Ok it's fixed
I stripped the 1052E down to a bare chassis and visually inspected the motherboard around the CH1 and CH2 input circuits. As it turned out I needn't have bothered ! Read on.....
1. Whilst disassembling the scope it was noted that the motherboard retaining screws and BNC connector nuts were loose. Whilst it crossed my mind that this could be the cause of the problem, I decided to investigate further and as the scope was already disassembled it gave me the opportunity to look at the build quality of the PCB.
2. Once the motherboard was extricated from the chassis, it was visually inspected in the region of the CH1 & CH2 input circuits with a USB microscope (damned useful bit of kit). No suspect joints were found. The general quality of component placement and soldering was very good though PCB ‘hygiene’ was not great with fingerprints and minor detritus visible.
3. The scope was partially re-assembled to the point that it could operate but the CH1 & CH2 screening cover was not refitted. A 'knock' test was carried out on every component in the CH2 input circuit and adjacent PCB. On the 2mV range It was possible to produce a ringing waveform if the PCB was struck hard (but not too hard
). It was noted that a similar ringing waveform could be induced into CH1. This tended to disprove this as a fault as it is more likely to be a normal microphonic response within a component in the input circuit.
4. Specific capacitors in the input circuits of CH1 & CH2 were checked for shock sensitivity. Two large pink capacitors did produce impulse type responses and this is likely to be piezo effect as already suggested by another member of this forum.
5. It was not possible to reproduce the original fault symptom as detailed at the start of this topic
The scope was carefully re-assembled and all screws and nuts were tightened to a decent torque. The scope was then placed in Self Calibration mode after which it was tested for impact response.
My DS1052E now behaves normally and it takes a significant thump on the case to cause any sort of response on either CH1 or CH2. Both channels provide similar ‘bump response’ on the 2mV range. This is of negligible amplitude and may be attributed to normal component impact sensitivity and related effects.
So it would appear that my scopes sensitivity to impact stemmed from loose screws and nuts !
It was in interesting little exercise with which to occupy my evening