Hi guys,
I'm very green when it comes to test and measurement equipment, and bought myself an entry level OWON SDS7102 to learn the ropes with. Whilst testing a Ham-It-Up converter, I tried measuring a 125Mhz crystal just for laughs, under the assumption the 100Mhz scope wouldn't manage to trigger.
To my surprise it managed it, and successfully measured the frequency pretty accurately. To get a reasonable trace I ended up setting the scope all the way down to a 2ns timebase, and only a 1k sample length
I'm struggling with the concept of how a scope with a 100Mhz front-end can accomplish this, am I misunderstanding what the 100Mhz limit really is, or do manufacturers build in some wiggle room?
Owon SDS7102 is named as 100MHz. It is promised in specs that risetime (10% - 90%) is less or equal 3.5ns.
In practice, every SDS7102 what I have tested (and I have tested lot of) have -3dB point around 150MHz to 170MHz when used 50ohm cable and external 50ohm feed thru termiantion in scope input. (exept 2mV/div range what have highly reduced BW and there -3dB "niminal" point is around 20MHz. (in practice +/- 5MHz)
If use probe with 1x setting it need note also that probe bandwidth is very higly reduced and in practice if want even some level accuracy it is better to stay well under 5MHz.
With single channel in use and if selected 10M memory it kan keep 1GSa/s down to 0.5ms/div horizontal speed.
With FFT it can use up to 500MHz (note signal attenuation over 100MHz frequencies)
here can also find some tests:
http://owon.freeforums.org/some-tests-with-fft-t8.htmlhttp://owon.freeforums.org/some-tests-with-new-sds7102v-version-2-6-and-2-8-2-t10.htmlalso here (partially same but something more.
http://owon.freeforums.org/This is main forum but due to some old issues with this forum there is "spare" forum owon.freeforums.org but also without registration there can see images)
There is now lot of not updated information due to very busy times with some projects.
Example that today FW version Owon waveform update rate is much better than previously.
Also there can find samplerate tables for different Timebase (time/div) and memory settings.
----
Of course probe GND wire is too long for fast rising (or high frequency) signals.
It may be around 180nH.
(straight wire, 0.3mm diameter and 150mm lenght inductance is around 200nH
with 100MHz its reactance is 125ohm, and with 200MHz of course then 250ohm.
This is not resistive, it is inductive reactance and then there are affecting also capacitances... result is example "ringing" with fast edges)