| tautech:
--- Quote from: barry14 on February 10, 2019, 11:36:11 pm ---You should use the 1v cal signal to check your probes as that is what it was designed for. The 1v level should be accurate to a few percent. The 10v gate signal is a signal that is high (about 10v) during the horizontal sweep and low otherwise. It is not intended as a calibration signal but for use in syncing the oscilloscope sweep to other equipment. You can certainly look at it once your probes have been adjusted using the cal signal. --- End quote --- Exactly this ^. :-+ |
| wa7son:
--- Quote from: wa7son on February 10, 2019, 08:49:13 pm ---I don't have any lying around, but I'll head to the store tomorrow to buy one and try it out --- End quote --- Bought a 4mm banana plug and used the CAL 1V port. It worked like a charm for getting the probes calibrated. Thanks for all the help :) |