OK, I got a cheap non-medical isolation transformer, and have had a chance to take it apart. What I found inside is pretty interesting, and somewhat confusing. I've also done some reading, which has confused me further.
The following are connected together:
- the input ground pin
- the chassis
- the output ground pin
- the secondary neutral
- a "mystery" green wire going into the xformer on the 2ndary side
So, this is clearly not what I want. Now, my first idea to modify this to be useful as a bench tool was to separate the ground pin on the output and separate the secondary neutral from ground. However, this doesn't work because the outlet provided for the output actually links the ground screw and the mounting screws to the chassis.
So, my second idea was just to separate the input ground and leave everything else as is. This gives me true isolation on the secondary, and it leaves the secondary "floating ground" and neutral at the same potential. BUT it also leaves the chassis of the transformer at the secondary "floating ground."
Is this the best possible arrangement? It seems not to me. I don't like that a fault between primary and the transformer chassis would go undetected until I touched it. I'm also not sure if the neutral and ground on the output should be shorted. But, from the perspective of a DUT plugged into a "real" outlet, this is exactly the situation (they are connected back at the service entrance) so maybe it's fine.
Second, what is that mystery wire going into the transformer? It looks as if it is tap on the winding, but I know it is not. Is it to some kind of cage around the secondary to manage capacitive coupling?
Finally, I'm considering connecting the secondary "floating ground" to the real input ground through a 10 Meg resistor, maybe a small cap, too. Is this a bad idea? Pointless?
Suggestions welcome!