There's nothing wrong with your observation.
measuring 11uA matches the 150pF, that is the current that flows through your body at 230V / 50Hz and the conditions you've described.
I doesn't make any difference if you touch the line phase directly or the 230V output of the earthed isolation transformer.
I expect you'd feel the same nothing if you touch the phase directly (No I don't recommend you doing this).
If you got zapped by touching the 230V line, I suppose there must have been different conditions than with your isolation transformer experiment, maybe you were wearing other shoes, or your skin was more sweaty, ...
I have touched in my life many times accidentally the 230V mains, and believe me, even with rubber sole shoes, dry hands and standing on wood floor, you get a decent shock.
That why i'm surprised that i don't feel anything when i touch the 230V output of the grounded isolation transformer.
That's the reason also of this topic, i do not want to kill myself, but i like to know, where the difference is between both power sources.
Wow, TN-S, you guys really do have 230vac referenced to earth!!! That is a sure killer if you screw up, unless protected with ground fault type gear. Here in the states the 240vac system is made from 120-N-120 and so the worst shock to earth is only 120vac not 230/240. When I worked at the hospital in the O.R.'s the systems were 120vac center-tapped with something like 2.2 megaohms from the center tap to earth and a LEM module to measure offset and current attempting to flow to earth. In the U.S. the N.U.E.C. specifies that the 'earth prong' of the three prong plug be oriented at the bottom of the outlet. In our operating rooms the plugs were oriented with earth prong on the top side. This was in case the very fine catheter
wires would accidentally fall down on an outlet it would likely hit the ground and be physically prevented from falling farther to the L1/L2 prongs.
Don't go grabbing live conductors!! There is no reset button for electrocution victims.
Prior before mid 80's (1986 i think) residential electric installations here in Belgium needed one RCD of 300mA for the entire installation.
From 1986 and above, there are 2 RCD's needed, one RCD of 300mA for the entire installation, and a 30mA RCD for the wet rooms
like basement, garage, laundryroom etc....
Here at my home, my installation is from 1975, and i have one RCD of 30mA for the entire house, and if wanted, i can use
a portable 10mA RCD when working on electrical devices.
About grabbing live wires, when i first touched the grounded isolation transformer, i was not directly, but with a 1M resistor between me and the live wire.
After that, i kept lowering the resistor value untill i reached 10K and still didn't feel anything.
That was the moment that i quickly tapped the live wire without resistor for the first time.
Therefore, if I touched only the ungrounded pole of the isolation transformer without completing a circuit I would not expect to feel anything. However, if I also completed a path to ground or the other pole of the transformer then I would be concerned.
You are right, when the isolation transformer is not grounded, touching only 1 pole, won't do anything (the main purpose of a isolation transformer).
But in theory, if you ground 1 pole of the isolation transformer, and you touch the second pole, you will get a zap.
But it seems that in my case, even then this is safe.
My only concerns about this is, and only in case of IF, that if i'm working on a device that have a hidden Line to chassis fault , that i can make the final path
between both poles of the isolation transformer if i touch the chassis and the second pole accidentally.
Then i'm getting the full 230V through my body, while if get a tinkle by only touching one pole, i get a warning so that i can't touch both poles at once.
Like a mentioned before, this grounding of the secundairy side of the isolation transformer, is for testing purposes only
and not ment to be kept connected all the time.
Grtz
DeLorean