That might figure as I am having trouble with region of the circuit, maybe a biasing issue. If I warm Q12 the counter works on a particular range. I changed it for a brand new one with the same issue. I believe Q12 itself is fine, but warming it *may* alter its biasing needs, perfectly normally? If I measure the collector voltage referenced to ground I see 0.22 volts. If I warm Q12 even slightly I get a slowly, smoothly rising voltage, collector to ground, and at about .52V the counter starts reading on Band 2. Continue to warm Q12 (within very conservative bounds, not mad hot

), I see up to 11.9V.
As Q12 is allowed to cool to ambient the collector to ground voltage drops smoothly and slowly again, at around 0.5V the counter stops seeing the 10MHz input at 50mV, and reads zeros again. By the time Q12 is back to ambient I see 0.22V again. I am admittedly out of my depth, but feel the problem is local to this section of the circuit.
Base V is about 11.66V Emitter V is about 12.07V