I had a 468 that behaved as if all of the attenuation factors were off by a factor of 3 (1 V/div was .3V/div). That was only on 1 channel though. Tracked that down to a dirty AC/Gnd/DC coupling switch.
This could be the probes, the calibrator output or dirty contacts in the attenuators. Not enough info to say which though.
Although I would not recommend it at this point, before a complete diagnosis has been made, cleaning the attenuator contacts is possible. Use only IPA, lookup and follow the Tek cleaning procedure exactly. Be extremely careful. It's easy to mess up the tiny gold contacts and render them useless. Do not under any circumstances use any other contact cleaner.