Hello there.
I already own a Fluke 8050A bench multimeter, and I'm very pleased with it. So that's why I decided to acquire myself another one. I found a good deal on one, and thus ordered it. Afterwards though, I've discovered that it's the 110V model, and since I live den Denmark, we have 230V mains. I then decided to crack that sucker open, and found on the input (A white wire (assuming live), a green/yellow (mains ground) and a black wire (assuming neutral). The white wire is wired to a hole with "110V" marked on it, and next to it is a unpopulated hole marked "230V" which my current meter is wired to.
I suspected that it would be that easy, since it draws so little power, so I didn't suspect that Fluke would go through the hassle to make two different transformers for the two regions.
The question is, is there anything else that I should be aware of, or should I just solder the white wire to the 230V hole and be good to go? I don't want to brick it with a stupid mistake.
Thanks in advance!