Got the urge for a large LED bench DMM.  Between the Fluke 8505A and the HP 3456A, setting going price aside, which would you choose and why?

I do like the possibility for a shunt module on the Fluke, the A/C module is not a big concern but would be nice.  I know the HP is just dead solid stable, I've not found too much info on stability for the Fluke, but I suspect its close to the same.  The Fluke is not quite as deep, and would fit table where I would place it better.  Just want a new toy to play, no specific need. 
 

Offline pelule

  • Frequent Contributor
  • **
  • Posts: 513
  • Country: de
  • What is business? It’s other people’s money
Re: Fluke 8505A vs HP 3456A
« Reply #1 on: January 25, 2018, 12:44:08 am »
I am proud owner of both HP 3456A and Fluke 8506A (similar to 8505A except it offers thermal RMS conversion).
Simple said. It depends...(on the options installed at the Fluke).

HP3456A
HP3456A is a DVM
it has GPIB
it measures AC voltage (RMS by square root method), DC voltage 2w Ohms and 4w Ohms.
It does not offer current, neither DC nor AC (even I have seen offers claiming it does, but that may defekt  :palm:).
HP 3456A offers a very high accuracy for a 6.5 digit meter and seems to have very low drift over time (see thread here).

8506A/8505A
I have all available modules (options) for my 8506A
so it has GPIB
it measures voltage DC & AC RMS (thermal conversion for 8506A only) 2w and 4w Ohms and DC Current.
It also offers a 7.5 digit resolution in the 10V DC range. :-+
Since 2014 I have the equipment (HP 3458A) to measure the drift over time.
My 8506A has similar low drift as my 3456A. :-+

In fact I never want to miss any of them - I am using both for GPIB bus controlled logging.
/PeLuLe
You will learn something new every single day
 
The following users thanked this post: kj7e

Offline pigrew

  • Frequent Contributor
  • **
  • Posts: 680
  • Country: us
Re: Fluke 8505A vs HP 3456A
« Reply #2 on: January 25, 2018, 01:24:12 am »
My 8505A has a bit of 60-Hz mains hum from its transformer. It's annoying, but doesn't seem to effect the measurements. Mine has the DC options + GPIB (no AC measurements). The 8506A is probably better for AC, anyway.

When I bought it off of eBay, its calibration seemed a bit off when compared to a 34401A with a fresh calibration. I adjusted it to match the 34401A fairly easily. I don't trust any of my equipment to measure better than 10 ppm, so I can't say what it's been doing since then.

The user interface is a bit unintuitive. Without using it often, you'll forget how to set up filtering or other math operations (for example).

I read that some models of the 3456A have a fan while others do not, so they may or may not be noisier.

-Nathan
 
The following users thanked this post: kj7e

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #3 on: January 25, 2018, 01:27:54 am »
Interesting about the fan on the 3456, Ill have to look in to that.

Took about 2 seconds and found some notes on the various series of 3456;
https://www.febo.com/pipermail/volt-nuts/2010-November/000626.html
« Last Edit: January 25, 2018, 01:29:47 am by kj7e »
 

Offline guenthert

  • Frequent Contributor
  • **
  • Posts: 706
  • Country: de
Re: Fluke 8505A vs HP 3456A
« Reply #4 on: January 25, 2018, 12:26:10 pm »
Only very early generations of the HP3456A had a fan.  Mine doesn't and it gets only modestly warm, not a big burner.  The HP3456A is very stable and has a very linear ADC, it can measure resistance up to 1GOhm and has a built-in calculator, which can be handy to remove offsets or gain errors, but otherwise offers few functions.  It is calibrated with externally accessible potentiometers, so there's no danger of calibration data loss due to a drained battery.  It's huuuge.  Compared to the case, the LED display is pitifully small, but bright and lasts for decades.  They truly don't make 'em, like they used to.  Watch out for US (60Hz, option -360) vs. European (50Hz, option -350) versions.

The Fluke 8506A with its thermal transfer ACV converter seems more desirable, but tends to cost more. :-//
Both are generally overkill and of interest only to volt nuts.  ;D
 

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #5 on: January 25, 2018, 06:47:15 pm »
I already have a 34465a and DMM7510 for the voltnutting.  Also have both an HP 3457a and HP 3478a.  Leaning toward the Fluke, not many to pick from on the used market right now, at least not clean with the options I would be looking for.  Where as the 3456a can be found with ease.
 

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 131
  • Country: th
Re: Fluke 8505A vs HP 3456A
« Reply #6 on: January 25, 2018, 07:17:30 pm »
Get a new table with space for both. Forget all options except GPIB for the Fluke.

I have seen your videos and your posts. You do not need another DMM. Both the HP and the Fluke are voltmeters. Obsolete voltmeters for people with odd interests.

The soul of the 3456A is that nothing compares to the engineering packed around an LM399. And the Fluke uses a refamp (which when pampered enough is good enough to be the reference of 732B).

Get one of each, check that they are good, print out the manuals, and put the meters in range 10V DC never to be touched again.

[just my 10.00000 dollars]

 
The following users thanked this post: kj7e

Offline wn1fju

  • Frequent Contributor
  • **
  • Posts: 553
  • Country: us
Re: Fluke 8505A vs HP 3456A
« Reply #7 on: January 26, 2018, 12:30:41 am »
One thing that hasn't been mentioned is calibration, if you are so inclined.  I have both the HP 3456A and the Fluke 8506A.  The 3456A is
quite simple to calibrate and involves turning a few pots (accessible from a removable section of the front panel).  The 8506A, however,
was a nightmare, at least for me.  You can do a "hardware" calibration, which involves turning pots, or a "software" calibration which stores
zeros and gains into nonvolatile memory.  The manual is clear as mud on the entire subject.  I ended up trying the software cal, but the
linearity seemed to suffer and was much worse than the 3456A.  So I ended up doing the hardware cal.  This restored the linearity of the
8506A to 3456A levels, and why not - after all, the hardware cal has you adjust a whole bunch of intermediate points in the A/D ladder. 
Anyway, I think I ended up doing both the hardware and software cals, maybe that was a mistake.  As I said, the manual was very confusing
in my opinion. 

Both DMMs are nice units.  But for serviceability, I like the HP better than the Fluke with its series of self-contained modules (for which you
need extenders).
 
The following users thanked this post: kj7e

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #8 on: January 26, 2018, 01:31:43 am »
One thing that hasn't been mentioned is calibration, if you are so inclined.  I have both the HP 3456A and the Fluke 8506A.  The 3456A is
quite simple to calibrate and involves turning a few pots (accessible from a removable section of the front panel).  The 8506A, however,
was a nightmare, at least for me.  You can do a "hardware" calibration, which involves turning pots, or a "software" calibration which stores
zeros and gains into nonvolatile memory.  The manual is clear as mud on the entire subject.  I ended up trying the software cal, but the
linearity seemed to suffer and was much worse than the 3456A.  So I ended up doing the hardware cal.  This restored the linearity of the
8506A to 3456A levels, and why not - after all, the hardware cal has you adjust a whole bunch of intermediate points in the A/D ladder. 
Anyway, I think I ended up doing both the hardware and software cals, maybe that was a mistake.  As I said, the manual was very confusing
in my opinion. 

Both DMMs are nice units.  But for serviceability, I like the HP better than the Fluke with its series of self-contained modules (for which you
need extenders).

That is a great point!
 
The following users thanked this post: kado

Offline edavid

  • Super Contributor
  • ***
  • Posts: 3381
  • Country: us
Re: Fluke 8505A vs HP 3456A
« Reply #9 on: January 26, 2018, 02:38:58 am »
8506A/8505A
I have all available modules (options) for my 8506A
so it has GPIB
it measures voltage DC & AC RMS (thermal conversion for 8506A only) 2w and 4w Ohms and DC Current.

Clarification:  The 8506A has only one slot free, so you can install either the ohms converter or the current shunts, not both at the same time.

My 8505A has a bit of 60-Hz mains hum from its transformer. It's annoying, but doesn't seem to effect the measurements.

My 8506A has this problem also.  It's bad enough that I have to turn it off when I'm not using it.


Also, the 3456A has a small, hard to read display (better than the 3457A though).
 

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #10 on: January 26, 2018, 08:55:38 pm »
Pulled the trigger on a 3456a, looks clean, appears to work.  The serviceability issue was the kicker.
 

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 131
  • Country: th
Re: Fluke 8505A vs HP 3456A
« Reply #11 on: January 26, 2018, 10:20:05 pm »
Very good. An amazing box. I bought one (a month ago) with HP calibration and anti-tampering stickers on the front panel and case screws.

I could have bought locally another 34401A just for the shipping cost of the 3456A. But I knew I wanted the 3456A for nostalgic reasons.

Since HP went Agilent around the year 2000, the meter hasn't been adjusted for some 17 years. When measuring references in range 10V DC, it tracks my best other instruments to within a few ppm (typically less then 20uV - no more than 50) from 20 - 30 degC.

An amazing machine. I'll take it to my grave  ::)



 
The following users thanked this post: kj7e, denimdragon

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #12 on: January 30, 2018, 05:13:30 am »
Had a few little issues off the bat with the new 3456a, consent resetting after power on.  Just needed the digital board to be reseated.  Literally blew out the cobwebs and gave it a good cleaning, came out not too bad looking.  Date code of 1984 on most of the parts.  Found the Ohms did not work, no current source, found F201 fuse broke (not blown), Ohms works fine now.  Had a Cal sticker from 2001 (edit: make that 1990) with the seals still in place.  Measures really darn close to my DMM7510  :-+









« Last Edit: January 30, 2018, 02:37:20 pm by kj7e »
 

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 131
  • Country: th
Re: Fluke 8505A vs HP 3456A
« Reply #13 on: January 30, 2018, 11:29:59 am »
 :-BROKE>:D x :-/O:-DMM :-+ :-+ :-+
 

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 131
  • Country: th
Re: Fluke 8505A vs HP 3456A
« Reply #14 on: January 30, 2018, 11:53:03 am »
I looked at PhotoBucket, but couldn't find a picture comparing the meters.
If you have the time to spare, please post some data (10v / a few temp points).

I don't see a calibration sticker on any of the two front panel screws. Have you adjusted the meter behind the front panel, or do you suppose it is within spec after 17 years?
 
The following users thanked this post: kj7e

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #15 on: January 30, 2018, 02:32:11 pm »
I looked at PhotoBucket, but couldn't find a picture comparing the meters.
If you have the time to spare, please post some data (10v / a few temp points).

I don't see a calibration sticker on any of the two front panel screws. Have you adjusted the meter behind the front panel, or do you suppose it is within spec after 17 years?

I removed all the stickers during the cleanup, here is what It looked like as received (sorry, last cal 1990);




I was rather displeased with the packing, the sender just tossed the meter in an oversized box with white styrophome peanuts, the meter pulverized the peanuts and they where embedded in every single hole.  The first hour I spent cleaning styrophome out of the meter.

No adjustments made yet and it was about 10uV (1ppm) off my DMM7510 at 10v last night!  I left it running overnight, Ill take some comparisons vs my recently calibrated 34465A and DMM7510.
« Last Edit: January 30, 2018, 02:45:30 pm by kj7e »
 

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #16 on: January 30, 2018, 04:48:41 pm »
All at 23 Deg C;
1mV


100mV


1V


10V


After running all night, about 8ppm high compared to the DMM7510. Not bad for not being touched in 27 years.
« Last Edit: January 30, 2018, 06:12:11 pm by kj7e »
 
The following users thanked this post: AG7CK

Offline guido

  • Regular Contributor
  • *
  • Posts: 207
  • Country: nl
Re: Fluke 8505A vs HP 3456A
« Reply #17 on: January 30, 2018, 06:57:00 pm »

Your 7510 seems a bit off  >:D

As to the 3456A. The power supply caps probably need changing and also you could encounter a bad ROM (or more). One of my devices had that, replaced them with EPROMS.
 
The following users thanked this post: kj7e, denimdragon

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #18 on: January 31, 2018, 03:09:03 am »
There, plenty of room  ;D

 

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 131
  • Country: th
Re: Fluke 8505A vs HP 3456A
« Reply #19 on: January 31, 2018, 04:30:28 am »
The meter looks very nice. (So does the bench and the LTZ1000 box in your videos. Wish it was me - my stuff is nothing but unfinished mess ...).

I think you have been lucky with this meter. First - the stickers show that it has not been vandalized. Secondly - the "722" means port 7 HP-IB (GPIB) and address 22. Your meter has probably had its former life (before it went into storage) in a serious test/measurement system controlled by a HP computer with HP-IB controller. Something like this:

http://www.hp9845.net/9845/tutorials/hpib/

I am impressed - 8.4ppm or so on the native 10v ADC in 27 years - that is some 0.31 ppm per year in average. It is a pity they truncated the 7th digit (see the manual p. 8-12 Table 8.5) from the ADC (while they did include a 7th digit from GPIB on the less stable 3457A and 34401A.)


 
The following users thanked this post: kj7e

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #20 on: January 31, 2018, 07:25:12 pm »
Only been playing with the 56 for two days now, I'm starting to wonder why more people don't have these?  What a great meter for the price.
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Fluke 8505A vs HP 3456A
« Reply #21 on: February 01, 2018, 08:13:24 am »
Shhh. Not too loudly, now. The prices will go up! ;D
TEA is the way. | TEA Time channel
 
The following users thanked this post: montemcguire, denimdragon

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1245
  • Country: au
  • Watts in an ohm?
Re: Fluke 8505A vs HP 3456A
« Reply #22 on: February 01, 2018, 11:47:29 am »
There, plenty of room  ;D


Nice bench there btw, reloading press on the right too !. Pistol cal ? judging be the size of the dies.
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #23 on: February 01, 2018, 01:24:21 pm »

Nice bench there btw, reloading press on the right too !. Pistol cal ? judging be the size of the dies.

On the Redding T7 press there, 338 Lapua and 458 SOCOM dies.  I also reload for 45/70 govt, 44mag, 475 Linebaugh, 45 ACP. I have two other presses not in view.
 

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1245
  • Country: au
  • Watts in an ohm?
Re: Fluke 8505A vs HP 3456A
« Reply #24 on: February 02, 2018, 03:33:29 am »
Jeez must get my eye ball re called  :D. Some decent chamberings there. Fan of the 45/70 too.  :-+
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #25 on: February 02, 2018, 04:57:56 am »
I knew I would need to address the power supply caps, ran for two days before one shorted out.  Ill have new caps ordered in the morning.



I greatly suspect its been many years since this particular instrument has seen any operation. 
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Fluke 8505A vs HP 3456A
« Reply #26 on: February 02, 2018, 05:23:49 am »
That was fast. But, better now than in the middle of a project.
TEA is the way. | TEA Time channel
 

Offline lowimpedance

  • Super Contributor
  • ***
  • Posts: 1245
  • Country: au
  • Watts in an ohm?
Re: Fluke 8505A vs HP 3456A
« Reply #27 on: February 02, 2018, 05:28:49 am »
I'll bet they made one long groan at that first power up  :-DD

Some HP3456a service notes I uploaded 2012 that I found at work, thread here;
https://www.eevblog.com/forum/testgear/hp3456a-service-notes-for-owners-info!/msg144820/#msg144820
Once all the caps in the PS are replaced the meter will most likely be still chugging away long after we are gone  :D.

Enjoy
The odd multimeter or 2 or 3 or 4...or........can't remember !.
 
The following users thanked this post: kj7e

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #28 on: February 02, 2018, 07:25:02 pm »
Got to love the old manuals.  This showed up today,



Caps ordered.  I followed TiN's lead and up'ed the size a bit;
https://xdevs.com/fix/hp3456a/#psu_fix
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Fluke 8505A vs HP 3456A
« Reply #29 on: February 04, 2018, 02:17:18 am »
That's a pretty good trick: They revised it a year after it was already printed. :-DD
TEA is the way. | TEA Time channel
 

Offline kj7eTopic starter

  • Frequent Contributor
  • **
  • Posts: 911
  • Country: us
  • Damon Stewart
Re: Fluke 8505A vs HP 3456A
« Reply #30 on: February 05, 2018, 10:24:35 pm »
That's a pretty good trick: They revised it a year after it was already printed. :-DD

I wondered about that too, but before I even saw it.

New caps, fixed and improved;


(Found this guy NOS on eBay)


(Both meters where cold)


Edit, Digikey part nubers for Rubycon 105C caps with correct lead spacing, a bit extra capacitance and voltage rating;
Index   Quantity   Part Number   Manufacturer Part Number   Description   Customer Reference   Backorder   Unit Price   Extended Price
1   3   1189-3012-ND   50ZLH1000MEFC16X25   CAP ALUM 1000UF 20% 50V RADIAL      0   1.62000   $4.86
2   3   1189-1422-ND   50YXJ2200M16X35.5   CAP ALUM 2200UF 20% 50V RADIAL      0   1.93000   $5.79
3   2   1189-4026-ND   63PX220MEFC10X16   CAP ALUM RAD      0   0.73000   $1.46
4   2   1189-4028-ND   63YXJ100M10X12.5   CAP ALUM 100UF 20% 63V RADIAL      0   0.52000   $1.04
5   1   1189-3906-ND   35MXG5600MEFCSN22X35   CAP ALUM 5600UF 20% 35V SNAP      0   4.23000   $4.23
                     Subtotal   $17.38

Now to let it run for a few weeks.
« Last Edit: February 06, 2018, 06:24:51 am by kj7e »
 

Offline bitseeker

  • Super Contributor
  • ***
  • Posts: 9057
  • Country: us
  • Lots of engineer-tweakable parts inside!
Re: Fluke 8505A vs HP 3456A
« Reply #31 on: February 06, 2018, 05:33:50 am »
Looking good, inside and out.
TEA is the way. | TEA Time channel
 

Offline cellularmitosis

  • Supporter
  • ****
  • Posts: 1111
  • Country: us
Re: Fluke 8505A vs HP 3456A
« Reply #32 on: April 15, 2018, 10:01:03 am »


I am impressed - 8.4ppm or so on the native 10v ADC in 27 years - that is some 0.31 ppm per year in average. It is a pity they truncated the 7th digit (see the manual p. 8-12 Table 8.5) from the ADC (while they did include a 7th digit from GPIB on the less stable 3457A and 34401A.)

Do you know where this truncation happens?  I took a look at the manual but couldn’t find anything obvious.  There is a serial link between the inguard processor and the outgaurd processor.  I wonder if that could be spied on to get more resolution?  The theory of operation section describes something about how a comparator check is only performed every 5th clock cycle, so it seems there’s two more bits worth of resolution hiding in there somewhere...
« Last Edit: April 15, 2018, 10:03:36 am by cellularmitosis »
LTZs: KX FX MX CX PX Frank A9 QX
 

Online alm

  • Super Contributor
  • ***
  • Posts: 2841
  • Country: 00
Re: Fluke 8505A vs HP 3456A
« Reply #33 on: April 15, 2018, 10:19:50 am »
If there is enough noise in the signal, you might be able to extract the extra resolution by just averaging readings. That's how Datron extracted an extra digit of resolution on some of their 7.5 digit meters. That would be easier than trying to hack the internal communication. Obviously this will slow down your effective sample rate.
 
The following users thanked this post: cellularmitosis

Offline cellularmitosis

  • Supporter
  • ****
  • Posts: 1111
  • Country: us
Re: Fluke 8505A vs HP 3456A
« Reply #34 on: April 15, 2018, 09:36:26 pm »
I figured out the seventh digit  8)

https://www.eevblog.com/forum/metrology/hp-3456a-7th-digit/

Also, my 3456A was also inoperative on arrival (-4000... on test).  As many others have experienced, my 33V rail was low (around 7V).  I checked all of the caps on the power supply board with an ESR meter and all looked ok except for C7, which had failed open circuit.  Just replacing that cap (47uF, 50V) was enough to get the meter operational again.
« Last Edit: April 15, 2018, 09:40:35 pm by cellularmitosis »
LTZs: KX FX MX CX PX Frank A9 QX
 
The following users thanked this post: edavid, kj7e, URI

Offline AG7CK

  • Regular Contributor
  • *
  • Posts: 131
  • Country: th
Re: Fluke 8505A vs HP 3456A
« Reply #35 on: April 16, 2018, 11:27:58 pm »

Do you know where this truncation happens?  I took a look at the manual but couldn’t find anything obvious.  There is a serial link between the inguard processor and the outgaurd processor.  I wonder if that could be spied on to get more resolution?  ....

Sorry - I never saw this post before today. It never occurred to me that the rounding was done outguard (because the manual talks about readings per second / transfer speed / reading size in bits between the inguard and outguard processors. Hence I never suspected they send unused information to the outguard processor).

Very good work. Thanks a lot.

 