I was looking for a Nixie tube DMM since some time, but they were generally overpriced (there is currently an auction at $380 on ebay for a 8300A
) ; I ended by finding a "not working" 8300A for only $5! Not sure it had any option (the seller was unable to tell me for sure), but I decided to buy it anyway, and I got a nice surprise: it has all the most useful options: the Ohm and mV option board + the AC volts board, great!
I am going to explain what was faulty and then show some pictures of this old Fluke 8300A Nixie tubes DMM
This Fluke 8300A was manufactured in april 1972, it's a 6 digits (120000 count), auto range that uses beautiful Nixie tubes for the display. It has a resolution of 1µV DC / 10µV AC / 10mOhms, with a fairly good accuracy. Detailed characteristics and schematics are available on Fluke site:
http://assets.fluke.com/manuals/8300A___imeng0400.pdfTime to try to repair it:I cared about selecting the right mains voltage, then powering it up: the display is working, but nothing changes on it, always displaying 00000 or 00001.
I quickly figured out that something wrong was happening with the front switches. Testing the switches' resistance shown some bad contacts (some of them where far above 1kOhms).
Disassembling these switches is a pain, they are all tied together, so I found another way: using some brake cleaner spray (some true contact cleaner would have been better, but I don't have any, and I already saved a radio potentiometer with this brake cleaner). Anyway, the brake cleaner worked, but it was not efficient enough for some of the contacts, so I decided to make pass a high current (several amps) through each contact, while manipulating the switches dozen of times. Before doing this, I carefully checked that I couldn't fry any low voltage component (like a Germanium transistor or diode), and I limited the voltage to 0.5V anyway.
The combination of the brake cleaner and the high current worked like a charm, and after this operation, all the switches contacts resistance were less than 1 ohm
Once reassembled, I powered the DMM (without the options boards), and ... it worked ... for a few minutes, until the first digit failed
(all numerals were lit at the same time on this 1st digit).
I checked the digital signals and voltages coming to the Nixie tubes: everything seemed to be OK, and only one numeral was electrically commuted at a time (though all numerals were lit). I have no knowledge on tubes, but I read on the internet that the Nixie tubes may lit all the numerals when they are powered with a too high voltage.
However the voltage is within the Fluke datasheed specs (190V) ... time to disassemble the DMM again (this DMM is pain to disassemble, HP has much better mechanic designs).
I found that the failing tube had been replaced with another model, hence the problem I guessed: the replacement model doesn't cope well with the 190V voltage?
Looking at the Fluke schematics, I figured out that the HV power supply and serie resistor is common to the 5 rightest tubes, but the sixth tube uses a separate resistor and driver (the 6th tube only display "1"). So I decided to exchange the 1st and the 6th tube, and increase a bit the resistor of the sixth tube: it worked! So the replacement model of the Nixie tube didn't support such high voltage, that's why all the numerals were lit ;-)
Notice the original Burroughs B-5870ST on the left (brown socket), compared to the replacement model: a National_Electronics_1970-0042
Now sitting on top of my HP 3457A, and quite in spec
(only had to tweak a really few settings and it was not used since year 2000 according to the seller!):
Posting some photos of the DMM:This is the top view of the DMM, without any option board
[continuing on next post]