One thing to keep in mind: two floating power supplies in series will generally be fine as long as you do not exceed their maximum common mode voltage relative to ground (don't go put two high voltage supplies in series, or several low voltage supplies in series, without checking this). But the device under test may not be happy with incorrect sequencing of power rails. For example the negative rail coming up before the positive rail, or the positive rail shutting down due to current limiting, while the negative rail does not. That's where two tracking voltage rails might be valuable. Some dual power supplies offer this, and some older power supplies also had rear terminal strips that allowed you two link two identical power supplies for tracking operation.
The cases where this would actually damage the device under test are probably fairly obscure, but keep in mind that there is more to this than just the power supplies.
… which is going to show that even an apparently simple topic could easily fill a book if we want to cover all possible eventualities and use cases
I’d like to expand on this.
The max. permissible common mode voltage is essentially the isolation of the PSU circuitry from the case and consequently mains earth ground. It is surprisingly hard to find any specifications for that, at least for the cheaper power supplies.
I did find some more detailed data only for a way more expensive PSU like the Rohde & Schwarz NGE 102/103. So I’ll use this as an example here:
Maximum voltage to earth 150V DC
Maximum reverse voltage 33V
Maximum inverse voltage 0.4V
Maximum permitted current in case of inverse voltage 3A
I figured this is a good opportunity to discuss specifications like this.
The max. voltage to earth is what has been referred to as common mode voltage before. So if this PSU is connected in series with some other PSU for increased total output voltage, then it should be made sure that neither of the output terminals of this PSU gets elevated more than 150V above ground potential (which will be the same potential as the PSUs metal case). I guess it will be next to impossible to find a PSU with even less isolation, specified or not, so as a general rule it is certainly safe to connect several PSUs in series up to at least 100V.
The max. reverse voltage indicates the point where reverse current starts to flow when an external voltage is applied to the outputs of the PSU, according to my earlier post with regard to paralleling PSUs. This specification is just 1V higher than the max. output voltage, hinting on a very efficient design with the input voltage only slightly higher than the max. output. This is a fairly conservative rating and I’m quite positive it does not mean the PSU will instantly blow up as soon as the reverse voltage exceeds 33V, just R&S don’t want to guarantee for anything in this case. Even so, with a little care it shouldn’t be too hard to meet this requirement.
The
inverse voltage specifications are for the
inverse voltage protection diode (I think I have incorrectly called it “
reverse voltage protection diode” in one of my earlier posts) connected in parallel with the output. It has a forward voltage drop of 0.4V (hinting on a Schottky diode) and can handle a max. current of 3A. So any imbalance in the current limiting of several PSUs connected in series should be less than 3A. As a general rule, I would keep them identical within 1A, which should be safe in any case, even if the reverse polarity protection diode happens to be just a 1N4004 or something similar.
Regarding the sequencing of power rails, it is quite obvious that we’re hardly able to turn them on/off at exactly the same time, when it’s two independent PSUs. Thankfully, I cannot think of a circuit requiring a symmetrical dual supply that would get damaged if one supply rail drops out. It just should not become reverse polarity, but that’s why we have the inverse polarity protection diodes in the PSU.
I think most dual output power supplies provide a tracking mode as described in an earlier post, and the master/slave principle does not only ensure identical settings on both channels, but also symmetrical output voltage in all operating modes, i.e. even when the current limit is active. So if one rail gets overloaded and reduces output voltage in order to maintain the current limit, the 2nd channel will reduce its output voltage accordingly.
There are indeed PSUs where several of them can be linked for building a tracking multi channel power supply. The Manson SSP-7080 SMPS is an example for this.
http://www.manson.com.hk/products/detail/178I own only one of these, hence haven’t tested the Master & Slave / link feature. Apart from that, I like this tiny little PSU a lot and it has proven to be reliable and convenient to operate.