Hey guys,
I have a Rigol DS1054Z and I've been playing around with some filters today. I was comparing the output of the first filter on CH1 versus the output of the second filter on CH2. I was getting "strange" output on the second filter, so I made the second filter equal to the first filter and to my surprise, the output was not equal on the scope.
Scope version information:
CH1 @ 1V per division and CH2 @ 1V per division triggering off CH3. Math A-B is on at 1V per division.
CH1 @ 2V per division and CH2 @ 2V per division triggering off CH3.
CH1 @ 500mV per division and CH2 @ 500mV per division triggering off CH3. I forgot to turn the math function on here but the A-B operation reflects the difference between to the two wave forms.
Note the difference in wave form. The probes are physically connected to the same point in the circuit:
I first suspected probe compensation so I checked both probes:
CH1 compensation:
CH2 compensation:
CH1 & CH2 & CH3 compensation side by side @ 1V per division:
CH1 & CH2 compensation side by side @ 500mV per division:
Note the difference in amplitude of the square wave. I tried two different probes connected to channel 2 but the same issue was present. I tried the self calibration feature of the scope but again the problem persisted. Just to be sure, I then tried switching CH2 with CH3 and the results were correct.
CH1 @ 1V per division and CH3 @ 1V per division triggering off CH2.
CH1 @ 500mV per division and CH3 @ 500mV per division triggering off CH2.
This issue only happens when switching from 1V per division to 500mV per division and lower on CH2. I can't decide whether this is a hardware or software bug since there are no relay clicks when you change from 1V to 500mV per division. Any ideas on what the issue could be?
« Last Edit: December 28, 2015, 02:59:22 pm by bitshift »
"It’s all fun and games until an innocent opamp gets hurt!" - Dave Jones
Online Fungus
Re: Rigol DS1054Z Software or Hardware Bug?
« Reply #1 on: December 28, 2015, 03:06:58 pm »
Might be something to do with the fact that the DS1054Z hardware doesn't support 500mV per division. You're not supposed to enable that 'feature'.
« Reply #2 on: December 28, 2015, 03:07:45 pm »
I believe that is 500uV not 500mV?
Online Fungus
« Reply #3 on: December 28, 2015, 03:18:52 pm »
Oops.  :palm:
Online Fungus
« Reply #4 on: December 28, 2015, 03:21:42 pm »
Channel 2 is AC coupled so I expect to see a capacitor charge/discharge waveform like that. What is channel 1 set to?
« Last Edit: December 28, 2015, 03:23:17 pm by Fungus »
« Reply #5 on: December 28, 2015, 03:22:53 pm »
Both AC coupled. I didn't mention it but on both AC and DC coupling the result is the same.
Online Fungus
« Reply #6 on: December 28, 2015, 03:33:46 pm »
What happens if you turn the 20MHz bandwidth limiter off?
Online Mark
« Reply #7 on: December 28, 2015, 03:39:25 pm »
What happens when you connect all the scope ground leads to ground and repeat the test?
« Reply #8 on: December 28, 2015, 03:39:53 pm »
The same response is acquired. This is with no bandwidth limit and both channels DC coupled.
« Last Edit: December 28, 2015, 03:43:19 pm by bitshift »
« Reply #9 on: December 28, 2015, 03:41:34 pm »
No difference
Offline Orange
« Reply #10 on: December 28, 2015, 03:45:15 pm »
Interesting puzzle.
It could be a faulty CH2 on your scope. It could also be a faulty CH2 probe.
Have you tried swapping the CH1 <> CH2 probe at the scope BNC connector ?
« Reply #11 on: December 28, 2015, 03:46:44 pm »
Interesting puzzle.
I tried swapping the probe from CH2 with CH1 and the again with CH3 but in both cases the problem remained.
« Reply #12 on: December 28, 2015, 03:50:00 pm »
Just for more context, CH1 and CH2 are the outputs of a filtered PWM signal. I'm trying to visualize the response of two different filters (but in the screenshots it's the same filter) to a step in the PWM duty cycle. CH3 is only a reference output so I can cleanly see when the PWM duty cycle changes. CH3 is not connected to CH1 and CH2 in any way.
« Last Edit: December 28, 2015, 03:51:39 pm by bitshift »
Offline Orange
« Reply #13 on: December 28, 2015, 03:52:12 pm »
Interesting puzzle.
I think your scopes CH2 hardware is damaged on the lower range. Probably a compensation C disconnected/open in the front-end.
Online Fungus
« Reply #14 on: December 28, 2015, 04:00:38 pm »
I agree.
It's a capacitance curve, something has a lot more capacitance than it's supposed to have. If it's only channel 2 then channel 2 is broken.
Do a factory reset to make sure it's not some obscure configuration. If it's still there then send it back.
« Reply #15 on: December 28, 2015, 04:09:05 pm »
Did a factory reset and the issue persists. I guess its time to test if Rigol will honor the warranty with the hacked firmware.
Thank you all for your help  :-+
« Last Edit: December 28, 2015, 04:55:33 pm by bitshift »
Offline alsetalokin4017
« Reply #16 on: December 28, 2015, 04:36:53 pm »
Remove the "hack" before you send it back !!  (And just for completeness sake, retest for the problem after you've removed the hack.)
:SYSTem:OPTion:UNINSTall
sent to the scope using your favorite method of sending SCPI commands.
Online Fungus
« Reply #17 on: December 28, 2015, 05:00:08 pm »
You can remove the hack... use an ethernet cable and install the Rigol software on the CD.
There's a command-line thingy in there somewhere. Open it and do ":SYSTem:OPTion:UNINSTall"
« Reply #18 on: December 28, 2015, 05:06:04 pm »
Thanks all  :-+
Just did a reset so back to a 1054Z. I tested again and the problem is still there. I will get hold of the dealer tomorrow and arrange for the unit to get sent back.
« Reply #19 on: January 14, 2016, 07:59:34 am »
I took the scope to the dealer and they checked it out and let me know that Rigol said they would give me a brand new unit. I received a unit today but the box was a little beat up/opened already. I'm skeptical that it is indeed a new unit. Is there a way I can lookup when the unit was manufactured or released based on the serial number?
« Reply #20 on: January 14, 2016, 11:39:21 am »
Further to the above, would it be wrong for me to insist on a sealed unit? I haven't had to time to inspect the instrument yet. I will do so thoroughly once I get home from work.
Online Fungus
« Reply #21 on: January 14, 2016, 01:38:08 pm »
Rigol is under no obligation so long as it's in at least as good condition as the one you returned.
(and the alternative might be to wait months for your personal 'scope to be fixed).
« Reply #22 on: January 14, 2016, 02:02:14 pm »
Understood, thanks. My inner skeptic took over for a while :scared: Busy testing the scope now and all seems perfect.
Offline fivefish
« Reply #23 on: January 14, 2016, 02:42:25 pm »
Good to hear.
And to answer your question, YES.
When you get back the scope, the hack isn't there though, but you'll have the latest firmware.