So something is way off, as the voltage specified for that step is way off.
The page attached is from the wrong manual. My serial number starts with 0995A. HP’s serial numbering scheme back then leaves a lot to be desired. The manual from HP’s archive part number 00428-90003 dated 1970, says it is for serial numbers starting with 995 and above. I assumed this was applicable to my unit.
A version of the manual of which I have a very poor, mostly unreadable copy of is 00428-90005 dated 1983 and says it is for serial numbers 0995A and later. 0995A vs. 995?!?!? Older units start with 131 or 601! That is the one I suppose is for my unit. Other discrepancies between instructions in the different manuals are some state disconnecting feedback while the other makes no mention of it or specifically says to reconnect it!!!!!!
I compared the alignment pages as best I can and they read practically the same, except the newer manual adds that the voltage expected for the test in paragraph 5-42 is 2.5-3.3v RMS. I am getting 102mv. Steps 5-36 to 5-40 are all good.
V5 tests fairly good though not super strong on my tube tester. The 12vdc feeding R49 is my measured rail of 11.7vdc. The 16.2 vdc shown for the resonant circuit is 15.8. But the waveform shown right above it which look like 10v p-p sine, is a jittery .5v p-p sine.
Again, the bottom half of the L7 core does not move. The coil is firmly attached to it and wires from the coil pass through the gap on the bottom half of the core. The top half is free to rotate when assembled but how they expected you to rotate it is a mystery. Either way, I don’t believe any setting of L7 is going to raise the voltage from 102mv to 2.5v.
Your knowledge of the circuitry leads me to believe you are already referencing the schematic, but if a full schematic is needed I’ll post it.
Thanks,
Jeff