I have run into a few measurement anomolies during some testing I am doing. These are specific to the Siglent SSA3021X Spectrum Analyzer (SA). I am posting this message to get some advice on what may be causing these problems.
The test setup is shown in figure 1. There is an FEI FE-5650 Rubidium oscillator sourcing a 10 MHz sine wave (the red box at the lower left corner of the figure). It is connected to a Rigol DS1104Z (middle of the figure) by a coax and BNC T-connector, which forwards the signal to another coax connected to the SA (upper right of the figure). The SA connects to the coax through a N to BNC adapter. I warmed up the Rubidium oscillator for an hour before making measurements.
Figure 1
I first looked at the peak power level of the sine wave. The plot has the following parameters: start Freq - 8 MHz; stop Freq - 12 MHz; RBW - 30 Hz; SA internal attenuator - 20 dBm; SA Ref level - 0 dBm; Peak detection is turned on as is the Peak Table (figure 2).
Figure 2
The first thing I noticed was the power of the 10 MHz signal kept decreasing. The 10 MHz signal power began at 3.52 and then decreased as the number of sweeps increased. For example, here is the first series I observed: 3.52, 3.51, 3.50, 3.48, 3.47, 3.46 .... I wondered if this had something to do with a problem in the SA, so I turned it off for 10 minutes and then turned it on again without changing the other parts of the setup. However, the first measurement after 2nd turn on was 3.41 dBm, so turning the SA off and on didn't help.
I then disconnected the coax from the SA and reconnected it while the setup was running. The power of the 10 MHz signal went up to 3.49 dBm, but then started to decrease as before. I disconnected the coax a third time and the 10 MHz power went up to 3.46. It then remained stable at 3.44-3.46 dBm.
I have no idea what caused this. Perhaps some sort of static charge accumulated at the SA input. Perhaps it has something to do with the coax cables, the Rigol DS1104Z inputs, the BNC T-connector, the N to BNC adapter or the Rubidium oscillator input. I am totally stumped by this behavior. It is the first issue concerning which I am seeking commment.
Next, I used the channel power option of the SA to measure the power in a 4 MHz band around the 10 MHz fundamental frequency. The results are shown in figure 3. The first thing to notice is the measured channel power is 3.66 dBM. I used
this calculator to compute the difference between the channel power and that of the 10 MHz signal power: -9.83 dBm. This should be the value of the noise power in the 4 MHz band around 10 MHz.
Figure 3
However, notice in figure 3 the value circled in red at the top right of the picture. This shows the value of the center frequence marker, which according to the figure is -18.83 dBm. Huh? Well, if you look at the second thing circled in red on figure 5, you see that when I used the Channel Power option, the SA changed the Detect Type from Peak Power to Average Power. OK, but the two figures are over 20 dB apart. Surely, averaging the power, rather than detecting the peak shouldn't decrease the measured value by that much. (Ignore the value circled in blue, it relates to something I document in my internal notes)
The next day, I decided to reproduce the peak power level test and then instead of selecting the channel power option to simply switch the detect mode to Avg. and see what happens. (I mention that these tests took place the next day, since I ran into the decreasing power level problem again. The peak power value is different than the 3.44-3.46 dBm observed the day before. In fact it was 3.51 dBm)
Figure 4 shows the results. It shows the 10 MHz power as -78.43 dBm. This is obviously wrong, so I turned the SA off and then on again. Before selecting the power spectrum plot, I switched the detect mode to Avg. The results were effectively the same.
Figure 4
Maybe I misunderstand the meaning of Avg. power detect mode. I thought it would average the values within a 30 Hz bin instead of using the maximum of those values. In any case, since the SA automatically switches to Avg power detect mode when using the Channel Power option, I can't get a simultaneous value for both the channel power and the center frequency power in order to compute the noise power (i.e., channel power - center frequency power).
Comments?