Dear experts
I recently bought a Rigol DP832.Everything was fine about the supply until today.While working today on bare foot,I accidentally touch the test leads at the back of my palm and felt some tingling feeling(At that time,all the supply outputs were "OFF").So i attached a electrical tester to the output test leads and the light was glowing!!!!
I received a two pin power cord with the power supply which has no Earthing pin(Schuko type).So I replaced the cord with a 3 pin cord and this time I tested with the electrical tester & the light was not glowing(Also No tingling sensation at the back of my palm).
I attached the dafault cord again and measured resistance(One of my fluke lead to one output lead of the supply & the other fluke lead to one of the input pin of the power cord ) for both Live & Neutral pin of the power cord.But each time,the meter showed "OL".
So my question,Is this normal or something is wrong with my supply??Any help is appreciated.