Author Topic: Calibration : Fluke 52 II Series thermometer  (Read 5071 times)

0 Members and 1 Guest are viewing this topic.

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Calibration : Fluke 52 II Series thermometer
« on: June 10, 2014, 02:27:25 pm »
I'm back home, and what we have here for today?



Guesses?  ;D
« Last Edit: June 11, 2014, 12:42:12 am by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 

Offline TiNTopic starter

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: Fluke 52 II Series performance
« Reply #1 on: June 10, 2014, 02:28:05 pm »
Fluke service manual with calibration procedure.
Fluke recommends 5700A calibrator, but we will use poor-man DC standard, EDC MV106.

Initial inspection, "before calibration" state.
On right side - ADC calibration fixture (two-copper terminal jack made from two thermocouple T jacks).



Cal sticker is tampered.



-82°C with tested K-type thermocouple hanging in +25C room temp ain't right.



Closeup for connector used for voltage source (EDC MV106 DC standard calibrated to Keithley 2001 with Tek cal 5.FEB.2014) connection to Fluke 52-2 inputs.
I used copper wire from UTP LAN cable.

0.0000v set - meter shows 20-30C...



Calibration A-D convertor by providing 80.0000mV and -8.00000mV to inputs.



-8.00000mV read as -7996.9 uV, which is in range for calibration requirements of –7995.0 to –8005.0.

Calibration reference point:



REF point offset is 3.8C.



After corrected offset:



"After calibration" Performance test values (K-Type Thermocouple) with icebath.
Channel 1 on left, channel 2 on right:

1. 0.0000 mVDC / 0.0 °C



2. 0.9193 mVDC / 23.0 °C



3. -5.7297 mVDC / -190.0 °C



4. 40.8853 mVDC / 990.0 °C



5. 48.8382 mVDC / 1200 °C



Same K-type probe reads correct +25.6°C now.



Thanks for watching :)



P.S.

Used gear:

Calibrated Keithley 2001 Multimeter (5.FEB.2014)
Keithley 2002 Multimeter (10.APR.2007)
Calibrated EDC MV106 DC Voltage standard +-30ppm/°C (8.FEB.2014)
Calibrated Keithley 2400 Sourcemeter (14.JAN.2014)
« Last Edit: June 11, 2014, 12:46:23 am by TiN »
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf