Fluke 87 V with built-in thermometer?
« on: April 24, 2017, 06:26:57 am »
When did they start putting built-in thermometers in Fluke 87 V multimeters?

It is the second item down on the features tab on the Fluke page:
http://en-us.fluke.com/products/digital-multimeters/fluke-87v-digital-multimeter.html#features
 

Offline IanB

  • Super Contributor
  • ***
  • Posts: 11891
  • Country: us
Re: Fluke 87 V with built-in thermometer?
« Reply #1 on: April 24, 2017, 06:39:57 am »
Every meter that can read the temperature of a thermocouple has a built-in thermometer. When determining the thermocouple reading the meter has to measure the temperature of the reference junction inside the meter, and to do this it needs an internal thermometer.

You can make a meter display the internal temperature by shorting out the input terminals instead of plugging in a thermocouple. Some meters will do this automatically when no thermocouple is plugged in, but many don't.
 

Offline joseph nicholas

  • Frequent Contributor
  • **
  • Posts: 408
  • Country: mx
Re: Fluke 87 V with built-in thermometer?
« Reply #2 on: April 24, 2017, 08:32:54 am »
Its seems a bit wierd.  Who asked the meter to compare 2 readings?  Is all you want is to have the meter read the value at the K type thermocouple which is a standard value across all K type thermocouple. 
 

Offline IanB

  • Super Contributor
  • ***
  • Posts: 11891
  • Country: us
Re: Fluke 87 V with built-in thermometer?
« Reply #3 on: April 24, 2017, 08:38:59 am »
The thermocouple asked the meter to compare two readings because that's how thermocouples work. A thermocouple does not measure the junction temperature, it measures the difference in temperature between two junctions. One junction is placed at the point to be measured, and the other junction is placed at some known reference temperature. The EMF generated by the thermocouple is a function of the difference in temperature between the two junctions. Therefore if the temperature of the reference junction is known the temperature of the measurement junction can be determined.
 

Offline mzacharias

  • Frequent Contributor
  • **
  • Posts: 709
  • Country: us
Re: Fluke 87 V with built-in thermometer?
« Reply #4 on: April 24, 2017, 09:49:53 am »
Every meter that can read the temperature of a thermocouple has a built-in thermometer. When determining the thermocouple reading the meter has to measure the temperature of the reference junction inside the meter, and to do this it needs an internal thermometer.

You can make a meter display the internal temperature by shorting out the input terminals instead of plugging in a thermocouple. Some meters will do this automatically when no thermocouple is plugged in, but many don't.

My 87V (acquired in2010) displays OPEn when the thermo probe is not plugged in. It has always been this way, and makes temperature readings perfectly well when the probe is inserted. I have had other meters which do have the internal temperature reference.
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #5 on: April 24, 2017, 10:42:16 am »
The Fluke 185 always shows cold junction temperature on the upper display.
On the main display, with decimal:
1. No probe, cold junction temperature
2. Probes open or shorted, cold junction temperature
3. K-type thermocouple, hot junction temperature

There is a procedure to zero the thermocouple voltage offset, by first immersing its bulb in an ice bath, to increase the accuracy of all further measurements.
« Last Edit: April 24, 2017, 10:48:25 am by Wytnucls »
 

Online tooki

  • Super Contributor
  • ***
  • Posts: 11505
  • Country: ch
Re: Fluke 87 V with built-in thermometer?
« Reply #6 on: April 24, 2017, 11:17:57 am »
Every meter that can read the temperature of a thermocouple has a built-in thermometer. When determining the thermocouple reading the meter has to measure the temperature of the reference junction inside the meter, and to do this it needs an internal thermometer.
I'm pretty sure that wasn't the OP's actual question. I'm pretty sure they just wanted to know when the Fluke 87 series gained the temperature feature.

When did they start putting built-in thermometers in Fluke 87 V multimeters?

It is the second item down on the features tab on the Fluke page:
http://en-us.fluke.com/products/digital-multimeters/fluke-87v-digital-multimeter.html#features
Am I right that your question wasn't about an internal thermometer for cold-junction compensation, but rather "built-in thermometer" meaning that it can take temperature measurements without a separate instrument? This was added to the 87V. Older versions of the 87 didn't have the thermometer feature.

 

Offline Zbig

  • Frequent Contributor
  • **
  • Posts: 927
  • Country: pl
Re: Fluke 87 V with built-in thermometer?
« Reply #7 on: April 24, 2017, 04:01:48 pm »
My 87V (acquired in2010) displays OPEn when the thermo probe is not plugged in. It has always been this way, and makes temperature readings perfectly well when the probe is inserted. I have had other meters which do have the internal temperature reference.

If it's able to measure the absolute temperature of the thermocouple's tip, it has to have the cold junction compensation inside the meter in one form or the other. There's no way around that - that's how thermocouples work. Without that, all your meter could tell you would be "the thing you're probing is x degrees hotter/colder than where I'm sitting".
 

Offline AxtmanTopic starter

  • Regular Contributor
  • *
  • Posts: 181
  • Country: us
Re: Fluke 87 V with built-in thermometer?
« Reply #8 on: April 24, 2017, 04:50:09 pm »
I'm sorry but I should have explained what lead up to my question.

I kept hearing and reading that the Fluke 87 V had a built-in thermometer.  This feature is even described on Fluke's model features page.  I bought my used Fluke multimeter several years and of course it did not come with a manual.  I know that Fluke does occasionally update the multimeter, for example they fixed the mobile phone interference issue.  So if figured that they added the built-in thermometer (and maybe other features) but did not change the model number.

Whenever I switched my multimeter to temperature range the display would indicate "OPEn", which I assumed was shorthand for, "Hey stupid!  Plug in a K-type probe!"  Nowhere in the online manual (May 2004 Rev.2, 11/08) does it indicate that by shorting the leads you can get temperature reading.  (Thanks for the tip IanB!)

I will write Fluke and ask them to revise their literature to reflect this feature, so other people like me do not get confused.

Thanks again for great discussion.
 

Offline Zbig

  • Frequent Contributor
  • **
  • Posts: 927
  • Country: pl
Re: Fluke 87 V with built-in thermometer?
« Reply #9 on: April 24, 2017, 06:31:39 pm »
Whenever I switched my multimeter to temperature range the display would indicate "OPEn", which I assumed was shorthand for, "Hey stupid!  Plug in a K-type probe!"  Nowhere in the online manual (May 2004 Rev.2, 11/08) does it indicate that by shorting the leads you can get temperature reading.  (Thanks for the tip IanB!)

I will write Fluke and ask them to revise their literature to reflect this feature, so other people like me do not get confused.

They haven't mentioned that in the manual because that's more of a "side-effect" (or "added bonus" if you will) than a feature, really.
 

Online tooki

  • Super Contributor
  • ***
  • Posts: 11505
  • Country: ch
Re: Fluke 87 V with built-in thermometer?
« Reply #10 on: April 24, 2017, 06:35:11 pm »
I think the only people being confused are those who are overthinking it because they know about cold-junction compensation.

When everyone says that the 87V has a built-in thermometer, all that's meant is "you can plug in a thermocouple to measure temperature", as opposed to needing a separate device. (Yes, the OPen display absolutely does mean "plug in your damned thermocouple!") The fact that you can short the input and see something approximating ambient temperature is neither a feature nor a bug, it's just a characteristic of thermocouple thermometers. There's NOTHING to indicate that some 87V implement the thermometer function differently, since they ALL use a type-K thermocouple. So don't think that some revisions have an internal cold-junction compensation and some don't.
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #11 on: April 24, 2017, 07:22:58 pm »
Some meters always display the cold junction temperature, whether a K-type thermocouple is connected or not. It is documented and is a feature.
In all meters, the measurement is taken close to the reference junction (ambient), with a thermistor, RTD, thermal diode or temp sensor IC.
Once the thermocouple is plugged in, that temperature is used for cold junction compensation.
The hot junction temperature can then be calculated, using voltage and look-up tables (or formulae) to compensate for non-linearity.

Shorting the probes, mimics a thermocouple with no induced voltage and the meter only displays the remaining cold junction temperature with its compensation (ambient).
« Last Edit: April 24, 2017, 07:49:33 pm by Wytnucls »
 

Online tooki

  • Super Contributor
  • ***
  • Posts: 11505
  • Country: ch
Re: Fluke 87 V with built-in thermometer?
« Reply #12 on: April 24, 2017, 07:50:41 pm »
We know this. The point is that that's not what Fluke is talking about when they say 'built-in thermometer'. All they mean is that you can plug a thermocouple probe in. Repeating this detail is only going to confuse our already-confused OP even further.
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #13 on: April 24, 2017, 08:06:53 pm »
'Built-in thermometer conveniently allows you to take temperature readings without having to carry a separate instrument'

The OP doesn't seem to know how temperature is measured with a multimeter.
Fluke wrote a confusing statement. One needs a separate instrument or tool called a thermocouple to measure the temperature of a DUT.
Shorting the probes, will mimic a thermometer to display ambient temperature pretty accurately, without the use of a thermocouple.
There is nothing to be confused about, once you know how the damned thing works!
 
 

Offline AxtmanTopic starter

  • Regular Contributor
  • *
  • Posts: 181
  • Country: us
Re: Fluke 87 V with built-in thermometer?
« Reply #14 on: April 24, 2017, 08:17:16 pm »
Obviously I am confused.  Thus the reason for my question.

I have another multimeter where when I turn the dial to (degF / degC) it would display the ambient room temperature.  When you instert a K-thermocouple you get a specific temperature for the item being measured.  I had mistakenly assumed that my Fluke meter (or a newer version) would operate the same way.

On occasion I have to be out on job site and I need to report weather conditions.  These measurements do not have to be super accurate and it is nice to know that my Fluke can give me ambient temperature.
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #15 on: April 24, 2017, 08:24:38 pm »
I understood that you were confused initially. if you're still confused now, let's just say that some meters can display ambient temperature directly and some can display it indirectly. The preferred method is to use a thermocouple and read whatever it says on the screen.
Enjoy your meters.
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Re: Fluke 87 V with built-in thermometer?
« Reply #16 on: April 24, 2017, 10:04:59 pm »
Some meters always display the cold junction temperature, whether a K-type thermocouple is connected or not. It is documented and is a feature.
In all meters, the measurement is taken close to the reference junction (ambient), with a thermistor, RTD, thermal diode or temp sensor IC.
Once the thermocouple is plugged in, that temperature is used for cold junction compensation.
The hot junction temperature can then be calculated, using voltage and look-up tables (or formulae) to compensate for non-linearity.

Shorting the probes, mimics a thermocouple with no induced voltage and the meter only displays the remaining cold junction temperature with its compensation (ambient).

Really?

My Fluke 287 show me 26,2C with short circuit leads and 24,7C with connected thermocouple to the end of this leads.  So, what is reference and for what this reference, when type K thermocouple junction voltage well know and constant more or less for all thermocouple probes?
What reference temperature (in mV for processor) in digital thermometers that supported more than one types external thermocouples?
Have you even tried connected to your meter Type K thermocouples, already being some time under high or low temperatures to compare with DMM environment temperature? 

IMHO, internal temperature that DDM show you with short circuit leads only for show user that temperature coefficient must be applied to the readings, when internal DDM temperature stabilized away from calibrated diapason 23+/-5C. It easy can be checked with just short circuit leads, no matter have you external thermocouple or not.
When you have external thermocouple, especially with fast response, internal and external temperature always will different with thermocouple connected and internal DDMs temperature can't be reference.
« Last Edit: April 24, 2017, 10:20:39 pm by serggio »
 

Online tooki

  • Super Contributor
  • ***
  • Posts: 11505
  • Country: ch
Re: Fluke 87 V with built-in thermometer?
« Reply #17 on: April 25, 2017, 12:27:44 am »
'Built-in thermometer conveniently allows you to take temperature readings without having to carry a separate instrument'

The OP doesn't seem to know how temperature is measured with a multimeter.
Fluke wrote a confusing statement. One needs a separate instrument or tool called a thermocouple to measure the temperature of a DUT.
…which is included. The instructions tell you to use the thermocouple to take a measurement.
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #18 on: April 25, 2017, 05:11:23 am »
We know that. It's not exactly built-in, is it?
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #19 on: April 25, 2017, 05:13:13 am »
They should have said: The meter comes with a separate thermocouple probe which allows temperature measurements.
« Last Edit: April 25, 2017, 05:18:17 am by Wytnucls »
 

Offline serggio

  • Regular Contributor
  • *
  • Posts: 149
  • Country: ru
Re: Fluke 87 V with built-in thermometer?
« Reply #20 on: April 25, 2017, 06:52:04 am »
Hi Wytnucls,
Could you explain please in details, for what we need 'could' junction measurement inside meter?
For meter both internal and external temperature unpredictable when it start  measurement.   
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #21 on: April 25, 2017, 07:19:55 am »
Dave has done a video about thermocouple:

 
The following users thanked this post: Marco1971

Online tooki

  • Super Contributor
  • ***
  • Posts: 11505
  • Country: ch
Re: Fluke 87 V with built-in thermometer?
« Reply #22 on: April 25, 2017, 11:34:56 am »
We know that. It's not exactly built-in, is it?
Do you have to carry around a separate thermometer device to plug the thermocouple into? No, you don't. So yes, it's built in.

They should have said: The meter comes with a separate thermocouple probe which allows temperature measurements.
A thermocouple alone doesn't give you temperature readings. You still need the thermometer circuitry inside the meter.



This is Fluke's complete bullet on the features tab of the product page: "Built-in thermometer conveniently allows you to take temperature readings without having to carry a separate instrument"

Are you just being willfully dense for the sake of being argumentative? Seems to be a pastime on this damned forum...  :palm:  |O
« Last Edit: April 25, 2017, 11:42:29 am by tooki »
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Fluke 87 V with built-in thermometer?
« Reply #23 on: April 25, 2017, 11:42:16 am »
Spare us the insults please. You're the one that's dense. Fluke's writing is what confused the original poster.
Your thin interpretation of it doesn't make it true.
 

Online tooki

  • Super Contributor
  • ***
  • Posts: 11505
  • Country: ch
Re: Fluke 87 V with built-in thermometer?
« Reply #24 on: April 25, 2017, 11:44:31 am »
Spare us the insults please. You're the one that's dense. Fluke's writing is what confused the original poster.
Your thin interpretation of it doesn't make it true.
Hardly. Everyone on the damned planet knows exactly what Fluke meant, except for you and the OP. Fluke's description makes perfectly clear what it does, since taking temperature readings necessarily requires some kind of probe. If it couldn't, they would have said "measures ambient temperature" so people wouldn't sue them for false advertising when they discovered that's all it could do.
 