When I tried the built in NUC, it is so bad that each time the NUC is performed, a different line pattern showed up.
I decided to investigate this as it had never been this bad on any thermocamera i v ever seen.
50 frame is taken right after NUC is performed, with a piece of cardboard against the lens.
randomly distributed line-pattern looks horrible.
another 50 frames with a low contrast scene are taken sequentially after cardboard CAL
without any post processing, with built-in NUC,
image looks like this.
take a average of 50 frame, noise reduced as expected by statistics
subtracted my own NUC (50 frame of cardboard), noise reduced much drastically, but un-uniform line is still there/
this conclude the inadequate built-in NUC algorithm, and could potentially conclude the drifting gain of horizontal ROIC with a bit more detail.