Hi all,
I know this thread is old, but I recently got my THS710 and based on this thread I build a new NiMh based battery pack.
I do not have the external charger, and after a night in the scope the battery is lukewarm, so it should be ok.
I have a question about the scope calibration l like to perform because there is a small offset in ch1.
This what I did so far:
Power down,
Open the device,
Remove the jumper,
Perform calibration steps,
They end with 'pass',
Power down the unit,
Open the unit,
Replace the jumper,
Close the unit,
Power back on.
However, in the utility menu the scope item does not show 'pass'.
What did I do wrong?
To be honest, I now realize I left the battery in...
Hoping for some feedback,
Leo