Author Topic: Calibration - How to Mathematically  (Read 2602 times)

0 Members and 1 Guest are viewing this topic.

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #25 on: March 02, 2024, 11:46:56 am »
Of course, you can calculate the inverse function that gives you the temperature from the voltage delivered by the thermocouple.

coeficients =[-3.23439565e-04  2.05876957e-02 -5.93682771e-01  2.55052633e+01 2.42295885e-01]

Program:
Code: [Select]
import numpy as np

voltages = [
   0.000, 0.039, 0.078, 0.117, 0.156, 0.195, 0.234, 0.273, 0.312, 0.352,
   0.391, 0.431, 0.470, 0.510, 0.549, 0.589, 0.629, 0.669, 0.709, 0.749,
   0.790, 0.830, 0.870, 0.911, 0.951, 0.992, 1.033, 1.074, 1.114, 1.155,
   1.196, 1.238, 1.279, 1.320, 1.362, 1.403, 1.445, 1.486, 1.528, 1.570,
   1.612, 1.654, 1.696, 1.738, 1.780, 1.823, 1.865, 1.908, 1.950, 1.993,
   2.036, 2.079, 2.122, 2.165, 2.208, 2.251, 2.294, 2.338, 2.381, 2.425,
   2.468, 2.512, 2.556, 2.600, 2.643, 2.687, 2.732, 2.776, 2.820, 2.864,
   2.909, 2.953, 2.998, 3.043, 3.087, 3.132, 3.177, 3.222, 3.267, 3.312,
   3.358, 3.403, 3.448, 3.494, 3.539, 3.585, 3.631, 3.677, 3.722, 3.768,
   3.814, 3.860, 3.907, 3.953, 3.999, 4.046, 4.092, 4.138, 4.185, 4.232,
   4.279, 4.325, 4.372, 4.419, 4.466, 4.513, 4.561, 4.608, 4.655, 4.702,
   4.750, 4.798, 4.845, 4.893, 4.941, 4.988, 5.036, 5.084, 5.132, 5.180,
   5.228, 5.277, 5.325, 5.373, 5.422, 5.470, 5.519, 5.567, 5.616, 5.665,
   5.714, 5.763, 5.812, 5.861, 5.910, 5.959, 6.008, 6.057, 6.107, 6.156,
   6.206, 6.255, 6.305, 6.355, 6.404, 6.454, 6.504, 6.554, 6.604, 6.654,
   6.704, 6.754, 6.805, 6.855, 6.905, 6.956, 7.006, 7.057, 7.107, 7.158,
   7.209, 7.260, 7.310, 7.361, 7.412, 7.463, 7.515, 7.566, 7.617, 7.668,
   7.720, 7.771, 7.823, 7.874, 7.926, 7.977, 8.029, 8.081, 8.133, 8.185,
   8.237, 8.289, 8.341, 8.393, 8.445, 8.497, 8.550, 8.602, 8.654, 8.707,
   8.759, 8.812, 8.865, 8.917, 8.970, 9.023, 9.076, 9.129, 9.182, 9.235,
   9.288, 9.341, 9.395, 9.448, 9.501, 9.555, 9.608, 9.662, 9.715, 9.769,
   9.822, 9.876, 9.930, 9.984, 10.038, 10.092, 10.146, 10.200, 10.254, 10.308,
   10.362, 10.417, 10.471, 10.525, 10.580, 10.634, 10.689, 10.743, 10.798, 10.853,
   10.907, 10.962, 11.017, 11.072, 11.127, 11.182, 11.237, 11.292, 11.347, 11.403,
   11.458, 11.513, 11.569, 11.624, 11.680, 11.735, 11.791, 11.846, 11.902, 11.958,
   12.013, 12.069, 12.125, 12.181, 12.237, 12.293, 12.349, 12.405, 12.461, 12.518,
   12.574, 12.630, 12.687, 12.743, 12.799, 12.856, 12.912, 12.969, 13.026, 13.082,
   13.139, 13.196, 13.253, 13.310, 13.366, 13.423, 13.480, 13.537, 13.595, 13.652,
   13.709, 13.766, 13.823, 13.881, 13.938, 13.995, 14.053, 14.110, 14.168, 14.226,
   14.283, 14.341, 14.399, 14.456, 14.514, 14.572, 14.630, 14.688, 14.746, 14.804,
   14.862, 14.920, 14.978, 15.036, 15.095, 15.153, 15.211, 15.270, 15.328, 15.386,
   15.445, 15.503, 15.562, 15.621, 15.679, 15.738, 15.797, 15.856, 15.914, 15.973,
   16.032, 16.091, 16.150, 16.209, 16.268, 16.327, 16.387, 16.446, 16.505, 16.564,
   16.624, 16.683, 16.742, 16.802, 16.861, 16.921, 16.980, 17.040, 17.100, 17.159,
   17.219, 17.279, 17.339, 17.399, 17.458, 17.518, 17.578, 17.638, 17.698, 17.759,
   17.819, 17.879, 17.939, 17.999, 18.060, 18.120, 18.180, 18.241, 18.301, 18.362,
   18.422, 18.483, 18.543, 18.604, 18.665, 18.725, 18.786, 18.847, 18.908, 18.969,
   19.030, 19.091, 19.152, 19.213, 19.274, 19.335, 19.396, 19.457, 19.518, 19.579,
   19.641, 19.702, 19.763, 19.825, 19.886, 19.947, 20.009, 20.070, 20.132, 20.193,
   20.255, 20.317, 20.378, 20.440, 20.502, 20.563, 20.625, 20.687, 20.748, 20.810
]

x = np.array(list(range(0, 400)))
y = np.array(voltages[0:400])
coef = np.polyfit(y, x, 4)
print(coef)

p = np.poly1d(coef)
for t in range(1, 400):
   print(f'temp: {t:3d}ºC  error:{p(voltages[t])-t:1.2f}ºC')


Result:
Code: [Select]
[-3.23439565e-04  2.05876957e-02 -5.93682771e-01  2.55052633e+01
  2.42295885e-01]
temp:   1ºC  error:0.24ºC
temp:   2ºC  error:0.23ºC
temp:   3ºC  error:0.22ºC
temp:   4ºC  error:0.21ºC
temp:   5ºC  error:0.19ºC
temp:   6ºC  error:0.18ºC
temp:   7ºC  error:0.16ºC
temp:   8ºC  error:0.14ºC
temp:   9ºC  error:0.15ºC
temp:  10ºC  error:0.13ºC
temp:  11ºC  error:0.13ºC
temp:  12ºC  error:0.10ºC
temp:  13ºC  error:0.10ºC
temp:  14ºC  error:0.07ºC
temp:  15ºC  error:0.06ºC
temp:  16ºC  error:0.06ºC
temp:  17ºC  error:0.05ºC
temp:  18ºC  error:0.03ºC
temp:  19ºC  error:0.02ºC
temp:  20ºC  error:0.03ºC
temp:  21ºC  error:0.01ºC
temp:  22ºC  error:-0.00ºC
temp:  23ºC  error:0.00ºC
temp:  24ºC  error:-0.02ºC
temp:  25ºC  error:-0.02ºC
temp:  26ºC  error:-0.02ºC
temp:  27ºC  error:-0.02ºC
temp:  28ºC  error:-0.05ºC
temp:  29ºC  error:-0.06ºC
temp:  30ºC  error:-0.07ºC
temp:  31ºC  error:-0.05ºC
temp:  32ºC  error:-0.07ºC
temp:  33ºC  error:-0.08ºC
temp:  34ºC  error:-0.07ºC
temp:  35ºC  error:-0.09ºC
temp:  36ºC  error:-0.08ºC
temp:  37ºC  error:-0.10ºC
temp:  38ºC  error:-0.10ºC
temp:  39ºC  error:-0.10ºC
temp:  40ºC  error:-0.10ºC
temp:  41ºC  error:-0.11ºC
temp:  42ºC  error:-0.11ºC
temp:  43ºC  error:-0.12ºC
temp:  44ºC  error:-0.13ºC
temp:  45ºC  error:-0.11ºC
temp:  46ºC  error:-0.13ºC
temp:  47ºC  error:-0.12ºC
temp:  48ºC  error:-0.13ºC
temp:  49ºC  error:-0.13ºC
temp:  50ºC  error:-0.12ºC
temp:  51ºC  error:-0.12ºC
temp:  52ºC  error:-0.12ºC
temp:  53ºC  error:-0.12ºC
temp:  54ºC  error:-0.12ºC
temp:  55ºC  error:-0.13ºC
temp:  56ºC  error:-0.13ºC
temp:  57ºC  error:-0.12ºC
temp:  58ºC  error:-0.13ºC
temp:  59ºC  error:-0.12ºC
temp:  60ºC  error:-0.13ºC
temp:  61ºC  error:-0.12ºC
temp:  62ºC  error:-0.11ºC
temp:  63ºC  error:-0.11ºC
temp:  64ºC  error:-0.13ºC
temp:  65ºC  error:-0.13ºC
temp:  66ºC  error:-0.11ºC
temp:  67ºC  error:-0.11ºC
temp:  68ºC  error:-0.11ºC
temp:  69ºC  error:-0.12ºC
temp:  70ºC  error:-0.10ºC
temp:  71ºC  error:-0.11ºC
temp:  72ºC  error:-0.10ºC
temp:  73ºC  error:-0.09ºC
temp:  74ºC  error:-0.10ºC
temp:  75ºC  error:-0.10ºC
temp:  76ºC  error:-0.09ºC
temp:  77ºC  error:-0.09ºC
temp:  78ºC  error:-0.09ºC
temp:  79ºC  error:-0.09ºC
temp:  80ºC  error:-0.07ºC
temp:  81ºC  error:-0.07ºC
temp:  82ºC  error:-0.08ºC
temp:  83ºC  error:-0.06ºC
temp:  84ºC  error:-0.07ºC
temp:  85ºC  error:-0.06ºC
temp:  86ºC  error:-0.05ºC
temp:  87ºC  error:-0.04ºC
temp:  88ºC  error:-0.05ºC
temp:  89ºC  error:-0.05ºC
temp:  90ºC  error:-0.04ºC
temp:  91ºC  error:-0.04ºC
temp:  92ºC  error:-0.02ºC
temp:  93ºC  error:-0.02ºC
temp:  94ºC  error:-0.02ºC
temp:  95ºC  error:-0.01ºC
temp:  96ºC  error:-0.01ºC
temp:  97ºC  error:-0.02ºC
temp:  98ºC  error:-0.01ºC
temp:  99ºC  error:0.00ºC
temp: 100ºC  error:0.01ºC
temp: 101ºC  error:-0.00ºC
temp: 102ºC  error:0.01ºC
temp: 103ºC  error:0.01ºC
temp: 104ºC  error:0.01ºC
temp: 105ºC  error:0.01ºC
temp: 106ºC  error:0.04ºC
temp: 107ºC  error:0.03ºC
temp: 108ºC  error:0.03ºC
temp: 109ºC  error:0.02ºC
temp: 110ºC  error:0.04ºC
temp: 111ºC  error:0.05ºC
temp: 112ºC  error:0.04ºC
temp: 113ºC  error:0.05ºC
temp: 114ºC  error:0.06ºC
temp: 115ºC  error:0.05ºC
temp: 116ºC  error:0.05ºC
temp: 117ºC  error:0.06ºC
temp: 118ºC  error:0.06ºC
temp: 119ºC  error:0.06ºC
temp: 120ºC  error:0.06ºC
temp: 121ºC  error:0.08ºC
temp: 122ºC  error:0.07ºC
temp: 123ºC  error:0.07ºC
temp: 124ºC  error:0.08ºC
temp: 125ºC  error:0.07ºC
temp: 126ºC  error:0.08ºC
temp: 127ºC  error:0.07ºC
temp: 128ºC  error:0.08ºC
temp: 129ºC  error:0.09ºC
temp: 130ºC  error:0.09ºC
temp: 131ºC  error:0.10ºC
temp: 132ºC  error:0.10ºC
temp: 133ºC  error:0.10ºC
temp: 134ºC  error:0.10ºC
temp: 135ºC  error:0.10ºC
temp: 136ºC  error:0.09ºC
temp: 137ºC  error:0.09ºC
temp: 138ºC  error:0.10ºC
temp: 139ºC  error:0.09ºC
temp: 140ºC  error:0.10ºC
temp: 141ºC  error:0.09ºC
temp: 142ºC  error:0.10ºC
temp: 143ºC  error:0.11ºC
temp: 144ºC  error:0.09ºC
temp: 145ºC  error:0.10ºC
temp: 146ºC  error:0.10ºC
temp: 147ºC  error:0.10ºC
temp: 148ºC  error:0.10ºC
temp: 149ºC  error:0.10ºC
temp: 150ºC  error:0.10ºC
temp: 151ºC  error:0.09ºC
temp: 152ºC  error:0.11ºC
temp: 153ºC  error:0.10ºC
temp: 154ºC  error:0.09ºC
temp: 155ºC  error:0.10ºC
temp: 156ºC  error:0.09ºC
temp: 157ºC  error:0.10ºC
temp: 158ºC  error:0.09ºC
temp: 159ºC  error:0.09ºC
temp: 160ºC  error:0.10ºC
temp: 161ºC  error:0.10ºC
temp: 162ºC  error:0.08ºC
temp: 163ºC  error:0.08ºC
temp: 164ºC  error:0.08ºC
temp: 165ºC  error:0.08ºC
temp: 166ºC  error:0.09ºC
temp: 167ºC  error:0.09ºC
temp: 168ºC  error:0.08ºC
temp: 169ºC  error:0.07ºC
temp: 170ºC  error:0.08ºC
temp: 171ºC  error:0.07ºC
temp: 172ºC  error:0.08ºC
temp: 173ºC  error:0.07ºC
temp: 174ºC  error:0.08ºC
temp: 175ºC  error:0.06ºC
temp: 176ºC  error:0.06ºC
temp: 177ºC  error:0.07ºC
temp: 178ºC  error:0.07ºC
temp: 179ºC  error:0.07ºC
temp: 180ºC  error:0.07ºC
temp: 181ºC  error:0.06ºC
temp: 182ºC  error:0.06ºC
temp: 183ºC  error:0.05ºC
temp: 184ºC  error:0.05ºC
temp: 185ºC  error:0.04ºC
temp: 186ºC  error:0.05ºC
temp: 187ºC  error:0.04ºC
temp: 188ºC  error:0.03ºC
temp: 189ºC  error:0.04ºC
temp: 190ºC  error:0.03ºC
temp: 191ºC  error:0.03ºC
temp: 192ºC  error:0.04ºC
temp: 193ºC  error:0.02ºC
temp: 194ºC  error:0.02ºC
temp: 195ºC  error:0.02ºC
temp: 196ºC  error:0.02ºC
temp: 197ºC  error:0.02ºC
temp: 198ºC  error:0.02ºC
temp: 199ºC  error:0.01ºC
temp: 200ºC  error:0.01ºC
temp: 201ºC  error:0.00ºC
temp: 202ºC  error:0.01ºC
temp: 203ºC  error:0.01ºC
temp: 204ºC  error:-0.00ºC
temp: 205ºC  error:0.01ºC
temp: 206ºC  error:-0.00ºC
temp: 207ºC  error:0.00ºC
temp: 208ºC  error:-0.01ºC
temp: 209ºC  error:-0.01ºC
temp: 210ºC  error:-0.02ºC
temp: 211ºC  error:-0.02ºC
temp: 212ºC  error:-0.02ºC
temp: 213ºC  error:-0.02ºC
temp: 214ºC  error:-0.02ºC
temp: 215ºC  error:-0.02ºC
temp: 216ºC  error:-0.02ºC
temp: 217ºC  error:-0.02ºC
temp: 218ºC  error:-0.03ºC
temp: 219ºC  error:-0.03ºC
temp: 220ºC  error:-0.04ºC
temp: 221ºC  error:-0.03ºC
temp: 222ºC  error:-0.04ºC
temp: 223ºC  error:-0.05ºC
temp: 224ºC  error:-0.04ºC
temp: 225ºC  error:-0.05ºC
temp: 226ºC  error:-0.04ºC
temp: 227ºC  error:-0.05ºC
temp: 228ºC  error:-0.05ºC
temp: 229ºC  error:-0.05ºC
temp: 230ºC  error:-0.06ºC
temp: 231ºC  error:-0.06ºC
temp: 232ºC  error:-0.06ºC
temp: 233ºC  error:-0.06ºC
temp: 234ºC  error:-0.06ºC
temp: 235ºC  error:-0.06ºC
temp: 236ºC  error:-0.06ºC
temp: 237ºC  error:-0.07ºC
temp: 238ºC  error:-0.07ºC
temp: 239ºC  error:-0.06ºC
temp: 240ºC  error:-0.07ºC
temp: 241ºC  error:-0.07ºC
temp: 242ºC  error:-0.06ºC
temp: 243ºC  error:-0.07ºC
temp: 244ºC  error:-0.06ºC
temp: 245ºC  error:-0.07ºC
temp: 246ºC  error:-0.07ºC
temp: 247ºC  error:-0.08ºC
temp: 248ºC  error:-0.07ºC
temp: 249ºC  error:-0.07ºC
temp: 250ºC  error:-0.08ºC
temp: 251ºC  error:-0.08ºC
temp: 252ºC  error:-0.08ºC
temp: 253ºC  error:-0.08ºC
temp: 254ºC  error:-0.08ºC
temp: 255ºC  error:-0.08ºC
temp: 256ºC  error:-0.08ºC
temp: 257ºC  error:-0.08ºC
temp: 258ºC  error:-0.08ºC
temp: 259ºC  error:-0.07ºC
temp: 260ºC  error:-0.08ºC
temp: 261ºC  error:-0.08ºC
temp: 262ºC  error:-0.07ºC
temp: 263ºC  error:-0.08ºC
temp: 264ºC  error:-0.08ºC
temp: 265ºC  error:-0.07ºC
temp: 266ºC  error:-0.08ºC
temp: 267ºC  error:-0.08ºC
temp: 268ºC  error:-0.07ºC
temp: 269ºC  error:-0.08ºC
temp: 270ºC  error:-0.08ºC
temp: 271ºC  error:-0.07ºC
temp: 272ºC  error:-0.07ºC
temp: 273ºC  error:-0.06ºC
temp: 274ºC  error:-0.08ºC
temp: 275ºC  error:-0.08ºC
temp: 276ºC  error:-0.08ºC
temp: 277ºC  error:-0.08ºC
temp: 278ºC  error:-0.06ºC
temp: 279ºC  error:-0.06ºC
temp: 280ºC  error:-0.06ºC
temp: 281ºC  error:-0.06ºC
temp: 282ºC  error:-0.07ºC
temp: 283ºC  error:-0.06ºC
temp: 284ºC  error:-0.06ºC
temp: 285ºC  error:-0.07ºC
temp: 286ºC  error:-0.05ºC
temp: 287ºC  error:-0.06ºC
temp: 288ºC  error:-0.05ºC
temp: 289ºC  error:-0.04ºC
temp: 290ºC  error:-0.05ºC
temp: 291ºC  error:-0.04ºC
temp: 292ºC  error:-0.04ºC
temp: 293ºC  error:-0.05ºC
temp: 294ºC  error:-0.04ºC
temp: 295ºC  error:-0.04ºC
temp: 296ºC  error:-0.04ºC
temp: 297ºC  error:-0.03ºC
temp: 298ºC  error:-0.03ºC
temp: 299ºC  error:-0.03ºC
temp: 300ºC  error:-0.03ºC
temp: 301ºC  error:-0.03ºC
temp: 302ºC  error:-0.03ºC
temp: 303ºC  error:-0.03ºC
temp: 304ºC  error:-0.01ºC
temp: 305ºC  error:-0.02ºC
temp: 306ºC  error:-0.02ºC
temp: 307ºC  error:-0.00ºC
temp: 308ºC  error:-0.01ºC
temp: 309ºC  error:-0.01ºC
temp: 310ºC  error:-0.00ºC
temp: 311ºC  error:-0.01ºC
temp: 312ºC  error:-0.00ºC
temp: 313ºC  error:0.01ºC
temp: 314ºC  error:0.00ºC
temp: 315ºC  error:0.01ºC
temp: 316ºC  error:0.01ºC
temp: 317ºC  error:0.02ºC
temp: 318ºC  error:0.01ºC
temp: 319ºC  error:0.01ºC
temp: 320ºC  error:0.02ºC
temp: 321ºC  error:0.02ºC
temp: 322ºC  error:0.02ºC
temp: 323ºC  error:0.03ºC
temp: 324ºC  error:0.03ºC
temp: 325ºC  error:0.03ºC
temp: 326ºC  error:0.05ºC
temp: 327ºC  error:0.04ºC
temp: 328ºC  error:0.04ºC
temp: 329ºC  error:0.04ºC
temp: 330ºC  error:0.05ºC
temp: 331ºC  error:0.05ºC
temp: 332ºC  error:0.05ºC
temp: 333ºC  error:0.06ºC
temp: 334ºC  error:0.05ºC
temp: 335ºC  error:0.06ºC
temp: 336ºC  error:0.05ºC
temp: 337ºC  error:0.06ºC
temp: 338ºC  error:0.07ºC
temp: 339ºC  error:0.06ºC
temp: 340ºC  error:0.07ºC
temp: 341ºC  error:0.07ºC
temp: 342ºC  error:0.08ºC
temp: 343ºC  error:0.08ºC
temp: 344ºC  error:0.07ºC
temp: 345ºC  error:0.07ºC
temp: 346ºC  error:0.07ºC
temp: 347ºC  error:0.07ºC
temp: 348ºC  error:0.08ºC
temp: 349ºC  error:0.09ºC
temp: 350ºC  error:0.09ºC
temp: 351ºC  error:0.09ºC
temp: 352ºC  error:0.08ºC
temp: 353ºC  error:0.08ºC
temp: 354ºC  error:0.09ºC
temp: 355ºC  error:0.09ºC
temp: 356ºC  error:0.08ºC
temp: 357ºC  error:0.09ºC
temp: 358ºC  error:0.08ºC
temp: 359ºC  error:0.09ºC
temp: 360ºC  error:0.08ºC
temp: 361ºC  error:0.09ºC
temp: 362ºC  error:0.08ºC
temp: 363ºC  error:0.08ºC
temp: 364ºC  error:0.09ºC
temp: 365ºC  error:0.07ºC
temp: 366ºC  error:0.07ºC
temp: 367ºC  error:0.08ºC
temp: 368ºC  error:0.08ºC
temp: 369ºC  error:0.08ºC
temp: 370ºC  error:0.07ºC
temp: 371ºC  error:0.07ºC
temp: 372ºC  error:0.07ºC
temp: 373ºC  error:0.06ºC
temp: 374ºC  error:0.06ºC
temp: 375ºC  error:0.05ºC
temp: 376ºC  error:0.05ºC
temp: 377ºC  error:0.04ºC
temp: 378ºC  error:0.03ºC
temp: 379ºC  error:0.02ºC
temp: 380ºC  error:0.02ºC
temp: 381ºC  error:0.01ºC
temp: 382ºC  error:-0.00ºC
temp: 383ºC  error:0.00ºC
temp: 384ºC  error:-0.01ºC
temp: 385ºC  error:-0.03ºC
temp: 386ºC  error:-0.03ºC
temp: 387ºC  error:-0.05ºC
temp: 388ºC  error:-0.05ºC
temp: 389ºC  error:-0.07ºC
temp: 390ºC  error:-0.07ºC
temp: 391ºC  error:-0.08ºC
temp: 392ºC  error:-0.10ºC
temp: 393ºC  error:-0.11ºC
temp: 394ºC  error:-0.12ºC
temp: 395ºC  error:-0.15ºC
temp: 396ºC  error:-0.16ºC
temp: 397ºC  error:-0.17ºC
temp: 398ºC  error:-0.20ºC
temp: 399ºC  error:-0.21ºC
« Last Edit: March 02, 2024, 11:51:44 am by Picuino »
 

Offline MrAl

  • Super Contributor
  • ***
  • Posts: 1450
Re: Calibration - How to Mathematically
« Reply #26 on: March 02, 2024, 07:10:26 pm »
There is a problem with this approach. The curve may be approximately logarithmic, but not an exact logarithm. In that case (which is most of the time) it is again necessary to use a polynomial.

Note that logarithms are also calculated with polynomials, usually by doing the transformation:
t=(x-1)/(x+1)

And then applying a polynomial:
log_10(x) = a1·t + a3·t^3 + a5·t^5 + a7·t^7 + a9·t^9
a1=0.868591718
a3=0.289335524
a5=0.177522071
a7=0.094376476
a9=0.191337714
for x in the interval [1/sqrt(10), sqrt(10)]
(From Abramowitz and Stegun)

Hi,

We do not use log's alone, we use log's and a polynomial fit.  That results in very low error.

Of course here is the formula for thermistors too which may be even simpler if the thermistor is a good one.
 

Offline MrAl

  • Super Contributor
  • ***
  • Posts: 1450
Re: Calibration - How to Mathematically
« Reply #27 on: March 02, 2024, 07:13:13 pm »
Many people already said helpful things about different approaches to fitting polynomials, piecewiese linear interpolation, splines etc. So all I would like to add is this: If you happen to know why and how the sensor response is non-linear, the best method is to fit the specific function that is rooted in the physical model instead of fitting a generic smoother. E.g. if you know the response is logarithmic, fit a log function, not a polynomial. But if you don't, go with splines/LOESS/polynomials/whatever.

Hi,

I had mentioned this as one trick in post #12.  I had done that several years ago and it results is a really good fit, within about 0.1 percent for the entire temperature range.  That means that if the resistance calculated was 100 Ohms, it could really be 99.9 Ohms to 100.1 Ohms.
It requires using logarithms combined with a polynomial fit.
 

Offline Nominal Animal

  • Super Contributor
  • ***
  • Posts: 6288
  • Country: fi
    • My home page and email address
Re: Calibration - How to Mathematically
« Reply #28 on: March 02, 2024, 09:13:42 pm »
If we limit ourselves to calibration, then the issue becomes one of interpolation between calibrated values.
That is, we don't have, nor do we need, any specific algebraic description of the behaviour of the system; we only have the calibration points we can trust.

The two common approaches are linear and cubic interpolation.  Here is an example implementation of a linear interpolation, using an array of CALIBRATION_POINTS pairs of input, value points:
Code: [Select]
// SPDX-License-Identifier: CC0-1.0
#include <stdlib.h> // Only for the test program part
#include <stdio.h>  // Only for the test program part
#include <math.h>   // Only for the test program part
#include <stdint.h>

// For the test program part
#ifndef   CALIBRATION_POINTS
#define   CALIBRATION_POINTS  512
#endif

// Sorted in input increasing order: calibrated_input[i] < calibrated_input[i+1]
uint16_t        calibrated_input[CALIBRATION_POINTS];
int32_t         calibrated_value[CALIBRATION_POINTS];
const uint32_t  calibrated_limit = CALIBRATION_POINTS - 1;

int32_t  calibrated(uint_fast16_t input)
{
    // Range of values is limited to the calibration range.
    // That is, this does *interpolation* only, not *extrapolation*.
    if (calibrated_input[0] >= input)
        return calibrated_value[0];
    if (calibrated_input[calibrated_limit] <= input)
        return calibrated_value[calibrated_limit];

    uint_fast16_t  imin = 0;                 // Minimum index, inclusive
    uint_fast16_t  imax = calibrated_limit;  // Maximum index, exclusive
    while (1) {
        uint_fast16_t  i = imin + (imax - imin) / 2;
        if (i == imin)
            break;
        if (calibrated_input[i] < input) {
            imin = i;
        } else
        if (calibrated_input[i] > input) {
            imax = i;
        } else {
            return calibrated_value[i];
        }
    }

    if (calibrated_input[imin] > input || calibrated_input[imin+1] < input)
        printf("Error: %d < %d < %d is not true!\n", (int)calibrated_input[imin], (int)input, (int)calibrated_input[imin+1]);

    // At this point, we have calibrated_input[imin]   < input
    //                    and calibrated_input[imin+1] > input.
    const uint_fast16_t dmin = calibrated_input[imin+1] - input;
    const uint_fast16_t dmax = input - calibrated_input[imin];
    const uint_fast16_t dsum = calibrated_input[imin+1] - calibrated_input[imin];
    // Note that by definition, dmin, dmax, dsum are 16-bit values,
    // so the following really only uses 48-bit integers.
    return (int32_t)( ( dmin * (int64_t)calibrated_value[imin]
                      + dmax * (int64_t)calibrated_value[imin+1] ) / dsum );
}

/*
 * Test program follows
*/

// Given x = 0..1, what should the sample value x be?
double w(double x)
{
    return pow(x, 1.5);
}

// Function to use for testing
double f(double x)
{
    return sqrt(x);
}

int main(void)
{
    // For testing, we use fixed point integer valus with 9 integer bits and 22 fractional bits,
    // with f(x) = sqrt(x), x = 0 .. 65535
    printf("# input[i] value[i] f(input[i])\n");
    for (uint_fast16_t  i = 0; i < CALIBRATION_POINTS; i++) {
        // Importance sampling: we emphasize low values
        const uint_fast16_t  x = round(65535.0 * w((double)i / (CALIBRATION_POINTS - 1)));
        const double         r = f((double)x);
        const int32_t        v = round(r * 4194304.0);

        calibrated_input[i] = x;
        calibrated_value[i] = v;
        printf("%-5u %-10d %16.12f\n", (unsigned int)x, (int)v, r);
    }
    printf("\n");


    printf("# x calibrated(x) f(x) relative_error\n");
    double  errmin = 0.0, errmax = 0.0;
    for (uint_fast16_t x = 65535; x > 0; x--) {
        uint32_t       v = calibrated(x);
        const double   d = (double)v / 4194304.0;
        const double   r = sqrt((double)x);
        const double err = (d - r) / r;

        if (errmin > err)
            errmin = err;
        if (errmax < err)
            errmax = err;
        if (errmin == err || errmax == err)
            printf("%5d %16.12f %16.12f %+.12f\n", (int)x, d, r, err);
   }

    return 0;
}
It includes a test program that creates a synthetic calibration table for fixed-point square roots.

The calibrated() function assumes the input values are in increasing order, and uses a binary search to look up the correct index.
Each iteration halves the index range sought after, so with 2048 values, it only does 11 iterations.  Even with 65535 values, it'd only do 15 iterations.  As you can see, it does have one index comparison (to find out when we have the index), and a <, =, > between the input at that index and the input we are trying to resolve to a value, so it is very, very fast and efficient.

Because I wish to emphasize that regular intervals of inputs is not needed –– in many cases the binary search is faster than the division used to map input value range to a regularly spaced points, especially if the microcontroller does not have a fast hardware division operation –– I used importance sampled square root as the example function here.  The arrays contain x and sqrt(x), the latter in fixed-point integer format with 22 fractional bits.  The input is 16-bit, and value 32-bit, so it takes 6 bytes per calibration point.  By default, the example program uses 512 calibration points (would correspond to 3k of Flash), but you can easily adjust it, just define CALIBRATION_POINTS when compiling (-DCALIBRATION_POINTS=300 for example).
The output is in two parts: the first part is the calibration table.  Note how the calibration inputs are not regularly spaced!
The second part compares the looked-up result to calculated result, and whenever the relative error increases prints the output, so you can see the relative errors this gives.

As sqrt() function slope starts steep, then becomes smaller, the linearly interpolated value will always be below the actual value.  (The occasional positive relative error is from rounding the floating-point value to an integer.)  If you used regularly spaced input values, the relative error in the small values would become unacceptable for this kind of functions.  As it is, the error is less than 0.04% for inputs above 256, but at 23 and smaller, exceed 1%.

Thus, the most important thing when using linear interpolation is to choose the calibration points so that connecting them with lines produces a good representation of the continuous values.
« Last Edit: March 02, 2024, 11:16:39 pm by Nominal Animal »
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #29 on: March 02, 2024, 10:23:00 pm »
Linear interpolation is usually very poor and only gives precision results using a very large number of points.

Cubic interpolation is more difficult to handle, but in contrast returns surprisingly good accuracy compared to linear. This can be used to significantly reduce the number of points in the table.

Reference for programming algorithms:
https://en.wikipedia.org/wiki/Neville%27s_algorithm
https://www.grad.hr/nastava/gs/prg/NumericalRecipesinC.pdf
https://apps.dtic.mil/sti/pdfs/ADA612943.pdf
https://www.sci.brooklyn.cuny.edu/~mate/nml/numanal.pdf
« Last Edit: March 02, 2024, 10:42:05 pm by Picuino »
 

Offline radiolistener

  • Super Contributor
  • ***
  • Posts: 3423
  • Country: ua
Re: Calibration - How to Mathematically
« Reply #30 on: March 02, 2024, 10:44:55 pm »
some time ago I was implemented 100000 points calibration for NWT-7 device :D, it works very good to compensate very non linear curve and allows to get pretty precise measurements on a cheap device. :)
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #31 on: March 02, 2024, 10:54:07 pm »
Of course, LUTs without interpolation are the simplest way that can be done and have the great advantage of not needing calculations and being very fast.

It is like random number generation. You can store a list of 100000 random numbers, (with the advantage of being really random) or perform a more complicated calculation to calculate pseudo-random numbers (with the advantage of taking less space).

----------------

Another way to interpolate a table is to store the coefficients for the several interpolating polynomials by table sections. You can place several relatively low degree (cubic) interpolating polynomials with coefficients already precomputed, to save you the somewhat computationally intensive (O(n2)) of Neville's algorithm.
« Last Edit: March 02, 2024, 10:57:24 pm by Picuino »
 

Offline Nominal Animal

  • Super Contributor
  • ***
  • Posts: 6288
  • Country: fi
    • My home page and email address
Re: Calibration - How to Mathematically
« Reply #32 on: March 02, 2024, 11:59:51 pm »
Here is the cubic interpolation corresponding to the linear interpolation I showed in reply #28.  It's not difficult at all, either.
Code: [Select]
// SPDX-License-Identifier: CC0-1.0
#include <stdlib.h> // Only for the test program part
#include <stdio.h>  // Only for the test program part
#include <math.h>   // Only for the test program part
#include <stdint.h>

// For the test program part
#ifndef   CALIBRATION_POINTS
#define   CALIBRATION_POINTS  128
#endif

// Sorted in input increasing order: calibrated_input[i] < calibrated_input[i+1]
float           calibrated_input[CALIBRATION_POINTS];
float           calibrated_coeff[CALIBRATION_POINTS - 1][4];
const uint32_t  calibrated_limit = CALIBRATION_POINTS - 1;

float calibrated(const float input)
{
    uint32_t  i;

    do {
        if (calibrated_input[0] >= input) {
            i = 0;
            break;
        }

        if (calibrated_input[calibrated_limit] <= input) {
            i = calibrated_limit - 1;
            break;
        }

        // Binary search.
        uint32_t  imin = 0;                 // Inclusive
        uint32_t  imax = calibrated_limit;  // Exclusive
        while (1) {
            i = imin + (imax - imin) / 2;
            if (i == imin)
                break;
            if (calibrated_input[i] > input) {
                imax = i;
            } else
            if (calibrated_input[i] < input) {
                imin = i;
            } else
                break;
        }
    } while (0);

    // We have calibrated_input[i] <= input < calibrated_input[i+1].
    const float  d0 = input - calibrated_input[i];
    const float  d1 = calibrated_input[i+1] - input;
    const float  d0d0 = d0 * d0;
    const float  d1d1 = d1 * d1;
    return d0d0*(d0*calibrated_coeff[i][0] + d1*calibrated_coeff[i][1])
         + d1d1*(d0*calibrated_coeff[i][2] + d1*calibrated_coeff[i][3]);
}

/*
 * Test program follows
*/

// Value and derivative used to calculate the coefficients
float calibrated_value[CALIBRATION_POINTS];
float calibrated_deriv[CALIBRATION_POINTS];  // Actual slope at the calibration point
float calibrated_slope[CALIBRATION_POINTS];  // Estimated slope at the calibration point

// Importance sampling: Given x = 0..1, what should the sample value x be?
double w(double x)
{
    return x*x;
}

// Function to use for testing
double f(double x)
{
    return sqrt(x);
}

// Derivative of the above function for testing
double df(double x)
{
    return 0.5 / sqrt(x);
}

int main(void)
{
    /*
     * Mapping calibration values to coefficients
    */

    // First, we obtain a set of calibration points, input sorted in increasing order.
    for (uint32_t  i = 0; i < CALIBRATION_POINTS; i++) {
        // Importance sampling: we have denser calibration points for smaller inputs.
        const double p = w((double)i / (double)(CALIBRATION_POINTS - 1));
        const double x = (1.0 - p)*1.0 + p*65535.0;

        calibrated_input[i] = x;
        calibrated_value[i] = f(x);
        calibrated_deriv[i] = df(x);
    }

    // If we do not know the derivative at each calibration point, we can calculate them
    // based on preceding and succeeding point for each calibration point.
    // Initial slope is toward the second calibration point.
    calibrated_slope[0] = (calibrated_value[1] - calibrated_value[0]) / (calibrated_input[1] - calibrated_input[0]);

    // Final slope is also similarly linear.  Thus, extrapolation will be purely linear.
    calibrated_slope[calibrated_limit] = (calibrated_value[calibrated_limit] - calibrated_value[calibrated_limit - 1])
                                         / (calibrated_input[calibrated_limit] - calibrated_input[calibrated_limit - 1]);

    // For the others, the slope is calculated from the preceding and succeeding slopes between the calibration points,
    // weighed by the input distances.
    for (uint32_t  i = 1; i < CALIBRATION_POINTS - 1; i++) {
        const double  dx0 = calibrated_input[i    ] - calibrated_input[i - 1];
        const double  dx1 = calibrated_input[i + 1] - calibrated_input[i    ];
        const double  dx  = calibrated_input[i + 1] - calibrated_input[i - 1];
        const double  dy0 = calibrated_value[i    ] - calibrated_value[i - 1];
        const double  dy1 = calibrated_value[i + 1] - calibrated_value[i    ];
        // dy0/dx0 is the slope between previous and current point, dy1/dx1 the slope between current and next point.
        // Weight is (1 - dx0)/dx = dx1/dx for the previous slope, and (1 - dx1)/dx = dx0/dx for the next slope.
        calibrated_slope[i] = (dx1/dx)*(dy0/dx0) + (dx0/dx)*(dy1/dx1);
        // Advanced models can adjust slopes to control the behaviour between calibration points,
        // when additional information is present.
        // It is most commonly done only for segments containing a peak in value, to control the overshoot.
    }

    // When calibration sample points and the slopes at each point are known, we can calculate the four coefficients.
    printf("# x f(x) df(x)/dx slope C0 C1 C2 C3\n");
    for (uint32_t  i = 0; i < CALIBRATION_POINTS - 1; i++) {
        const double  dx = calibrated_input[i + 1] - calibrated_input[i];
        const double  dx2 = dx*dx;
        const double  dx3 = dx*dx2;
        const double  c0 = calibrated_value[i + 1] / dx3;
        const double  c3 = calibrated_value[i] / dx3;
#ifdef  USE_SLOPE
        const double  c1 = 3*c0 - calibrated_slope[i+1] / dx2;
        const double  c2 = 3*c3 + calibrated_slope[i] / dx2;
#else /* USE_DERIV */
        const double  c1 = 3*c0 - calibrated_deriv[i+1] / dx2;
        const double  c2 = 3*c3 + calibrated_deriv[i] / dx2;
#endif

        calibrated_coeff[i][0] = c0;
        calibrated_coeff[i][1] = c1;
        calibrated_coeff[i][2] = c2;
        calibrated_coeff[i][3] = c3;

        printf("%18.12f %15.12f %15.12f %15.12f  %15.12f %15.12f %15.12f %15.12f\n",
               calibrated_input[i], calibrated_value[i],
               calibrated_deriv[i], calibrated_slope[i],
               calibrated_coeff[i][0], calibrated_coeff[i][1],
               calibrated_coeff[i][2], calibrated_coeff[i][3]);
    }
    printf("\n");

    printf("# x calibrated(x) f(x) absolute_error relative_error\n");
    for (float x = 1.0f; x < 65536.0f; x += 1.0f) {
        const float    c = calibrated(x);
        const double   r = f(x);
        printf("%18.12f %18.12f %18.12f %+.12f %+.12f\n", x, c, r, c - r, (c - r)/r);
    }

    return 0;
}
The calibrated_input[] array consists of the CALIBRATION_INPUT input values, and calibrated_coeff[] array consists of the four coefficients of each of the CALIBRATION_INPUT-1 segments (starting at the corresponding input value).  Extrapolation uses the initial and final segments.

The calibrated() function again uses a binary search to find the segment that contains the input value.  It then uses the numerically stable version of the cubic interpolation using the four coefficients (and the input values defining the segment).  This time, I used a slightly different structure for the function, with an outer do { ... } while (0) loop only for early break-out (when the segment index i has been discovered), but as you can see, it is very similar to the one for linear interpolation.

The test program uses the known derivative of \$\sqrt{x}\$, \$\frac{d \sqrt{x}}{d x} = \frac{1}{2 \sqrt{x}}\$, but if you define USE_SLOPE at compile time (e.g. -DUSE_SLOPE command-line parameter to gcc, clang, or icc), it estimates the slope for each calibration point using the previous and next calibration points.  The four coefficients for each curve segment is as listed in reply #9.  Because each sample point takes 4+16 bytes (compared to 2+4 bytes in the linear interpolation example), this time the default number of calibration points is just 128 (for 2560 bytes of Flash).  Also, both input and resulting values are now floats.  I could have implemented this in fixed point, but it would have looked much messier, because multiplying a*b where they have fa and fb fractional bits, and the result has fc < fa+fb bits, is (a*b)>>(fc-fa-fb), and the intermediate type has to be large enough to hold the entire product; plus correct rounding can look odd (adding 1<<(fc-fa-fb-1) (if result is positive, subtract if result is negative) before the shift is applied), so I wanted to focus on the important bits.  The cubic polynomial part itself isn't onerous either way, because it is just 8 "high-precision" multiplications and five additions or subtractions.

Even though I used importance sampling here as well, relative error in the interpolated value is still rather big for small inputs.  Even heavier emphasis (shorter segments at small inputs) would be useful, because the relative error becomes ridiculously small at higher values.  (After all, float only has about seven to eight significant digits; a relative error of ±0.00000003 or less means the value is within half an ULP in a float, i.e. as correct as you can get with float.  The actual values are computed at double precision –– I normally prefer to go as high as possible, long double or _Float128 (via libquadmath in C), but kept to the familiar double here for clarity.)

Note: I am not saying that piecewise cubic or linear approximation is the sensible way to deal with square root operation!  It is much better solved using even naïve Newton-Raphson, for example, where you iterate xx/2 + C/x to solve x = sqrt(2*C) starting with x=C (or with the Doom trick), although many other approaches are available, including binary ones.  I'm trying to show the look-up instead, and show how to define the cubic coefficients based either on preceding and succeeding calibration points for each point, or on the derivative of the value at each calibration point.
« Last Edit: March 03, 2024, 12:02:59 am by Nominal Animal »
 

Offline MrAl

  • Super Contributor
  • ***
  • Posts: 1450
Re: Calibration - How to Mathematically
« Reply #33 on: March 03, 2024, 06:53:03 pm »
Linear interpolation is usually very poor and only gives precision results using a very large number of points.

Cubic interpolation is more difficult to handle, but in contrast returns surprisingly good accuracy compared to linear. This can be used to significantly reduce the number of points in the table.

Reference for programming algorithms:
https://en.wikipedia.org/wiki/Neville%27s_algorithm
https://www.grad.hr/nastava/gs/prg/NumericalRecipesinC.pdf
https://apps.dtic.mil/sti/pdfs/ADA612943.pdf
https://www.sci.brooklyn.cuny.edu/~mate/nml/numanal.pdf

Hi there,

Not sure where you got that idea from that linear interpolation usually gives poor results.  It depends highly on the curve type.  With some curves it is going to give very good results.

You can measure the efficiency of the technique though for any curve by comparing the interpolated results for many samples with a given finite table to a table with an increased number of samples stored in it.  In one case, using a table of 100 values and linear interpolation resulted in matching the results of a table of 45000 entries and no interpolation.  That's a 450 to 1 savings in memory.  Of course without question a cubic interpolation could work even better maybe pumping that up even to 4500 to 1, but then the run time calculations take longer.
 

Offline Nominal Animal

  • Super Contributor
  • ***
  • Posts: 6288
  • Country: fi
    • My home page and email address
Re: Calibration - How to Mathematically
« Reply #34 on: March 03, 2024, 11:17:49 pm »
Bivariate interpolation, where you have a function \$z = f(x,y)\$ and some calibrated points \$(x, y, z)\$ can be divided into two major classes: One has regular grid, and the other has random points, in \$x, y\$.

The regular rectangular grid one is easiest, and both bilinear and bicubic interpolation is used.  For bilinear, you find the cell containing the point; the four points then have coordinates \$(x_0, y_0, z_{00})\$, \$(x_1, y_0, z_{10})\$, \$(x_0, y_1, z_{01})\$, and \$(x_1, y_1, z_{11})\$.  The relative coordinates within the cell are calculated first, then the bilinearly interpolated \$z\$:
$$\begin{aligned}
u &= \frac{x_1 - x}{x_1 - x_0} \\
v &= \frac{y_1 - y}{y_1 - y_0} \\
z &= (1-v)\bigl( (1-u) z_{00} + u z_{10} \bigr) + v \bigl( (1-u) z_{01} + u z_{11} \bigr) \\
~ &= z_{00} + u (z_{10} - z_{00}) + v (z_{01} - z_{00}) + u v ( z_{11} - z_{10} - z_{01} + z_{00} ) \\
\end{aligned}$$
although the first version for the calculation of \$z\$ is numerically stable, whereas the second is not.  (I included it to show that it indeed is bilinear: there is a constant coefficient, a linear coefficient for each variable, and a cross coefficient for the product of the two input variables.  That kind of equation can always be turned into a more numerically stable form, similar to the upper one of the two.)

For regular rectangular grid and bicubic interpolation, there are two variants.  A bicubic patch uses 4×4 points, or eight additional cells surrounding the current one; or you can calculate partial derivatives along each axis for each point, similar to slopes in reply #32 for univariate cubic interpolation.

When each cell is triangular, barycentric coordinates are often used.  Each (2D) point inside a triangle is identified by three coordinates, \$u\$, \$v\$, \$w\$, where \$u + v + w = 1\$.  This means that only two really only need to be specified, since the third can always be calculated (say, \$w = 1 - u - v\$).
The vertices of the triangle are always at coordinates \$(1, 0, 0)\$, \$(0, 1, 0)\$, and \$(0, 0, 1)\$, regardless of the shape of the triangle.  The center (average of the vertex coordinates) is always \$(1/3, 1/3, 1/3)\$.  If the \$z\$ at the respective vertices gets values \$z_u\$, \$z_v\$, and \$z_w\$, then the interpolated value is
$$\begin{aligned}
z &= u z_u + v z_v + w z_w = u z_u + v z_v + (1 - u - v) z_w \\
~ &= z_w + u (z_u - z_w) + v (z_v - z_w) \\
\end{aligned}$$
where yet again, the first version is numerically stable.  The same interpolation formula applies to \$x\$ and \$y\$ too.  Solving for the inverse involving \$x\$ and \$y\$ for \$u\$, \$v\$, and \$w\$, we find
$$\begin{aligned}
D &= x_w (y_u - y_v) + y_w (x_v - x_u) + x_u y_v  - x_v y_u  \\
u &= \frac{ x (y_v - y_w) - y (x_v - x_w) + x_v y_w - x_w y_v }{D} \\
v &= \frac{ y (x_u - x_w) - x (y_u - y_w) + x_w y_u - x_u y_w }{D} \\
w &= \frac{ x (y_u - y_v) + y (x_v - x_u) + x_u y_v - x_v y_u }{D} \\
\end{aligned}$$
where \$D\$ is twice the triangle area.  Again, only two of \$u\$, \$v\$, \$w\$ need to be solved, because the third is trivial to calculate.  If any of them is less than \$0\$ or greater than \$1\$, point \$(x, y)\$ is outside the triangle.
For interpolation among random points in \$x, y\$, the points are first converted to a triangular mesh.  The valid interpolation area is formed by the convex hull of the points (as if you stretched a rubber band around the points).  For each triangular cell, any of the six choices for labeling the vertices is valid.

In the one-dimensional case, we could use binary search to find the correct segment.  For triangular meshes in general, most time will be spent locating the correct triangle which contains point \$(x, y)\$.  Because of this, it is easier to "walk" along the mesh, and generate e.g. isocurves: a polygon describing constant values of \$z\$ in the \$x, y\$ plane.  Looking up individual points is always dominated by the time it takes to find the correct triangle.  For walking along the mesh, one typically just records the three triangles sharing sides with the current one, forming a bidirectional connectivity graph between the triangles.  Only when the "walk" path exactly crosses a vertex is a search necessary (among all triangles sharing that vertex), but as the triangles around a vertex will always form a connected ring in the connectivity graph, finding the correct triangle is a linear search among the triangles sharing the common vertex.  (The search can be made faster by using the 2D equivalent of the vector cross product to compare the walk direction and the direction of the other two vertices for each triangle in \$x, y\$ coordinates, as the product sign depends on their relative orientation, clockwise or counterclockwise.  For the correct triangle, the walk direction will be between the two.)

There are fast methods of converting a random set of 2D points into a mesh, the most commonly used one being Delaunay triangulation.

The underlying model can be easily extended to higher number of dimensions, to trilinear and tricubic interpolation.  Then, the triangle is replaced by its generalization, simplex, which is a tetrahedron in three dimensions, adding one new coordinate axis for each new dimension.  Because of the relative slowness of looking up the correct simplex cell, most commonly regular rectangular grids are used in higher number of dimensions.
Because the number of calibration points in \$D\$ dimensions where you have \$K\$ samples along each Cartesian axis means a total of \$K^D\$ points are needed, and this very rapidly grows to stupid number of points, polynomial and other functions tend to be used instead.  There, least squares fitting a smaller number of full-dimensional calibration points is very useful.  You can use tools like Scipy, Matlab/Octave for multi-variate fitting.

One scheme that I have seen is to start with the random calibration points, then sample the "slow" triangle mesh to generate a regular rectangular grid.  There is typically enough information in the triangular mesh to provide good estimates for bicubic interpolation (surface tangents, or slopes along each grid axis), yielding a smooth and continuous interpolation with acceptable memory use, if the underlying surface being modeled is smooth and continuous itself.

Discontinuities in the interpolated function are the hardest to deal with.  Consider e.g. \$f(x) = \frac{1}{x - 2}\$, which has a discontinuity at \$x = 2\$.  If you approach it from below, it tends to negative infinity; if you approach it from above, it tends to positive infinity.  For one-dimensional functions you can discover the discontinuities and define the function piecewise, but for 2D and higher dimensions, it becomes very complicated, and generally not worth even trying.
« Last Edit: March 03, 2024, 11:19:28 pm by Nominal Animal »
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #35 on: March 04, 2024, 06:16:51 pm »
Hi there,

Not sure where you got that idea from that linear interpolation usually gives poor results.  It depends highly on the curve type.  With some curves it is going to give very good results.

You can measure the efficiency of the technique though for any curve by comparing the interpolated results for many samples with a given finite table to a table with an increased number of samples stored in it.  In one case, using a table of 100 values and linear interpolation resulted in matching the results of a table of 45000 entries and no interpolation.  That's a 450 to 1 savings in memory.  Of course without question a cubic interpolation could work even better maybe pumping that up even to 4500 to 1, but then the run time calculations take longer.

Obviously:
   Linear interpolation is faster (x10).
   Linear interpolation occupies more memory (x10).

   Cubic interpolation is slower (by a factor of about 10 if not precomputed).
   Cubic interpolation occupies less memory (by a factor of about 10).


A good linear table requiring 100 points can be solved with 10 points and cubic interpolation, probably with better accuracy.

The speed is quite fast for a current micro. If the cubic interpolation coefficients are precomputed, we are talking about doing 4 multiplications and 4 additions for each interpolation.
« Last Edit: March 04, 2024, 06:31:10 pm by Picuino »
 

Offline max_torque

  • Super Contributor
  • ***
  • Posts: 1294
  • Country: gb
    • bitdynamics
Re: Calibration - How to Mathematically
« Reply #36 on: March 04, 2024, 06:55:04 pm »
The CRITICAL fact of importance here is simply "how accurate do you need to be?"

People often go off down rabbiut holes and spend ages coming up with complex solutions to problems that simply don't require it!

For example, the solution to say a temperature sensor NTC calibration characteristic for a greenhouse temperature sensor (+- 1degC is more than good enough) is very very different to the solution required for say a complex organic chemistry thermostat, where +-0.05degC might be required.

Knowing where you are going is critical to picking the correct route!


Chances are, unless you are doing something very clever and complex, a basic linear interpolation from a look Up Table with carefully chosen index nodes is all that will be required (more nodes where there is a larger "acceleration" of the signal, less where it's more constant)
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #37 on: March 04, 2024, 07:36:09 pm »
The final accuracy is not very decisive. The NTC calibration can be solved with 100 points and linear interpolation and the other thermostat with 1000 points and linear interpolation.
You could also solve the NTC calibration with 10 points and cubic interpolation and the thermostat calibration with 100 points and cubic interpolation.
The method of chosen index nodes is valid for any interpolation, both linear and cubic.

With temperature (and many other measurements) speed is not important considering how fast any micro calculates today.

One thing is certain: if you are not good at math, the best thing to do is make a LUT. If you're up to the basics, linear interpolation.
If you know first year college math, you can handle cubic interpolation and beyond. if you are an expert in mathematics, you will be able to deal with rational interpolation, the most difficult and capable of approximating all types of curves.
You also have to take into account the time you will spend on understanding and applying methods that are becoming more and more complex.
« Last Edit: March 04, 2024, 08:02:57 pm by Picuino »
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #38 on: March 04, 2024, 07:52:56 pm »
How to calibrate a RTD (Platinum Resistance Thermometry), very accurate termometer, with ITS-90 standard.
Fluke: https://s3.amazonaws.com/download.flukecal.com/pub/literature/3498460B_EN_w.pdf
Bureau International des Poids et Mesures: https://www.bipm.org/documents/20126/41773843/Guide_ITS-90_5_SPRT_2021.pdf

 

Offline Nominal Animal

  • Super Contributor
  • ***
  • Posts: 6288
  • Country: fi
    • My home page and email address
Re: Calibration - How to Mathematically
« Reply #39 on: March 04, 2024, 10:53:58 pm »
Cubic interpolation is slower [than linear interpolation] (by a factor of about 10 if not precomputed).
No, those figures are not representative, even if you consider 10 as "an approximate order of magnitude".

Let's look at the precomputed implementations first.

Numerically stable linear interpolation uses \$x_i\$, \$x_{i+1}\$, \$A_i\$, and \$B_i\$, so that
$$y = (x_{i+1} - x) A_i + (x - x_i) B_i$$
which is exactly two multiplications and three additions or subtractions.  The precomputed coefficients are
$$A_i = \frac{y_i}{x_{i+1} - x_i}, \quad B_i = \frac{y_{i+1}}{x_{i+1} - x_i}$$
Note that there are one less pair of coefficients than there are calibration points.

Numerically stable cubic interpolation uses \$x_i\$, \$x_{i+1}\$, \$A_i\$, \$B_i\$, \$C_i\$, and \$D_i\$, so that
$$\begin{aligned}
y &= (x_{i+1} - x)^3 A_i + (x_{i+1} - x)^2 (x - x_i) B_i + (x_{i+1} - x)(x - x_i)^2 C_i + (x - x_i)^3 D_i \\
~ &= (x_{i+1} - x)^2 \bigl( (x_{i+1} - x) A_i + (x - x_i) B_i \bigr) + (x - x_i)^2 \bigl( (x_{i+1} - x) C_i + (x - x_i) D_i \bigr) \\
\end{aligned}$$
which is exactly eight multiplications and five additions or subtractions.  There are three basic ways to precompute the coefficients; using four consecutive points \$(x, y)\$, or using the derivative at each point, or using an estimate of the derivative at each point derived from the previous and next points relative to the current one.  The last two I've shown earlier in this thread.

So, with precomputed coefficients, the cubic interpolation is something like 4× the cost of linear interpolation, plus the same overhead for locating the correct segment (binary search performing it in log2(segments) time complexity).  The difference is therefore much less than a factor of four.  It is also ridiculously fast, involving so few arithmetic operators, that even function call overhead is a significant fraction of the overall time taken per call on many architectures.  (The exception is on 8-bit architectures where you use software emulated floating-point.  Fixed point is much better suited for this there.)

Not precomputing the coefficients does not save you much.  In addition to the values of \$x\$, you need either one (linear interpolation without precomputing, cubic interpolation without precomputing using each set of four consecutive points for each cubic polynomial), two (precomputed linear interpolation, cubic interpolation without precomputing using value and tangent at each point), or four (precomputed cubic interpolation) coefficient per point.
Linear interpolation without precomputing is
$$y = \frac{(x_{i + 1} - x) y_i + (x - x_i) y_{i + 1}}{x_{i+1} - x_i}$$
which uses two multiplications, one division, and four additions or subtractions.  The time needed for the division dominates the time taken by this computation, except on those rare-ish architectures with division as fast as multiplication.  On most architectures, it takes roughly the same time as cubic interpolation with precomputed coefficients, because the division is just so darn slow compared to multiplications and additions and subtractions.

Cubic interpolation using four consecutive points has a few dozen terms, and I've never optimized it, because it just isn't very useful (in the numerically stable form, which is what I want to use).  I can compute the exact formula, if someone wants it, though.

Cubic interpolation using the derivative (known or estimated) at each point, so that we have \$x_i\$, \$y_i\$, and the derivative or slope \$\gamma_i\$, at each point.  Then, the interpolation is
$$y = \frac{ (x_{i+1} - x)^2 \biggl( (x_{i+1} - x) y_i + (x - x_i) \bigl( 3 y_i + \gamma_i (x_{i+1} - x_i ) \bigr) \biggr)
+ (x - x_i)^2 \biggl( (x - x_i) y_{i+1} + (x_{i+1} - x) \bigl( 3 y_{i+1} - \gamma_{i+1} (x_{i+1} - x_i ) \bigr) \biggr)
}{(x_{i+1} - x_i)^3}$$
which can be done using 11 multiplications, one division, and 12 addition and subtractions, if you use some temporary variables, in the numerically stable form.  Even here, the division is a significant part of the time cost.  Still, it is no more than five times the cost of the linear interpolation.

Because importance sampling makes a huge difference in accuracy of the piecewise approximation, all these have the exact same cost of locating the correct segment in \$x\$, and it does take significant time compared to the actual arithmetic operations done to interpolate within the found segment.  One would expect something like a factor of 2 to 3 difference in practice between linear and cubic interpolation without precomputed coefficients; and that is very little when considering the much better fit one can get with piecewise cubic curves.  On architectures where division takes several times the time a multiplication does, cubic interpolation with precomputed coefficients can be faster than linear interpolation without precomputed coefficients.
Essentially, the precomputed coefficients can eliminate the division, half of the additions and subtractions, and a third of the multiplications: it trades increased memory use for fewer arithmetic operations.

Exact speed differences depend on the relative speed of the division compared to multiplication, and how large a fraction of the overall time does the binary search take to find the correct segment.

To repeat: you do not want to require \$x\$ to be regularly sampled, because importance sampling, i.e. picking more points where where the behaviour changes or is more complex, always yields better results than regular sampling, both in memory use and runtime taken.

Note that the formulas used to precompute the coefficient are not so complicated one could not do them on a microcontroller.  That is, you could easily reserve a portion of Flash for the coefficient table, and let the user set individual calibration points.  Just keep them in sorted order according to \$x\$.  When the user is satisfied, you could compute the coefficients, and update the Flash to reflect the new calibrated coefficients.  There is no need to do the precalculation offline, because the coefficient calculations are done only once per calibration point, and even if using software-emulated double-precision floating-point on an 8-bit microcontroller, don't take enough time for a human to notice, until there are hundreds of the calibration points.
« Last Edit: March 05, 2024, 06:51:01 pm by Nominal Animal »
 
The following users thanked this post: Picuino

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #40 on: March 05, 2024, 01:55:45 pm »
The factor I wrote (about 10) was to compare the linear interpolation with the cubic interpolation without precalculating the coefficients (Neville's algorithm) and it was an approximation, not an exact figure. My argument was trying to point out the slower speed of cubic interpolation, with the advantage of memory reduction.
Thanks for your clarification.

---------

On the other hand normally we are not going to have a whole table of calibration values available. In many cases we are going to have 3, 4 or 5 calibration points at most. In that case it is best to use an interpolating polynomial of maximum degree and choose well the calibration points.
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #41 on: March 05, 2024, 02:14:24 pm »
A very interesting application note from TDK for calibrating NTC thermistors for temperature measurement:
https://www.tdk-electronics.tdk.com/download/2999750/74d6ef4ba23ccb21f85f2df587a15eff/ntc-thermistors-an.pdf

It compares interpolation with 2 calibration points (blue), Steinhart-Hart equation with 3 calibration points (orange) and Lookup Table method (green).
« Last Edit: March 05, 2024, 02:21:14 pm by Picuino »
 

Offline Nominal Animal

  • Super Contributor
  • ***
  • Posts: 6288
  • Country: fi
    • My home page and email address
Re: Calibration - How to Mathematically
« Reply #42 on: March 05, 2024, 06:40:08 pm »
In many cases we are going to have 3, 4 or 5 calibration points at most. In that case it is best to use an interpolating polynomial of maximum degree and choose well the calibration points.
Yes.  In some cases the polynomial itself is known or specified, for example the 12th degree polynomial for RTDs at -259.35°C to 0°C and the 9th degree polynomial at 0°C to 962°C as specified in ITS-90, and the problem is to adjust the coefficients so that at the calibrated input values, the known result is obtained.  The math problem then is that there are not enough calibration points to fix the coefficients.

In practice, using an RTD like a Pt100 or Pt1000 sensor, the software problem is to convert an ADC reading to a temperature, i.e. we have a function T(x) where x is the ADC reading.  The function itself is a 12th degree polynomial, so we'd need 12 calibration points to fix the coefficients.  Partially "adjusting" the coefficients from only a smaller number of calibration points is mathematically extremely complicated, and not worth trying.

Instead,  a two-point calibration is often used, which assumes the shape of the polynomial is the one as specified in ITS-90, and only scale and offset need adjustment.  In other words, we assume that there are no nonlinearities in the resistance-to-digital measurement, with T(x) describing the correct behaviour, but requiring a linear correction: offset and scale.

If we have defined the temperature as a function of ADC reading as (piecewise)
$$T(x) = \sum_{i=0}^{12} C_i x^i = C_0 + C_1 x + C_2 x^2 + C_3 x^3 + \dots + C_{11} x^{11} + C_{12} x^{12}$$
then the two-point calibration uses
$$T(A_0 + A_1 x) = \sum_{i=0}^{12} C_i x^i = C_0 + C_1 x + C_2 x^2 + C_3 x^3 + \dots + C_{11} x^{11} + C_{12} x^{12}$$
If the two calibration temperatures are \$T_0\$ and \$T_1\$, with \$T(x_0) = T_0\$ and \$T(x_1) = T_1\$ (which you can use a binary search in \$x\$ because the function \$T\$ is monotonic even though it is a high-degree polynomial), then
$$\left\lbrace \begin{aligned}
A_0 + A_1 x_0 &= T_0 \\
A_0 + A_1 x_1 &= T_1 \\
\end{aligned} \right . \quad \iff \quad \left\lbrace \begin{aligned}
A_0 &= \frac{T_0 x_1 - T_1 x_0}{x_1 - x_0 } \\
A_1 &= \frac{T_1 - T_0}{x_1 - x_0} \\
\end{aligned} \right .$$
with the actual function \$T\$ always being unchanged, and the linear correction applied to the ADC reading before evaluating \$T\$.

If you wanted to apply the linear correction directly to the polynomial coefficient, I can show how that is done, but I recommend against it.  \$T(x)\$ tends to not be particularly numerically stable –– it is okay because it is mostly linear, so the higher-degree coefficients are small adjustments to the linear value ––, and applying the linear correction directly to the polynomial coefficients combines them in an even less numerically stable way.  Applying the linear correction to \$x\$ as above, on the other hand, yields acceptable results.  That is, for ADC reading \$x\$, the corresponding temperature is \$T(A_0 + A_1 x)\$.

Interestingly, this works the same even when T(x) was defined using piecewise cubic or piecewise linear segments, which can be done efficiently using integer or fixed-point arithmetic; useful when floating-point emulation is deemed too slow.

Technically speaking, if you have the theoretical behaviour modeled as a polynomial or a piecewise linear or piecewise cubic or any other type of function, you can apply the linear two-point calibration correction the same way.  You only need to find a way to obtain \$x\$ for each calibration point corresponding to the user-specified \$T\$.
For piecewise-defined monotonic functions, the same binary search can be used to find the segment containing the value \$T\$ (aka \$y = f(x)\$).  If the points at the end of the segment are \$x_i, y_i\$ and \$x_{i+1}, y_{i+1}\$, \$y = f(x)\$ solved for x is
$$x = \frac{(x_{i+1} - x_i) y + x_i y_{i+1} - x_{i+1} y_i}{y_{i+1} - y_i}$$
For the cubic interpolation, considering the complexity of the solution, I recommend doing a binary search within \$x_i\$ and \$x_{i+1}\$ to the desired resolution (each iteration adds one bit of precision to \$x\$) instead, considering this is only needed during calibration, and only needs to be sufficiently fast.

In the general case, for monotonic functions, you can solve for x using a binary search in calibrate(x).  It won't be the fastest possible option, but it will be fast enough for calibration, even at say 64 bits of precision.
« Last Edit: March 05, 2024, 06:57:33 pm by Nominal Animal »
 

Offline Tylerhowarth

  • Newbie
  • Posts: 3
  • Country: us
Re: Calibration - How to Mathematically
« Reply #43 on: April 25, 2024, 11:43:20 am »
The final accuracy is not very decisive. The NTC calibration can be solved with 100 points and linear interpolation and the other thermostat with 1000 points and linear interpolation.
You could also solve the NTC calibration with 10 points and cubic interpolation and the thermostat calibration with 100 points and cubic interpolation.
The method of chosen index nodes is valid for any interpolation, both linear and cubic.

With temperature (and many other measurements) speed is not important considering how fast any micro calculates today.

One thing is certain: if you are not good at math, the best thing to do is make a LUT. If you're up to the basics, linear interpolation.
If you know first year college math, you can handle cubic interpolation and beyond. if you are an expert in mathematics, you will be able to deal with rational interpolation, the most difficult and capable of approximating all types of curves.
For absolute beginners, I recommend using specialized services that focus on assisting schoolchildren and students. For myself, I've found https://essays.edubirdie.com/math to be particularly reliable, as it hasn't let me down yet. However, guys, remember that practice makes perfect.
You also have to take into account the time you will spend on understanding and applying methods that are becoming more and more complex.
Thanks for the tip on LUT! I was struggling with linear interpolation before, and it wasn't working out. Could you please share how accurate the results are with LUT and which type of LUT you recommend?
« Last Edit: May 01, 2024, 11:53:33 am by Tylerhowarth »
 

Offline Picuino

  • Frequent Contributor
  • **
  • Posts: 765
  • Country: 00
    • Picuino web
Re: Calibration - How to Mathematically
« Reply #44 on: April 26, 2024, 07:52:45 am »
It depends. What function do you want to approximate and with what accuracy?
 

Offline MrAl

  • Super Contributor
  • ***
  • Posts: 1450
Re: Calibration - How to Mathematically
« Reply #45 on: April 26, 2024, 01:39:08 pm »
Hello again,

Just to note, there are more numerical interpolation methods than there are stars in the sky  :phew:
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo