Author Topic: Yokogawa GS610 source measure unit service manual needed => have it => useless!  (Read 11397 times)

0 Members and 1 Guest are viewing this topic.

Online _Wim_Topic starter

  • Super Contributor
  • ***
  • Posts: 1517
  • Country: be
It could be that the unit now just needs calibration after this repair, but according to the service manual this needs to be performed automatically over GPIB with a Agilent 3458A. This is not something I have or want to buy, so I am “hoping” something is still broken…   :-//
 

Offline dl1640

  • Regular Contributor
  • *
  • Posts: 222
  • Country: cn
the GS is adjusted by 3458A with its built-in program automatically. i believe your local Yokogawa service center can do the job.
a pair of output modules might be defective along with analog board which you suspect.
 

Online _Wim_Topic starter

  • Super Contributor
  • ***
  • Posts: 1517
  • Country: be
Yeach, I am sure they can, but that will be way to expensive if I see the quote for the serivce manual and defect evaluation. So far I have been using the unit now as is, maybe one day I will buy a 3458A myself to perform the calibration...
 