OK, Tequipment.net seem to be willing to take it back. The question is do I ask for a replacement when they get them (presumable the "improved"(?) Rev3 board one) or do I look for something else?

The bugs discovered by Dave and forum were:

And mine exhibited all three. So what is fixed in the ones with the Rev3 board? All three? Just #1 (I see they put a larger heatsink)? To be honest #2 and #3 bugged me as much if not more. Not sure about #2 but I have the feeling #3 could not be fixed by a firmware update. Would greatly appreciate it if someone with a Rev3 unit does a quick check or points me to the right direction if this was already established.

BTW, has anyone seen later than Rev3 boards yet on that power supply?
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: EEVblog on September 19, 2013, 03:53:53 am
BTW, has anyone seen later than Rev3 boards yet on that power supply?

No, but I have been told by Rigol that the Rev 3 board was a result of my video. They had already fixed the heatsink issue in Rev 2 with the bigger heatsink, or at least they thought they had, but they have made it bigger again in the Rev 3 because they confirmed my resetting issue and the Rev 2 heatsink wasn't big enough for the extremes of the mains ranges.

The power on spike has been fixed I believe.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Electro Fan on September 19, 2013, 07:12:34 am
BTW, has anyone seen later than Rev3 boards yet on that power supply?

No, but I have been told by Rigol that the Rev 3 board was a result of my video. They had already fixed the heatsink issue in Rev 2 with the bigger heatsink, or at least they thought they had, but they have made it bigger again in the Rev 3 because they confirmed my resetting issue and the Rev 2 heatsink wasn't big enough for the extremes of the mains ranges.

The power on spike has been fixed I believe.

That would be very good!  The power on spike and the overheating (and the readout) were notable issues but if they have been corrected the product should be very popular and generate good sales for Rigol.  Congrats to Rigol for taking the corrective actions and congrats to EEVblog and all the forum members who helped provide feedback to Rigol.  (Maybe some recent buyers can confirm all 3 items have been fixed?) 

PS, any word on what caused the power on spike?
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: EEVblog on September 19, 2013, 07:25:03 am
(Maybe some recent buyers can confirm all 3 items have been fixed?) 

AFAIK, no rev 3 boards have hit the shelves yet.

Quote
PS, any word on what caused the power on spike?

No word.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: olsenn on September 19, 2013, 06:04:26 pm
Is there a way to tell if the boards we are rockin' are rev 1 or rev 2, without taking the case off? Such as by included firmware version or calibration date etc.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Marc M. on September 19, 2013, 09:52:26 pm
Is there a way to tell if the boards we are rockin' are rev 1 or rev 2, without taking the case off? Such as by included firmware version or calibration date etc.
I was able to see the board rev marking (V2.0) on my 832 thru the vent holes on the side.  I could also see the heatsink was located right next to the caps same as the one Dave reviewed.  It was interesting that my S/N was less than 40 units away from the one that Dave reviewed.  I received my unit on Monday, Sept. 16th so I was surprised to have a unit that close given the time that's passed between Dave's order and mine not to mention being half a world away.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: motocoder on September 25, 2013, 04:09:46 pm
AFAIK, no rev 3 boards have hit the shelves yet.

A few weeks ago, TEquipment.net told me they had a shipment of "fixed" DP832 arriving within 3 - 6 weeks. I placed an order for one, and just today received an email indicating the order was processed. Will post back here with what I get.

Can someone tell me where to look / best way to determine which Rev it is?
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Corporate666 on September 25, 2013, 06:03:41 pm
Maybe I'm confused - or not...


The "original" DP832 that Dave took apart with the bad thermal design - that is actually "rev2" as listed on the board, isn't it?  We are calling it Rev1 because it was the first unit we've seen, but it's actually Rev2 per the markings on the PCB.

Rev3 (what we're calling Rev2) is a design with an updated heat sink.

Rev4 (what we're calling Rev3) is an update on the heat sink fixed unit which includes an even bigger heat sink, and a fix for the power on spike.


Is that correct?


So the "Fixed" units that TE is talking about actually have the Rigol Rev3 (our Rev2) board... and there is an even newer Rigol Rev4 that hasn't hit the pipeline yet.


Right?
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: motocoder on September 26, 2013, 05:18:06 am
Maybe I'm confused - or not...


The "original" DP832 that Dave took apart with the bad thermal design - that is actually "rev2" as listed on the board, isn't it?  We are calling it Rev1 because it was the first unit we've seen, but it's actually Rev2 per the markings on the PCB.

Rev3 (what we're calling Rev2) is a design with an updated heat sink.

Rev4 (what we're calling Rev3) is an update on the heat sink fixed unit which includes an even bigger heat sink, and a fix for the power on spike.


Is that correct?


So the "Fixed" units that TE is talking about actually have the Rigol Rev3 (our Rev2) board... and there is an even newer Rigol Rev4 that hasn't hit the pipeline yet.


Right?

I sure hope not, as it means I'll be sending this one back immediately.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: EEVblog on September 26, 2013, 05:20:29 am
IIRC, there was a 2.1 (or some dot number) after my one, so the final fixed one Rev 3. But don't quote me on that.
In any case we'll know when we see the actual board, as it should have an even bigger heatsink that that shown on the one I posted.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Corporate666 on September 26, 2013, 05:29:34 am
Maybe I'm confused - or not...


The "original" DP832 that Dave took apart with the bad thermal design - that is actually "rev2" as listed on the board, isn't it?  We are calling it Rev1 because it was the first unit we've seen, but it's actually Rev2 per the markings on the PCB.

Rev3 (what we're calling Rev2) is a design with an updated heat sink.

Rev4 (what we're calling Rev3) is an update on the heat sink fixed unit which includes an even bigger heat sink, and a fix for the power on spike.


Is that correct?


So the "Fixed" units that TE is talking about actually have the Rigol Rev3 (our Rev2) board... and there is an even newer Rigol Rev4 that hasn't hit the pipeline yet.


Right?

I sure hope not, as it means I'll be sending this one back immediately.

Could you please check the power on spike and low power reading bug when you get it?  If those are fixed, it should be golden.  It could be that the heat sink fix Dave saw was just a preview of an intermediate board that never ended up being used to fix the problem, because Rigol saw his video and then decided to actually fix it in a more 'serious' manner (and the power on/reading bug).  I think if you don't see any of the 3 known bugs, you should be good (and then I can send mine back to Rigol for fixing!).
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: motocoder on September 27, 2013, 01:53:06 am
Could you please check the power on spike and low power reading bug when you get it?  If those are fixed, it should be golden.  It could be that the heat sink fix Dave saw was just a preview of an intermediate board that never ended up being used to fix the problem, because Rigol saw his video and then decided to actually fix it in a more 'serious' manner (and the power on/reading bug).  I think if you don't see any of the 3 known bugs, you should be good (and then I can send mine back to Rigol for fixing!).

Absolutely. I'll check and post here straight away. I still haven't gotten a shipping notice / tracking number yet, though.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: JohannP on September 28, 2013, 02:15:35 pm
OK, I have an DP832A on order with Tequipment since the 8th July 2013. After Dave's video I have put this order on hold until Rigol have sort out their problems. My instructions to Tequipment were not to cancel the order but only to ship when they are sure the all the problems were shorted.

Now I see the shipping date for my order will be the 24 October 2013 only. So hopefully Rigol and Tequipment are confident that the problems are shorted, or at least I hope so as I have paid good money for this PSU in July all ready and is still waiting patiently for it.

Best regards

Johann.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 10, 2013, 02:41:29 pm
I just got notified mine shipped from tequipment.net so I should be able to report the changes in a couple of days.  These are the new "fixed" variety.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: LoyalServant on October 10, 2013, 03:55:52 pm
Do you guys know what they are doing for folks that have < rev 3 boards?
I was really looking at getting one of these but Dave's findings sort of put me off.
If they sort it then great.. but I don't want one that will tank on me.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 10, 2013, 03:59:32 pm
Do you guys know what they are doing for folks that have < rev 3 boards?
I was really looking at getting one of these but Dave's findings sort of put me off.
If they sort it then great.. but I don't want one that will tank on me.

My guess is you'd have to contact Rigol support and see if they will do something.  Or report a thermal issue and see.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Sparky on October 10, 2013, 04:30:36 pm
Do you guys know what they are doing for folks that have < rev 3 boards?
I was really looking at getting one of these but Dave's findings sort of put me off.
If they sort it then great.. but I don't want one that will tank on me.

Several people have already asked Rigol and TEquipment.net in the USA about this.  Read specifically the last 6 posts on this page:
https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/405/ (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/405/)

Anyway, as per what grego said immediately above your post, there is apparently a new fixed revision shipping, so we will hopefully have lots of reports here soon of the revision.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 13, 2013, 06:12:59 pm
Just a quick peek inside the new DP832 -- mind you I just got mine and the stupid analog board (bottom board) is failing self-test so I might have to poke around but the heat sinking loots pretty much totally different now.  Apologies on the pictures, I only had my cell phone handy.

This is board rev 2.10 dated 2013-03-27.

Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: marmad on October 13, 2013, 08:20:34 pm
Just a quick peek inside the new DP832 -- mind you I just got mine and the stupid analog board (bottom board) is failing self-test so I might have to poke around but the heat sinking loots pretty much totally different now.  Apologies on the pictures, I only had my cell phone handy.

This is board rev 2.10 dated 2013-03-27.
This looks like the first revision with a slightly-larger heatsink, as shown here in Dave's photo (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg284232/#msg284232).
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Corporate666 on October 13, 2013, 09:17:01 pm
Has anyone in the USA heard about the recall?

I contacted Rigol USA and heard nothing back.  Then I read that TE was saying Rigol were waiting on fixed boards, and IIRC said Rigol will be fixing defective units at no cost.  It's been quite a few weeks and we've heard nothing else, although clearly updated units are now shipping.

What's the process for getting the recall done for USA customers?
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Sparky on October 13, 2013, 09:47:49 pm
Has anyone in the USA heard about the recall?

I contacted Rigol USA and heard nothing back.  Then I read that TE was saying Rigol were waiting on fixed boards, and IIRC said Rigol will be fixing defective units at no cost.  It's been quite a few weeks and we've heard nothing else, although clearly updated units are now shipping.

What's the process for getting the recall done for USA customers?

I submit a form on a Rigol website related to the DP800 series upgrade, but haven't heard anything yet.  See this post (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg293370/#msg293370), and the link within.

New units, and ones which have had board changes already, are possibly all coming from China to Rigol's warehouses for distribution.  I'm guessing Rigol NA are still working on getting the logistics of the recall worked out, and discussing with distributors how it will take place.

Also, updates for US customers are likely "low priority" because of the lower 110V AC mains.  Because of the lower voltage, we're less likely to experience extreme temperature on the LM317 or the resetting issue.  I've never had a reset on my DP832, and I don't recall anyone else in US posting that they experienced a reset.  Probably the first batches of new boards are going out to countries where the issue is more critical.


Edit: Also this post (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg293538/#msg293538); we just have to be patient a bit :)
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Sparky on October 13, 2013, 09:56:51 pm
Just a quick peek inside the new DP832 -- mind you I just got mine and the stupid analog board (bottom board) is failing self-test so I might have to poke around but the heat sinking loots pretty much totally different now.  Apologies on the pictures, I only had my cell phone handy.

This is board rev 2.10 dated 2013-03-27.
This looks like the first revision with a slightly-larger heatsink, as shown here in Dave's photo (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg284232/#msg284232).

Hey marmad!

It's beginning to look like the larger heatsink (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg299111/#msg299111) might not have made it into the board updates?!?  :-//

According to this Rigol page (http://beyondmeasure.rigoltech.com/acton/fs/blocks/showLandingPage/a/1579/p/p-0011/t/page/fm/0), the newest board revision is 2.1. 
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 13, 2013, 10:06:22 pm
2.10 is the board revision I just got on Friday.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: marmad on October 13, 2013, 10:08:27 pm
It's beginning to look like the larger heatsink (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg299111/#msg299111) might not have made it into the board updates?!?  :-//
Well, in the post you link to, John South mentions the larger heatsink versions will be leaving China around Oct. 17th - so still in the future. Perhaps it will be another 3-4 months or so before that size is found as standard.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Sparky on October 13, 2013, 10:13:17 pm
2.10 is the board revision I just got on Friday.

Yep, so it's definitely the updated version, I was just expecting to see even larger heat sinks (like shown in the links above)...so, bit of a surprise...

The sense wiring doesn't look any different.  Can you comment on the "out of spec" result when shorting the terminals?  See here (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg301588/#msg301588)
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Sparky on October 13, 2013, 10:17:54 pm
It's beginning to look like the larger heatsink (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg299111/#msg299111) might not have made it into the board updates?!?  :-//
Well, in the post you link to, John South mentions the larger heatsink versions will be leaving China around Oct. 17th - so still in the future. Perhaps it will be another 3-4 months or so before that size is found as standard.

Yep, so it's definitely the updated version, I was just expecting to see even larger heat sinks (like shown in the links above)...so, bit of a surprise...

Good point marmad!  Not so surprising after all :)
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: leafi on October 14, 2013, 10:12:11 pm
I just got my DP832. I will have to take a look. It is still in the box and just arrived today.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: iceisfun on October 14, 2013, 10:20:27 pm
Where is the best place to order from to make sure you get the newest revision hardware?

I have been wanting one of these PSU units

Right now I have a ebay Mastech PSU and it works great, but its time to upgrade it. For $200 the Mastech unit was a work horse and never really failed me.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 14, 2013, 10:24:04 pm
Where is the best place to order from to make sure you get the newest revision hardware?

I have been wanting one of these PSU units

Right now I have a ebay Mastech PSU and it works great, but its time to upgrade it. For $200 the Mastech unit was a work horse and never really failed me.

If you are in the US tequipment.net has the Rev 2.10 boards as I pictured above.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: centon1 on October 17, 2013, 03:23:10 pm
TEquipment
DP832 F/W 00.01.06
S/N DP8C1532xxxxx
Board revision V02.10 (2013-03-26)

Programmable Triple Output DC Power Supply
Less than $ 500.00 CAD. Half the price of a decent smartphone.
Something you can fix yourself, should you have to.
Works for me. I'm in!
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 17, 2013, 09:06:58 pm
Well I can confirm they definitely didn't fix the output spike on powerup with the V2.10 motherboard.

I've attached scope shots from all three channels (sequential) measuring on powerup.

Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: rbola35618 on October 18, 2013, 08:32:40 pm
Grego,

Any chance of measuring the current of the spike by placing a 1 ohms resistor. If there is not current, then it might not be an issue.

Robert
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 18, 2013, 10:20:12 pm
Grego,

Any chance of measuring the current of the spike by placing a 1 ohms resistor. If there is not current, then it might not be an issue.

Robert

Sure I can set that up.

Regardless, it's a problem.  There should be nothing on the outputs when the supply is turned on - especially since all the outputs default to off (as they should).
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Electro Fan on October 18, 2013, 11:09:45 pm
Any chance someone with any power supply can show us some screen shot's similar to Grego's (in terms of amplitude and time) that shows their power supply exhibiting a more flat response during turn-on (with the output turned off)?  Maybe someone with an Agilent or other highly regarded power supply?  Thanks
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Svuppe on October 19, 2013, 02:27:44 pm
Any chance someone with any power supply can show us some screen shot's similar to Grego's (in terms of amplitude and time) that shows their power supply exhibiting a more flat response during turn-on (with the output turned off)?

Well, I guess that is my cue to register on this forum, after being lurking around here for a while.
This is a screen capture of my Agilent U8002A turning on:

Mikael
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 19, 2013, 02:54:05 pm
Any chance someone with any power supply can show us some screen shot's similar to Grego's (in terms of amplitude and time) that shows their power supply exhibiting a more flat response during turn-on (with the output turned off)?

Well, I guess that is my cue to register on this forum, after being lurking around here for a while.
This is a screen capture of my Agilent U8002A turning on:

Mikael

Wow - transient spikes there too.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Svuppe on October 19, 2013, 03:59:05 pm
Here is the same measurement of my U8002A, but with a 680 ohm resistor across the output (couldn't find any smaller).
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: tnt on October 19, 2013, 04:15:46 pm
[quote vauthor=grego link=topic=20841.msg312679#msg312679 date=1382134812]
Regardless, it's a problem.  There should be nothing on the outputs when the supply is turned on - especially since all the outputs default to off (as they should).
[/quote]

Well, without load and just connected to the 1M input of a scope, anything could cause this, just induction in the wires, or even capacitive coupling between two disconnected relay contacts could realistically transfer such spikes.

But all those have no currents behind them, as soon as you connect them to something real, a real load, they'll be crushed ...
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: alm on October 19, 2013, 04:24:16 pm
Depends on what that something real is. What if it's connected to an MCU that represents a load of a few mA, or even a high-impedance input of an MCU of FPGA? Bench supplies are not just used for driving motors, sometimes they might just be used as reference voltage for an ADC.

That said, this spike might very well be just a common-mode spike (unless someone verified this).
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Rufus on October 19, 2013, 04:30:42 pm
Wow - transient spikes there too.

What do you expect when you switch mains with a contact that is going to arc and probably bounce onto almost any piece of equipment?

The transient will likely be mostly common mode coming through stray capacitance between the mains side and isolated power side. Your scope will see some of it as differential because of the huge impedance difference between the probe ground and tip and the scope is connected to the same earth as the power supply though its mains lead.

I doubt you will find a power supply with a real mains switch that doesn't appear to produce similar transients.

I thought the problem that Dave noticed was some real voltage lasting for 10s of ms on the output indicating the power supply regulator circuit or control circuit has a power up transient problem and some real voltage could do some real damage to a connected circuit.

When a circuit has multiple power supplies that may appear at different times or rise at different rates it is difficult to predict what will happen during power up and power down transients.

Title: Re: What is actually fixed in Rigol dp832(a) Rev3? - A summary
Post by: arvidj on October 26, 2013, 03:04:11 pm
I want to order an DP832 from TEquipment and would like to know what to expect when I open the box.

I've read this thread and am trying to summarize what it says. Feel free to provide corrections.

Original issues:

1) the overheating heatsink / voltage regulator

This appears to have been an issue with the Rigol version 2.0 boards. Rigol responded to issue with a revision 2.1 board which appears to be what is used in the units that are currently being delivered.

John South suggests Rigol will have "yet another revision" ... Re: EEVblog #512 - Rigol DP832 Bad Design Investigation Reply #433 (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg299111/#msg299111) that would have begun shipping on October 17th but no one has reported having received a unit with this board installed.

There have been several "mitigate the risk yourself" solutions suggested including replacing the LM317 with a  switching style three terminal equivalent (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg289944/#msg289944), putting a bigger heatsink on the existing LM317 and removal of the LM317 from the board and remotely mounting it on a much larger heatsink (https://www.eevblog.com/forum/blog/eevblog-512-rigol-dp832-bad-design-investigation/msg294419/#msg294419).

It has been suggested that the "larger\remote heat sink" solution ... be it by the DP832 owner or Rigol ... may not be a good long term solution as the localized temperature of the LM317 may get quite high even with a large heat sink. The "correct" solution would be to reduce the amount of power the LM317 is being asked to dissipate by lowering the input voltage, a solution that is best accomplished by Rigol. 

2) the messed up ammeter readout when current is under 12mA or so

I have not been able to find any reports of this issue having been resolved.

3) the output voltage spike when turning on the unit

I have not been able to find any reports of this issue having been resolved but if I understand the discussion in theses posts ... Reply #38 (https://www.eevblog.com/forum/testgear/what-is-actually-fixed-in-rigol-dp832(a)-rev3/msg312996/#msg312996),  Reply #39 (https://www.eevblog.com/forum/testgear/what-is-actually-fixed-in-rigol-dp832(a)-rev3/msg313000/#msg313000) and  Reply #40 (https://www.eevblog.com/forum/testgear/what-is-actually-fixed-in-rigol-dp832(a)-rev3/msg313005/#msg313005) there seems to be some thoughts that the current behavior [no pun intended] is satisfactory.

The "mitigate is yourself" solutions would seem to be "do not have anything plugged into the supply" when you turn it on or an outboard switch between the channel terminals and you device ... both of which seem to be fraught with human error possibilities ... especially if you are old like me :)

4) the current sense wires for channel 2 and 3

This was discovered in Re: Another issue with the Rigol DP832 power supply: sense wires carrying 3 amps! (http://test|https://www.eevblog.com/forum/testgear/another-issue-with-the-rigol-dp832-power-supply-sense-wires-carrying-3-amps!/msg285150/#msg285150) but I have not seen any information related to Rigol addressing the issue.

The "mitigate the risk yourself" suggestions are to take the front panel off and putting a heavy piece of wire between the channel 2 and 3 negative binding post connections ... though no one has reported actually doing that ... or doing the 'heavy but ugly wire' between the negative channel 2 and 3 posts on the outside of the front panel.

Corrections to this summary are encouraged.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Electro Fan on October 26, 2013, 06:17:50 pm
Any chance someone with any power supply can show us some screen shot's similar to Grego's (in terms of amplitude and time) that shows their power supply exhibiting a more flat response during turn-on (with the output turned off)?

Well, I guess that is my cue to register on this forum, after being lurking around here for a while.
This is a screen capture of my Agilent U8002A turning on:

Mikael

Mikael, Thanks for posting the image(s).  Looks like your power supply might be the leader in terms of having only two spikes (under no load) both with relatively low amplitudes and short durations. 

Maybe someone can post images from a power supply with an even cleaner turn-on?

Or maybe there is no such thing as a perfectly clean turn-on (with no load) with a typical hobbiest to medium price power supply (or even a high end power supply)?

Maybe a clean turn-on can only happen with at least some token load attached?

Looks like the jury is still out on this stuff?

EF 
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Svuppe on October 26, 2013, 07:57:28 pm
I just tried again, but this time with probe and ground both connected to the negative terminal of the psu. And I got pretty much the same picture. I believe Rufus is right in the "spikes" being common mode bouncing.
I should have tried a differential measurement with two probes, but that will have to wait until tomorrow. I am on the way to bed now.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Electro Fan on October 26, 2013, 08:30:33 pm
From the Agilent U8001A / U8002A User’s and Service Guide - notice the comments about "Wait for a few seconds for the output of the power supply to settle."  Not sure if this is relevant to the overall discussion of turn-on spikes.  (Also note the instruction to add 0.01 Ohm current shunt monitoring resistor across the output to be tested.)

Constant Current (CC) Verification
Current Programming and Readback Accuracy
This test is to verify that the current programming and
readback accuracy are within published specifications. The
accuracy of the current monitoring resistor must be 0.1% or
better.
Procedures:
1 Power- off the power supply and connect a 0.01 ? current
shunt monitoring resistor (RM) across the output to be
tested and a DMM across the current monitoring resistor
(RM).
2 Power on the power supply.
3 When the power supply is in limit mode, program the
output voltage to maximum programmable value and the
current to 0 A.
4 Enable the output.
5 Wait for a few seconds for the output of the power supply
to settle.  Make sure that the power supply is in CC mode.
6 Divide the voltage drop (DMM reading) across the current
monitoring resistor (RM) by its resistance to convert to
amps and record this value (IO). This value should be
within the limit of 0 A ±20 mA.
7 When the power supply is in meter mode, record the
current reading displayed on the front display of the
power supply. This value should be within the limit of
IO ±20 mA.
8 Disable the output.
9 When the power supply is in limit mode, program the
output current to full rated value, i.e. 3 A for U8001A and
5 A for U8002A.
10 Enable the output.
11 Wait for a few seconds for the output of the power supply
to settle.  Make sure that the power supply is in CC mode.
Service Guide 5
U8001A/U8002A User’s and Service Guide 77
12 Divide the voltage drop (DMM reading) across the current
monitoring resistor (RM) by its resistance to convert to
amps and record this value (IO). This value should be
within the limit of:
• U8001A: 3 A ±30.5 mA
• U8002A: 5 A ±37.5 mA
13 When the power supply is in meter mode, record the
current reading displayed on the front display of the
power supply. This value should be within the limit of:
• U8001A: IO ±30.5 mA
• U8002A: IO ±37.5 mA
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: Svuppe on October 27, 2013, 10:11:45 am
I should have tried a differential measurement with two probes, but that will have to wait until tomorrow.

Now I've done the two probe test. One probe to the positive output and the other to the negative. Both referenced to ground.
The math (bottom) line is the difference between the two channels. Same scale.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: LaurenceW on October 27, 2013, 12:12:39 pm
Well, I think that proves the point, exactly. All of the screen shots I am seeing here are not "real" voltages appearing across the output terminals of the power supply, but simply electromagnetically coupled noise from the mains switch. I can recreate that, with my 'scope probe simply lying unconnected on my bench, while switching on other kit!

A few screenshots from my own power supply - an old school analogue TTI PL330P. All single-ended measurements.

Under no-load conditions, the output voltage is reached, somewhat quicker.

I do go back to an earlier observation about the Rigol power supply design. Dissipating - was it around 7 Watts? - great chunks of power in an internal 5V regulator is just bad Karma. Sticking a whopping great heat sink on it in the hope that nobody will notice, is not good Karma.

Finally, I think it was Bill Gates who once said, "Never buy a piece of software until it reaches at least V3.0" And if any of you remember Windows 2, you'll know he was right. :palm:
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: grego on October 27, 2013, 12:36:46 pm
Well, I think that proves the point, exactly. All of the screen shots I am seeing here are not "real" voltages appearing across the output terminals of the power supply, but simply electromagnetically coupled noise from the mains switch. I can recreate that, with my 'scope probe simply lying unconnected on my bench, while switching on other kit!

A few screenshots from my own power supply - an old school analogue TTI PL330P. All single-ended measurements.
  • First shot shows the output under a 100R load, when set to 5V and the mains input is switched on. Good, clean rise in the output over 10mS, which is only one half of a mains cylce! The little step at 1V moves about, depending on when in the mains cycle the AC switch is thrown. This is not significant.
  • Second shot shows the same conditions, but set for 24V. A similar result, as expected. No dramas.
  • Final shot shows same 24V output setting, but this time enabled by the DC output switch (AC mains inout on, throughout this time). On my supply, the DC output switch is REAL, not electronic. There are a variety of turn-on spikes, but I would see the same result from a 24V battery, to be honest! Different results, each time. Notice the time duration, here. It is all over in a few uS. This is a mixture of switch bounce, and AC currents flowing into the inductive and capacity load of assorted bench leads. Can look exciting, but it is really no issue. If you have ciruit loads that might be disturbed by this sort of power supply, you've got a bigger design problem with your circuit! The power supply itself is not a problem, in this case.

Under no-load conditions, the output voltage is reached, somewhat quicker.

I do go back to an earlier observation about the Rigol power supply design. Dissipating - was it around 7 Watts? - great chunks of power in an internal 5V regulator is just bad Karma. Sticking a whopping great heat sink on it in the hope that nobody will notice, is not good Karma.

Finally, I think it was Bill Gates who once said, "Never buy a piece of software until it reaches at least V3.0" And if any of you remember Windows 2, you'll know he was right. :palm:

I agree entirely.  I think it's significantly less of an issue for us in the US since we're one 110 vs 240 but it's still a bad design if it's supposed to support all the voltages worldwide as an input.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: alm on October 27, 2013, 01:53:14 pm
Now I've done the two probe test. One probe to the positive output and the other to the negative. Both referenced to ground.
The math (bottom) line is the difference between the two channels. Same scale.
Note that the residual signal may very well be due to the fairly limited CMRR of this setup, especially at higher frequencies.
Title: Re: What is actually fixed in Rigol dp832(a) Rev3?
Post by: arvidj on October 28, 2013, 01:31:11 pm
As a noob I would like clarification of the previous "turn on voltage" discussion as it related to the Rigol DP832.

If I am understanding several of the posts related to the Agilent U8002A and TTI PL330P there is no real "turn on voltage" issue. What is seen on the scope is simply electromagnetically coupled noise and (a) not caused by the power supply and (b) is not an issue.

Can this conclusion be extended to the behavior exhibited by the Rigol DP832 or is additional testing required?

As a totally separate issue, I have revised my "summary" to indicate that simply putting a larger heat sink on the LM317 is a questionable solution as the localized temperature of the LM317 may be extremely high even with a very large heat sink. The correct solution is to reduce the total amount of power the LM317 is being asked to dissipate.

As usual, corrections to my naive and inexperienced conclusions are appreciated.