I’m not entirely sure if frequencies below 1Hz still qualify as just ‘low’
Anyway, the SA44 is not useful below 1Hz anymore. Not because the datasheet says so, but the lowest resolution bandwidth is 0.1Hz and quite obviously spectrum analysis becomes problematic when the signal frequency approaches the resolution bandwidth.
For challenges like this, I had to pull another weapon, it’s the PicoScope 4262. Once again a little David that isn’t afraid to match with the various Goliath boat anchors out there
The lowest analysis bandwidth is 100Hz, but we can zoom in as much as we like (x100 in this example) and take a closer look at the 0-1Hz segment.
Just to make sure that David cannot lose, the RBW has been set to some 725uHz, resulting from a 524284 point FFT thus 381.5uHz frequency step and 725uHz RBW with the Blackman-Harris window.
Dual channel FFT? Well, why not?!
• Channel A : 1Vrms 100mHz sine wave
• Channel B : 1Vrms 200mHz sine wave
One sweep takes close to 44 minutes, so should be in the same ballpark as the HP35665.
Pico 4262 FFT-1Hz_2
Harmonic distortion of the input signals is below -70dB and all the harmonics are clearly visible.
Note the dynamic range (113dB) and SNR (110dB) in this measurement.
Also keep in mind that this is just 1% of the whole picture. We can adapt the frequency window by changing position and zoom factor anytime we like. The screenshot below uses the same zoom factor, but different frequency areas. The 100mHz signal for instance has still some odd harmonics even above 27Hz. For the 200mHz signal, the range 1-2Hz is shown, where we can see the last harmonic at 1.6Hz.
Pico 4262 FFT-1Hz_3