Finally reassembled the scope and plugged it into the dim bulb unit, switched on, bulb glowed dimly for a second and went out, the power led and others on the front flashed about 5 times and went out then nothing. Switching off and on again, produced exactly the same results except that the 100w bulb now does not glow??
There was no arcing, bang or magic smoke at all.
As I had to rebuild the heater connections in the PSU it is possible that I have got the connections to the heater reversed bearing in mind that F14 connection to the heater also goes to the HV board? Could this be the problem??
Maybe, the heater supply is polarised and need be around the right way.
Otherwise sounds good.
Try it with a lower wattage bulb and if all is well it should do similar but not go completely out.
With the tester in circuit you might like to check the SMPS is working with another scope......no connection ahla near field probe style.
Look for it oscillating in the 20 -40 KHz range.
I would love to do that but as I don't have a near field probe, I can't.
I have also bypassed the bulb with same results and talking with 9a4wy he suggested that I measure the voltages each time I switch on and here the results
Measured at the test socket for voltages (P2700} and had to break out my analogue Avo for this as my DMM was unreadable with the constant voltage changes and I got the following results
Expected Actual
-12v -3v
-5v -2.5v
+5v +2.5v
+12v +5v
+115v +60v
+57v +20v
All of these were only present for a second and decayed to zero with in a couple of seconds.
So its common on all rails.
We have also checked the resistance to ground for each of the rails at the same location and the results are as follows:-
-12v 75.5ohms
-5v 257ohms
+5v 440ohms
+12v 202ohms
+115v 36.92k
+57v 32.93k
so we can rule out short circuits.
He also recommends recapping the whole PSU so that is my next task but I need to order up another couple 470uf 25v as I'm short.
So thats where I'm presently at.
Picked up a lovely Hitachi V525 scope this week almost mint condition apart from a (rusty handle, been in its carry bag all its life and the probves still in sealed bags), and a slight problem on channel 2, the volts per division are not changing on screen, timebase and both perfectly ok on channel 1, but that one can sit while I refurb this Hameg.