189 and it is now back to life !!
Thanks for reporting back.  Too often, we never get feedback from the original poster.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #61 on: November 28, 2013, 01:12:52 am »
By the way, I just replaced the supercap on my Fluke 189 and it is now back to life !!
Thanks for reporting back.  Too often, we never get feedback from the original poster.

+1 , thanks too, and this thread will be good resource and its proven, and maybe useful for someone that experiencing the same problem.  :-+

JohnDee, since you're the thread starter, suggesting to update the thread title (edit the 1st post's title) by adding two words "287" and "289", hopefully these will help improve the search result.

Meanwhile, since this is a proven fix, maybe its time to stalk cheap & dead Fluke 189 on ebay ?  >:D

« Last Edit: November 28, 2013, 04:35:36 pm by BravoV »
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount coin cell
« Reply #62 on: November 28, 2013, 01:50:31 am »
Meanwhile, since this is a proven fix, maybe its time to stalk cheap & dead Fluke 189 on ebay ?  >:D
Just remember that the modern Flukes are all closed case calibration.  So dead modern Flukes do have a risk of being uncalibrated after being repaired.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #63 on: November 28, 2013, 01:56:54 am »
Meanwhile, since this is a proven fix, maybe its time to stalk cheap & dead Fluke 189 on ebay ?  >:D
Just remember that the modern Flukes are all closed case calibration.  So dead modern Flukes do have a risk of being uncalibrated after being repaired.

C'mon retiredcaps, I guess probably your finger will be itchy to pull the trigger after seeing an offer for a dead 189 but with good physical for a price say at $50 or less ?  Worth the gamble isn't it ? >:D

I'm pretty sure ModemHead will do it.  :-DD

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount coin cell
« Reply #64 on: November 28, 2013, 04:45:26 am »
C'mon retiredcaps, I guess probably your finger will be itchy to pull the trigger after seeing an offer for a dead 189 but with good physical for a price say at $50 or less ?
:-X  :-X  ;)
 

Offline Napalm2002

  • Regular Contributor
  • *
  • Posts: 152
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #65 on: November 28, 2013, 03:10:09 pm »
I love my 289. I know battery life sucks but I feel that it's worth a few batteries every two weeks to have that much power at your fingertips.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #66 on: November 28, 2013, 03:49:19 pm »
I love my 289. I know battery life sucks but I feel that it's worth a few batteries every two weeks to have that much power at your fingertips.

How old is your 289 ?

Not sure though, maybe the leaking cap is also leaking the current and drains the battery even its turned off ? As we know this 280 series never turned off galvanically by design.  :-//

Offline Napalm2002

  • Regular Contributor
  • *
  • Posts: 152
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #67 on: November 29, 2013, 12:42:06 am »
I bought my 289 used about a year ago. I don't get to use it often and I do not have to replace batteries every two weeks. I was commenting on previous posts about the non desirable battery life. I do feel that no matter how often that I need to replace the batteries that it is worth it to have soo much diagnostic power in te palm of your hand.

Snake dr out...
 

Offline smoothtalker

  • Regular Contributor
  • *
  • Posts: 76
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #68 on: November 30, 2013, 10:51:38 pm »
I'm just wondering, why doesn't anyone consider sending it back for warranty claim?? isn't it lifetime warranty??? I've never sent in my fluke but i purchase them for that reason. LOL

and if everyone send in.. fluke will surely notice this "design problem" and improve?
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #69 on: November 30, 2013, 11:32:34 pm »
I'm just wondering, why doesn't anyone consider sending it back for warranty claim?? isn't it lifetime warranty???
See Fluke's definition in attached picture taken from the 187/189 manual.  The  89IV/189 are discontinued.

In addition, some of us buy used/abused Flukes which do not officially qualify.
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #70 on: December 01, 2013, 01:22:51 pm »
I'm just wondering, why doesn't anyone consider sending it back for warranty claim?? isn't it lifetime warranty???

For me personally?  It took me < a few minutes to fix my 289, vs filing with Fluke, mailing the unit in, waiting for them to fix it, shipping back etc.

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #71 on: December 01, 2013, 01:30:07 pm »
I'm just wondering, why doesn't anyone consider sending it back for warranty claim?? isn't it lifetime warranty???

For me personally?  It took me < a few minutes to fix my 289, vs filing with Fluke, mailing the unit in, waiting for them to fix it, shipping back etc.

+1

Also there is no guarantee that Fluke will replace it with better cap than this problematic one, and even though you still can make the same claim again in next few years, meanwhile you suffer from the high battery drain while the cap is rotting in those period.   :-\

My relatively new 287 with date code April 2013 already shows the early rotting sign at the cap.  >:(

Offline ElektroQuark

  • Supporter
  • ****
  • Posts: 1244
  • Country: es
    • ElektroQuark
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #72 on: December 01, 2013, 02:39:04 pm »
But by repairing aren't you voiding the warranty?

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #73 on: December 01, 2013, 02:46:32 pm »
But by repairing aren't you voiding the warranty?

Of the hundreds that I've deployed, never once a bad Fluke, saving weeks of RMA time is worth the risk IMO.

Offline neslekkim

  • Super Contributor
  • ***
  • Posts: 1305
  • Country: no
Re: Fluke 189 with leaking surface mount coin cell
« Reply #74 on: December 01, 2013, 03:34:33 pm »
I love my 289. I know battery life sucks but I feel that it's worth a few batteries every two weeks to have that much power at your fingertips.

I bought my 289 at an auction in june (was not used), and havent even changed the batteries that came with it, it doesnt get used a lot, but at least, the batteries is not draining..
 

Offline nukie

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #75 on: December 02, 2013, 02:01:33 am »
I have two 187 and a 189 of different manufacture batch. No leaky caps. No battery drain. Battery life has been great with Eneloops. I never have to worry about it.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #76 on: December 02, 2013, 02:03:42 am »
I have two 187 and a 189 of different manufacture batch. No leaky caps. No battery drain. Battery life has been great with Eneloops. I never have to worry about it.

I thought 187 doesn't have that cap ? See post #15.

Offline nukie

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #77 on: December 02, 2013, 02:20:44 am »
Yes I know 187s have no caps, what I mean was the 187s and 189 have similar battery consumption and I have replaced my 189 cap. It wasn't leaky or anything but it sure looks pretty rusty when I  got it last time.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #78 on: December 02, 2013, 02:26:07 am »
Is that the new replaced cap ? I can see there is a (M) Mashusita logo on it, are you using the same type ?

Isn't that Mashusita brand product is no longer produced and replaced by Panasonic brand ?

Offline nukie

  • Frequent Contributor
  • **
  • Posts: 799
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #79 on: December 02, 2013, 02:29:21 am »
I bought it from rockby.com.au many years ago but it seems they don't have it anymore.

I bought some Panasonic FM, FR caps from Digikey recently they are still using the "M" logo. I have some bigger  EC series caps have the full "Panasonic" printed on them but the M logo would save them printing cost on caps that cost a few cents.

I found exact replacement here for $1
http://www.westfloridacomponents.com/J544APK02/0.1uF+5.5V+Memory+Backup+Capacitor+Panasonic+EECS0HD104.html
Disregard the specs on the webpage it's entered incorrectly the part no. ending with 104 is 0.1F

Capacitance: 0.10uF would not be a supercap. It would be a very poorcap.
« Last Edit: December 02, 2013, 02:42:48 am by nukie »
 

Offline smoothtalker

  • Regular Contributor
  • *
  • Posts: 76
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #80 on: December 03, 2013, 04:17:00 am »
Also there is no guarantee that Fluke will replace it with better cap than this problematic one, and even though you still can make the same claim again in next few years, meanwhile you suffer from the high battery drain while the cap is rotting in those period.   :-\

My relatively new 287 with date code April 2013 already shows the early rotting sign at the cap.  >:(

Dave should really do a video on this issue. Fluke needs to do something. It's actually quite a shame to Fluke's reputation that this is happening.
« Last Edit: December 03, 2013, 04:21:32 am by smoothtalker »
 

Offline CSmith

  • Supporter
  • ****
  • Posts: 23
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #81 on: December 03, 2013, 04:51:35 am »
I have the Fluke 189, and I bought the "upgraded" Fluke 289 when it came out. Am I alone in that I find myself preferring the Fluke 189? I believe that I find the boot time of the Fluke 289 to be excessive.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #82 on: December 03, 2013, 04:57:25 am »
Also there is no guarantee that Fluke will replace it with better cap than this problematic one, and even though you still can make the same claim again in next few years, meanwhile you suffer from the high battery drain while the cap is rotting in those period.   :-\

My relatively new 287 with date code April 2013 already shows the early rotting sign at the cap.  >:(

Dave should really do a video on this issue. Fluke needs to do something. It's actually quite a shame to Fluke's reputation that this is happening.

Agree, and the worst part is this problem is affecting Fluke's high-end models which I think deserved more attentions, especially 287 or 289 are belong to top tier models, and they are not cheap.  :(

Already raised this issue to Dave at my post #14, and also PM-ed him, but no reply yet though.

A few more similar PMs from some of you that have the same concern to Dave probably will raise his attention, just a suggestion.

Also I think this issue alone has good and enough materials to make single video out of it, especially involves more details like measuring the off current, ultra cap leak, or ask people who are affected to do the same measurements to gather much larger data and get a clue what is happening and etc.

If I had a chance, I will measure the off state current consumption on my both 287s to see if there are big differences, and hope not.  >:(
« Last Edit: December 03, 2013, 05:02:41 am by BravoV »
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #83 on: December 06, 2013, 05:55:28 pm »
Add Fluke 1653 with potential bad Panasonic gold EN cap to list as per

https://www.eevblog.com/forum/testgear/fluke-1653-installation-tester-repair-help/
 

Offline smoothtalker

  • Regular Contributor
  • *
  • Posts: 76
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #84 on: December 06, 2013, 09:41:00 pm »
Someone needs to tell Fluke about this! I can't find any contact details on Fluke website after they change it.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #85 on: December 08, 2013, 08:19:58 am »
Someone needs to tell Fluke about this! I can't find any contact details on Fluke website after they change it.

Don't think they will give a damn.  :(

If friends ask me about 280 series, looking at current situation, I will tell them to go Agilent straight a away. >:(

Looking at the past experience with 87V model which was vulnerable to GSM phone signal, that Dave raised up while ago, looks like Fluke reacted quickly. Again, I think raising this to Dave and hopefully he will consider to make a video on this once he sees enough people are "whining" about it.  >:D

Offline apelly

  • Supporter
  • ****
  • Posts: 1061
  • Country: nz
  • Probe
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #86 on: December 08, 2013, 09:37:32 am »
Don't think they will give a damn.
Correct.
 

Offline smoothtalker

  • Regular Contributor
  • *
  • Posts: 76
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #87 on: December 08, 2013, 02:27:12 pm »
Someone needs to tell Fluke about this! I can't find any contact details on Fluke website after they change it.

Don't think they will give a damn.  :(

If friends ask me about 280 series, looking at current situation, I will tell them to go Agilent straight a away. >:(

Looking at the past experience with 87V model which was vulnerable to GSM phone signal, that Dave raised up while ago, looks like Fluke reacted quickly. Again, I think raising this to Dave and hopefully he will consider to make a video on this once he sees enough people are "whining" about it.  >:D

Actually Fluke took quite a while over the GSM issue. Don't seem very sincere though. Fluke seems to be becoming more passive in everything recently.

I do have contact of Fluke's Marcom manager. Unfortunately she is not related to engineering. We need your powers Dave! Where are you Dave!
 

alm

  • Guest
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #88 on: December 08, 2013, 03:36:44 pm »
Has anyone actually shipped a meter with leaking cap to Fluke and was refused repair?
 

Offline Nermash

  • Frequent Contributor
  • **
  • Posts: 257
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #89 on: December 09, 2013, 07:29:41 am »
Maybe we should ask Martin from mjlorton channel to do a video about this... He has a 287 which is old enough to demonstrate this issue.

Last time it took Fluke arround 18 months to issue a firmware update for 287 (problem causing shutdown when logging, in presence of high intensity IR source around data port).

This would probably take the same amount of time, and the solution would probably be the same as with the 87V GSM issue: "send it for calibration and we will fix it for free".
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #90 on: December 09, 2013, 02:31:18 pm »
The difficulties, cost and resources needed in fixing issues just by updating the firmware, is nothing when compared to fixing an issue that needs Fluke to replace that sick cap from complaining owners, maybe also unsold products withdrawal that are affected, new cap thorough testing, procurement and etc. 
« Last Edit: December 09, 2013, 02:37:16 pm by BravoV »
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #91 on: December 09, 2013, 11:20:19 pm »
Does someone with a non gmail/yahoo/hotmail (in case Fluke blocks these addresses) address want to try contacting

duane.smith@fluke.com

He is Senior Sale Support at Fluke.  I have seen a lot of his articles.  His linkedin profile.

http://www.linkedin.com/pub/duane-smith/11/434/436
 

Offline neslekkim

  • Super Contributor
  • ***
  • Posts: 1305
  • Country: no
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #92 on: December 10, 2013, 08:28:02 am »
fluke didn't block my gmail when contacting them atleast, I mailed them because i got my 289 with and note to register for getting probes, the link at fluke was dead, but they responded, and month later I got my probes.
 

winluk

  • Guest
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #93 on: January 03, 2014, 07:59:58 am »
Just had a look in my 289 - no corrosion on the cap.
I bought it in mid 2013 but the setup menu / info says it was built 25 Jun 2010 but calibrated 25 Apr 2013.
The board is Rev. 015.

Cheers
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #94 on: January 03, 2014, 08:53:36 am »
Does someone with a non gmail/yahoo/hotmail (in case Fluke blocks these addresses) address want to try contacting

duane.smith@fluke.com
Martin did a video and will send an email to Duane.

 

Offline Huluvu

  • Regular Contributor
  • *
  • Posts: 91
  • Country: de
    • ECM Home
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #95 on: January 03, 2014, 09:19:43 am »
Just watched Martins Video and checked the behavior of my Fluke 189 which shows the same residues on the super-cap surface.
I found this remark in the Datasheet from Panasonic (page 20):

Quote
7. Brown extraction
Brown extraction may occur around the sealing part depending on use conditions. This extraction is electrically nonconductive. Therefore, it does not influence in the electric characteristics.

For myself I will monitor the part periodically if something changed, but it is not that dramatically as described by the OP
i'm more concerned about leaking batteries by storing a meter for a long time period.
"Yeah, but no, but yeah, but no..."
 

Offline mimmus78

  • Supporter
  • ****
  • Posts: 676
  • Country: it
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #96 on: January 03, 2014, 09:50:18 am »
Mine 287 is from 2007 or 2008 and supercap is corroded.

I think this should be changed ...

 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #97 on: January 03, 2014, 09:51:01 am »
Just had a look in my 289 - no corrosion on the cap.
I bought it in mid 2013 but the setup menu / info says it was built 25 Jun 2010 but calibrated 25 Apr 2013.
The board is Rev. 015.
That is weird, what is the current cal count ? If its still 1 (one), then this must be strange, manufactured in 2010 but with 2013 factory cal date is quite long time.  :o

Are you sure you're not mistaken by the bootloader's date at the info screen ?

Also "assuming" the marking at the cap represents the cap's date code, then your cap there is quite new (made at 2013), while at my 287 is manufactured in April 2013 (check post #14), and with the cap's date code at 2012.

Here my 287 info screen and it's box which stated its manufactured at April 2013, and yes, both serial no (blurred here  :P) are identical at the meter and at the box.



« Last Edit: January 03, 2014, 10:01:31 am by BravoV »
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16276
  • Country: za
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #98 on: January 03, 2014, 09:56:40 am »
Bootloader build is the firmware build date, it would have been programmed later than that, and the cal date says it was finally assembled and tested probably the day before and then left running overnight for burn in prior to calibration. AFAIK Fluke uses an ATE bed to do cal, so the date would be the time of manufacture within a week at most before, and the first cal is the factory one.
 

winluk

  • Guest
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #99 on: January 03, 2014, 09:57:44 am »
Quote
That is weird, what is the current cal count ? If its still 1 (one), then this must be strange, manufactured in 2010 but with 2013 factory cal date is quite long time.  :o

Are you sure you're not mistaken by the bootloader's date at the info screen

Calibration Date: 25/04/13
Calibration Counter: 1
Board ID: 3
The rest is same as per your screen...

 |O  I'm DOPE!!!
The 'built' part is for the 2 liner... "ARM bootloader..... built 08:38:05, Jun 25 2010"  :)   
 

Offline mimmus78

  • Supporter
  • ****
  • Posts: 676
  • Country: it
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #100 on: January 03, 2014, 09:58:11 am »
I think this "built" date is relative to the build date of the boot loader (when it was compiled) not to the meter "motherboard".
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #101 on: January 03, 2014, 10:03:44 am »
Quote
That is weird, what is the current cal count ? If its still 1 (one), then this must be strange, manufactured in 2010 but with 2013 factory cal date is quite long time.  :o

Are you sure you're not mistaken by the bootloader's date at the info screen

Calibration Date: 25/04/13
Calibration Counter: 1
Board ID: 3
The rest is same as per your screen...

 |O  I'm DOPE!!!
The 'built' part is for the 2 liner... "ARM bootloader..... built 08:38:05, Jun 25 2010"  :)

Well, just see at the bright side, at least you now have few months to enjoy the shiny ultra cap before its starting to poo out around at the outer shell like mine.   :-DD  :palm:

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #102 on: January 03, 2014, 07:31:47 pm »
after watching Martins video

I had a look in my 289 and it is also having some leakage on the capacitor  :--
Although that looks like an easy to fix issue, I must say, that is not what I expected to see in my Fluke. 
 

Offline Terabyte2007

  • Frequent Contributor
  • **
  • Posts: 527
  • Country: us
  • It is purpose that created us... That defines us..
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #103 on: January 03, 2014, 10:32:32 pm »
Looks like I am cracking open my 289 this weekend!
Eric Haney, MCSE, EE, DMC-D
Electronics Designer, Prototype Builder
 

Offline CSmith

  • Supporter
  • ****
  • Posts: 23
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #104 on: January 03, 2014, 11:17:26 pm »
I opened up two of my Fluke 189's and two of my Fluke 289's, and their supercaps were all in pristine condition despite the passage of time, so some of the supercap's must be up to the challenge. I wonder if it is a supercap manufacturing quality issue such that there were bad batches, or whether there might be environmental influences that help spur them to go bad.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #105 on: January 04, 2014, 05:01:59 am »
I opened up two of my Fluke 189's and two of my Fluke 289's, and their supercaps were all in pristine condition despite the passage of time, so some of the supercap's must be up to the challenge. I wonder if it is a supercap manufacturing quality issue such that there were bad batches, or whether there might be environmental influences that help spur them to go bad.

Interesting, how old are those 189s and 289s ?

Offline CSmith

  • Supporter
  • ****
  • Posts: 23
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #106 on: January 05, 2014, 11:26:01 pm »
I opened up two of my Fluke 189's and two of my Fluke 289's, and their supercaps were all in pristine condition despite the passage of time, so some of the supercap's must be up to the challenge. I wonder if it is a supercap manufacturing quality issue such that there were bad batches, or whether there might be environmental influences that help spur them to go bad.

Interesting, how old are those 189s and 289s ?

By purchase date, the 189's were acquired in 2002/12/05 and 2003/06/05.
By initial calibration date, the 289's were first calibrated in 2008/5/15 and 2009/6/15.
 

Offline mjlorton

  • Contributor
  • Posts: 44
  • Country: us
  • www.YouTube.com/mjlorton
    • Solar Power and Electronic Measurement Equipment
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #107 on: January 07, 2014, 12:43:02 am »
After posting a video on the issue I emailed Duane (Fluke). He has just responded with the following:

On 1/6/2014 6:40 PM, Smith, Duane wrote:> We are aware of the failures of this particular part. We are working with the vendor in hopes of determining why these particular parts are leaking.
> Customers who have units either where the cap has failed or experienced leakage can get their units repaired under warranty at their closest Fluke service facility.
>   Since this part is soldered to the pcb, customer replacement is not recommended as it will void the units warranty.
>
> The capacitor specification says that the capacitor should draw <150 uA after 10 minutes.
> If there were one measurement to take, it would be to measure battery draw with the unit off after applying  the battery voltage for 10 minutes. The
> current draw on failed caps will rise after 10 minutes.
>
> The good news is the supercap is only in the unit to maintain the units date and time. It has no other purpose. It does not support instrument memory or logged /saved readings.
> After we hear back from our component vendor we will determine next steps needed if any but for the short-term customers who own units affected by this part can have it replaced
>   under warranty through Fluke service centers worldwide.
>
> Cheers, Duane

« Last Edit: January 07, 2014, 04:29:29 pm by mjlorton »
www.YouTube.com/mjlorton
Solar Power and Electronic Measurement Equipment
 

Offline ModemHead

  • Frequent Contributor
  • **
  • Posts: 731
  • Country: us
  • No user-serviceable parts inside.
    • Mr. ModemHead
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #108 on: January 07, 2014, 01:06:33 am »
Thanks for looking into this and posting the response.  That explains why I have to reset the clock every time I change batteries, but never lose the logged data.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #109 on: January 07, 2014, 05:43:29 am »
Martin, thanks for the update, really appreciate it !  :-+


>
> The capacitor specification says that the capacitor should draw <150 uA after 10 minutes.
> If there were one measurement to take, it would be to measure battery draw after 10 minutes. The
> current draw on failed caps will rise after 10 minutes.
>

Great info ! Now, I'm going to measure mine.

Curious on the time, why 10 minutes ?     Edit : NVM, found it in the cap datasheet.
« Last Edit: January 07, 2014, 05:45:40 am by BravoV »
 

Offline Huluvu

  • Regular Contributor
  • *
  • Posts: 91
  • Country: de
    • ECM Home
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #110 on: January 07, 2014, 07:06:29 am »
> The good news is the supercap is only in the unit to maintain the units date and time. It has no other purpose. It does not support instrument memory or logged /saved readings.

What about removing the supercap permanently from the Meter?
I rarely use the logging capability.
"Yeah, but no, but yeah, but no..."
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #111 on: January 07, 2014, 07:16:27 am »
> The good news is the supercap is only in the unit to maintain the units date and time. It has no other purpose. It does not support instrument memory or logged /saved readings.

What about removing the supercap permanently from the Meter?
I rarely use the logging capability.

Or maybe replace it with just ordinary lower capacitance cap, but good and reliable one ?

jucole

  • Guest
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #112 on: January 08, 2014, 01:29:04 pm »
Someone posted a link to a Panasonic datasheet with what looked like to be the same super-cap. type;   in the datasheet it specified the max. current draw to be 10uA and just wondered what the draw was on the Flukes and also the operating voltage;  there seem to be 2 types which look very similar ; the EN and EM types but have different properties;  for example the EN has a current limit of 10uA and can be reflowed upto 250c for 5 secs according the information at the start of the document but the EM type seems like it can only be reflowed to 240c, but yet has no max. current draw specified.  It's interesting to note that the specification for the EN type mentioned the reflow of 240c but that conflicts with the information at the beginning of the document.

Did someone get the EN and EM types mixed up maybe?  Or maybe they reflowed the EM type to the EN temperature described at the top of the document?

Looking at Mikes DMM, with the wide tabs it looks like a Type A,  but is that an EN or EM?
« Last Edit: January 08, 2014, 03:20:13 pm by jucole »
 

Offline bernroth

  • Regular Contributor
  • *
  • Posts: 126
  • Country: de
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #113 on: January 08, 2014, 09:52:32 pm »
Just for the statistics: The problem is not only with the multimeters, the Fluke 1653 installation testers are affected too by this problem. I suspect the newer models 1653B/1654B using the same cap!

I recently bought two Fluke 1653 installation testers and while checking them internally I found exactly the same type of surface mount capacitors on the front panel leaking like hell. The testers are not the newest but honestly this must not happen.
I replaced them with two TH supercaps (rated 0.25F@5V) and everything is back working normal.
« Last Edit: January 08, 2014, 09:54:41 pm by bernroth »
 

Offline mjlorton

  • Contributor
  • Posts: 44
  • Country: us
  • www.YouTube.com/mjlorton
    • Solar Power and Electronic Measurement Equipment
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #114 on: January 09, 2014, 01:17:03 pm »
Further feedback from Duane / Fluke:
"As a follow up to this discussion, I've attached the datasheet for the supercap. If you review Note 1 toward the bottom of the page, it makes note of "brown deposited materials found around the sealing area" and the fact that this material will not affect the electrical performance of the part. So, in essence in some situations this leakage could be found as normal under some environmental conditions. Regardless, we plan on still working with the vendor to clarify this further."
www.YouTube.com/mjlorton
Solar Power and Electronic Measurement Equipment
 

jucole

  • Guest
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #115 on: January 09, 2014, 01:30:42 pm »
I've attached the datasheet for the supercap.
Thanks Martin for clarifying the actual part number used.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #116 on: January 10, 2014, 09:21:22 am »
Even though its harmless, is it fair to expect from a high end Fluke's DMM series that use component that doesn't poo ?  :(

Its very disappointing, especially they already knew it from the beginning.  :--

Offline eb4eqa

  • Regular Contributor
  • *
  • Posts: 98
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #117 on: January 10, 2014, 02:36:13 pm »
Hi all,

I found leackage on the top of the cap in my 289 (5 years old). I simply removed it since I don't change batteries that frequently but I do want to prevent further damage to the PCB.

I use eneloops in it and they last for a long time with the use I make of the meter. I am a big fan of those batteries, highly recommended.

Thanks from bringing this up,
Roberto EB4EQA
 

Offline CSmith

  • Supporter
  • ****
  • Posts: 23
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #118 on: January 10, 2014, 04:25:11 pm »
I really do appreciate Fluke's continued use of common AA batteries in their handheld meters, especially compared to Agilent's (Keysight?) use of AAA batteries that do not last very long at all in these more sophisticated meters.
 

Offline ron

  • Regular Contributor
  • *
  • Posts: 87
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #119 on: January 10, 2014, 06:44:26 pm »
What corrosion?  What "corrosive" material is in a cap?  Remember that the original poster thought it was a battery cell.

I just checked my oldest 189 and sure enough the cap had obviously leaked.  I used a tooth pick to remove the most of the dried brown crud (Latin term for dielectric?) and wiped the remaining residue off with a q-tip dampened with IPA.  Looks clean now.  Under a 20x loupe, I see no pitting or "corrosion" on the cap case.   The residue is harmless except it indicates a failed or failing cap.
Yes, this cap has a problem, but it isn't going to corrode the pcb or anything else.
Don't panic.  :)

 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #120 on: January 11, 2014, 11:31:22 am »
What corrosion?  What "corrosive" material is in a cap?  Remember that the original poster thought it was a battery cell.

I just checked my oldest 189 and sure enough the cap had obviously leaked.  I used a tooth pick to remove the most of the dried brown crud (Latin term for dielectric?) and wiped the remaining residue off with a q-tip dampened with IPA.  Looks clean now.  Under a 20x loupe, I see no pitting or "corrosion" on the cap case.   The residue is harmless except it indicates a failed or failing cap.
Yes, this cap has a problem, but it isn't going to corrode the pcb or anything else.
Don't panic.  :)

If cap already oozed out so many of it's internal material, let alone this goo if its harmless, but then this raised few questions :
  • Does the leakage current still at < 150uA as new one ?
  • If the internal leakage is increasing, then how much ? Will the cap suck the battery charge significantly even the DMM is at OFF position, since its basically connected all the time to the battery.
Still don't have time to measure my 287s off current.  :'(

Offline ron

  • Regular Contributor
  • *
  • Posts: 87
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #121 on: January 11, 2014, 08:09:02 pm »

If cap already oozed out so many of it's internal material, let alone this goo if its harmless, but then this raised few questions :
  • Does the leakage current still at < 150uA as new one ?
  • If the internal leakage is increasing, then how much ? Will the cap suck the battery charge significantly even the DMM is at OFF position, since its basically connected all the time to the battery.
Still don't have time to measure my 287s off current.  :'(


The cap apparently failed open(or nearly open), as would be expected.  Current draw in off position is <70uA.  This meter does forget time when batteries are removed -- a non issue for me. 
I checked another 189 and 2 289's and they are OK.  Using enerloops, battery life in all of my X89 meters is much better than spec.
 

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #122 on: January 27, 2014, 05:45:00 am »
Hi,

I'll have to tear into it and check, but I bought a 189 off ebay recently. When it arrived it was dead. Replaced the batteries, and it came right up.
Curious now if the supercap is the reason.
 

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #123 on: January 27, 2014, 05:24:05 pm »
Had the creeping crud...   removed the cap entirely. Cleaned up the mess on the board. Not sure I will be installing a replacement. Never use logging storage. Oh, date/time...  maybe I will replace!
 

Offline true

  • Frequent Contributor
  • **
  • Posts: 329
  • Country: us
  • INTERNET
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #124 on: January 29, 2014, 07:10:56 am »
> The good news is the supercap is only in the unit to maintain the units date and time. It has no other purpose. It does not support instrument memory or logged /saved readings.

What about removing the supercap permanently from the Meter?
I rarely use the logging capability.

Or maybe replace it with just ordinary lower capacitance cap, but good and reliable one ?

I wrote this earlier in the thread, but I did just this and it works fine.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #125 on: January 29, 2014, 07:33:34 am »
Is there any significant change on the battery's life after the replacement ?

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #126 on: January 29, 2014, 04:44:33 pm »
Besides my personal 189, we have one here at work that is used in the RF Lab daily. I inquired as to it's history of battery life, the main user says he thinks it typically lasts over one year. We have not however done an inspection of this one yet . I'm a bit confused on the exact purpose of the SuperCap. Does it's usefullness only come into play during battery swaps, when there is no other voltage present to maintain the memories? Mine is without one now, but if I just power off, and back on, it still seem to retain memory.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #127 on: January 29, 2014, 04:51:50 pm »
That cap's purpose is just to retain the dmm clock during batteries replacement, according to Fluke.

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #128 on: January 30, 2014, 05:07:00 pm »
Thanks...   that's good to confirm,  I was coming to that conclusion. I think I'll just leave it out. Obviously it caused more problems for Fluke, than it solved.
 

Offline tony3d

  • Frequent Contributor
  • **
  • Posts: 393
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #129 on: January 30, 2014, 06:31:15 pm »
According to CDI The life expectancy seems to be 2.8 years. That is dependent on several factors, one humidity must remain below 50%. Kinda strange for a meter meant to last 10 years!
 

Offline stuartk

  • Regular Contributor
  • *
  • Posts: 102
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #130 on: February 13, 2014, 04:11:03 pm »
I cracked the case on my 289 purchased in 2007 and found the leaking capacitor. It was giving me intermittent loss of timekeeping.

I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

They have replaced the cap, changed my scuffed up battery cover, changed the batteries and re-calibrated it, with a certificate and documentation.

All for free.  :)

Kudos to Fluke Canada and thank you.

Stuart
 

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #131 on: February 13, 2014, 04:20:54 pm »
Has anyone found the Fluke part number for this capacitor?   When one has a meter repaired under warranty, is this information given?  I'd expect not, but since these aren't available in the US in quantities less than 2000, I was considering ordering from Fluke.  However, Fluke will do no parts research - one has to know the part number.

 

Offline stuartk

  • Regular Contributor
  • *
  • Posts: 102
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #132 on: February 13, 2014, 08:39:54 pm »
No they didn't. They only said replaced capacitor C145...
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #133 on: February 13, 2014, 08:46:50 pm »
Has anyone found the Fluke part number for this capacitor?

Search back a few pages, I replaced mine with a $2 part from Digikey.

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #134 on: February 13, 2014, 08:56:47 pm »
Has anyone found the Fluke part number for this capacitor?

Search back a few pages, I replaced mine with a $2 part from Digikey.

I saw that.  It looks like that's the only alternative for a non-warranty unit.

My 189 has the larger capacitor pictured earlier in this thread.  It seems to be OK, but I didn't unsolder it to look underneath.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #135 on: February 14, 2014, 02:59:06 am »
I cracked the case on my 289 purchased in 2007 and found the leaking capacitor. It was giving me intermittent loss of timekeeping.

I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

They have replaced the cap, changed my scuffed up battery cover, changed the batteries and re-calibrated it, with a certificate and documentation.

All for free.  :)

Kudos to Fluke Canada and thank you.

Stuart

Stuart, closeup photo of the newly replaced cap please.

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #136 on: February 14, 2014, 04:34:29 am »
I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

All for free.  :)
Hi Stuart,

Are you the original owner of the meter and did you purchase it from an authorized dealer?

edit: The other advantage of sending it back to Fluke for repair, especially if you owned the meter for a few years, is that you might get calibration thrown in for free.
« Last Edit: February 14, 2014, 04:43:35 am by retiredcaps »
 

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #137 on: February 16, 2014, 06:13:30 pm »
I cracked the case on my 289 purchased in 2007 and found the leaking capacitor. It was giving me intermittent loss of timekeeping.

I sent it off to Fluke Canada. Apparently they have a lifetime warranty up here in the frozen north.

They have replaced the cap, changed my scuffed up battery cover, changed the batteries and re-calibrated it, with a certificate and documentation.

All for free.  :)

Kudos to Fluke Canada and thank you.

Stuart

Stuart, closeup photo of the newly replaced cap please.


Probably would void the calibration :)
 

Offline stuartk

  • Regular Contributor
  • *
  • Posts: 102
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #138 on: February 17, 2014, 12:38:48 am »
Yep,

I'm the original owner. I was going to crack the case, but they put the calibration sticker and a tamper sticker over the case.

I think they were trying to send a message to me after I sent them a picture of my leaking cap! ;D

I suspect I might mess up the calibration if I crack it in anyhow....

Stuart
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #139 on: February 17, 2014, 05:20:33 am »
I'm the original owner. I was going to crack the case, but they put the calibration sticker and a tamper sticker over the case.
Well, thanks for reporting back.  I wouldn't break the seals either if it were my meter.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #140 on: February 17, 2014, 08:17:14 am »
+1 , thanks for the update.

I guess we'll have to wait at least until next year to be able to peek the replaced cap in Stuart's 289.  :P

Offline SoundTech-LG

  • Frequent Contributor
  • **
  • Posts: 788
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #141 on: February 18, 2014, 01:28:56 am »
One, or two things seem pretty certain. One funky Super Cap does not ruin the Fluke reputation. Fluke equipment holds it's value, particularly the handhelds. They get re-sold for quite respectable sums on a regular basis. This, in contrast to the junk meters that you toss in the trash.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #142 on: February 18, 2014, 02:00:14 am »
Agree, apart from the ultra-cap problem, have to admit this 28x series build quality is excellent.

I was very impressed when seeing Martin tested his 287 in freezing cold in the chill box, the measurement result was still spot on with it's spec at operating temp -20 °C to 55 °C.  :-+
« Last Edit: February 18, 2014, 02:05:38 am by BravoV »
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #143 on: February 18, 2014, 12:41:27 pm »
hi guys..
i have a 1 year old fluke287 and it consumes battery much faster..i opened the case and i saw the same cap corroding.. i wonder if a 0.06F supercap i have could be a good replacement?
Always learn how to break and fix things electronics!
 

Offline kaz911

  • Super Contributor
  • ***
  • Posts: 1052
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #144 on: February 27, 2014, 09:17:16 am »
hi guys..
i have a 1 year old fluke287 and it consumes battery much faster..i opened the case and i saw the same cap corroding.. i wonder if a 0.06F supercap i have could be a good replacement?

Don't fix them. I just handed a 289 and a 287 into Fluke and they will be repaired on warranty despite being quite old :) Fluke has "Limited World Wide" warranty on them - and they have a list of serial numbers for which repair is free of charge. As I only paid $200 for my 289 since it was not perfect - I'm now very happy that has been fixed on warranty as well.

And to quote the Calibration girl - 287/289's are really well made apart from the Cap issue - so they are usually in spec/calibration even after 6+ years - they have so far not seen a working 287/289 that was NOT within spec if it had not been "mishandled"
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #145 on: February 27, 2014, 09:45:12 am »

Don't fix them. I just handed a 289 and a 287 into Fluke and they will be repaired on warranty despite being quite old :) Fluke has "Limited World Wide" warranty on them - and they have a list of serial numbers for which repair is free of charge. As I only paid $200 for my 289 since it was not perfect - I'm now very happy that has been fixed on warranty as well.

And to quote the Calibration girl - 287/289's are really well made apart from the Cap issue - so they are usually in spec/calibration even after 6+ years - they have so far not seen a working 287/289 that was NOT within spec if it had not been "mishandled"

well the problem is i cant find any fluke service center here in Philippines :--..so im thinking the best thing to do is to replace the cap myself just like others did..i love my fluke but i hate the way it consumes the battery :palm:
« Last Edit: February 27, 2014, 09:47:44 am by elex_enthusiast »
Always learn how to break and fix things electronics!
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #146 on: February 27, 2014, 10:36:28 am »
well the problem is i cant find any fluke service center here in Philippines :--..so im thinking the best thing to do is to replace the cap myself just like others did..i love my fluke but i hate the way it consumes the battery :palm:

Are you sure the cap is bad, the 28x is a battery hog even with a good cap? 

Mine was drawing something like 10mA when powered off, if you have the means you might test yours.  If you look back in this thread there is a spec, in the uA is what it should draw when working correctly.

Offline kaz911

  • Super Contributor
  • ***
  • Posts: 1052
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #147 on: February 27, 2014, 10:53:05 am »

Don't fix them. I just handed a 289 and a 287 into Fluke and they will be repaired on warranty despite being quite old :) Fluke has "Limited World Wide" warranty on them - and they have a list of serial numbers for which repair is free of charge. As I only paid $200 for my 289 since it was not perfect - I'm now very happy that has been fixed on warranty as well.

And to quote the Calibration girl - 287/289's are really well made apart from the Cap issue - so they are usually in spec/calibration even after 6+ years - they have so far not seen a working 287/289 that was NOT within spec if it had not been "mishandled"


well the problem is i cant find any fluke service center here in Philippines :--..so im thinking the best thing to do is to replace the cap myself just like others did..i love my fluke but i hate the way it consumes the battery :palm:

I handed mine into a distributor - not Fluke direct. So any distributor can verify warranty recalls

Mine has been off to Holland for repair as the local guys does not do any repairs. All at no cost to me.
 

Offline kaz911

  • Super Contributor
  • ***
  • Posts: 1052
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #148 on: February 27, 2014, 10:56:22 am »
and the quick way to check CAP problem - put in new batteries - check time and date is fine. Leave on for a couple of minutes. Shut it down. Remove the batteries - count to 10 - plug batteries back in. If it asks for time/date - you have a super-cap issue.
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #149 on: February 27, 2014, 12:20:05 pm »
yes it does ask to reset the time/date..
Mine was drawing something like 10mA when powered off, if you have the means you might test yours.

unfortunately i have no other reliable meter i can use to measure its current draw but im pretty sure the cap is bad..
« Last Edit: February 27, 2014, 01:48:10 pm by elex_enthusiast »
Always learn how to break and fix things electronics!
 

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #150 on: February 27, 2014, 01:03:47 pm »
Just had a look in my 287 - some signs of corrosion on the cap

Great.  I tried to order the Elna part to replace my capacitor (small type, like Mike's picture above) and Digi-Key and Mouser are  out of stock.

The Panasonic part is only available in quantities of 2000 or more.

Are those the only choices?

For the larger cap used in the 189 and 89-V,  is the parts situation any better?    I may have missed it, but I haven't seen any recommended part numbers to replace this one.

Do I need to unsolder it and check underneath, or would leakage be visible from the top?

Maybe I'll make a Wiki page about this to summarize all the information and ask Dave to add it.
« Last Edit: April 02, 2014, 02:00:13 am by Excavatoree »
 

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #151 on: February 27, 2014, 03:00:14 pm »
Have the same problem.
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #152 on: February 27, 2014, 07:10:33 pm »
Great.  I tried to order the Elna part to replace my capacitor (small type, like Mike's picture above) and Digi-Key and Mouser are  out of stock.

Just get this instead: http://www.digikey.com/product-detail/en/DSK-3R3H224U-HL/604-1020-1-ND/970232

Very slightly taller, but with 0.22F instead of the original 0.20F, and same footprint as the original.
 

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: Fluke 189 with leaking surface mount coin cell
« Reply #153 on: February 27, 2014, 08:00:15 pm »
Great.  I tried to order the Elna part to replace my capacitor (small type, like Mike's picture above) and Digi-Key and Mouser are  out of stock.

Just get this instead: http://www.digikey.com/product-detail/en/DSK-3R3H224U-HL/604-1020-1-ND/970232

Very slightly taller, but with 0.22F instead of the original 0.20F, and same footprint as the original.


Why not this?

http://export.farnell.com/taiyo-yuden/pas414hr-va5r/cap-super-60000uf-3-3v-smd/dp/2112777

 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #154 on: February 28, 2014, 01:19:45 am »
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #155 on: February 28, 2014, 02:04:59 am »
Are those the only choices?

Whats wrong with the CDE cap that I proposed earlier at post #38 ? DK part no 338-3477-ND and its available.

Panasonic cap is not even in the same league when it comes to reliability and long term life compared to this one.

Though you will need to bend it's pin to fit in the foot print, but not a big deal imo.
« Last Edit: February 28, 2014, 02:09:06 am by BravoV »
 

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Re: Fluke 189 with leaking surface mount coin cell
« Reply #156 on: February 28, 2014, 03:47:31 am »
Are those the only choices?

Whats wrong with the CDE cap that I proposed earlier at post #38 ? DK part no 338-3477-ND and its available.

Panasonic cap is not even in the same league when it comes to reliability and long term life compared to this one.

Though you will need to bend it's pin to fit in the foot print, but not a big deal imo.

Ah, I forgot.  That's one reason I thought it might be better to put a summary of such information in one place.  When it's spread out over several pages in a thread, things can get lost.
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount coin cell
« Reply #157 on: February 28, 2014, 04:04:29 am »

60000uf is only .06F, about 30% as much capacitance as the original.
[/quote]

will it affect the performance of the meter? i have the same cap stocked on my bin and i thought of using it to replace the bad cap on my meter..
Always learn how to break and fix things electronics!
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount coin cell
« Reply #158 on: February 28, 2014, 05:03:08 am »
will it affect the performance of the meter?

No, that cap's purpose is just as a backup power for the dmm's clock & calendar during the batteries swapping period.

Not a very serious problem imo, especially you rarely use the logging feature.

Just make sure the new replacement cap has low leakage, other wise it will drain the batteries power even its powered off, as the example of the original cap does when its deteriorated.  :-[

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #159 on: February 28, 2014, 05:46:52 am »
I think it’s not a problem to leave it out, if you don’t need the clock and calendar when swapping batteries.
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Re: Fluke 189 with leaking surface mount coin cell
« Reply #160 on: February 28, 2014, 06:05:44 am »
will it affect the performance of the meter?

No, that cap's purpose is just as a backup power for the dmm's clock & calendar during the batteries swapping period.

Not a very serious problem imo, especially you rarely use the logging feature.

Just make sure the new replacement cap has low leakage, other wise it will drain the batteries power even its powered off, as the example of the original cap does when its deteriorated.  :-[

yeah i seldom use its logging feature..i only use the meter to measure parameters of the projects im working in my bench..i just want to be sure that replacing the bad cap with a lower capacitance wont affect the performance of my meter since i dont have any reference service manual..does anyone here did the same thing??

i was wondering why fluke didnt use a standard CMOS batterry instead of those shitty caps  to back-up the date/time of the meter :-//.. does it makes any difference?
Always learn how to break and fix things electronics!
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #161 on: February 28, 2014, 10:36:26 am »
The cap has no effect on meter performance.  It only holds the date/time/backlight settings, no cal data.

Offline t95mwp

  • Newbie
  • Posts: 2
  • Country: us
Hi folks,I replaced mine capacitor on 189 after it died and never come back alive. Anything else known wrong with this meter,yes I installed capacitor right and new fresh batteries  :). Thanks
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
The cap has no effect on meter performance.  It only holds the date/time/backlight settings, no cal data.
indeed.. i have replaced mine with a 0.06F bnew cap and it worked fine..thanks for the advice..this time the date/clock remains after i remove the battery for 10-15 sec...
Always learn how to break and fix things electronics!
 

Offline elex_enthusiast

  • Regular Contributor
  • *
  • Posts: 108
  • Country: ph
  • Everything has user-salvageable parts inside!!!!
Hi folks,I replaced mine capacitor on 189 after it died and never come back alive. Anything else known wrong with this meter,yes I installed capacitor right and new fresh batteries  :). Thanks

maybe the supercap is not an issue with your dmm..have you checked the board..?
Always learn how to break and fix things electronics!
 

Offline Marvin

  • Contributor
  • Posts: 45
  • Country: ee
This is interesting - I received my brand new 289 with manufacturer date of 30. Sept 2013 and it has calibration count of 2! It was not 100% official channel but a reliable seller from eBay and it really looks like it was brand new in box - and the box serial/manufacturing date stickers and all match. Could it be that it was manufactured on that date and then when the capacitor issue became public a week later they already refitted the fix in it or this is just some strange coincidence it coming out of factory with 2 calibrations? Can't open it up to see the soldering on the capacitor as the case is closed with a tamperproof sticker and I don't want to invalidate it.

Got to love friends who visit USA once and a while and agree to bring stuff over the pond in their bags :)
 

Offline t95mwp

  • Newbie
  • Posts: 2
  • Country: us
Hi folks,I replaced mine capacitor on 189 after it died and never come back alive. Anything else known wrong with this meter,yes I installed capacitor right and new fresh batteries  :). Thanks

maybe the supercap is not an issue with your dmm..have you checked the board..?

It was bad definitely,it started with would not turn on all the time and by changing batteries would fix the problem for the while then  finally died.Board is super clean no damage of leaking and I installed new capacitor that was purchased form digi-key and recommended here on the board.
 

Offline smoothtalker

  • Regular Contributor
  • *
  • Posts: 76
  • Country: us
This is interesting - I received my brand new 289 with manufacturer date of 30. Sept 2013 and it has calibration count of 2! It was not 100% official channel but a reliable seller from eBay and it really looks like it was brand new in box - and the box serial/manufacturing date stickers and all match. Could it be that it was manufactured on that date and then when the capacitor issue became public a week later they already refitted the fix in it or this is just some strange coincidence it coming out of factory with 2 calibrations? Can't open it up to see the soldering on the capacitor as the case is closed with a tamperproof sticker and I don't want to invalidate it.

Got to love friends who visit USA once and a while and agree to bring stuff over the pond in their bags :)

your best bet is to check with fluke directly! email them the serial number.

Anyway i don't think the super capacitor affects the cal. it's job is just to retain the dates etc as mentioned previously. 
 

Offline Marvin

  • Contributor
  • Posts: 45
  • Country: ee
If you open the case it is instant calibration void (technically that is, of course if you have a new sticker etc you can close it and put a new sticker on it) - that's what I'm thinking - the manufacturing date on the box match the calibration date in the unit but the count is 2. Maybe they opened it before sending it out and fixed something? So it is officially factory calibrated but 2 times.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Marvin, an intriguing new Fluke 289 with 2 cal counts.

Any chance taking the photo of the tamper proof seal, or does it look similar to this example here, and also whats the firmware version it has in there ?
« Last Edit: April 01, 2014, 04:59:26 pm by BravoV »
 

Offline Marvin

  • Contributor
  • Posts: 45
  • Country: ee
Here you go. I  bought the cable too and changed the owner name if you are wondering if it came like that - nope.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Marvin, thanks for the photos, looks legit to me, and its also has the latest firmware too.  :-+

Pretty sure its also came with the Fluke's factory calibration paper that has the exact date printed as in your meter info, right ? Example from my 287 -> HERE.

Well, at least you still have quite some time before the cap start to poo  :P since its quite new, congratulation, enjoy that good DMM.

PS : Suggesting NOT to use alkaline cells as it will leak as it happened to mine  >:(, good NiMH cells work just fine like Eneloop.

Offline Marvin

  • Contributor
  • Posts: 45
  • Country: ee
Yes, just like that. Only added LoZ and LoOhm measurements. Date/serial matches.
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
Interesting that yours shows 96% memory like other people reported. Mine shows 100% but I think I have the supercap problem since it loses the date and time each time I change batteries. It also drains batteries fast until it shows only two bars and then stays there (have eneloops now). I don't know if a problem supercap increases current drain :(
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
I don't know if a problem supercap increases current drain :(

Yes, a shorted cap will contribute to battery drain.  Mine would be completely dead in about a week without use on Alkaline batts.

Offline TVman

  • Frequent Contributor
  • **
  • Posts: 260
  • Country: us
  • Life is A bunch of people staring at computers. ;)
You need to replace that capacitor! :scared:
Otherwise it will be  :-BROKE forever!
Yeah, I play Minecraft!
But I'm on here more because I learn more. :D
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
dr Diesel,
Did you repair it yourself or sent it to Fluke?
I purchased mine second hand.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
It was bad definitely,it started with would not turn on all the time and by changing batteries would fix the problem for the while then  finally died.
Check the battery contact strips to make sure they are making contact with the pcb.  You might have bent them while putting the back cover on?

See modemhead's pictures here.

http://mrmodemhead.com/blog/wp-content/gallery/f189_00/F189_026.JPG

If they seem okay, see if the PCB is getting any voltage from the 4 AA batteries?
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
dr Diesel,
Did you repair it yourself or sent it to Fluke?
I purchased mine second hand.

Yup replaced myself, no need to wait weeks/months for Fluke and a $2 part, took 2 minutes.

Disclaimer, you'll likely void your warranty if there is any left.  If you search back on this thread I posted the Digikey part number I used.

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
Got RMA for my Fluke 289 today. I'm shipping to Everett Washington tomorrow. They said that turnaround is 5 days, I can live with that. Never sent a unit to Fluke before, I want to see what happens.
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Casinada, units I've sent in to them in the past were cleaned and calibrated. They have pretty nice service.
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Casinada, what type of damage did you claim at the RMA declaration ? Just bad cap ?

Too bad its sealed, really curious what type of cap Fluke used for the replacement.  ::)


Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
I sent the Fluke 289 on April 2nd and they shipped it back on April 17th. I should be getting it back sometime next week.
I called and told them that the each time I changed the batteries I had to reset the time and date and that it was draining the batteries faster than normal otherwise the DMM was working great. It must be something very common as they didn't ask any questions and generated an RMA. Originally they told me that the turnaround was 5 business days but when I called Tech support on Tuesday they told me that the backlog was around four weeks. I'll report further when I get the unit back. So far the service has been top notch. :)
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
I got the Repaired Fluke 289 today:
Service detail:
Problem description: Does not hold time, drains battery.
Repair description: Diagnosis: confirmed the problem
Replaced C145, Ready for cal

The calibration was verified with a Fluke 5520A Calibrator and found in tolerance.
They sent a calibration certificate but without values :(
Overall great service  :-+
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
Well, C145 is the supercap, so that solves that.



Looks like you got a nice turnaround time. Glad to hear it worked out for you.
 

Offline Jay_Diddy_B

  • Super Contributor
  • ***
  • Posts: 2733
  • Country: ca
Hi,

I can report two DIY fixes of Fluke 289s.

Unit 1


Symptoms: unit works fine, all reading correct, kills the batteries in a few days, even if switched off.

Findings:
Measured current consumption around 30 mA when off. Removed the CAP C145, off current dropped to 45uA

Connected C145 to power supply, set 3.3V  with mA meter in series. Test confirmed bad supercap.

Fix: Replaced the super capacitor with Panasonic Coin Super Capacitor 0.33F 5.5V EEC-S0HD334H

(This is larger than the original but it can be made to fit)


Unit 2


Symptoms: unit works fine, all reading correct, loses time and data if batteries are removed for about 10 minutes

Findings:
Measured current consumption around 2 mA when off. Removed the CAP C145, off current dropped to 45uA

Connected C145 to power supply, set 3.3V  with mA meter in series. Test confirmed bad supercap.

Fix: Replaced the super capacitor with Panasonic Coin Super Capacitor 0.33F 5.5V EEC-S0HD334H

(This is larger than the original but it can be made to fit)

I think that Panasonic was optimistic in rating the original part at 3.3V. Typical values are 2.7V for a single capacitor and 5.5V for the parts that internally have two capacitors in series.

Unit 3

This is a unit that I purchased new and has very few hours of use. The current consumption of this meter was 45uA when off.

Regards,

Jay_Diddy_B







 
The following users thanked this post: Marco1971

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
Good job Jay.
Why didn't you send them to Fluke? At least they do a Calibration verification after the repair. :)
 

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
... current consumption around 30 mA when off.

Wow .. 30mA !!!!  :o

Those darn cap are sucking the batteries power so bad even the meter is turned off, I guess they will be sucked dried in days.

Btw, did you measure the capacitance of those goo oozing bad caps ? How much left ?

JDB, thanks for reporting this.  :-+

Offline xwarp

  • Frequent Contributor
  • **
  • Posts: 367
  • Country: us
Fix: Replaced the super capacitor with Panasonic Coin Super Capacitor 0.33F 5.5V EEC-S0HD334H

(This is larger than the original but it can be made to fit)


What had to be done for the new cap to be made to fit?

If this cap is nothing more than being used to keep the time in memory for the time it takes to replace batteries, then can a lower voltage be used if the size is closer to the original?
 

Offline WVL_KsZeN

  • Regular Contributor
  • *
  • Posts: 92
  • Country: nl
Never use a lower voltage, only a lower capacity. Actually, you want to use a cap with a higher voltage to make it last longer.
 

Offline Jay_Diddy_B

  • Super Contributor
  • ***
  • Posts: 2733
  • Country: ca
Good job Jay.
Why didn't you send them to Fluke? At least they do a Calibration verification after the repair. :)

Several reasons:

1) I am not the original owner. The new one which has not failed (yet?) is mine from new. One of the used units is mine the other belongs to a friend.

2) I want to see if the meters had the problem reported here.

3) The super capacitor does not impact calibration.

4) It was quicker to fix than return to Fluke.

5) If Fluke replaced C145 with the same part, the problem will return.

6) I can do a performance check, I can't call it a calibration because my equipment has not been calibrated recently. I have the following:

Fluke 5101B - Multifunction calibrator
Fluke 732A - DC Voltage standard
Datron 1281 - 8.5 digit meter

Fluke would like to find them out of calibration, so they can charge you for a calibration.

7) I have no reason to believe that the meters were out of calibration.

To xwarp

I will post a picture the next time I have it apart. It is quite easy, you trim the leads and solder the thru-hole part to the SMD pads.

To BravoV,

I did not measure the capacitance of the bad caps, they was so much leakage current it would be hard to do. If I had to, I would measure the discharge time with a resistive load. But these caps were dead, expired, permanently damaged, useless, beyond hope...  ;D

I think that this is correct:

 :-DMM  :-DMM for two fixed 289s  :-+

Regards,

Jay_Diddy_B
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
Thank you for Replying Jay,
I purchased mine used and they never asked me anything but the serial number of the unit. As soon as I told them the problem they issued an RMA.
I'm sure they know what is going on with the supercap so they're probably using a better part (don't know because it has the stickers now so I don't want to open it). Every time they open an instrument for repair they do a calibration if necessary. On my 289 they plugged it to a 5520A and it was still within specifications. I wonder how they test temperature and capacitance.
I'm trying to repair a Fluke 5102B...progress is very slow. It would help to have the extender boards. There are lots of revisions of this calibrator :(
I try to repair everything myself but in this case sending the unit to Fluke was part of the experience :)
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
I wonder how they test temperature and capacitance.

If you look in the calibration manual (page 16-17) you find

if you do not have a calibrator you can just look at a Type K data table and try to simulate the Input Voltages:
0°C = 0V
100°C = 4,09623mV
1000°C = 41,27561mV

and for capacitance verification they only check one value (5nF see page 19)
« Last Edit: May 14, 2014, 01:07:39 pm by quarks »
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
 

Offline Excavatoree

  • Frequent Contributor
  • **
  • Posts: 901
  • Country: us
Thank you for Replying Jay,
I purchased mine used and they never asked me anything but the serial number of the unit. As soon as I told them the problem they issued an RMA.

I doubt they'd repair my 189-II, however.  It's pretty obvious I'm not the original owner.  I suppose I could get a 287 LCD mask for it, but I suspect the serial number is flagged as well.   

However, I could be wrong.  Has anyone tried it?
 

Offline casinada

  • Frequent Contributor
  • **
  • Posts: 599
  • Country: us
Excavatoree,
Just Call and open a Ticket. They will ask for your name, address, email, serial number.:)
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Excavatoree,
Just Call and open a Ticket. They will ask for your name, address, email, serial number.:)
@Excavatoree, identify yourself as the owner of the staircase Fluke photo.  That might buy you some goodwill?
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
@Excavatoree, identify yourself as the owner of the staircase Fluke photo.  That might buy you some goodwill?
Wow, just looked it up... stairway to heaven? ;)
 

Offline malgailany

  • Newbie
  • Posts: 2
  • Country: us
Excavatoree,

I wrote my experience with fluke, check it out
http://mjlorton.com/forum/index.php?topic=531.msg4429#msg4429

Cheers!
M. AlGailani
 

Offline tsmith35

  • Frequent Contributor
  • **
  • Posts: 265
  • Country: us
I wrote my experience with fluke, check it out
http://mjlorton.com/forum/index.php?topic=531.msg4429#msg4429

Wow, that's pretty cool. Looks like a new meter. :)
 

Offline ivaylo

  • Frequent Contributor
  • **
  • Posts: 661
  • Country: us
Just got mine back too. Same thing, back cover replaced, calibrated, they even put fresh batteries. Haven't looked inside yet, but now these stories about replacing PCBs and all are tempting me to peel the seal... Very good experience over all!
 

Offline malgailany

  • Newbie
  • Posts: 2
  • Country: us
I'm interested in knowing the differences between board ID1 and board ID3. I took some photos to my old board (ID1) before I send it to service.
I don't want to peel my sticker to look inside. It would be nice if someones share photos for board ID3.
M. AlGailani
 
The following users thanked this post: bc888

Offline erazor

  • Newbie
  • Posts: 7
I have an issue with the clock drifting in my Fluke 289, see here: https://www.eevblog.com/forum/testgear/fluke-289-clock-drift/
User retiredcaps suggested this might be related to this capacitor issue. To see if this is in fact the case, maybe a user could check behaviour before and after capacitor replacement. Thanks.
 
The following users thanked this post: Marco1971

Offline Crazyasianskillz

  • Newbie
  • Posts: 3
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #203 on: December 25, 2014, 10:11:49 pm »
Sorry to start up an old thread, but I'm considering purchasing a new fluke 287 from an authorized dealer. Does anyone know if fluke has resolved the leaky cap issue in 2014 builds?
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #204 on: December 25, 2014, 10:35:34 pm »
Does anyone know if fluke has resolved the leaky cap issue in 2014 builds?

No change or resolution has been reported.

Offline Crazyasianskillz

  • Newbie
  • Posts: 3
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #205 on: December 26, 2014, 04:01:23 pm »
I gave fluke tech support a call this morning inquiring about this issue. They claimed that this issue had been handled and that newer meters should not have this issue. Can anyone confirm this? I'm going to go ahead with the purchase. Im planning to use this for microcontroller development and msic tasks around the house and garage.
 

Offline JoeO

  • Frequent Contributor
  • **
  • Posts: 527
  • Country: us
  • I admit to being deplorable
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #206 on: December 26, 2014, 04:36:33 pm »
I gave fluke tech support a call this morning inquiring about this issue. They claimed that this issue had been handled and that newer meters should not have this issue. Can anyone confirm this? I'm going to go ahead with the purchase. Im planning to use this for microcontroller development and msic tasks around the house and garage.
What did Fluke mean by "newer meters".  Call them back and ask for the serial number that the fix was put in.
Then when you buy a meter, ask them for the serial number of the meter. 
There is no other way to guarantee that you will receive a "fixed" meter unless you do this.
The day Al Gore was born there were 7,000 polar bears on Earth.
Today, only 26,000 remain.
 

Offline Crazyasianskillz

  • Newbie
  • Posts: 3
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #207 on: December 26, 2014, 05:56:42 pm »
I asked when the changed was implemented but couldn't get any additional information. The only description that was given was that it was "handled and should not be an issue anymore". However I received a follow up email from info-fluke@fluke.com claiming it was the first time the tech had heard of the leaking capacitor issue and had no database record. The disparity in the responses is a little concerting.

I'll do a check to see what the current draw is when the unit is powered down.From what I've seen it appears the nominal draw should be in the micro amp range, if it's in the milli range then the unit would most likely be affected. I'm ordering the unit off of amazon prime so hopefully they stock newer built units.

I'm hoping my incoming unit isn't affected, but they do come with a lifetime warranty just in case. I've got a 189 unit on my bench at work and doesn't appear to have symptoms. I'd just like my personal meter to be flawless as well.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #208 on: December 26, 2014, 06:42:45 pm »
However I received a follow up email from info-fluke@fluke.com claiming it was the first time the tech had heard of the leaking capacitor issue and had no database record. The disparity in the responses is a little concerting.
It is not surprising to hear that Fluke support personnel don't know about the issue or that you are getting conflicting answers.

Quote
I'm hoping my incoming unit isn't affected, but they do come with a lifetime warranty just in case.
You should check to make sure Amazon is an authorized Fluke dealer because there was controversy about that in the past.

https://www.eevblog.com/forum/testgear/amazon-con-is-not-a-fluke-authorized-reseller/

You should read the fineprint regarding lifetime warranty in the 289 manual.  The word lifetime doesn't mean lifetime in the way an ordinary person would interpret it.

In addition, Danaher needs to make money long term in order to actually be around to provide "lifetime" warranty.
 

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #209 on: December 26, 2014, 10:13:15 pm »
My story with the meter.
After I’ve read the forum, I opened mine and found the capacitor leaking. I desoldered it and ordered this one, http://www.digikey.com/productsearch/en?vendor=0&keywords=EDLRD224   
as per Reply #38 on this forum.

 When it came I didn’t have time to change it so I left it over. The meter hadn’t the capacitor in as people around here, claimed it is only for time and date so I didn’t care. But when I needed it I powered it up, a very short message notified me about memory- time, something like that (it was fast to make a notice) and then died.

It never started up again. Because I had to continue I left it, and continued with another meter. After 3-5 days I had time to look at the meter, the meter was dead, I measured the batteries to find out that they were also completely dead.
Wow I said I found the problem so I changed the batteries, nothing at all, the meter was dead even the power switch did not respond, turning itself green. Again I left it, Next day, searched for the problem to find that the new batteries were completely dead at 0,5 volts. Something was draining them completely.

So I soldered the capacitor back and as it has a limited lifetime guarantee I sent it to my local distributor mentioning that the meter is completely dead, he took it and sent it the fluke service agent. 2 months later the meter arrived back full functional, calibrated and stickers around the meter.

 No charge, meter has a limited lifetime guarantee, so why try it yourselves? Just send it back.



IMG_9835 by sotron100, on Flickr

IMG_0591 by sotron100, on Flickr

IMG_0407 by sotron100, on Flickr
« Last Edit: December 26, 2014, 10:14:52 pm by sotos »
 

Offline kitaismith

  • Newbie
  • Posts: 2
Just opened my Fluke 287 as well, lo and behold I have it too..:



I called the service center 01603256620, they told me they are not aware such a problem existed and further more, since the capacitor
is corroded it is not longer under warranty. To add salt to the wound, the lady on the phone added that it'll take about £240 for it to be repaired..... :-DD

I asked her to repeat that like about 5 times..
I've fired off an email to the UKServiceDesk@fluke.com explaining the situation and what I've just been told, still waiting for reply.
So we'll see how it goes...
 
The following users thanked this post: Marco1971

Offline Ericho

  • Regular Contributor
  • *
  • Posts: 52
  • Country: be
Quote
I called the service center 01603256620, they told me they are not aware such a problem existed and further more, since the capacitor
is corroded it is not longer under warranty. To add salt to the wound, the lady on the phone added that it'll take about £240 for it to be repaired..... :-DD

I asked her to repeat that like about 5 times..
I've fired off an email to the UKServiceDesk@fluke.com explaining the situation and what I've just been told, still waiting for reply.
So we'll see how it goes...

There goes the extra money you pay for warranty, service and quality. Fluke  :--

I'm not gonna bother with them, Its takes less time to fix it myself then 1 Phone call with them of a visit to the local shop.
 

Offline kitaismith

  • Newbie
  • Posts: 2
Yeah...I know it would've been easy to repair it myself but since it should be under warranty and doing it myself will void the warranty , I might as well use it (in theory at least).
Anyways, I got a reply back from the email I sent out..:

Quote
We have forwarded this information to the higher management and our Manager has agreed to bring the unit in under Warranty.
Can you kindly fill out the attached RMA form so we can create an Returns number for you.
 
Hope to hear from you.

I filled out the attached RMA form, emailed it back to them and they've emailed a Service Order Form to me. That'll be included with the multimeter
when I send it back.

All in all, looks like they've decided to do the right thing and to be honest, apart from the lady I spoke to earlier on the phone, the guys I emailed about it got everything sorted
real quick so I'm gonna reserve my judgement till everything's sorted.
 I'll send it back tomorrow and keep you guys posted.
 :)
« Last Edit: March 06, 2015, 05:37:43 pm by kitaismith »
 

Offline mjlorton

  • Contributor
  • Posts: 44
  • Country: us
  • www.YouTube.com/mjlorton
    • Solar Power and Electronic Measurement Equipment
Fluke respond to the issue of corrosion of the capacitor on the  Fluke 287/289:

Original video (2 January 2014) - T4D #88 - Fluke multimeters with leaking / corroded surface mount supercap:

On Forum: http://mjlorton.com/forum/index.php?topic=531.30

Fluke email:
On 4/16/2015 11:22 AM, Wright, Dan wrote: Hello Martin
 I’ve done some investigating on the issue of the Fluke 287/289 and the corrosion of the capacitor.  Our team has looked in to this at length and arrived at the following:
#   Yes, there are instances where the capacitor is showing visible sign of discoloration, corrosion and what may be described as “leakage.”
#   Yes, this is an issue that will be addressed in the next design and production update taking place next month.
#   Despite the visual signs, there is no degradation to the performance of the meter.  The component in question provides backup power to the internal clock and should it fail, the only thing affected will be time and date settings after a battery change.
#   If a user finds that the component has made the meter unusable, the user should contact a Fluke Service Center so the product can be evaluated.
Best regards
Dan
www.YouTube.com/mjlorton
Solar Power and Electronic Measurement Equipment
 

Offline shane124

  • Newbie
  • Posts: 2
  • Country: us
Just replaced the bad cap in my 289. Not very difficult and works great! :)
Used the one from Digikey 338-3477-ND. MFG# EDLRD224H3R6C 220MF
3.6V.

Bought it used. Est. Around 2009 Model. (The meter).
« Last Edit: May 03, 2015, 03:55:46 am by shane124 »
 
The following users thanked this post: Marco1971

Offline Gonzakpo

  • Newbie
  • Posts: 3
Any news about this issue? I'm considering buying the 287 but this is kind of a showstopper. I mean, the only reason I want to buy a Fluke it's because its reliability. But if that is gone... then maybe I should be looking somewhere else.
 
The following users thanked this post: Chris56000

Offline dadler

  • Supporter
  • ****
  • Posts: 851
  • Country: us
Has anyone had this actually affect the performance of their meter?
 

Offline Lightages

  • Supporter
  • ****
  • Posts: 4314
  • Country: ca
  • Canadian po
If Fluke says it doesn't affect the meter, I would tend to believe it. Does it look good that they won't replace the thousands of bad looking components? I think it hurts their reputation a bit but I wouldn't worry about it. Leave it until/if it causes a problem and they will fix it then.
 

Offline Jay_Diddy_B

  • Super Contributor
  • ***
  • Posts: 2733
  • Country: ca
Has anyone had this actually affect the performance of their meter?

I had the problem on two meters. It did not effect the measurement accuracy. It impact the life of the AA batteries. I posted my experience in this message:

https://www.eevblog.com/forum/testgear/fluke-189-with-leaking-surface-mount-coin-cell/msg443374/#msg443374

Regards,

Jay_Diddy_B
 

Offline Muxr

  • Super Contributor
  • ***
  • Posts: 1369
  • Country: us
My guess is the supercap is used for the Real Time Clock. It allows the RTC to survive the battery change. I am assuming RTC is important for the data logging feature.
 
The following users thanked this post: Marco1971

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Martin's latest video has some of the feedback and discussions he had with Fluke at the 26 to 30 minute mark regarding the IR and super capacitor.

 

Offline miguelvp

  • Super Contributor
  • ***
  • Posts: 5550
  • Country: us
One of my co-workers back on the late 90s used to work for Boing (unmanned vehicle Dark Star) he told me that they used to get rain inside their buildings  :scared:
 

Offline SeanB

  • Super Contributor
  • ***
  • Posts: 16276
  • Country: za
One of my co-workers back on the late 90s used to work for Boing (unmanned vehicle Dark Star) he told me that they used to get rain inside their buildings  :scared:

Reminds me when I was in training, and the only room available was the one with the big half roof skylight. Good thing was it was the dry season, and I put in a fan heater on a time switch, to heat the room for 2 hours before we came back, and to turn off when we left in the morning. Was interesting to be able to do astronomy from my bed at night. Was not fun though that the showers invariably had no hot water, especially as it was winter in Pretoria. Used to wake up at 2AM to go to a shower and have that hot shower luxury. Plus was the mains was always between 250 and 265V, due to the area having legacy 250VAC mains transformers, dating from the early days of electrification in South Africa, in the supply, and harmonising the area to 220V was considered too expensive in replacing all of them simultaneously. Great for heating, though lights had a really short but brilliant life.
 

Offline firehopper

  • Frequent Contributor
  • **
  • Posts: 408
  • Country: us
I have a fluke 289 with the rusty looking cap, it wont hold time or date when you pull the batteries. so what should I do? replace it myself or send it in..
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
replace it myself or send it in..
If you are the original owner and bought it from an authorized Fluke dealer, I suggest sending it in for the following reasons:

1) you might get a complete new pcb with the latest modifications and firmware
2) you are likely to have the 289 (either repaired or new pcb) with its calibration checked for "free"
3) shipping within the USA even with a tracking number should be a reasonable $10 to Fluke

The calibration check normally costs between $50 to $100 and you get it back knowing it is within its accuracy specifications across all ranges.
« Last Edit: June 08, 2015, 06:07:52 pm by retiredcaps »
 

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
I agree with retiredcaps.  When I tried to change mine it got bricked, I had to send it to the service center and after the repair it was also calibrated, for free.
 

Offline firehopper

  • Frequent Contributor
  • **
  • Posts: 408
  • Country: us
I'm not the original owner.. I bought it used off ebay. I did register the serial number on flukes website, on fluke's advice.  so no idea if I should still send it back, like I said remove batteries and it forgets time and date.



replace it myself or send it in..
If you are the original owner and bought it from an authorized Fluke dealer, I suggest sending it in for the following reasons:

1) you might get a complete new pcb with the latest modifications and firmware
2) you are likely to have the 289 (either repaired or new pcb) with its calibration checked for "free"
3) shipping within the USA even with a tracking number should be a reasonable $10 to Fluke

The calibration check normally costs between $50 to $100 and you get it back knowing it is within its accuracy specifications across all ranges.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
If you are not the original owner, then Fluke, technically, has no obligation to fix it under warranty for you.

Here are the options that I see.

1) If the date and time setting are not required, just ignore the problem and leave the capacitor as is.  Alternatively,  you can remove the capacitor and the unit will function fine.  Note, I can't remember if it is in this thread or this model, but at least one person who removed the capacitor couldn't get the meter to power on again.  So there is a non zero chance of screwing something up if you are not careful (lift a solder pad, short something out, zap components/pcb with ESD, etc).

2) Fluke built its service reputation in the past, as reported here and in other forums, by providing fantastic service to the customer.  Sometimes, but not all, the service centre will fix the problem regardless of the terms and condition of the warranty for free.

So you could take a chance and send it to Fluke (make sure you get a tracking number + insurance) for around $10 USD and see if Fluke will fix it under warranty and ignore the fact that you bought it from ebay second hand.  It is my understanding, and you should confirm by talking to a customer service representative, that if you do not qualify for warranty repair that Fluke will contact you with an $$ estimate to fix your meter.  Then you can decide if the price is reasonable or not.  If you choose not to get it fixed due to cost, then Fluke will send it back to you unrepaired.  I believe Fluke will pay return shipping so at most you are out the original shipping costs.

Again, you may want to talk and verify with a Fluke customer service representative.  If you choose the second option, please report back so we know the outcome.

PS. I am well aware of negative posts about Fluke customer service.  I find that by talking nicely to a person and treat them the way you expect to be treated, that you will get a lot better result.
 

Offline Kibi

  • Supporter
  • ****
  • Posts: 385
  • Country: england
PS. I am well aware of negative posts about Fluke customer service.  I find that by talking nicely to a person and treat them the way you expect to be treated, that you will get a lot better result.

I totally agree that good manners will get you a long way.
However, politeness and good manners don't get you anywhere with Fluke UK. Their representatives are polite, well mannered and sound professional. I treat them over the telephone as they treat me, but unfortunately that is the last you hear from them. They never once got back to me about my spare parts request on any of the three occasions that I telephoned their service centre. I also sent the same request over email, but I got no response either. In my opinion, this is very poor service on the part of Fluke UK.
Hopefully their other branches provide a better service.

I went elsewhere and then got my parts within days.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
However, politeness and good manners don't get you anywhere with Fluke UK.
My professional experience/industry is not the T&M so I don't know how it works in T&M, but when I call a company for help, I'm usually assigned a case number.  That case isn't closed until I say so.

In the past coule of years, I have made contact with Fluke Canada and Extech USA.  Neither provided me with a case number.  Without a case number, I have no way to escalate to a supervisor and when I get passed onto another customer service representative, there is no history or case notes.
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
So you could take a chance and send it to Fluke (make sure you get a tracking number + insurance) for around $10 USD and see if Fluke will fix it under warranty and ignore the fact that you bought it from ebay second hand. 
@firehopper, see this thread wrt to above comments.

https://www.eevblog.com/forum/testgear/fluke-287(9)-leaking-super-cap/msg692664/#msg692664
 

Offline D. Head

  • Contributor
  • Posts: 29
  • Country: nl
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #231 on: August 13, 2015, 06:10:25 pm »
Hello Everyone,
this is what the supercap from my Fluke 289 looks like.

Is this "bad" enough to send it back to Fluke?
(i live in the area of Eindhoven)

Thanks
« Last Edit: August 13, 2015, 06:14:01 pm by D. Head »
 

Offline retiredcaps

  • Super Contributor
  • ***
  • Posts: 3575
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #232 on: August 13, 2015, 07:30:10 pm »
Is this "bad" enough to send it back to Fluke?
The cap is supposed to keep time and date.  If you take the batteries out and it doesn't properly keep time and date, then you would have a valid claim for warranty service (free).  Take a look at my post 225 above again for why you might want to send it in.
 

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #233 on: August 19, 2015, 07:57:22 pm »
Is this "bad" enough to send it back to Fluke?
The cap is supposed to keep time and date.  If you take the batteries out and it doesn't properly keep time and date, then you would have a valid claim for warranty service (free).  Take a look at my post 225 above again for why you might want to send it in.

I dont recall from the manual how much time it supposed to hold the time and date settings? wanna see how is my unit holding up without taking it apart
 

Offline dr.diesel

  • Super Contributor
  • ***
  • Posts: 2214
  • Country: us
  • Cramming the magic smoke back in...
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #234 on: August 19, 2015, 08:10:21 pm »
Measure the parasitic power draw.  When mine went bad it was almost 10mA, lost the time/date instantly if batteries removed.

Offline Muttley Snickers

  • Supporter
  • ****
  • Posts: 2340
  • Country: au
  • Cursed: 679 times
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #235 on: December 23, 2015, 08:14:24 am »
The super capacitor in my 289 looks a little bit crusty but the meter still seems to work well and I have listed the current consumption details below for reference. I recently opened and inspected my 189 again and it is still in immaculate condition in relation to the super capacitor, I have owned this meter since brand new and as mentioned elsewhere on the forum I religiously remove the batteries after each use and have always done so.

Fluke 289 Current Consumption

Power Off, Logged for 20 mins.....32uA
Power On, Ohms Open, BL Off......22mA
Power On, Ohms Open, BL Low......36mA
Power On, Ohms Open, BL High......56mA

Firmware Version 1.12/V0.88 (Now upgraded to 1.16)
Calibration Date: 25/07/2010. Calibration Count: 1.  Board ID: 1.

« Last Edit: August 14, 2018, 02:50:25 pm by Muttley Snickers »
 

Offline Fonnesbek

  • Newbie
  • Posts: 8
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #236 on: December 23, 2015, 11:16:03 am »
Hi

I have just replaced the super cap in my 189 I also have a 289 but the cap looked ok in that, so I will not change it at this moment.
I replaced the cap with a EDLRD224H3R6C which I got from amazon.co.uk
http://www.amazon.co.uk/gp/product/B017V3X02S?ref_=pe_385721_37038051_TE_3p_dp_1
It can properly be bought cheaper but shipping rate and time was an important factor.

I have attached a few pictures of the replacement, the last one is of the super cap in my 289.

/Fonnesbek
 

Offline Helix70

  • Supporter
  • ****
  • Posts: 289
  • Country: au
  • VK4JNA
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #237 on: February 13, 2016, 04:09:22 am »
My 287 supercap failed in a different way to most on here. Mine won't hold charge. I took it out, attached it to 3.3V and it immediately jumped to 3.3V drawing 40uA. This current slowly decays over about 2 minutes until it hits about 3uA. I set the meter to >10G input impedance and disconnected the supply, and in 5 seconds the voltage on the supercap dropped to about 2V, then 1.8V after 30 seconds, drawing < 0.02uA.

Needless to say, the meter did not hold date and time, and the corrosion on mine was minimal (ignore the flux on the leads, I put that there).

« Last Edit: February 13, 2016, 04:12:46 am by Helix70 »
 

Offline toni31

  • Contributor
  • Posts: 19
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #238 on: February 20, 2016, 08:58:19 pm »
I'm considering purchasing a new fluke 287 from an authorized dealer
does anyone know if fluke has resolved the leaky cap issue
 

Offline Helix70

  • Supporter
  • ****
  • Posts: 289
  • Country: au
  • VK4JNA
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #239 on: February 21, 2016, 12:37:04 am »
All fixed. Got a 0.1F supercap from RS - a Kemet FCS0H104ZFTBR24. Cost $4 for two delivered.

I had to snip the lower left plastic corner off to make it fit around the SOT23 part, otherwise it fit very neatly. The shroud fits over it without issues.

Tried it for 15 minutes without batteries, all ok. Hopefully with the increased voltage rating this one won't die or leak!
 
The following users thanked this post: Marco1971

Offline hansibull

  • Regular Contributor
  • *
  • Posts: 108
  • Country: no
I just desoldered this 0.33F 5V Matsushita super cap from some almost-unused equipment. Can/Should I just whack this in, or should I get one that's almost identical to the original one. And how about the polarity, what is + and what is - on the PCB?

 

Offline VK5RC

  • Supporter
  • ****
  • Posts: 2672
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #241 on: November 23, 2016, 11:44:22 am »
A recent purchase of a 289 view from the inside, looks like they have redesigned the board a little and the Cap has been replaced by a battery??
Whoah! Watch where that landed we might need it later.
 
The following users thanked this post: Marco1971

Offline Daruosha

  • Regular Contributor
  • *
  • Posts: 181
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #242 on: February 21, 2017, 04:56:11 pm »
I just found this thread and wondered what is your PCB rev? I purchased two fluke 289 and production date of the first one is 2015 and the other one is 2016.

Anyone would suggest to break the warranty and tear them apart for further inspection?
 

Offline evava

  • Regular Contributor
  • *
  • Posts: 172
  • Country: cz
Hello everyone,

I also got Fluke 289 from ebay lately.
And I have got two questions:

1. when changing batteries, it actually "holds" time (and date), by which I mean time (and date) "freezes" until new batteries are put in and then time runs again from the point where I removed batteries, but delay remains.
So, it "holds" time, but time is not running forward when batteries are not in the meter - is it OK?
Or is it supposed that time counter continues when batteries are removed?
(comsumption when meter is off is about 50uAmps, which is OK probably)

2. on Lo Ohm (50 Ohms) range, with shorted input posts, readings slowly decreases from about 60 ohms to the 0 Ohms during ten (fifteen) seconds and through 0 Ohms continues to the minus Ohms readings slowly and still very slowly increases minus Ohms readings, just like in that video
 
and maybe never stops completely (it is not my movie nor mine instrument but the same symptom).
If it is normal behaviour, then it seems to me that 50 Ohm range is not any good?
Or is my meter broken?

« Last Edit: April 28, 2017, 09:29:50 am by evava »
 

Offline todd_fuller

  • Contributor
  • Posts: 37
  • Country: us
I recently acquired a Fluke 287/FVF and was quite curious if this was something I needed to be worried about. This was purchased from Amazon. Serial # 3567XXXX, 7/16 build date, rev 17 board.

As others show, cap is now replaced with a battery.


Off-topic: anyone know if the J5 header is is the UART to the IR? It would be interesting to add a BT module here.
 

Offline M. András

  • Super Contributor
  • ***
  • Posts: 1014
  • Country: hu
measured my 289 in stanby today. consumes over 600uA as my testo clamp says OL back to a chinese meter it says 650uA if you can belive to them in terms of accuracy, the was bought back in 2012 so its still well within warranty, will see what the distributor says in the next days
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
My 287 after replacing super caps consumed 26,5 uA (238 uW)  in standby (OFF) mode.
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Both 189 and 289 have the ongoing cRapacitor problem which I cleaned up with some isohol
and may swap out one day when I get the right part and patience to do without killing it
(since Fluke in Oz want receipts n s***, to verify first ownership to fix a -FLUKE PROBLEM- for an 'unimportant' long time loyal Fluke customer) 

Strangely enough, the battery life on both has been great for over 6 months now, on or off, backlight on, logging stuff etc just using ANY weird name brand cheapies

i.e. same set of cheap batteries going well for 6 months, and no leaks yet.

Am I lucky, or 6 months use on a set of cheapies is no big deal?

I haven't been removing or blocking the batteries either btw (lazy, forgetful, busy, whatever) and the meters are still going fine


The math is that 5 or 6 dollars in total to swap out twice a year on batteries for each feature packed meter sounds like a good deal,
and may be beat any leakage happening too if doing so every 6 months

 :-//
« Last Edit: May 28, 2017, 03:44:17 am by Electro Detective »
 
The following users thanked this post: Marco1971

Offline Dwaine

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: ca
Yeah, I think your right about the header being for the IR.  Slap in a Bluetooth and the device could be used better for data logging.
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Yeah, I think your right about the header being for the IR.  Slap in a Bluetooth and the device could be used better for data logging.
And for internal noise too  ;) Or you'll do completely EMI testing with all Bluetooth module's modes? Or will changing firmware for sending "sleep" signal for bluetooth module?
 


Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
A recent purchase of a 289 view from the inside, looks like they have redesigned the board a little and the Cap has been replaced by a battery??
New one revision 017 is pretty interesting... If this is real battery, how long it will support memory without main batteries installed in meter? What will happening when backup battery will completely dead? They will change battery or entire PCB? :)
For super caps mentioned time for keeping data without main batteries installed on meter is about 4 hour.
Also for super capacitors shelf life time is about from 1000 h to several years.
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
A recent purchase of a 289 view from the inside, looks like they have redesigned the board a little and the Cap has been replaced by a battery??
New one revision 017 is pretty interesting... If this is real battery, how long it will support memory without main batteries installed in meter? What will happening when backup battery will completely dead? They will change battery or entire PCB? :)
For super caps mentioned time for keeping data without main batteries installed on meter is about 4 hour.
Also for super capacitors shelf life time is about from 1000 h to several years.

This is Fluke's job to solve and FIX FOR FREE  -UNCONDITIONALLY-  FOR ANY OWNER,

look at the LUDICROUS prices they charge for decent meters  :-+ with SILLY problems  :-- that DON'T always need a board replacement and a 'yeah it works' -calibration- 
   :-//
« Last Edit: May 28, 2017, 09:09:59 am by Electro Detective »
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #253 on: May 28, 2017, 09:06:28 am »
So other manufactures do not have problem with super caps?
Yes, Fluke offering "lifetime" warranty and probably will solve problem. Past PCB revisions for 280 series used Panasonic EN series supercaps with just 500h shelf life  :-- this series has been discounted by Panasonic.
Probably best way is using battery that will able changed by customer. I do not see any reason for using battery soldered directly to PCB.
« Last Edit: May 28, 2017, 09:11:11 am by serggio »
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
So other manufactures do not have problem with super caps?
Yes, Fluke offering "lifetime" warranty and probably will solve problem. Past PCB revisions for 280 series used Panasonic EN series supercaps with just 500h shelf life  :-- this series has been discounted by Panasonic.
Probably best way is using battery that will abble changed by customer. I do not see any reason for using battery soldered directly to PCB.

BRAVO!   :clap:  this is EXACTLY what I believe too.  :-+

It is SHEER DUMBASS to replace one SOLDERED leaking component with another SOLDERED soon to leak battery component inside an overpriced CAT 111/1V meter !!!  :palm: :palm:
that may do WORSE DAMAGE in the future,

usually ONE DAY after the lifetime warranty has ENDED.   |O

Whoever pays the bills and wages at Fluke needs to word up the engineers YESTERDAY   :=\ :=\  to look inside their CAD computers  :-/O and check out the 'user replaceable' BIOS backup battery arrangement on the main board,  :o :o
that's been in use across TWO CENTURIES of consumer electronics!!!!!!      :palm:  :palm:

« Last Edit: May 28, 2017, 09:57:37 pm by Electro Detective »
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Whoever pays the bills and wages at Fluke needs to word up the engineers YESTERDAY   :=\ :=\  to look inside their CAD computers  :-/O and check out the 'user replaceable' BIOS backup battery arrangement on the main board,  :o :o
that's been in use across TWO CENTURIES of consumer electronics!!!!!!      :palm:  :palm:
It's not BIOS  :) RTC and probably temp offset, contrast and some user variables. http://media.fluke.com/documents/287-289_mveng0200.pdf
To be honestly, I do not see big problem with super capacitor except of shorting life of batteries. Even if you do not have warranty, you can easy to replace it oneself. I replaced super caps in my meter oneself just for save my time.
At any case they have new revision for PCB, let's see what will going on with
 

Offline Dwaine

  • Frequent Contributor
  • **
  • Posts: 299
  • Country: ca
That's true.  The existing separate Bluetooth module really nice.
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Whoever pays the bills and wages at Fluke needs to word up the engineers YESTERDAY   :=\ :=\  to look inside their CAD computers  :-/O and check out the 'user replaceable' BIOS backup battery arrangement on the main board,  :o :o
that's been in use across TWO CENTURIES of consumer electronics!!!!!!      :palm:  :palm:
It's not BIOS  :) RTC and probably temp offset, contrast and some user variables. http://media.fluke.com/documents/287-289_mveng0200.pdf
To be honestly, I do not see big problem with super capacitor except of shorting life of batteries. Even if you do not have warranty, you can easy to replace it oneself. I replaced super caps in my meter oneself just for save my time.
At any case they have new revision for PCB, let's see what will going on with

My point was to put a shake proof secured 'user replaceable battery' rather than a stupid leaky underated cRapacitor or dumbass soldered in battery that may leak too,
whose insides may be half fried by the soldering process to help along the inevitable.  -is not rocket science-  :palm:


« Last Edit: May 28, 2017, 10:11:40 pm by Electro Detective »
 

Offline Tim5000

  • Contributor
  • Posts: 12
  • Country: gb
Hello all,
I recently discovered my Fluke 189 had a leaking supercap. It also struggled a little turning on sometimes (may or may not be related).
I was aware of this thread so I contacted Fluke UK and they immediately said "It's under limited warranty - just send it in." I queried charges and they said I'd have to pay postage...great I thought they sound like they might replace the cap!

So I sent it in.
After a couple of weeks I've received the quote...

Please be advised that the meter is out of service hence no parts available.
We are quoting you for an replacement unit Fluke-289 at the repair cost which includes 50% discount off the pricelist.
Please be advised that the 50% discount is only applicable for trading in your old unit.

-Repair price : 268.00GBP (End customer price)   
-Calibration type : Traceable Calibration with Data  (OPTIONAL ON REPLACEMENT UNIT)
-Calibration price : 87.50GBP (End customer price)
-Investigation fee if quotation is rejected : 46.00GBP (End customer price)
-Shipping charge : 10.00GBP

All ex VAT of course.

To me that sounds like a pretty poor show from Fluke.
The offer of:
1) Buying a 289 from them (which I don't want) at a price that is more than I want to spend.
2) Calibrating my meter for £105 + postage but retaining the fault as "no parts available"
3) Paying £55 + postage to have it sent back to me.

I even mentioned that there's a very long thread about the design issue of the supercap and she didn't sound like it was a big shock.
I've absolutely no problem paying for these things when told upfront but the very nice lady on the phone didn't indicate these options when I asked what the options could be.
Has anyone else had any better luck from Fluke (UK) on this matter? I'm wishing I'd just replaced it myself now :(

Thanks
Tim
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Are you eligible?

Quote
Lifetime is defined as seven years after Fluke discontinues manufacturing the product, but the warranty period shall be at least ten years from date of purchase. *(Lifetime Warranty applies to products manufactured after October 1996). The warranty does not cover manuals, fuses, disposable batteries, damage from neglect, misuse, contamination, alteration, accident or abnormal conditions of operation or handling, including failures caused by use outside of the product's specifications, or normal wear and tear of mechanical components. This warranty covers the original purchaser only and is not transferable. This warranty covers the LCD for 10 years only (state-of-the-art for LCDs). To establish original ownership proof of purchase is required (20, 70, 80, 170, 180 and 280 models). THE FOLLOWING PRODUCTS ARE INTENDED FOR SALE AND USE IN THE CHINESE AND HIGH GROWTH MARKETS ONLY AND AVAILABLE FOR WARRANTY PROTECTION ONLY IF SOLD, PURCHASED AND USED SOLELY FROM AN AUTHORIZED DISTRIBUTOR WITHIN THESE TERRITORIES: 101, 106, 107, 15B+, 17B+ and 18B+. THE WARRANTY FOR THESE PRODUCTS IS VOID IF THE PRODUCT IS SOLD, PURCHASED OR USED OUTSIDE OF THESE TERRITORIES, OR FROM UNAUTHORIZED DISTRIBUTORS AND NO WARRANTY COVERAGE, SERVICE, REPAIR OR REPLACEMENT WILL BE PROVIDED.

If not, ask them return your meter, replace capacitor oneself and be happy with your trusted meter.
 

Offline Tim5000

  • Contributor
  • Posts: 12
  • Country: gb
Hi serggio, yes I believe I am eligible. As I mentioned I also queried it with the support lady.
What's frustrating is she could have quite easily told me "we can't repair them" but she didn't - she instead told me to send it in as it's covered under the limited warranty.
And I phoned twice to check before going ahead with it.

But now to get it back I have is to pay £55 + postage. Again if that had been mentioned upfront when I asked about fees I wouldn't have a complaint - but it absolutely wasn't mentioned.
My concern is why didn't they just tell me over the phone that they won't repair these meters? To my mind not being upfront when asked is just theft.

I would be interested if anyone else has dealt with Fluke UK specifically on the matter of this design issue.
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Hi Tim5000,
If you still eligible for lifetime limited warranty (no 10 year left from date of purchase), I suppose they should did repair/replace your meter for free.
You still responsible for postage fees but not for repair cost.
Try contact their US support by mail for clarify this situation or head their service department in UK.
Girl on reception just for bla-bla-bla, she is not responsible for any decisions and not check any service bulletin during phone conversation.
« Last Edit: June 07, 2017, 07:15:42 am by serggio »
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
It must be obvious by now that Fluke are either dropping the ball too often,

or their service agents are not worded up properly about how Fluke support works,

or both!

Tim5000s situation described above sounds like a feeble attempt to relieve him or her of more money
without fixing a WELL KNOWN meter issue that is Fluke's part blunder, not the CUSTOMERS!  :palm:

To add more SALT to the wound they want Tim5000 to dump the 189 and fork out difference money and get a 289,
which bears NO RESEMBLANCE in functionality,    [[SIZE!!!]]     and convenience to a 189, even though many of the specs are similar

i.e. the 189 is a sportscar, the 289 a pimped up pickup ute. 

What the FLUKE is going on?

The competition is watching...note more orange Keysights, green Mastechs and red Brymens and Uni-Ts out there !  :scared:

let's not forget those little rippa BLUE meters that are quickly gaining favour too   :-+


« Last Edit: June 24, 2017, 08:02:49 pm by Electro Detective »
 

Offline iamjanco

  • Newbie
  • Posts: 6
  • Country: us
Hi everyone, realize this thread is about the 289 supercap issue, hoping my query is somewhat relevant (I'd rather not invest in a current Fluke, if they're no longer the brand they used to be).

I've been considering purchasing a 289 FVF package for use in computer benching given its logging capabilities, but haven't been able to locate much in the way of whether Fluke has properly addressed the leaking supercap issue in the newer releases. I'm also a bit concerned about battery life during longer test intervals and am open to other alternatives that offer similar functionality. My needs are focused on short and longer interval voltage and thermal measurements and am not adverse to a smaller bench unit (space is limited), even a used one, though I'd prefer to keep pricing within the same range.

Info and suggestions would be greatly appreciated.
 

Offline Tim5000

  • Contributor
  • Posts: 12
  • Country: gb
Hi, There was a post earlier in the thread I think that seemed to show the cap had been designed out of newer models and a battery has instead been designed in...

With regards to my issue with Fluke UK, I'm just waiting for my meter back.
I wasn't happy with the offer of a 289 - I think they should at least have replaced the cap if they didn't have any boards.
But apparently fluke don't have any parts...I'm not sure I believe that but it is what it is!
They did at least waiver the "investigation fee" so I went with that option.

Mostly I was dissapointed they didn't straight up say that they didn't/couldn't/wouldn't do anything - it would have saved both mine and their time and money and I'd have got it done by now!

So all in all I'm a little bit lighter of pocket for double shipping fees and will investigate fixing it myself when it arrives.


 
The following users thanked this post: theirishscion, Electro Detective, iamjanco

Offline technogeeky

  • Frequent Contributor
  • **
  • Posts: 555
  • Country: us
  • Older New "New Player" Player Playa'
Hi, There was a post earlier in the thread I think that seemed to show the cap had been designed out of newer models and a battery has instead been designed in...

With regards to my issue with Fluke UK, I'm just waiting for my meter back.
I wasn't happy with the offer of a 289 - I think they should at least have replaced the cap if they didn't have any boards.
But apparently fluke don't have any parts...I'm not sure I believe that but it is what it is!
They did at least waiver the "investigation fee" so I went with that option.

Mostly I was dissapointed they didn't straight up say that they didn't/couldn't/wouldn't do anything - it would have saved both mine and their time and money and I'd have got it done by now!

So all in all I'm a little bit lighter of pocket for double shipping fees and will investigate fixing it myself when it arrives.

As fucking well they should have waived the fee! You already knew what the problem was! They know what the problem is. What is there to investigate?

This whole situation is absurd.

I don't currently own any Fluke multimeters, but I certainly aspired to one day own what I thought was such a high-end meter. But now, I can't see myself making such a purchase. It's probably been obvious for a long time to some that they aren't fundamentally better; and now it's obvious that they can indeed be worse.

It's a shame, really. They were a "great" American brand.
 
The following users thanked this post: Electro Detective

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
You just need to find someone local (clued and honest) that can sort out the cRapacitor issue (without causing other issues) if you don't want to stuff about DIY.

Mine still suffer from the issue but are working 100%

Both the 189 and 289 are still killa meters to this day, and will continue to be so,  :-+


unless Volts, Amps and Ohms become obsoleted in the near future.   ;D

 

Offline Tim5000

  • Contributor
  • Posts: 12
  • Country: gb
Thanks for the replies.
I've got a couple of meters - the 189 that's getting on and then I've also got a 2016 87V that I got for a great price on ebay virtually new. The 87V is good - I like it, but I still reach for the 189.
As I told Fluke - this was my personal meter but as a consultant engineer at my work when it comes to buying new equipment I can and will give an honest account of my experience with their support.
I even pointed them at this forum and the 87V quality forum to highlight that a lot of people are losing faith in them...but they just weren't open to moving at all.
Apparently the limited support ran out in April (2 months out) so they couldn't do anything but of course were very sorry!

So..when the meter comes back I'll order the parts and get it repaired myself, it doesn't look hard and I can't leave it knowing it's there - I just wanted to use Fluke's "legendary" service.
I'm not happy - but lessons learned and people will always vote with their feet.

Tim

 
The following users thanked this post: theirishscion

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
I use the 189 and 87V combo sometimes, it's a bench test no brainer   :-+

Just watch out for the 87V AC default return when measuring DC current and switching from MA to uA,
I've got a taped on sticky note to remind me   

 

Offline ketimoi

  • Contributor
  • !
  • Posts: 10
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #269 on: September 30, 2017, 05:12:59 am »
This thread is so useful no doubt about that. I own a fluke multimeter after reading this best multimeter review https://totalguide.org/best-multimeter/. Overall it is an good multimeter. But, this month I am seeing some problem with it. I am reading whole thread to figure out my problem.
« Last Edit: October 01, 2017, 04:49:49 am by ketimoi »
.
 

Offline Tim5000

  • Contributor
  • Posts: 12
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #270 on: September 30, 2017, 04:07:37 pm »
This thread is so useful no doubt about that. I own a fluke multimeter and seeing some problem with it. I am reading whole thread to figure out my problem.
Hi ketimoi, good news is you'll get a lot of help here I'm sure.
And being in the US everything I've read is that Fluke is very much more interesting in doing "the right thing".

I've actually only just got around to revisiting my Fluke 189 meter with this issue so I've ordered what a few others have mentioned - a Panasonic EECS0HD334H.
Will update when I try and fit it -  it should be arriving this week.

EDIT: Confirmed that it fitted (with a little trimming of the tabs as it's physically bigger), and time is now held whilst changing batteries.
« Last Edit: October 03, 2017, 09:09:28 pm by Tim5000 »
 

Offline Marco1971

  • Contributor
  • Posts: 29
  • Country: it
Hi, all...my 89 IV and 189 are in the same condition (pristine)...only in the 89 IV there are some signs of leaking. Both from year 2000.
Any issue if intead of 0.1 F i'll use one supercap with another value of capacitance (e.g 0.22 or 0.33 F) ? I just ordered four of this...



Many thanks.

Marco1971.
« Last Edit: March 23, 2018, 09:13:23 am by Marco1971 »
 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5879
  • Country: ca
For the moment in my 189

I have removed the supercap, it was rusted

I dont care about the clock, the meter works very well,  i did find an physical equivalent at digikey, 0.22f at 5 volts at 1.52$ can, ordered 5 of them  but they are out of stock, back order for 2 month ...
 

Offline bayati

  • Contributor
  • Posts: 15
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #273 on: January 05, 2019, 07:01:29 pm »
Hi
I'm sorry to bringing up an old thread and I know it has been talked a lot about this issue but please bear with me because I'm seriously confused.

I have 2 fluke 289. both have corroded cap and they can't hold the time when the battery is removed. I measured it's consumption right after installing the batteries. It draws 10 mA at first but after 1 min it stops at 20-30 uA (I think it charges the cap). Both working perfectly and I don't have any problems in measurements and I think battery consumptions are normal.

Note: I can't send them to Fluke

-- Now my questions:
1. Two people mentioned that after removing the cap their meters broke. How common is that? It can be related to other issues not removing cap?
2. Many substitute caps mentioned. 0.1F 3.3v, 0.22F 5.5v, 0.33F 5.5v? In super caps using higher capacity is dangerous or higher voltage?
3. The meter's firmware is 1.12. Is it better to change the super cap after updating to 1.16 or before that?
4. Should I wait for the cap to get discharged completely then removing them or it does not matter?
5. This question may look absurd but one person mentioned just using hot air to change the cap, not soldering iron?!!! but he did not mentioned why! (I think using soldering iron is better because using hot air, I thing it messes up the calibration)
6. The meters are calibrated in 2011. I don't care about the date but I'm worried that the battery corrosion expands or shorts. what do you thing? should change it?

Thank you in advance.
« Last Edit: January 05, 2019, 10:08:08 pm by bayati »
 
The following users thanked this post: Marco1971

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5879
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #274 on: January 06, 2019, 03:25:13 am »
On my 3x 189  they had an 0.1 farad supercap,  changed them to 0.33 farad at 5.5 volts,  bended the legs to fit pcb pads, i desoldered one side at the time,  none of them broke afterwards,  while wainting for the caps,   i did use them with no problems ??

I did use anti static protocol, not shorted anything in case the supercaps where charged, mine were corroded on top.
« Last Edit: January 06, 2019, 02:47:59 pm by coromonadalix »
 
The following users thanked this post: bayati

Offline Daruosha

  • Regular Contributor
  • *
  • Posts: 181
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #275 on: January 06, 2019, 09:33:55 am »
Hi
I'm sorry to bringing up an old thread and I know it has been talked a lot about this issue but please bear with me because I'm seriously confused.

I have 2 fluke 289. both have corroded cap and they can't hold the time when the battery is removed. I measured it's consumption right after installing the batteries. It draws 10 mA at first but after 1 min it stops at 20-30 uA (I think it charges the cap). Both working perfectly and I don't have any problems in measurements and I think battery consumptions are normal.

Note: I can't send them to Fluke

-- Now my questions:
1. Two people mentioned that after removing the cap their meters broke. How common is that? It can be related to other issues not removing cap?
2. Many substitute caps mentioned. 0.1F 3.3v, 0.22F 5.5v, 0.33F 5.5v? In super caps using higher capacity is dangerous or higher voltage?
3. The meter's firmware is 1.12. Is it better to change the super cap after updating to 1.16 or before that?
4. Should I wait for the cap to get discharged completely then removing them or it does not matter?
5. This question may look absurd but one person mentioned just using hot air to change the cap, not soldering iron?!!! but he did not mentioned why! (I think using soldering iron is better because using hot air, I thing it messes up the calibration)
6. The meters are calibrated in 2011. I don't care about the date but I'm worried that the battery corrosion expands or shorts. what do you thing? should change it?

Thank you in advance.

1- Removing the caps wouldn't do any harm. Fluke themselves  confirmed the purpose of that cap is to keep the clock.
2- Higher caps ratings are OK (except in high ripple filtering where ESR is important)
3- Firmware version doesn't matter.
4- If the corrosion process has been started, remove the cap ASAP.
5- Removing an SMD cap can be done in many many ways. Just do it the way it suites you and you feel comfortable enough to do it. If you do it properly with small enough air nozzles, hot air wouldn't do any damage and drift.
6- Yes, Change them.

 
The following users thanked this post: Marco1971, bayati

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #276 on: January 06, 2019, 01:14:59 pm »
I’m one of the guys that after removing the capacitor the meter refused to boot. I sent it to Fluke and they changed the capacitor and repaired the meter free of charge but I had to wait 1 month as I remember.

PS. I repair pcbs for a living and know how to solder and desolder through hole and smd components.
« Last Edit: January 06, 2019, 01:20:53 pm by sotos »
 
The following users thanked this post: bayati

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5879
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #277 on: January 06, 2019, 02:54:55 pm »
@sotos  Errors can happen to experimented techs, an sudden discharge at the wrong place, bad manipulation  etc ...

Glad Fluke repaired it,  sad they haven't issued a official recall for this supercap, "they" know its a problem for a long time.

I used Panasonic  eec-s0hd334h      pn: p11064-nd at Digikey

But they are being phased out ... i have 3 left in case ...
 
The following users thanked this post: Marco1971, bayati

Offline bayati

  • Contributor
  • Posts: 15
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #278 on: January 06, 2019, 04:09:23 pm »
Fluke has a big unofficial market in Iran (just like apple) but no official service. I think i should take a risk and do it. I try to discharge the super cap with a resistor before removing it. wish me luck :-/O
 

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #279 on: January 06, 2019, 05:02:11 pm »
When you remove the capacitor don’t try to turn it on to see if it can work without the cap. I suggest you remove it put a new one in and then turn it on.
I did this to mine and didn’t boot, I tried it without the cap. Maybe it doesn’t like it.
 
The following users thanked this post: bayati

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5879
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #280 on: January 06, 2019, 05:18:37 pm »
@ sotos   

My 189 worked pretty well without the supercap, I did not use any time logging feature and or set a clock

I wont say for the 289 series,  apparently there may be an real battery instead of an supercap ??  Some teardown photos showed that.
 
The following users thanked this post: Marco1971

Offline say

  • Contributor
  • Posts: 28
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #281 on: January 06, 2019, 06:58:03 pm »
Could you please post a link of those photos.
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 11709
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #282 on: January 06, 2019, 07:15:23 pm »
I was working with a friend that has a 189 a month ago.   I went to use his meter and it was dead.  He had just put fresh batteries in.  With permission, I took it apart and sure enough the cap was going bad.  No need to discharge it, it was flat.   These meters are getting old.
 
The following users thanked this post: Marco1971, bayati

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #283 on: January 07, 2019, 12:06:53 am »
I'm still sitting on the fence about removing the stupercrapacitors on the 189 and 289  :-[

I would prefer to remove the caps and leave a blank space there,
but without killing the meter/s and or affecting their logging functions

or have the charger circuit/firmware/software go nuts with "Dude, where's my cap?!!"   ???


Reset of the clock on battery changes is a trivial matter compared to having leak prone caps and coin batteries lurking in there

Both meters working great, I really don't want to kill em
 

Would an electrically isolated soldering iron or station with a wee tiny tip be ok to get this done?
or should the iron be earth/ground referenced?

I don't want to mess with hot air, nor do I wish to encourage any ESD zaps during a simple heat and remove operation

i.e. I want them GONE without the    'SOL-Game Over'    thing flashing on the meter display 

« Last Edit: January 07, 2019, 08:32:12 am by Electro Detective »
 
The following users thanked this post: bayati

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5879
  • Country: ca
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #284 on: January 07, 2019, 12:58:01 am »
i've done the supercap change on a anti static bench mat, with an normally earth grounded  hakko soldering post, nothing fancy, if you're not so sure, discharge the supercap with an 100 ohms resistor ...


For the supposed battery in a 287 289   link : https://www.eevblog.com/forum/testgear/fluke-189-with-leaking-surface-mount-coin-cell/225/

On the pcb it is marked BT1  :  https://www.eevblog.com/forum/testgear/fluke-189-with-leaking-surface-mount-coin-cell/?action=dlattach;attach=311290;image

My 189 had c xxx marked on the pcb,  the confusion would be : it is a supercap used as a small battery ??
 
The following users thanked this post: Electro Detective, bayati

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #285 on: January 07, 2019, 08:56:41 am »

Mine both look like un-shaven, not so super capacitors

I reckon owners of later model 289, with the internal backup battery instead of the supercRapacitor, will be having similar adventures in a few years time,
especially if the battery sort of internally shorts itself or changes polarity, leaks, or just goes weird... :-//  as I've seen happen on PC Bios backup batteries   :o

Why oh why did Fluke not fit a CR2032 battery style socket,
someplace where vibrations and leakage won't cause drama  ???

or silicone bodge it down...  :horse:

These are their flagship meters FFS (For Flukes Sake) and 289 is still current,
this silly business does not instill repeat buyer confidence, especially at those 'forget mates rates' prices...

 
 
The following users thanked this post: bayati

Offline bayati

  • Contributor
  • Posts: 15
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #286 on: January 09, 2019, 10:16:24 am »
Just to confirm that Fluke could do much better job, it is a picture of Samsung E250 motherboard. The board is made in 2006. It has a ~3.4v lithium battery at least 13 years old. As you see, no leaking and I read 3v on it right now. It is as small as the supercap that is used in the 189/289/287 (if not smaller).
« Last Edit: January 09, 2019, 10:22:51 am by bayati »
 
The following users thanked this post: Electro Detective, Marco1971

Offline SteinarN

  • Newbie
  • Posts: 1
  • Country: no
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #287 on: January 29, 2019, 10:59:43 pm »
My first post on this forum.

I've had a 287 for years, but from the start it was draining the batteries in a month or two just by sitting on the shelf, clock also lost its time.
Tonight i was searching this problem and I came over a couple threads on this forum. I have identified the supercap on the pcb and ordered 4 of these:
https://no.mouser.com/ProductDetail/555-DSK3R3H224U-HL

I hope this will cure the illness of my othervise excellent meter.

I have a few small projects I am working on and I hope I can get some help from you for those projects when I start on them later :)

Steinar
 

Offline say

  • Contributor
  • Posts: 28
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #288 on: January 30, 2019, 09:10:44 pm »
It should. Good luck
 

Offline bayati

  • Contributor
  • Posts: 15
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #289 on: January 31, 2019, 12:48:43 pm »
Hi guys.
I changed my 289's supercap.

My meter is built in 2011 Firmware 1.12. Before sharing my experience you should know that Fluke 289 is about 700$ to 800$ in Iran and as it is illegal to be exported to Iran, it is very rare and there is no warranty for it.

First I searched for a replacement cap and the best I could find in Iran was MGC 5.5V 0.33F (size wise not the best choice but the best for me). I bought it for 1$. It is a Japanese Ricoh cap so I believe it will serve for a long time. 

I opened the meter and I was reading 0.670v on the old supercap. I shorted it but I still read 0.670 . I disordered it anyway and soldered the new cap with soldering iron set on 330c. 
It boot with no problem but the hart attack happened when I checked the meter. It was reading too high!! :scared:
 
I opened it and thanks to the resent video of Dave, I thought maybe the shield is touching the negative lead of the cap as it is too big. I put tape on it and now it works perfectly.   :-DMM

So i confirm that changing the cap does not break the meter. Just be ESD safe and do not turn it on before putting the new cap (I talked with someone who has removed the cap and it's working without it).
Tnx to all.

P.S. In the last picture the leads has been zeroed, that’s why it reads few counts down

P.S. look at the inductor above the supercap!!! I didn’t even touch it and I’m sure that I’m the first one who has opened it. It should be happened at the assembly. :box: :--

*If the pictures don't open use this link: https://photos.app.goo.gl/nmGjZ31ACf3h1owY8











« Last Edit: January 31, 2019, 04:23:23 pm by bayati »
 
The following users thanked this post: Electro Detective

Offline bayati

  • Contributor
  • Posts: 15
  • Country: ir
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #290 on: February 01, 2019, 07:55:46 pm »
As we are talking about supercaps, you can use the following pdf (datasheet) or pictures to estimate the lifetime of your supercap.
based on this equation mine should last for ~6.5 years.

https://www.mouser.com/ds/2/88/EDL-19670.pdf
« Last Edit: February 01, 2019, 09:13:36 pm by bayati »
 

Offline FrankE

  • Regular Contributor
  • *
  • Posts: 188
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #291 on: February 03, 2019, 01:25:23 pm »
The Supercap failed on my Fluke 289 so I sent it back to Fluke in the Netherlands in 2016 under an RMA for a warranty repair. I think it took about a week round trip,
I wouldn't have known the Supercap was the issue if it wasn't for a thread that year on here.
Every Friday I check all my batteries,replace/recharge as necessary. I also power on anything with a Supercap Friday and Tuesday (though I assume the 289s supercap is energised direct from the AA cells from before the switch. I noticed last Friday the time needed reset and the personal details have been wiped.

Changing the supercap is a warranty repair so shouldn't cost if still in warranty. The repair won't affect calibration and Fluke normally won't do a new factory calibration.
 
The following users thanked this post: Marco1971

Offline agdr

  • Supporter
  • ****
  • Posts: 105
  • Country: us
    • agdr Audio
10 year "limited" Fluke 289 warranty
« Reply #292 on: February 05, 2019, 10:31:28 pm »
So today I finally get around to calling Fluke's repair department here in the US about my two 289's that seem to have the supercap issue (both forget the date and time when the batteries are removed).   I learn that the 289's have a 10 year warranty.  The warranty repair does come with the non-data type calibration.  Maybe their policy on calibration with repairs varies from country to country..

Also: I asked - when 289s go back to Fluke for repair you don't need to include the test leads, batteries, or any case.  Just the meters.  Shipping both via FedEx Ground today was $18.  If I had fully insured probably would have been more like $26.  Lol tried UPS first and they literally could not process the return address. Pffftt.  There are some building numbers in front of part of it, and the UPS system didn't show that exact building.  That is why I stopped using UPS for outbound a loooong time ago.

I hope this information helps others! 
« Last Edit: February 07, 2019, 12:00:04 am by agdr »
 

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #293 on: February 06, 2019, 12:30:29 am »
I recently had my 289 in for a faded display issue in low ohms and possible supercap issue.  Spoke with Fluke TS to confirm the work that was performed and the rep stated no pay, no calibration.  I received the unit back yesterday.  Fluke replaced the mobo and included a calibration certificate.  All expense covered under warranty except shipping to Fluke.

Changes from my original unit
Board ID was 3 replaced with 5
Build date 6/2010 replaced with 11/2017
FW v1.16 replaced with v1.41

I purchased this unit last month on fleabay.. forgot to tell them I wasn't the original owner  :-DD
 
The following users thanked this post: FrankE, Marco1971

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au

I got into one of those rare lab moods early today  |O 
and decided to takes my chances and de-solder the supercrapacitor from both the 189 and 289,
with no intention of replacements same or better,

and damn the torpedoes...   >:(

Got everything set up ready to unsolder, ensure no ESR drama, thinking about somehow safely discharging any residual charge from the supercaps first,

and... WHOA! wait a minute!   ???


I went another way:   >:D

I CAREFULLY used my sharp sidecutters to cut the top metal tab off the supercap,
slowly lifted back the supercap to about 90 degrees max,
bent it back and forth a few times, supporting the back so no pressure on the PCB solder area,
till the bottom supercap tab snapped off clean

Cleaned up with isopropyl alcohol > Job done  :clap:


Both meters working perfect, set their clocks, and no more suss supercrapacitor issues to think about,
or rant on and Fluke bash at 189/289 posts 


FWIW: the supercap tab remnants can be unsoldered any time in the future if I    (no chance!  :-- )
or the next owner wants to wack in a supercap, and resume the saga I just ended



i.e. No Fluke was harmed by this procedure to remove a malignant tumor   :phew:


Happy Fluke Camper again   :-DMM   :-DMM

 

Offline Tim5000

  • Contributor
  • Posts: 12
  • Country: gb
FW v1.16 replaced with v1.41

Firmware v1.41 - interesting! I wonder what the changes in that are.
The latest available on their website is 1.16: https://www.fluke.com/en-gb/support/software-downloads/fluke-287-and-289-multimeter-firmware-update
 
The following users thanked this post: Marco1971

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
FW v1.16 replaced with v1.41

Firmware v1.41 - interesting! I wonder what the changes in that are.
The latest available on their website is 1.16: https://www.fluke.com/en-gb/support/software-downloads/fluke-287-and-289-multimeter-firmware-update

If your unit is working well as it was designed, I wouldn't worry about the firmware changes between v1.16 and v1.41
The major problems have already been addressed.  It does boot fast and the batteries last a long time though... :-DD
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au

(Reader Alert: Tired and perhaps one beer too many zone...  :popcorn: )


Would soldering in a small PCB coin battery holder be a proposition, so as to 'plug' in a supercapacitor (with its tabs cut)
giving the meter owner the options of replacement and or removal any time

or perhaps a pair of wires soldered in that tiny spot,
and the supercap and or holder parked elsewhere in a bigger or blank space on the PCB and properly secured. 

-----------------

Here's another brain bender:

is there a simple component addon to act as a one way valve so the supercapacitor can be charged with just enough voltage/current to keep the clock alive for a few minutes during a battery swap,
but ensure there is no way any voltage/current or a short or freaky polarity reversal can kick back at the charge source.

Maybe it's in there already, I don't know  :-//  and haven't gone that far to work it out (no schematic+lazy)

Faulty, underated, temperamental, or shorted supercaps and or the charging circuit overdriven into a short, may have caused grief to 189, 289 and 89IV owners over the years, 
awarding an undeserved bad rep to these fine meters

Fluke could/should have handled this matter a LOT better than just dumping it on the owners to sort it out, surfing for info at forums and Youtube,
or hoping the issue will eventually time out and blow away

or cash strapped battlers can go without, sledge the Fluking meter  :horse:  and wake up to affordable Aneng or better still Brymen/EEVblog meters  8)
and corporats toss them in the dumpster and buy cartons of new Flukes  :-DMM :-DMM
   

 :=\

 

Offline coromonadalix

  • Super Contributor
  • ***
  • Posts: 5879
  • Country: ca
@ Electro

This is too complicated for nothing,  a new supercap at 2-4 $ usd and voila, your troubles goes a way for a long time
 
The following users thanked this post: ModemHead

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
@ Electro

This is too complicated for nothing,  a new supercap at 2-4 $ usd and voila, your troubles goes a way for a long time


Hi mate, I'm aware of that option for a quick fix  :-+   but sometimes you can forget to check up on a multimeter for a long time, especially if the batteries last,

AND.. maybe have bad luck (that's me  :'( ) with a faulty or leaky brand or batch of supercapacitor AGAIN,

or worse a freaky ESR zap or CAL nuke during the unsolder/soldering  :scared:

and be back to the beginning of the drama, or worse off  |O

These meters are too expensive to easily replace, and no real qualified local techs near me to fix them... for any amount less than the price of a new meter  ;D



I'm leaving them out for now as the meters are working great and no more thinking about what's going on inside or what may happen    :phew:

and I don't mind updating the logger time clock once every few months (or longer) on battery change

and due to Daylight Savings Time here, you have to set the clock back or forward by one hour  :-//   twice a year anyway

 
« Last Edit: April 15, 2019, 02:08:13 am by Electro Detective »
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Three years ago I replased leaked cupercap in my Fluke 287 on Elna DSK-3R3H224U-HL. It still working well and keep time on my meter for four hour. That is enought for recharging Eneloop batteries that I use with my meter. Eneloop is really cool solution for power this meter.
I do not see that supercaps is real issue.
 
The following users thanked this post: Daruosha, Marco1971

Offline rauldm

  • Regular Contributor
  • *
  • Posts: 155
  • Country: mx
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #301 on: September 13, 2019, 04:51:35 am »
Hi, I have a 289 rev 15, the fucking meter works very good, but it has corrosion on cap 145, Fluke says in its web site, the 42830001 series it comes with new firmware "1.41", and all meters with  lower serial numbers can't update to this firmware, I supposed the new Fluke, it comes with new PCB board and general improvements, someone knows about this? the cap leakage is solved? another good meter with same features? I have the u1282a, but is slow in cap and ohm modes, the metrawatt is good option? or only is stupid expensive?
Regards.
 

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #302 on: September 13, 2019, 05:17:06 am »
Hi, I have a 289 rev 15, the fucking meter works very good, but it has corrosion on cap 145, Fluke says in its web site, the 42830001 series it comes with new firmware "1.41", and all meters with  lower serial numbers can't update to this firmware, I supposed the new Fluke, it comes with new PCB board and general improvements, someone knows about this? the cap leakage is solved?
I have a Fluke 289 with FW v1.41 and the serial starts with 3826####.  My 289 was sent in for repair for the low ohms fade issue in January 2019.  Fluke replaced the board and calibrated the unit prior to shipping it back.  I asked about the supercap issue in the original work order and was told that is was no longer a problem.
 
The following users thanked this post: Marco1971, rauldm

Offline rauldm

  • Regular Contributor
  • *
  • Posts: 155
  • Country: mx
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #303 on: September 13, 2019, 04:10:29 pm »
Hi, I have a 289 rev 15, the fucking meter works very good, but it has corrosion on cap 145, Fluke says in its web site, the 42830001 series it comes with new firmware "1.41", and all meters with  lower serial numbers can't update to this firmware, I supposed the new Fluke, it comes with new PCB board and general improvements, someone knows about this? the cap leakage is solved?
I have a Fluke 289 with FW v1.41 and the serial starts with 3826####.  My 289 was sent in for repair for the low ohms fade issue in January 2019.  Fluke replaced the board and calibrated the unit prior to shipping it back.  I asked about the supercap issue in the original work order and was told that is was no longer a problem.

The first board with ohms issues, do you know what revision number was?
Your new board you know what revision is?
if Fluke puts new firmware from 42830001 meter series I think your new board is last revision.

Really I searched another sustitute for this meter, but I think another meter is not gathed all features for same price range.

Thanks for your answer, really I like this meter, I have 87v,28ii and 117, but this meter for me is better than others, even my others meter are just complement for my 289.
 

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #304 on: September 13, 2019, 04:40:07 pm »
In the meter info screen, the Board ID was 3 and was built on 6-26-10 with FW v1.16 / v.88.

My replacement Board ID is 5, built on  11-28-17 with FW v1.41 / v.88.

I noticed on the RMA that the serial # remained the same.  Fluke probably just updated the number in the board for administrative purposes.

Service was fantastic from Fluke. One email and a phone call to resolve the defect.  I bought my meter used on ePray for a great price and Fluke covered it under warranty. No hassles, new meter with calibration certificate.  Can't beat that.

Yikes!  I just realized the same info you requested is in a previous post #299 by me.  My memory is failing...

Sent from my Moto Z (2) using Tapatalk
« Last Edit: September 13, 2019, 04:49:46 pm by mr.fabe »
 

Offline rauldm

  • Regular Contributor
  • *
  • Posts: 155
  • Country: mx
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #305 on: September 13, 2019, 06:52:06 pm »
Ok, fluke eliminates the capacitor 145 on new board? How solves the problem? New firmware has it improves?
 

Offline mr.fabe

  • Regular Contributor
  • *
  • Posts: 171
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #306 on: September 16, 2019, 07:49:50 pm »
I don't know what is on the new board but was told by Fluke that the supercap problem is not an issue.  I have no reason to doubt Fluke and if there were a problem with losing the date/time then it would be covered under warranty.  The supercap problem is a few years old and Fluke mentions that they were going to make changes on future revisions of the 287/289.
« Last Edit: September 16, 2019, 08:00:33 pm by mr.fabe »
 

Offline rauldm

  • Regular Contributor
  • *
  • Posts: 155
  • Country: mx
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #307 on: September 18, 2019, 07:15:58 pm »
I talked with Fluke support about this capacitor, I attached datasheet of capacitor.  In datasheet it comes this:
1.- Re-flow soldering method available (250 °C)
Note : 1. When the capacitor is being used in a high temperature and high humidity environment for a long period, brown deposited
materials might be found around the sealing area. However, the materials are insulators and will not affect any electrical
characteristics of the capacitor.
2. Do not use refl ow soldering when cell voltage is above 0.3 V.

Really I dont know is or not recomendable an user change this capacitor, I think a meter is a tool in that you trust your life, if you don't doing the procedure correctly perhaps will not be trustworthy again.
 
The following users thanked this post: BravoV

Offline CDaniel

  • Frequent Contributor
  • **
  • Posts: 411
  • Country: ro
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #308 on: September 18, 2019, 08:56:56 pm »
The model 187 doesn't have that supercap fitted ... so it is just for not losing the logging data when the batteries are consumed . A user that doesn't do logging every day could use the meter ( at least 189 ) very well with that cap just removed if it goes bad .
« Last Edit: September 18, 2019, 08:59:30 pm by CDaniel »
 
The following users thanked this post: Marco1971

Offline BravoV

  • Super Contributor
  • ***
  • Posts: 7547
  • Country: 00
  • +++ ATH1
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #309 on: September 19, 2019, 03:41:05 am »
2. Do not use refl ow soldering when cell voltage is above 0.3 V.

Hmm ... the devil is in the details isn't it ?

I guess the manufacturers need to babysit short the cap before hand if use reflow method.  ::)

Learned something today, thanks.  :clap:

Offline Marco1971

  • Contributor
  • Posts: 29
  • Country: it
@ Electro

This is too complicated for nothing,  a new supercap at 2-4 $ usd and voila, your troubles goes a way for a long time

Not much long time...two and half years for me and my Fluke 287.

Marco1971
 

Offline paddyred

  • Newbie
  • Posts: 7
  • Country: gb
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #311 on: December 22, 2020, 03:33:29 pm »
Anyone know what the difference is on these new 289 boards? Rev5? I just sent mine in for repair after the supercap issue (produced 2016) and fluke just sent me a BNIB one with new leads, accessories and all the trimmings. New one made August 2020.  ;D ;D ;D
 

Offline Xandinator

  • Contributor
  • Posts: 34
  • Country: 00
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #312 on: December 10, 2021, 10:36:42 am »
Unfortunately haven't been so lucky with their customer service. Another 289 confirmed working w/o cap tho.
 

Offline rivest

  • Contributor
  • Posts: 10
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #313 on: September 27, 2022, 08:50:17 am »
Fluke 189 working without cap here.
 

Offline DarkskyZ

  • Newbie
  • Posts: 2
  • Country: fr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #314 on: November 07, 2022, 01:13:27 pm »
Hi there, just a quick first post here after reading about everything about this supercap mess on these Flukes. This particular post made me act on it and I am glad I did.

I got my 289 second hand for around 200€ about 2 years ago as I needed some recording capability for a one time job. I am using it on my bench ever since as a very nice bit of kit.
But there is no way I would agree to pay Fluke France 316.5€ + 30€ for S/H they are asking for a repair while in the mean time they told me that "lifetime warranty" is only available to original owner.

I also like (not) how their definition of " lifetime" is their own view on what lifetime is.  :-//

So... the DIY, soldering iron way I went. It has been an easy 2 minutes job with easy access to the soldering points even for me being less than competent with a soldering iron in hand.  :phew:

I prefer to have peace of mind with this thing not leaking all over inside the meter even if that means having to set date and time from time to time.

Anyway, thank you all for confirming that a 289 doesn't need to have its supercap to work fine. That made me confident to remove it and I confirm it just works.  :)
« Last Edit: November 07, 2022, 04:48:59 pm by DarkskyZ »
 

Offline BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1191
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #315 on: November 07, 2022, 01:53:58 pm »
If removing the super cap means that it would lose time when you remove the AA batteries then it's not a problem. The date and time only needed when you do the recording.
 

Offline DarkskyZ

  • Newbie
  • Posts: 2
  • Country: fr
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #316 on: November 07, 2022, 04:51:33 pm »
Yes, that is all there is to it.
Except for hardcore everyday logging people that is not a problem.
I would assume those people probably don’t rely on battery operated meters anyway so, to me at least, that supercap (when working and not leaking) is more of a convenience than a need.
« Last Edit: November 07, 2022, 06:05:39 pm by DarkskyZ »
 

Offline BeBuLamar

  • Super Contributor
  • ***
  • Posts: 1191
  • Country: us
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #317 on: November 07, 2022, 06:04:11 pm »
Lukily my 289 (2017) and 287(2022) don't have the leaky capacitor yet. But if it ever does I would simply remove the cap. I got the new 287 because someone beat the hell out of one of the 287's in the company I worked for. My boss had it laying in the junk box so I asked him to give it to me. I sent it in to Fluke and they gave me the brand new one under warranty.
 

Offline Electro Detective

  • Super Contributor
  • ***
  • Posts: 2715
  • Country: au
Re: Fluke 189 with leaking surface mount supercap (also Fluke 287, Fluke 289)
« Reply #318 on: January 16, 2023, 10:03:13 am »
Revisiting this post:

As previously stated, in peaved frustration, and willing to reset clocks on AA battery changes, I removed the superCRAPicitors from both Fluke 189 and 289 two years ago,
and they have been flawless.

I forgot about it till I came across this post.

Seriously folks, either fit a new capacitor, or lose the bearded leaky one,
or.. leave it in there at your risk  :scared:
and toss coins the meter does not go weird with a shorted or halfass behaving cap stressing the meter's smarts. Your call  :popcorn:


I sincerely thank the many members who commented across many posts about this so that I could make up my mind which way to roll back then

The Disclaimer is just because my quick fix hacknique worked for me, YMMV

These meters are absolute keepers for me, especially now that I don't leave AA batteries in there during non use periods to leak!


Surely there's a special lake in Hell for battery industry bean counters, and their shareholder dividend junkies   >:D




 
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf