Author Topic: Wiltron 561 Scalar Network analyser ramp error  (Read 2237 times)
Offline matt09Topic starter
« on: December 12, 2018, 09:51:33 pm »
Hi all,
Recently acquired this bit of kit, not really my area of expertise so unsure where to go with it.
I have substituted the dead NiCad battery but I still get the error:
TEST FAILED
(A4) warning - ramp not calibrated
CH2 LED flashes on power up which does indicate an A4 - sweeper interface error.
Does anyone have experience with kit of this era?
Have glanced through the service manual but it doesn't mention this and cannot find a copy of the regular manual.
Cheers
« Last Edit: December 12, 2018, 09:53:06 pm by matt09 »
Re: Wiltron 561 Scalar Network analyser ramp error
« Reply #1 on: December 14, 2018, 11:23:51 pm »
Anyone?
« Reply #2 on: December 22, 2018, 03:29:14 pm »
bump!
« Reply #3 on: January 01, 2019, 02:42:50 pm »
New years bump, someone must have used these?
« Reply #4 on: January 04, 2019, 06:31:42 pm »
Please guys   :-//
Offline nonlinear
  • Country: ru
« Reply #5 on: January 05, 2019, 11:37:54 am »
Ramp is a sawtooth frequency sweep signal (usually 0-10V) generated either by analyzer itself or by external frequency sweeper.
In your case there are 2 variants:
1. Analyzer is configured to use external ramp and misses it.
2. Internal ramp generator is faulty.
Look through the user and service manuals (both are freely available - Google for them).
« Reply #6 on: January 06, 2019, 08:14:46 pm »
Yes I know the board is for sawtooth generation, I've looked at both manuals but neither seem to give specific information regarding a procedure to verify the unit via test points. This would imply there is a third type of test and validation manual, though I can find no mention of it.
Offline nonlinear
  • Country: ru
« Reply #7 on: January 07, 2019, 04:57:30 pm »
Ramp generation is a part of the scan cycle.
So at least is the scan cycle itself working?
Any LED blinks each cycle?
Any blank/channel_switch strobes on the outputs?
Will analyzer accept external ramp signal (0-10V with potentiometer)?