I've got myself CMU200 with a purpose of using it for pre-compliance emission testing.
I'm trying to make sense of the numbers I reading, and there is something I don't understand. If setup the the generator to output level of -35dBm, should not I be getting that number from the Analyzer meter?
Both in and out are connected to the same RF2 port, so the connection between generator and analyzer is internal.
Now, as you see on the attached pics, the reading of power is about 10dBm higher than the output of the generator. And it is always about 10dBm higher, at various levels of the generator.
I have tried RF_Mode manual, and different values for RF_Attenuation, which don't make any difference.
I would appreciate any advice. Either I'm missing something in the setup of the CMU200, or I'm missing something more important, or the unit requires adjustment...