I spent a little more time on the camera and software today.
The control interface is a little clunky as no keyboard keys are used to replicate the cameras buttons and joystick interfaces. It is all mouse driven.
I have tested the camera with an SDCard and all is well. Images are stored on the SDCard so that option is enabled.
The camera has four levels of noise reduction as already stated. The best for everyday use is the 'Weak' filtering option as it does not noticing you impact upon imaging fast moving objects. The highest setting is definitely intended for static objects and such is stead in the manual. The filtering reduces the noise content of the images when small temperature spans are used. The camera is capable of 2C spans and with filtering switched off there is quite a lot of random noise content. This drops markedly when the filter is set to 'Weak' strength.
I noted in the imaging settings that there was an 'HQ interpolation' option that was selected. This may have been causing some of the weird artefacts in the original images. The cameras image output is upscaled by a factor of 2 before being displayed on the PC. This remains the case no matter what the condition of HQ interpolation. More investigation will be carried out in this area later.
The two streaming modes now make more sense to me. As I previously stated, the streaming is either 'RGB' or 'IR-Data'. In the RGB mode the images presented on the PC software window are literally the RGB video data that would normally be presented to the cameras LCD display or viewfinder (on the Variocam version). That is why the menu overlays are seen. This mode appears to provide a less pleasant image as it is literally an RGB video feed rather than raw, unadulterated radiometric data being read and displayed.
In the IR-Data mode of operation, the camera outputs all of the radiometric data needed to create an image on the PC. No camera generated GUI data is present. In this mode, the camera radiometric image data is used to create a clean and true thermal image in the PC window. The thermal span etc is shown as it is also supplied in the radiometric data stream. This data may be recorded as a data stream, video stream or as still images. The recorded data may be analysed using the VarioAnalyser program that is able to read the Radiometric file type directly. The VarioAnalyzer program also enables the user to save images in common file types such as JPEG, BMP and TIFF.
Various analytical tools are available n the camera and in the PC Analysis software. These include, but are not limited to Thermal profiling along a defined line, and an image Histogram. The analysis software is simp,e enough to use and even provides report writing tools. As true Radiometric data is used, the user may add temperature measurements to the images. Image palette, spans and centre temperature may also be changed to suit the uses needs.
The camera and software contains some interesting palettes and I must say I like some of the ones that I had not seen before. I do not see any option to create new custom palettes but it may be possible to add them to the analysis software. I will need to look into that.
The camera has behaved flawlessly throughout the testing and is a good performer. Clean images and very accurate temperature measurements. The Close-up lens is a bit specialist and limits the use of the camera to that of an inspection tool and thermal microscope. It images modern high density electronics with ease though.
I will hopefully find time to take a look at the combined aperture plate and FFC shutter tomorrow. It sounds like the stepper motor is having difficulty moving the aperture plate and so the drive current limit kicks in. The camera seems happy enough in other respects though.
I ordered another couple of Lemo connectors for the other camera today. Ouch that hurt in the wallet ! They are essential though so I just had to bite the bullet.
Later I shall set up the RS232 control link instead of using the Firewire. I can the. Send commands from the control software and capture them at the serial port. With some effort I should be able to capture many of the cameras control commands. Sadly not all though, as I feel sure there will be commands that the control software does not produce. The command set for FLIR PM series cameras are extensive and these cameras are likely similar.
I may, or may not, build a RS232C remote control for these cameras It would certainly be easier than using the mouse driven PC software GUI. I need to build a remote control for my PM series cameras first though as that is a higher priority.
I have not decided on the future of these cameras yet. I already own the FLIR A40 and recently purchased the genuine Close-up lens for that camera. I may keep one of these Jenoptik cameras and sell the other as you only need so many close-up cameras in the lab
I think I got quite a bargain with these Jenoptik cameras. Luck was on my side and good fortune smiled upon me. To find the proper remote control and analysis software on an agents web page, when such is not available from Jenoptik, was a real surprise and it was the difference between a good buy and a disaster !
Fraser