Author Topic: Test/review of DMM/Clamp UNI-T UT210E  (Read 3348 times)

0 Members and 1 Guest are viewing this topic.

Offline HKJTopic starter

  • Super Contributor
  • ***
  • Posts: 2904
  • Country: dk
    • Tests
Test/review of DMM/Clamp UNI-T UT210E
« on: September 09, 2017, 04:06:07 pm »

For the full review see: http://lygte-info.dk/review/DMMUNI-TUT210E%20UK.html



    Volt and frequency
        AC volt can measure up to about 2kHz (RMS will not work at the frequency).
        AC volt with VFC reduces bandwidth to 400Hz and limits range to 200.0V & 600V
        All ranges starts on AC
        Input impedance is 10Mohm to 11Mohm on AC and DC
        Buzzer sounds on over voltage (600V)
    Current
        Buzzer sounds on over current (100A).
        All ranges starts on AC
        As usual clamp meters are not very precise and at 0.5A and 5A it was about 2% out.
    Ohm, Continuity, diode and capacity
        Ohm is 1V open and 0.33mA shorted
        Continuity is fast (Below 4ms).
        Continuity beeps when resistance is below 50ohm.
        Continuity is 1V open and 0.33mA shorted
        Diode range uses 3.3V, max. display is 2.2V at 0.5mA, max. current is 1.4mA shorted
        11000uF takes about 9 seconds to measure.
    Miscellaneous
        Current consumption of meter is 7.2mA (13.2mA with backlight) in current mode
        Current consumption of meter is about 1.7mA (7.7mA with backlight) in other modes.
        Meter works down to 2.2V where it turns off, battery symbol show at 2.5V.
        Backlight only works down to about 2.6V where it is fairly dim.
        Both current and voltage readings are stable with changing battery voltage.
        The meter usual need a couple of display update to reach the final value.
        Viewing angle is good.
        Display updates around 3 times/sec
        Backlight will automatic turn off in about 15 seconds.
        Will automatic turn power off in about 15 minutes.
        Weight is 174g without accessories, but with batteries.
        Size is 174 x 62 x 32mm
    Probes & clamp
        Probe resistance 37mOhm for one.
        Probe wire is 90cm long.
        Clamp can handle up to ΓΈ17mm







Conclusion

This meter is nice for measuring 100's of mA or more without any voltage drop, using it for lower DC currents requires careful use of the zero button.
It has basic DMM functions, but it missing frequency and low currents.
It looks like it has real input protection and the CAT rating may be correct.


 
The following users thanked this post: DaneLaw

Offline don.r

  • Frequent Contributor
  • **
  • Posts: 740
  • Country: ca
Re: Test/review of DMM/Clamp UNI-T UT210E
« Reply #1 on: September 09, 2017, 04:35:33 pm »
Very nice for a sub $20 clamp meter. UNI-T is definitely improving some of their low-end products nicely.
 

Offline H.O

  • Frequent Contributor
  • **
  • Posts: 812
  • Country: se
Re: Test/review of DMM/Clamp UNI-T UT210E
« Reply #2 on: September 09, 2017, 04:44:17 pm »
For more interesting reading about this meter see this 17 pages long thread.
 