MHS-5200A DDS Generator Frequency Calibration
« on: February 06, 2018, 04:32:56 pm »
I have replaced the reference oscillator in my MHS-5200A with a stable 16.000000 MHz TCXO.  Got it dialed in to find what someone else has discovered...there must be a factory calibration value stored in memory somewhere.

Ref Osc. = 16.000000 MHz
Channel 2 Setting to achieve 0 beat with 10MHz WWV = 10.001131 KHz
Frequency Counter readout (Ch 2 loop) = 10.001207 MHz

I tried dropping the frequency of the TCXO down to see if I could correct it but of course it doesn't have that much range.

Has anyone discovered a USB/serial command, front panel key combination, or internal jumper that allows you to adjust the "factory" calibration?

I e-mailed the factory (in English) the week before Chinese New Year and asked them how to do it.  I will be holding my breath until they reply...

I will post a picture of the mod later, involved removing the XTAL and 2 associated caps, and tapping 3.3V off another cap to feed the TCXO I purchased.
 
The following users thanked this post: paf, Wrenches of Death

Offline tardsatTopic starter

  • Newbie
  • Posts: 3
  • Country: us
Re: MHS-5200A DDS Generator Frequency Calibration
« Reply #1 on: February 15, 2018, 09:16:06 pm »
OK, I figured it out.

Below is a picture of the connections I made to connect the 16.000000MHz TCXO. 

But here is the BIG news...with the power off, press and hold the SET button while powering on the MHS-5200A.  WALA!  A calibration menu appears.  I haven't had time to write up the details, but for me a 5000 frequency number gives the best result, although I don't think the ratio is able to directly work with 16.0000000MHz, so you may want to bend your reference a little to get exactly on.  Of course, I am into measuring frequency within .1Hz at 20MHz, so you may not need to be that exact.  Maybe someone else can play around and help me out.  There is also voltage level settings in the calibration menu.

Bad news for the frequency counter...displays 10.000074 on the self test when fed with the 10.0000000, regardless of cal setting, but that's still really useful.  I think that's in the coding, and it is consistent (1.000007MHz for 1.000000MHz in).

If you hold the top left button (CH1/2) while powering on I think it factory resets.   

Just get a 3.3V 16.000000MHz TCXO and connect the 3 leads as show below.  I don't have a picture of the "after" with me here at work.  Once I was able to "un-calibrate" the factory error correction I have a dandy little agile generator, and it has great resolution and accuracy in the area below 1 MHz!  It stays within 1 Hz at 10MHz through a range of room temperatures.  I use it to verify my GPSDO and other references aren't wobbling, and also to drive other radios reference crystals.



 