Author Topic: Rigol DM3058E temperature measurement?  (Read 233 times)
Offline Dave92F1
  • Country: hr
    • Nerdfever
« on: September 10, 2018, 01:28:19 am »
Supposedly this DMM has presets for a Type K thermocouple. http://int.rigol.com/File/TechDoc/20151201/DM3058_DataSheet_EN.pdf even says "Thermocouple Cold Terminal compensation is built in the multimeter" (sic).
Anybody here know how to use this? I read the manual and I can't figure it out.
I've done:
  Sensor>Property>Type>Thermocouple
  Sensor>Property>Unit>Degrees C
When I hit "DONE" it says "Please set Para first" (which makes sense, I haven't told it it's a Type K yet).
The manual says to press "TC>Type" and "TC>JUNC", but I can't find these in the menus.
« Last Edit: September 10, 2018, 01:34:40 am by Dave92F1 »
Offline xrunner
  • hp>Agilent>Keysight>?
Re: Rigol DM3058E temperature measurement?
« Reply #1 on: September 10, 2018, 01:57:16 am »
Go to Sensor - Load - select file KITS90 - Read. I just tried it with a type K thermocouple and it works.  :)
The other method you were doing let's you input your own parameters for the sensor.
I am a Test Equipment Addict (TEA) - by virtue of this forum signature, I have now faced my addiction
The following users thanked this post: Dave92F1
Offline Dave92F1
  • Country: hr
    • Nerdfever
« Reply #2 on: September 10, 2018, 02:00:59 am »
That wasn't in the manual.
But it works!!!  ;D ;D ;D
Offline xrunner
« Reply #3 on: September 10, 2018, 02:07:50 am »
Just think of us as the manual within the manual.  :-DD