I have a need for pre-compliance testing of a product I am developing.  I plan to purchase a Siglent 3000 series Spec An.    Of course by itself, it is useless for EMC testing without the sensors and accessories needed to measure stuff.  One of those thing I will need is a decent bi-conic antenna.    The commercial ones seem to be priced in gold.
I can design and build an antenna but the problem is then to calibrate it to a reasonable degree of accuracy for pre-compliance testing.  One procedure is to use a calibrated antenna as a signal source to test and calibrate the antenna under test (AUT).    But that requires one of those golden calibrated antennas.
I propose to overcome the need for a golden calibrated antenna by building and testing two identical, but unknown antennae.   The idea is simple (which probably means others have already thought of it and done it).
Step 1.  Design/Build two identical antenna.
Step 2.  Connect cables / attenuators etc to the Spec An.
Step 3.  Normalise the test rig by doing a sweep with the antennae replaced with a short coax link.
Step 4.  Insert antennae and do the tests.
Step 5.  Halve the test results.
This is illustrated in the attached drawing.
The attenuators would provide wide band 50ohm closely matched to the Spec An impedance to miminise resonance in the test rig.
The antennae would need to be mounted far enough apart to be in the far field.
The test location would need to be rf quiet and without reflections.
The null link between the attenuaters would need to be as short as possible (an adapter).
Both legs of the test rig (Tx and Rx sides) should be identical for all sorts of reasons.
So the test results would then simply need to be halved to get calibrated performance.
If each antenna had 5dB gain, the test curve would show 10dB.
If each antenna had a 20db loss, the test curve would show 40dB.
So by halving the figures measured by the Spec An, a calibration curve would be  produced.
The method can be checked at a single frequency by using two identical dipoles whose performance can be mathematically calculated.  A ring-in for a golden calibrated antenna.
So the outcome of testing two identical but unknown antennae would be a known calibration curve.  In addition, there would be a spare antenna with its own calibration curve.
Some might think that building 2 antenna when you only need one is a waste.  Building 2 antenna is not much more work than building one when time taken to design, buy parts and test are accounted for.  The 2nd antenna could then be sold or kept as a spare.
I would be interested to know if anyone else has tried this and if so, what were the results?
Dazz
Dazz
Over Engineering: Why make something simple when you can make it really complicated AND get it to work?
Offline Jay_Diddy_B
Re: Howto Calibrate DIY EMC pre-compliance antenna??
« Reply #1 on: March 27, 2018, 01:59:58 am »
Dazz,
Welcome to the forum !!
If you go to this website:
http://everyspec.com/MIL-STD/MIL-STD-0300-0499/MIL-STD-461A_8679/
You can download MIL-STD-461A, make sure you get the A version. starting on page 25 is a detailed set of drawings for making the bi-conical antenna. If you stick to the dimensions it should have the antenna factors published in the document.
Here is one of the drawings:
There are instructions in the MIL-STD-461 for measuring antennas, using two identical antenna.
I built one of these antenna and found that it was 'next to useless' without a proper anechoic chamber. There was too much background noise.
I found an open TEM cell to be really useful and works well.
Regards,
Jay_Diddy_B
« Last Edit: March 27, 2018, 02:02:04 am by Jay_Diddy_B »
« Reply #2 on: March 27, 2018, 02:36:13 am »
Hi
Thanks for the reference.  I live 30 minutes from a rural site located away from the city RF noise in a  deep valley. There is a cell site about 2km away.  Apart from that, it should be relatively RF quiet.  A biconic should work.
I hadn't considered a TEM cell.  It would be easier and cheaper to build than a biconic.  I will investigate that option further.
Do you have plans or design method?
Dazz
Dazz
« Reply #3 on: March 27, 2018, 03:08:28 am »
Hi
I have found a standard that includes a design  SAE J1113-25 for automotive EMC testing.
I found a paper that analyses the performance of that design in detail. http://ieeexplore.ieee.org/document/950717/ That should include a lot of detail useful for any DYIers.
TEMS seem to be used more for immunity testing rather than emission testing.
Dazz
Offline Jay_Diddy_B
« Reply #4 on: March 27, 2018, 03:22:19 am »
Dazz
Dazz,
Try:
https://pdfs.semanticscholar.org/7b3c/c56771978fb7dab27ee4ccc31d67f7d3167f.pdf
If you google images 'open tem cell' you can get some ideas.
Regards,
Jay_Diddy_B
Offline EE-digger
« Reply #5 on: March 27, 2018, 04:14:06 am »
Dazz,  I'm assuming you know what band you need but I have to say I hadn't pulled the biconic out to use in a number of years.  Most problems with low to moderate speed clocks and switching regulators have tended to be in the 100-200MHz,  400-500MHz or 900MHz and a bit higher.  The switching regulators are all running in the 250kHz to 2MHz range but harmonics, especially those from poor layout on switching nodes, tend to show up exceeding limits in the first band I mentioned with various display clock harmonics in the second band.
So, for the most part, I use a 200MHz to 1.3GHz log periodic of fairly compact size (it's an older Emco, don't have the model handy).  For all else, I just wait until we do a pre-scan or final tests at a local test lab.
Offline hendorog
« Reply #6 on: March 27, 2018, 06:58:12 am »
Hi Dazz,
I think you are neglecting the path loss between the antennas in your idea?  There will be a chunk of free space path loss between the two antennas compared with the coax adapter which you normalised against.
You could make a couple of identical 'reference' antennas though and then compare other antennas to them. The measurement results won't be absolute as you don't know the precise performance of the reference antennas. However you might be able to get a rough idea of their performance if the reference antennas were a design which you can model or simply find on the internet.
Offline tautech
« Reply #7 on: March 27, 2018, 07:45:08 am »
Dazz
These could be well worth looking at for a cheap and characterized EMC probe set:
https://www.eevblog.com/forum/testgear/new-affordable-emc-probe-set-at-a-bargain-price-from-ariel-rocholl's-lab/
Avid Rabid Hobbyist
The following users thanked this post: SaabFAN
« Reply #8 on: March 27, 2018, 08:47:59 am »
Hi
At this stage I am trying to map out the gear I need to do budget pre-compliance.  My local EMC test facility is a few hundred kilometers away so even a pre-scan is a major exercise in time and cost.  It makes economic sense for me to spend money on test equipment rather than waste it on failed compliance tests.
I has been 20 years since I last did EMC testing but I was designing/testing buildings, not equipment.  I am out of date with everything and my current test gear is old and limited.  I don't have any EMC instruments.
I'd really like to get the Siglent 3032 but I probably only actually need the 3021 with the EMI add-on.
I can make some H and E probes easily enough.
My products use lots CE compliant (MeanWell) 90W 5V SMPS.  I will need to look at load side noise getting onto the SMPS input side   so I am sure I will need a LISN.
The TekBox CISPR 16 LISN unit looks like a good candidate .
I like the idea of a log period antenna simply because it will be a lot easier to store that a TEM cell or biconical.    I would really like the antennas to be calibrated so I can get reasonably close to the test lab results before it gets to the test lab.
I can make up some RF current probes.
Transient limiter for the Spec An
RF pre-amp.
Cables/adapters
chattering relay (RF noise generator)
That is my shopping/make-it list so far.
Not sure if I need an ESD tester.  My product is in an all plastic custom moulded enclosure.   The metal chassis is not exposed.
I have found recommended lists here: http://www.interferencetechnology.com/assembling-low-cost-emi-troubleshooting-kit-part-1-radiated
-emissions/
and here: https://interferencetechnology.com/assembling-low-cost-emi-troubleshooting-kit-part-2-immunity/
I think I have the essentials on my list.
I haven't yet determined what testing I must do.  My product is a collection of CE compliant units that are wired together.  The exception is a custom interface board with no clocks or other noise generators and a high powered batter charger. The battery charger unit is a strong candidate for RF noise because of the high currents and high ripple factor (60%@20A).
Dazz
« Reply #9 on: March 27, 2018, 09:03:37 am »
Hi Dazz,
I thought of that. If the antennas are far enough apart to be in the far field, the free space loss can be calculated.  Far from perfect.  A pair of accurately made dipoles operated at their design frequency can be used to confirm the loss.  There is a lot of data on dipole performance.  Somewhat closer to perfection.   More data can be obtained by pairing a dipole (at the design freq) with the AUT.     No where near as good as a proper lab test but cheap and better than no test.
Dazz
« Reply #10 on: March 27, 2018, 09:48:26 am »
Hi
Google found the Antenna test procedures here : www.diamondeng.net/library/AntennaMeasurement.pdf
Lots of info but a quick read indicates all methods described use a golden calibrated antenna.  A detailed read may reveal more.
Dazz
Dazz
Offline hendorog
« Reply #11 on: March 29, 2018, 06:14:16 am »
Hi Dazz,
Sorry for the delay, I didn't notice the reply notification for some reason.
Yep that is exactly what I was implying for the DIY reference antennas. I haven't specifically tried it, but you could do quite a few tests on your DIY dipole references with the SA+TG to see how close it is to the theoretical build. For example you could rotate one around while keeping the other one static and work out what the pattern looked like. And you could measure the return loss to work out the resonant frequency compared to your goal.
For a dipole you would need a balun to get reliable measurements, which you could also build and and test.
Offline SWR
  • Without engineering science is just philosophy.
« Reply #12 on: March 29, 2018, 07:10:42 am »
Thats what I bought and it's a really fine instrument.
You can max it out to 3,2GHz and all the options pretty easily if you're into that.
Instead of messing with antennas i use common mode chokes like the ones described here:
https://www.edn.com/electronics-blogs/the-emc-blog/4429621/Using-current-probes-to-estimate-E-fields
It's a much simpler solution and in my experience it gets close enough for useful pre compliance if you chose a high frequency ferrite. I cut a long SMA cable and pull out the inner wire a distance from the cut end. Then I wrap them round the ferrite and solder the ends together. The final touch are two 3D printed shells with grooves that keep the spacing of the coil turns in place. It's possible to get >50ohm up beyond 1GHz. :-+
You should never go down on equipment!
« Reply #13 on: March 29, 2018, 09:06:23 am »
Hi
Someone somewhere had to start with an uncalibrated antenna before they ended up with a calibrated version.  There are probably papers published  in the first half of last century by very smart researchers detailing the methods.
[Edit:  It turns out that someone has done that .  The Mil-Std461A referenced by Jay_Diddy_B includes a 2-antenna calibration procedure.]
Unfortunately the SSA3032 is the reason I own the 2.5litre version of my car rather than the 3.0litre engine.    The 3 litre engine is really nice but doesn't add any useable  capability for the greater cost.  The savings I can make in buying the 3021 I can spend on accessories that add significant capability.  I would get bigger bang for buck.  The choice is a personal one.  There is no right  or wrong answer that fits everyone.
The DIY chokes/current sensors are on my shopping list.
Dazz
« Last Edit: April 03, 2018, 08:30:17 am by dazz1 »
Dazz
Offline hendorog
« Reply #14 on: March 29, 2018, 03:27:32 pm »
I'm pretty sure that the SSA3021 is hackable up to be an SSA3032 - if you want to. So you can have the best of both worlds there.
« Reply #15 on: April 03, 2018, 08:40:53 am »
I'd be very surprised if the SSA3021 had the hardware to go to the same max freq as the SSA3032 even if a software hack allowed it. If I was producing software for a SA, I'd use identical software for all models with separate configuration files.   That would reduce software production and maintenance costs.
An alternative option for me would be to buy the SSA3032 but sacrifice the software options.  EMC testing has been done for many years without the option of helper software apps.
Dazz
Offline hendorog
« Reply #16 on: April 03, 2018, 09:04:56 am »
I did a quick search and it does seem to be possible. It sounds like you are right on the software side as it's the config which is changed. I don't have one but the hackability of them does sound like a big selling point for a hobbyist.
Have a look here, where someone comments that their 3021 behaves like a 3032:
https://www.eevblog.com/forum/testgear/hack-of-sigllent-spectrum-analyzer-ssa3021x/msg1261248/#msg1261248
Offline SWR
« Reply #17 on: April 03, 2018, 09:11:12 am »
Well - I think you'll have to concider yourself very surprised then. :wtf:
I converted my SSA3021 to 3,2GHz and enabled all the option while I was at it ... and it's all "Honky-dory" as they say.
You basically just have to rename a file so the analyzer can't find it. :-/O
« Last Edit: April 03, 2018, 09:16:38 am by SWR »
« Reply #18 on: April 03, 2018, 09:49:59 am »
It doesn't surprise me that the software allows it.  It doesn't surprise me that the hardware appears to work.  I would be surprised if a side by side comparison between a hacked 3021 and genuine 3032 showed the same performance.   That would require identical hardware on both.
Dazz
Offline SWR
« Reply #19 on: April 03, 2018, 01:28:19 pm »
They chose to go with one hardware and make the variation in software instead like many other manufacturers do.
It wouldn't make much sense to sell upgrades to a crippled hardware. They would face several challenges:
- Having to procuce separate PCBs and not gaining the benefit of low hardware variance.
- Dealing with unsatisfied customers that bought the upgrades without getting full performance.
It's all about doing the variants as late in the production process as possible.
A label and a software config is the optimum choise from a production point of view.
« Reply #20 on: April 22, 2018, 08:56:54 am »
Hi
I have been looking at the connection between the antenna and coax feeder.  I am aware that coax can be directly connected to the antenna cones.  This makes for a physically asymmetric antenna (including the antenna feed) which is likely to distort the antenna pattern.
I think a 1:1 balun would be a good idea.  This will allow the antenna feed to exit radially to avoid distorting the antenna pattern.   I found this research paper digitalcommons.calpoly.edu/cgi/viewcontent.cgi?article=1282&context=eesp.    The results indicate that the current transformer type balun has good performance.  The test frequency went to 1.5GHz.  I like measured data.
I don't know where to get suitable tiny ferrites.
Dazz
Dazz
Offline RFDUK
  • Embedded antenna design services
    • RFDesign
« Reply #21 on: April 22, 2018, 06:47:14 pm »
I went through the same search in the early 1990's setting up an EMC lab on a limited budget.
The solution I found is probably still a good answer to your question today.
In the early days of commercial EMC testing (as apposed to the established military EMC standards of the time), the FCC developed a 'reference dipole' for absolute field strength testing.
The guy responsible for the balun design was Willmar Roberts. The balun used simple lengths of 50 ohm cable.
'Roberts Balun' antennas cost a fortune in commercial sets like EMCO 3121, but can be manufactured for very little.
They are so simple that if built exactly to the drawings they are self calibrated to a very low uncertainty.
If you Google Roberts Balun reference dipoles hopefully the drawings are out there. They were initially published in an FCC document.
Your idea of building two antennas is a good one for sure. That's what I did with the Roberts Baluns, you need 4 baluns (and 4 sets of dipoles) to cover the range 30 MHz to 1 GHz.
By connecting two baluns for the same frequency range back to back (balanced connections), you can test the transmission performance integrity. The plots I still have somewhere looked great.
Testing 2 unknown antennas back to back poses the problem you mention of reflections. That isn't easily accommodated unfortunately.
Once you have a set of 'reference dipoles' you can calibrate other antennas you make, for example Biconical or Bilog etc. My homemade dipoles are still in use today 25 years later.
For higher frequencies very low cost PCB Log Periodic antennas are available from Kent at http://www.wa5vjb.com.
Don't be tempted by other products, Kent's antennas are the real deal.
Hope that's interesting, good luck, Martyn.
Weak signal comms specialist. Very low noise amplifier & precision calibrated noise source manufacturer. Embedded antenna design services. http://www.g8fek.com  http://www.rfdesignuk.com
« Reply #22 on: April 23, 2018, 01:01:55 am »
Hi
Thanks for the info.  I will look that up.
The nice thing about simple dipoles is that both the theory and practice have been so thoroughly characterised.  There can't be much left unknown about them.
What I found really interesting about the baluns in the link was the tiny size and the very fine Cu wire used.  I would have thought that this would present a major impedance mismatch.
I was planning to use Cat6 twisted pair (100R impedance) 2x in parallel (50R).  That would require a much larger ferrite core.  That's one of the reasons I like measured data because it eliminates the BS.
A Kent antenna is on my shopping list.
Dazz
« Reply #23 on: April 23, 2018, 01:09:47 am »
Dazz
Offline RFDUK
    • RFDesign
« Reply #24 on: April 23, 2018, 08:41:15 am »
Hello Dazz, it's great to see that doc again thanks! That's the one I used. The photo shows what I did using plastic boxes and cable conduit parts. BNC sockets were glued into the conduit ends. The lowest frequency antenna rods extend to about 60MHz. I had extension wires with crock clips to get to 30 MHz.
I made 2 of each balun and the swept performances fitted the doc frequency response very well.
I made a bicon at the time too. A friend milled a center piece from Delrin and turned aluminium bosses for each end of the rods. I noticed the bicon diagram posted earlier had connecting parts in the rods. I just bent the aluminium rods and that worked fine. The balun was a wire transformer on low permeability ferrites. I found a low priced commercial 30MHz - 2GHz Bilog on ebay about 15 years ago so the homemade bicon is out of service now.
It's all ways useful to have proper reference dipoles available as a sanity check against broadband antennas, I wouldn't want to be without them.
Bicons need careful calibration below the natural resonance of about 70MHz for the larger ones. The VSWR is poor below the natural resonance and it's best to put a 6dB pad at the antenna end to remove K factor drift associated with mismatch to the cable.
Martyn.
« Last Edit: April 23, 2018, 08:42:47 am by RFDUK »
Offline RFDUK
    • RFDesign
« Reply #25 on: April 23, 2018, 09:27:10 am »
Forgot to mention, we use the SSA3032 since January and can highly recommend it, great value. That's used more for radio dev work here.
Our main EMC work horse up to 1.8 Ghz for 20 years is an HP 8591EM. Has a card with EU and FCC limit lines. Screen dumps to PC using HP software. Would highly recommend that too if you can find a good one at the right price. If you are not into fixing faulty RF test equipement, the Siglent 3 year warranty perhaps weighs against a 20 year old machine that may prove unreliable though.
Sounds like you'll be needing CISPR specified quasi peak and average detectors as a minimum, but as you said in an earlier post, the EMC bells and whistles that can come expensive are not must haves. Normal resolution bandwidths of 10KHz for CISPR 9 and 100KHz for 120 are not too much of an issue for pre-compliance tests either.
Martyn.
« Reply #26 on: April 23, 2018, 09:36:00 am »
Hi
Decent used test equipment here is rare, and because it is rare, people who have it hold onto it, which makes it rarer.
One of the reasons I stay away from ebay is so I don't have to look at the pages and pages of test equipment I can't buy.  The cost of shipping and the risk of damage in transit is too high.
Dazz
« Reply #27 on: April 23, 2018, 10:20:55 am »
Hi
I found the attached design document used for the EDGES project.  It includes the design calculations.  I am wondering about the feasibility of using this type of balun on a biconic.  Roberts balun was made for dipoles so the design would need to be recalculated and verified for a biconic.
I don't have access to suitable simulation software to do similar analysis.
Dazz
« Last Edit: April 23, 2018, 10:24:39 am by dazz1 »
Dazz
Offline RFDUK
    • RFDesign
« Reply #28 on: April 23, 2018, 11:04:47 am »
The Roberts balun design not ideal for a biconical as it has a limited operating bandwidth of just over one octave.  Biconical requires 3 or 4 octaves.
Biconical impedance I forget, it depends on the cone angles.
Found thread Topic on this forum: Correctly calculating impedance of a biconical antenna and impedance matching.
Lots of stuff on the web. If you like maths ... http://archives.njit.edu/vol01/etd/1960s/1968/njit-etd1968-006/njit-etd1968-006.pdf
Martyn.
Offline philpem
  • That Sneaky British Bloke
« Reply #29 on: April 24, 2018, 07:06:52 am »
The UK has gone much the same way. The T&M scalpers are widely using ebay to buy equipment, which combined with their asking prices, has massively inflated the cost of even the most clapped-out old bit of "major repair needed" test gear.
You can still get deals, but you need to be ready to play the long game...
Phil / M0OFX -- Electronics/Software Engineer
"Why do I have a room full of test gear? Why, it saves on the heating bill!"
Offline Gandalf_Sr
« Reply #30 on: April 24, 2018, 10:31:53 am »
... For higher frequencies very low cost PCB Log Periodic antennas are available from Kent at http://www.wa5vjb.com.
Wow, not only are those PCB antennas really inexpensive, the References page is amazing - it has the original article by Hidetsugu Yagi on his antenna design dated 1926!
If at first you don't succeed, get a bigger hammer
« Reply #31 on: April 24, 2018, 11:37:57 am »
Hi
I think so too.  I have brought two of them as part of my EMC equipment fit out.
Dazz
Dazz
Offline Gandalf_Sr
« Reply #32 on: April 24, 2018, 12:19:44 pm »
Hi
Dazz
Do you have them yet?  I'm trying to figure out which SMA connector fits the PCBs that take them.
[EDIT] I emailed Kent and he replied - the PCB SMA footprint has 0.2" hole spacings so the TE SMA connectors from Digikey (Pt# A97594-ND) should work and are under $2 each.
« Last Edit: April 24, 2018, 12:49:04 pm by Gandalf_Sr »
« Reply #33 on: April 24, 2018, 08:07:40 pm »
Hi
I just paid extra to get the connectors included loose.  They also have an option to solder them on for you.
Dazz
Dazz
Offline Gandalf_Sr
« Reply #34 on: April 25, 2018, 12:54:08 am »
Oh, I didn't see that option - too late the PCBs have already shipped to me.  Anyway, I order from Digikey all the time so I'll order them tomorrow.
« Reply #35 on: April 26, 2018, 10:08:40 am »
Hi
I have been looking at precision dipoles.  More than a few implement the Richards balun in solid metal rather than coax cable.  https://avalontest.com/tdk-rf-solutions-dp-1800-precision-dipole-antenna/
This is a metalwork project.
Dazz
Offline charliedelta
« Reply #36 on: April 28, 2018, 01:14:41 am »
Roberts antenna construction attached.
A good reference about antenna calibration.
Theres some excellent information on the NPL web page that you can download after registering. You will certainly understand why these antennas cost so much after reading how these antennas are calibrated.
http://www.npl.co.uk/publications/calibration-and-use-of-antennas,-focusing-on-emc-applications.
« Reply #37 on: April 28, 2018, 05:15:33 am »
Hi
I find the NPL doc is fascinating reading, not so much for the highly detailed analysis but for the quaintness.
It refers to a photograph  as a "plate".   That type of printing process disappeared decades ago.
The photo of the NPL OATS showing a large 30m x 60m ground plane next to a residential area.  Fig 11 shows the high level of ambient RF.    The most perfect ground plane located on the most imperfect site.   Like setting up a hearing clinic at a rock concert.
This is not a criticism of the content. It is rare to find such a broad coverage of practical antenna design backed up by experience, data and maths.  Nice.
This looks to be a document written by an old fashioned expert in a well worn dust coat passing on his wisdom before retiring.
What is missing is details of how to replicate their calculable antennas.
With modern automation methods it should be possible to do in minutes what might have taken a day to do manually.  Calibrating an antenna should be cheap and easy now.
Dazz
« Last Edit: April 28, 2018, 05:25:42 am by dazz1 »
Dazz
« Reply #38 on: April 28, 2018, 09:01:58 am »
Hi
I have taken a closer look at the NPL calculable antenna design and it would be relatively easy to build.  The main barrier is the cost of a quality WB hybrid coupler like these ones: http://www.richardsonrfpd.com/Pages/Product-End-Category.aspx?productCategory=10136    You would only need 1 coupler for all dipole frequencies
The Minicircuits versions are a lot cheaper but with significantly less BW.
There is also a DIY discussion here: https://www.eevblog.com/forum/rf-microwave/directional-coupler-150mhz-buy-or-diy/
The thing I can't figure out is that Figure 1 specifies an orthogonal coupler (90degrees).  I think it should be a 180deg coupler.
« Last Edit: April 28, 2018, 09:11:22 am by dazz1 »
Dazz
« Reply #39 on: April 28, 2018, 09:56:00 pm »
Well enjoy the "quaintness" of the NPL document because it wont be long before it is privatized, outsourced and you have to pay for the documents. Much like the current standards that are funded by taxpayers but  you cant access because you have to pay again. I am sure you are familiar with that there in Australia!
Practical construction methods are included in many of the standards.
 I would download a copy of the MIL461G standard which includes a lot of   dimensions needed to build test antennas. Likewise ANSI 63.4  and many of the CISPR standards which give concise details of building the loops for measuring light fixtures  etc etc. The CISPR 25 automotive EMC standards especially the earlier copies had tons of information on building striplines and other test antennas and equipment. There was even details on how to build the conducted emission clamps. Its out there but you need to dig. Not many people like "sharing" the latest CISPR standards. There was also a lot of information published in the NBS and NIST documents and  these documents   had designs for everything needed to do make EMC test equipment. They also included all the maths  associated with calculating and building them. They were good old days when governments freely gave such good information.  Anyway those days are over, all our middle names are PAY according to the many governments of the world.
« Reply #40 on: April 29, 2018, 06:39:56 am »
Hi
I am designing the mechanics of a biconic antenna.  The attached drawing shows the cone at the base of the antenna. I plan to have an M5 screw going down the centre (not shown). The plan is to support the cone with a PTFE bush (not shown).   See photo.
What I haven't figured out is the best method of connection the balun to the M5 screw.   The Mil-Spec uses a large bolt.   I am thinking that I could use a brass screw and thin down the end so I can solder the balun wire to the end of the screw as shown in the drawing.  Alternatively I could make something like a dome nut and connect that to the balun.
How is this normally done for a biconic??  There are very few photos if inside the antenna termination box on Google.
« Last Edit: September 27, 2022, 01:03:02 am by dazz1 »
Dazz
Offline coppercone2
« Reply #41 on: March 11, 2019, 12:53:08 pm »
i saw an antenna build like this out of washers, bolts, etc.. no real milling or casting. the guy complained he had very poor performance (on some old radio forum)
Kinda wary but I will try to c ast those holder bits soon