How is a meters "Counts" determined? That is, the value of, say, 6000 counts doesn't correlate to a 12-bit (4096) or 14-bit (16384) value so just where does the count value come from?
Brian
Brian,
A lot of DMMs don't standard off the shelf ADCs. Some use a DMM chipset and others like the Keithley 2001 use a multi-slope ADC implemented in a FPGA.
So they are not constrained by powers of 2.
A lot of the early low cost DMMs were based on the Intersil ICL7106 or ICL7107 which were 2000 counts. Actually 0 -1999 with the sign bit they are +/- 1999 counts
This was described as 3 1/2 (three and a half digits)
Then there were 4 1/2 digit meters.
Then some marketing people came up with 6000 count or 60000 counts some times called 3 3/4 digits or 4 3/4 digits.
These are useful because you get more resolution when measuring common voltages like 24 Vdc, 48 Vdc and 230 VAC.
Be careful, don't confuse resolution and accuracy. You can have 5 1/2 digit meters that are more accurate than some 6 1/2 digit meters.
Regards,
Jay_Diddy_B