Does the following link work? Post-mortem photos.
https://photos.app.goo.gl/Jt4soTyt2fnhkUp1AI measured the fuses. 2 Ohm and 1.2 Ohm according to my £10 stand in.
I measured the input resistor network, 890K Ohm.
Given where the rotary knob melted, if it melted during the "incorrect use", it would have been over on the A / mA portion of the dial. There are visible marks and damage to the solder mask in that area.
However. That melted spot is also directly under the battery compartment. (EDIT Just on the complete otherside of the board Paul!) The meter has already suffered cell leakage damage (which is when I oddly lost the continuity buzzer), but the meter was cleaned and worked for 3 years since.
I did also replace a pair of Duracells in there that "felt" a bit waxy to the touch and a few flakes of white crust... about a month ago.
Is it likely the "incorrect use" was just coindidentally on the slow progression of battery acid eating the rotatory dial structure?
Why then the strong smell of anti-sceptic. I mean it "stank" on opening. Very obvious. Is that what battery acid eating rotary dial and solder mask smells like?
On warranty, given the previous leakage damage and the meter being 5 years old. The best I'd foresee is just being nice and asking Extech if they want the meter back for fault mode diagnosis. I doubt I'd get a replacement.