I think the Fluke 87V is really a good meter and want to see if a third one would be better
10 (9.6%)
This testing is pointless! Please STOP damaging these meters!
8 (7.7%)

Total Members Voted: 103

Author Topic: Handheld meter electrical robustness testing.  (Read 417786 times)

0 Members and 1 Guest are viewing this topic.

Offline 3db

  • Frequent Contributor
  • **
  • Posts: 328
  • Country: gb
Re: Handheld meter electrical robustness testing.
« Reply #2425 on: March 06, 2018, 10:17:40 am »
I'm sure Joe has a few guns too.  >:D
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 12678
  • Country: nz
  • NZ Siglent Distributor
    • Taupaki Technologies Ltd.
Re: Handheld meter electrical robustness testing.
« Reply #2426 on: March 06, 2018, 12:12:57 pm »
I'm sure Joe has a few guns too.  >:D

I could shoot it with my fuse powered cork gun.    :-DD
Piezo ignitor and Methanol upgrade required ?
Avid Rabid Hobbyist
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2427 on: March 06, 2018, 09:52:23 pm »
After seeing the bent shield in my brand new Fluke 87V, a member with a slightly older 87V took theirs apart and provided me with this picture.  I had hoped the meter I received was a one off mistake but it appears they are bending the shield and no one is catching it.  I have not yet done anymore with the 87V and it's still sitting in parts.  The shield looks like it is designed to sit flat.  I can place it flat in the board.   The two tabs seem to be setup correctly.    Maybe they cut their QC staff to make more money?  Maybe it's supposed to be bent?   

Too bad these large companies have no presence in these groups.   If it were Brymen, I would just ask them and based on all my previous experiences with them, would have an answer in a day.  I never found a contact at Danaher/Fluke.
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2428 on: March 06, 2018, 11:09:20 pm »
Maybe it's deliberate.

Does it press against the back of the case like a spring? Does it look like there's any reason to do that?

I can place it flat in the board.

Doesn't necessarily mean it's supposed to be like that.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2429 on: March 06, 2018, 11:14:16 pm »
Dave's running a live switch cycle on his second channel.  I watched it for a few minutes.  Tried to calculate the resistance with the formulas shown but looks like its 1.7K ohms the one way  a little better the other.  Guessing I am missing something as I think he said they could only read up to 10 ohms or so.   So much computing power and they resort to sticky notes.   :palm:

He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:

If that's fiberglass dust then it's an abrasive, just sayin'.
 

Offline ChrisLX200

  • Supporter
  • ****
  • Posts: 462
  • Country: gb
Re: Handheld meter electrical robustness testing.
« Reply #2430 on: March 06, 2018, 11:22:40 pm »
Dave's running a live switch cycle on his second channel.  I watched it for a few minutes.  Tried to calculate the resistance with the formulas shown but looks like its 1.7K ohms the one way  a little better the other.  Guessing I am missing something as I think he said they could only read up to 10 ohms or so.   So much computing power and they resort to sticky notes.   :palm:

He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:

If that's fiberglass dust then it's an abrasive, just sayin'.

I don't think it's fiberglass - solder mask more likely. In normal use the meter would be picked up and put down potentially dislodging the dust anyway after every use or so - you cannot replicate normal usage easily. These sorts of test can only ever be a rough guide to wear rate, and you would need to test many to get some sort of average.
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2431 on: March 06, 2018, 11:57:58 pm »
Maybe it's deliberate.

Does it press against the back of the case like a spring? Does it look like there's any reason to do that?

I can place it flat in the board.

Doesn't necessarily mean it's supposed to be like that.

It's held in place by the screws.  It does not appear to spring off the back of the case.  It looks like it was designed to fit flat. There are no clearance issues I see with that would prevent it from sitting flat.  I see no reason for it to be bowed other than poor quality control.
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2432 on: March 07, 2018, 12:24:18 am »
Dave's running a live switch cycle on his second channel.  I watched it for a few minutes.  Tried to calculate the resistance with the formulas shown but looks like its 1.7K ohms the one way  a little better the other.  Guessing I am missing something as I think he said they could only read up to 10 ohms or so.   So much computing power and they resort to sticky notes.   :palm:

He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:

If that's fiberglass dust then it's an abrasive, just sayin'.

I don't think it's fiberglass - solder mask more likely. In normal use the meter would be picked up and put down potentially dislodging the dust anyway after every use or so - you cannot replicate normal usage easily. These sorts of test can only ever be a rough guide to wear rate, and you would need to test many to get some sort of average.

I was under the impression that this round of testing was done live, non-stop, 50K half cycles, no cleaning with the shim installed correctly.  I only saw the last few minutes so if you saw him cleaning it again, that's a problem.   If your pulling that out of your ass and don't really know, we should ask Dave.   

I don't need to run more of the same meter to see the difference in wear.   Between the pictures and resistance metric, it's good enough for me.   Some have certainly cut deeper than mask.  Then there is that detent spring, it's not always the contacts that are the problem.       

Then again, I have been told many times that I can't draw a conclusion from the transient tests I run because the sample size is one. 
« Last Edit: March 07, 2018, 11:37:28 am by joeqsmith »
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2433 on: March 07, 2018, 12:47:32 am »
It's held in place by the screws.  It does not appear to spring off the back of the case.  It looks like it was designed to fit flat. There are no clearance issues I see with that would prevent it from sitting flat.  I see no reason for it to be bowed other than poor quality control.

So basically the metal tabs in the middle of the shield just weren't in their PCB holes?

Oh, dear.
« Last Edit: March 07, 2018, 12:49:33 am by Fungus »
 

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 2457
  • Country: us
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #2434 on: March 07, 2018, 02:50:51 am »
I foresee the used test gear market will start categorizing 87Vs between pre- and post- a specific serial#... That or photographs of the back of the meters will become almost mandatory.

IMHO every reaction is sparked by a previous action - as I said before, due to the amount of competition on the 20k count segment I honestly believe the 87V's sales figures can't support the margins it once had.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2435 on: March 07, 2018, 03:36:04 am »
IMHO every reaction is sparked by a previous action - as I said before, due to the amount of competition on the 20k count segment I honestly believe the 87V's sales figures can't support the margins it once had.

Sure, but they're not lowering the prices and I can't believe the cost of manufacturing them is going up.

This is just bean counting of the very worst kind. John Fluke must be doing about 12 RPM right now.
 

Offline floobydust

  • Super Contributor
  • ***
  • Posts: 1590
  • Country: ca
Re: Handheld meter electrical robustness testing.
« Reply #2436 on: March 07, 2018, 12:12:40 pm »
After seeing the bent shield in my brand new Fluke 87V, a member with a slightly older 87V took theirs apart and provided me with this picture.  I had hoped the meter I received was a one off mistake but it appears they are bending the shield and no one is catching it.  I have not yet done anymore with the 87V and it's still sitting in parts.  The shield looks like it is designed to sit flat.  I can place it flat in the board.   The two tabs seem to be setup correctly.    Maybe they cut their QC staff to make more money?  Maybe it's supposed to be bent?   

Too bad these large companies have no presence in these groups.   If it were Brymen, I would just ask them and based on all my previous experiences with them, would have an answer in a day.  I never found a contact at Danaher/Fluke.

The clearances from shiield to the HV input circuit parts looks bad?
The leads on the MOVs', 1k resistor etc. there should be a insulating barrier ?
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2437 on: March 07, 2018, 12:35:24 pm »
After seeing the bent shield in my brand new Fluke 87V, a member with a slightly older 87V took theirs apart and provided me with this picture.  I had hoped the meter I received was a one off mistake but it appears they are bending the shield and no one is catching it.  I have not yet done anymore with the 87V and it's still sitting in parts.  The shield looks like it is designed to sit flat.  I can place it flat in the board.   The two tabs seem to be setup correctly.    Maybe they cut their QC staff to make more money?  Maybe it's supposed to be bent?   

Too bad these large companies have no presence in these groups.   If it were Brymen, I would just ask them and based on all my previous experiences with them, would have an answer in a day.  I never found a contact at Danaher/Fluke.

The clearances from shiield to the HV input circuit parts looks bad?
The leads on the MOVs', 1k resistor etc. there should be a insulating barrier ?
You know, it does look bad in that picture they sent me.  There's that tab on the backside that should hold it away from the board and I have not looked but I don't think the shield extends down far enough.  That said, nothing says they did not change the shield design.  There meter is older than mine.  This same person has an older version of the meter and they said the shield was made differently in the earlier models.  They said they would provide me with a picture of that one for comparison.  I will post it once I have it.   

It could be a Gossen and have no shield and be unstable as you move your hand near it, expose it to RF, or have relays that change states when exposed to a magnetic hanger.  All preventable with a good shield design.    The things we can learn with a sample size of one! 

How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Offline EEVblog

  • Administrator
  • *****
  • Posts: 26639
  • Country: au
    • EEVblog
Re: Handheld meter electrical robustness testing.
« Reply #2438 on: March 07, 2018, 03:52:28 pm »
Dave's running a live switch cycle on his second channel.  I watched it for a few minutes.  Tried to calculate the resistance with the formulas shown but looks like its 1.7K ohms the one way  a little better the other.  Guessing I am missing something as I think he said they could only read up to 10 ohms or so.   So much computing power and they resort to sticky notes.   :palm:

He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:

No I did not.
The latest test was 50,000 cycles with no interruptions. You are welcome to watch the entire 23 hours video once Youtube has finished processing it, assuming it actually can process a 23hr live stream.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2439 on: March 07, 2018, 06:07:51 pm »
He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:

No I did not.

What happened at the 12 minute mark?



Quote: "A bit of spit on that..."

The latest test was 50,000 cycles with no interruptions.

Fair enough.
« Last Edit: March 07, 2018, 08:02:27 pm by Fungus »
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2440 on: March 08, 2018, 12:05:01 am »
He also cleaned off the accumulated dust/debris after every bunch of cycles.  :scared:

No I did not.

What happened at the 12 minute mark?

Quote: "A bit of spit on that..."

Fair enough.
Previous test was all.
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2441 on: March 08, 2018, 12:09:18 am »
These are the pictures showing the older Fluke 87V shield design.   I would think the stamped shield would be less costly to produce.  I have asked for a picture showing the stamped shield from the backside to see if they changed the profile. 
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2442 on: March 08, 2018, 01:16:42 pm »
It appears their older 87V's shield is identical to the one supplied with mine.  It sure looks like their tab was not pushed through the slits when the shield was installed.  It's too bad Fluke reps don't hang out here.  It seems like something they would want to know about, then again, it seems like it has been a problem for a long time so hard to believe the supervisor's and QC are not aware of it.   
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online IanB

  • Super Contributor
  • ***
  • Posts: 8996
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2443 on: March 08, 2018, 01:30:04 pm »
It appears their older 87V's shield is identical to the one supplied with mine.  It sure looks like their tab was not pushed through the slits when the shield was installed.  It's too bad Fluke reps don't hang out here.  It seems like something they would want to know about, then again, it seems like it has been a problem for a long time so hard to believe the supervisor's and QC are not aware of it.

You don't need to test a product for quality if lack of quality does not affect your sales. It's not cost-effective.
I'm not an EE--what am I doing here?
 

Online joeqsmith

  • Super Contributor
  • ***
  • Posts: 4712
  • Country: us
Re: Handheld meter electrical robustness testing.
« Reply #2444 on: March 08, 2018, 02:03:51 pm »
It appears their older 87V's shield is identical to the one supplied with mine.  It sure looks like their tab was not pushed through the slits when the shield was installed.  It's too bad Fluke reps don't hang out here.  It seems like something they would want to know about, then again, it seems like it has been a problem for a long time so hard to believe the supervisor's and QC are not aware of it.

You don't need to test a product for quality if lack of quality does not affect your sales. It's not cost-effective.
True but wouldn't take much more than the line supervisor looking at one and showing the worker how to properly install them.  If their supervisor can't handle something that basic, I would say they have the wrong person in that position.  But it may be cheep labor.  Or maybe unsupervised.  Surly the workers must think, that part should fit flat and maybe I should try and slid that tab into that slit....  Crazy...
How electrically robust is your meter?? http://www.youtube.com/channel/UCsK99WXk9VhcghnAauTBsbg
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2445 on: March 09, 2018, 12:16:51 am »
You don't need to test a product for quality if lack of quality does not affect your sales. It's not cost-effective.

Yep. If large corporations are buying these by the truckload without ever looking inside then what's the point? Got a problem? send it back.

Fluke can probably absorb 50% returns under warranty and still make a profit on these things.

(Maybe more: What's the BOM? About $50 I imagine... what's the next most expensive thing after the fuses and the precision resistors? The leads?)
 

Online rsjsouza

  • Super Contributor
  • ***
  • Posts: 2457
  • Country: us
    • Vbe - vídeo blog eletrônico
Re: Handheld meter electrical robustness testing.
« Reply #2446 on: March 09, 2018, 11:18:19 am »
You don't need to test a product for quality if lack of quality does not affect your sales. It's not cost-effective.

Yep. If large corporations are buying these by the truckload without ever looking inside then what's the point? Got a problem? send it back.

Fluke can probably absorb 50% returns under warranty and still make a profit on these things.

(Maybe more: What's the BOM? About $50 I imagine... what's the next most expensive thing after the fuses and the precision resistors? The leads?)

What was the cost to have the custom IC made?  Burden?  Certifications?
If it follows other meters from Fluke, it is probably a MSP430 with clipped JTAG and a custom P/N. I don't recall ever seeing a larger MSP430 ROM masked. If that is correct, the cost for the customization itself is no different than a commercial part, only the volume counts.
Vbe - vídeo blog eletrônico http://videos.vbeletronico.com

Oh, the "whys" of the datasheets... The information is there not to be an axiomatic truth, but instead each speck of data must be slowly inhaled while carefully performing a deep search inside oneself to find the true metaphysical sense...
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 12678
  • Country: nz
  • NZ Siglent Distributor
    • Taupaki Technologies Ltd.
Re: Handheld meter electrical robustness testing.
« Reply #2447 on: March 09, 2018, 12:11:11 pm »
If it follows other meters from Fluke, it is probably a MSP430 with clipped JTAG and a custom P/N. I don't recall ever seeing a larger MSP430 ROM masked. If that is correct, the cost for the customization itself is no different than a commercial part, only the volume counts.

I doubt that the front end is something you program with JTAG?  I expect it to be a full custom, mixed mode IC developed exclusively for Fluke, possibly by Fluke.  I doubt it's just a remarked device but I don't know.  We could ask a Fluke rep but they don't seem to hang around the forums. 

Starting to work on Part 4.
Starting to work on Part 4. repairing it again ?  :)
Avid Rabid Hobbyist
 

Offline Cliff Matthews

  • Supporter
  • ****
  • Posts: 1467
  • Country: ca
    • General Repair and Support
Re: Handheld meter electrical robustness testing.
« Reply #2448 on: March 09, 2018, 12:38:57 pm »
Ha! Speaking of repairs.. Bigclive just made a video on Fluke recalling the SM100/200/300 plug-in testers. After Joe pops the brains out of this one, me thinks John Fluke Sr will be turning.
 

Online Fungus

  • Super Contributor
  • ***
  • Posts: 8466
  • Country: 00
Re: Handheld meter electrical robustness testing.
« Reply #2449 on: March 09, 2018, 08:01:09 pm »
I doubt that the front end is something you program with JTAG?  I expect it to be a full custom, mixed mode IC developed exclusively for Fluke, possibly by Fluke.  I doubt it's just a remarked device but I don't know.

Design/development costs for the chip will have been recovered a long time ago.

 