It's a 485 with option 4 (EMI filtering) and option 78 (P11 phosphor). It's got the CRT mesh filter, which makes the blue seem even more brilliant! It's great for pictures (as that's it's intended purpose), but I didn't realize just how unfriendly P11 phosphor is on the eyes. It plays tricks with my vision and drives my eyes out of focus if I'm looking at it for more than just a couple seconds.
Sometimes that works. The vertical pots on mine are beyond that unfortunately.
I've had success with "repairing" these pots. It's just time consuming is all. Here's a quick run-down of what I've done:
1- remove vertical board, desolder pots, disassemble them and remove the part of the pot that holds the wiper. I like to put the rest of it back together to maintain orientation, and so I don't lose anything....
2- soak the wiper assembly in 99% IPA for a bit (usually an hour or so). After the initial soak, grab one of the leads with pliers and "whip" it to fling out the IPA.
3- repeat #2 two or three more times
4- Blast out with compressed air - I use canned air so I don't have to worry about the moisture & oil that comes out of an unfiltered compressor.
5- Reassemble, install, and test.
I don't clean the lube out from around the shaft unless it's dirty. I did clean it out on an intensity pot assembly once and the dielectric grease that I used to replace it just didn't seem to allow smooth operation. It's not noticeable on the display, but when I went to adjust the grid bias and Z axis, the darn thing was so stiff that I had to fiddle with it for some minutes before I got it to the correct voltage. Maybe some contaminants remained behind before I re-lubed it, maybe I used the wrong stuff, or maybe I used too much - but now I just leave the original lube in there and only clean out the wiper assembly. Fortunately I've only had to do this 3 or 4 times, but it would seem I'm about to be doing it again with some of the "sealed" Allen Bradley type vertical position pots. I'd rather do that than dish out $30 for a pair of "new" ones or raid another scope for replacements, though!!
It's a lot easier to do if they're the A-B type with the four slotted screws holding it together. It can be performed on ones with the plastic pins as well, you just have to either glue or melt it back together afterwards. The Bourns pots with aluminum dowel pins I haven't attempted yet, but I'm confident those can be done as well - it could be as simple as finding some screws and nuts small enough to replace the aluminum dowels.