At this point I will add one question to the experts.
nukie wrote that he need to measure a pulsed source.
At the users manual of my own Metrawatt MA4S,
its says ....... calibrated for sinusoidal 50Hz , still capable for square waves ( add 11% at your readings)
and triangle waves ( remove 3% from your readings) .
And the question is, are all the analog meters capable to measure all sort of waves ?
Basically.FET & Vacuum Tube Voltmeters read Average values for AC,but have the dial calibrated for RMS of a
sine wave,just like ordinary analog meters.
Some special units do have circuitry to read True RMS,but are very rare.
Others,rarer still,are calibrated in peak values,but this is just a special scale on what is really an average reading meter.
The advantage with these sort of meters is that,like a DMM,they have a very high input
impedance,& when connected across the normally much lower impedances of typical circuitry
have very little effect upon that circuit.
VK6ZGO