I understand the SDS1104X-E can be upgraded to 200MHz in firmware, which is an attractive feature.
I just paid the extra and bought the SDS1204x-E.
It has a few quarks, but I have been able to work around the nicely. Operational learning curve isn’t too bad. Email tech support is excellent and they seem to take software suggestions and pass then on to their engineering team.
I measure and image Ultrasonic pulser images. I found that this 200MHz s one compares very favorably to a $8k Tektronix TDS-3034 scope. These pulses range from -90volts to -500volts with ~8ns fall times and 80ns - 230ns pulse width (measured at -10% down from 0%, not at 50%.
You can not change the pulse width measurement reference levels from 50% to what you need (-10% in my case), so you have to use the vertical cursors to properly measure pulse width. I have been told my need is in the software engineering hand now. Using a 100:1 probe the only difference in meAsurement compared to the expensive Tektronix is about 1ns higher fall time reading. The Siglent will report 9ns where the Tek reports 8ns. The specification To compare to is typically “<15ns”, so that is not an issue.
Except for my need for variable pulse width measurement reference points, I find it works damn close to Tek TDS3000 series scopes costing a lot, lot more.
It’s display is really nice and it is very lite in weight, perhaps too lite - lol. If I really wanted to spend $8000 to $11000 for a new scope, I would get a Tek TDS3034 or aTDS3054, but being cheap, I’ll take the $700 Siglent.
Oh and using VB and VBA programming environment, I find the Siglent network communications much, much faster than GPIB!
Dave
Sent from my iPhone using Tapatalk